Temperature controller for oil immersion objectives

classic Classic list List threaded Threaded
♦
♦
3 messages Options
Joseph Georgeson Joseph Georgeson
Reply | Threaded
Open this post in threaded view
♦
♦
|

Temperature controller for oil immersion objectives

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Greetings all,

I would like to purchase a temperature controller for our inverted microscope (IX-83) for use with high magnification oil immersion objectives. Ideally I would like to use a stage-only, peltier-based temperature controller for both heating and cooling (~10-40C). This is for sub-micron scale objects, so the heating/cooling unit must not generate vibrations which I would expect from fan-cooled units.

Does anyone have experience with this? Any recommendations for products?

Thanks,
Joe
Dan Focht Dan Focht
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Temperature controller for oil immersion objectives (commercial response)

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Joe


Bioptechs has seen a significant trend toward micro-environmental systems that provide transitional temperature shifts for gene expression among other things.
The main factors other than temperature have been the speed at which the temperature changes and, as you pointed out, vibration.  
The approach we have taken is to do the temperature transfer with two, external, thereby vibration isolated, Peltier controlled recirculating systems each at a different temperature.  
We then set up a TTL activated solenoid switching system that provides the following conditions:
1. Temp A to the scope and back to itself while Temp B cycles back to itself
2. Temp B to the scope and back to itself while Temp A cycles back to itself
We have documented a 10 - 40 degree transition in less than 30 seconds excluding the objective, that takes a little longer.

On the scope we provide a fluid optical cavity such that the specimen on a coverslip has its own isolated space that is separated from the circulating temperature controlled fluid by a 0.5mm window.  If you are using high N.A. objectives, we use a temperature transfer collar on the objective.
This provides highly efficient heat transfer to the specimen and enables both reflected fluorescent and transmitted modes of microscopy.  
For additional information contact Dan at Bioptechs. 724-282-7145.  

Note, I would not recommend any type of system that relies on peripheral heat exchange due to the long equilibration times and thermal gradients that occur through peripheral heat propagation.  


Dan



> On Jul 28, 2016, at 7:11 AM, Joseph Georgeson <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Greetings all,
>
> I would like to purchase a temperature controller for our inverted microscope (IX-83) for use with high magnification oil immersion objectives. Ideally I would like to use a stage-only, peltier-based temperature controller for both heating and cooling (~10-40C). This is for sub-micron scale objects, so the heating/cooling unit must not generate vibrations which I would expect from fan-cooled units.
>
> Does anyone have experience with this? Any recommendations for products?
>
> Thanks,
> Joe

Dan Focht
Bioptechs, Inc.
3560 Beck Rd.
Butler PA 16002
V724-282-7145
F724-282-0745
Toll Free 877 lIVE-CELL (548-3235)
[hidden email]
jerie jerie
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Temperature controller for oil immersion objectives

In reply to this post by Joseph Georgeson
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Joe,

I have co-developed such a device 10 years ago, it is made available
through https://emblem.de/technologies/instruments/
Fast response mini stage temperature controller
We have used it to control temperature sensitive allels, it shifts between
Tech data sheet: http://195.34.172.78/downloads/5/EMBLEM_Ref_396.pdf
publication doi:10.1016/j.febslet.2007.03.052
<http://dx.doi.org/10.1016/j.febslet.2007.03.052>

Comercial disclaimer: inventors receive a small compensation from device
sales.

Abraços,
Jens

Dr. Jens Rietdorf, visiting scientist @ center for technological
development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de Janeiro
Brasil.

On Thu, Jul 28, 2016 at 8:11 AM, Joseph Georgeson <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Greetings all,
>
> I would like to purchase a temperature controller for our inverted
> microscope (IX-83) for use with high magnification oil immersion
> objectives. Ideally I would like to use a stage-only, peltier-based
> temperature controller for both heating and cooling (~10-40C). This is for
> sub-micron scale objects, so the heating/cooling unit must not generate
> vibrations which I would expect from fan-cooled units.
>
> Does anyone have experience with this? Any recommendations for products?
>
> Thanks,
> Joe
>