To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, I need to image fixed cryosections of brain tissue on an Olympus TIRF microscope. My guess is that I can't use conventional specimen slides (ie Fisher Superfrost +) to collect tissue. Has anyone had any experience collecting tissue on Coverslips, and preparing them for TIRF. Thank you! Adele Tufford, McGill University |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Adele Tufford, The coverslip is part of the TIRF imaging apparatus, which produces total internal reflection at the coverslip-specimen boundary. The evanescent wave “leaking” into the specimen, within 100nm immediately above the coverslip, excites the fluorophone to produce fluorescent emission. Therefore, you have to make sure that your tissue specimen attaches to the coverslip properly. It does not matter what slide you use. Cheers Gary G Li, PhD On Sun, Nov 18, 2012 at 2:25 PM, Adele Tufford <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello, > > I need to image fixed cryosections of brain tissue on an Olympus TIRF > microscope. > My guess is that I can't use conventional specimen slides (ie Fisher Superfrost > +) to collect tissue. Has anyone had any experience collecting tissue on > Coverslips, and preparing them for TIRF. > > Thank you! > > Adele Tufford, McGill University |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Adele Tufford, Further to my answer to your last posting, due to microtome damage to cells on the surfaces of the tissue section, I am not sure how much useful info you can get by using TIRF. I would recommend confocal laser scanning microscopy or widefield fluorescent microscopy with deconvolution for your application. Regards Gary G Li, PhD On Sun, Nov 18, 2012 at 4:20 PM, Gary G. Li <[hidden email]> wrote: > Hi Adele Tufford, > The coverslip is part of the TIRF imaging apparatus, which produces > total internal reflection at the coverslip-specimen boundary. The > evanescent wave “leaking” into the specimen, within 100nm immediately > above the coverslip, excites the fluorophone to produce fluorescent > emission. Therefore, you have to make sure that your tissue specimen > attaches to the coverslip properly. It does not matter what slide you > use. > Cheers > Gary G Li, PhD > > On Sun, Nov 18, 2012 at 2:25 PM, Adele Tufford > <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello, >> >> I need to image fixed cryosections of brain tissue on an Olympus TIRF >> microscope. >> My guess is that I can't use conventional specimen slides (ie Fisher Superfrost >> +) to collect tissue. Has anyone had any experience collecting tissue on >> Coverslips, and preparing them for TIRF. >> >> Thank you! >> >> Adele Tufford, McGill University |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I might be stating the obvious, but for TIRF to work well, your sample mounting medium must also be aqueous in order for the TIR effect to occur (strict TIRF requires a refractive index mismatch at the sample/coverslip interface). That's also assuming you really are interested in seeing what happens in the regions closest to the coverslip. If want to simply illuminate your sample (all depths) with laser light in a widefield/epifluorescence condition using a TIRF illuminator, it will work fine for that too. Now, having said that I recently observed that the TIR effect can occur (disappearance of out-of-focus light) with a fixed monolayer sample of cells (fluorescently labeled membranes) mounted in mowiol - but just barely. I'm not sure what the refractive index of mowiol is, but it must have been such that the critical angle corresponded to a a back-aperture radial position just inside the inner edge of the objective (the NA was 1.47 in that case). Cheers, John Oreopoulos Research Assistant Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca On 2012-11-18, at 5:56 PM, Gary G. Li wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Adele Tufford, > Further to my answer to your last posting, due to microtome damage to > cells on the surfaces of the tissue section, I am not sure how much > useful info you can get by using TIRF. I would recommend confocal > laser scanning microscopy or widefield fluorescent microscopy with > deconvolution for your application. > Regards > Gary G Li, PhD > > > On Sun, Nov 18, 2012 at 4:20 PM, Gary G. Li <[hidden email]> wrote: >> Hi Adele Tufford, >> The coverslip is part of the TIRF imaging apparatus, which produces >> total internal reflection at the coverslip-specimen boundary. The >> evanescent wave “leaking” into the specimen, within 100nm immediately >> above the coverslip, excites the fluorophone to produce fluorescent >> emission. Therefore, you have to make sure that your tissue specimen >> attaches to the coverslip properly. It does not matter what slide you >> use. >> Cheers >> Gary G Li, PhD >> >> On Sun, Nov 18, 2012 at 2:25 PM, Adele Tufford >> <[hidden email]> wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hello, >>> >>> I need to image fixed cryosections of brain tissue on an Olympus TIRF >>> microscope. >>> My guess is that I can't use conventional specimen slides (ie Fisher Superfrost >>> +) to collect tissue. Has anyone had any experience collecting tissue on >>> Coverslips, and preparing them for TIRF. >>> >>> Thank you! >>> >>> Adele Tufford, McGill University |
| Free forum by Nabble | Edit this page |