Hi all,

I'm happy to report we were able to undo the damage.

It took 5-7 rounds of alternating applications of EtOH/water (5-10s exposure) followed by wiping with lens paper.

Before: https://imgur.com/p38eD22
After: https://imgur.com/a/WWCPxHH
In action: https://imgur.com/a/yfR5Oez

Thank you to everyone who provided suggestions for solvents/protocols as well as creative punishments.

Best,
Will


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steven Nedellec
Sent: Friday, February 8, 2019 7:10 AM
To: [hidden email]
Subject: Re: What happened to this objective?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi William

We encountered the same problem with all of our Nikon Oil Immersion objectives. Apparently, the new immersion oil provided by Nikon seems to be "aggressive" if not perfectly cleaned after use. Even our Nikon contact in France agreed with the fact that is oil may be very good optically but can be a little hazardous when used on a multi users device. Apparently they don't have any surrogate...

We tried to clean the best we could with water and then very little amount of acetone, but it's risky. And it does not work when there are too much spots.

If other people noticed the same problem with Nikon oil, it should be interesting to share on this topic.

best

Steven Nedellec

MicroPICell facility manager

Center Of Excellence Nikon

Nantes University, france

________________________________

This e-mail message (including any attachments) is for the sole use of
the intended recipient(s) and may contain confidential and privileged
information. If the reader of this message is not the intended
recipient, you are hereby notified that any dissemination, distribution
or copying of this message (including any attachments) is strictly
prohibited.

If you have received this message in error, please contact
the sender by reply e-mail message and destroy all copies of the
original message (including attachments).