To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, Did the various searches, but nothing seems particularly versatile. What we have seen is a fixed hole in the center of the top of a chamber, with little to no play, even with the sealing gasket. Does anybody have a favorite/s they could suggest? Thanks, as always. -- Best, Gary Laevsky, Ph.D. Confocal Imaging Facility Manager Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 |
|
Graham Wright-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Gary, We had a look into this last year (although haven't yet committed) and had the Chamlide UM system suggested to us (no commercial interest): http://www.chamlide.com/product/product_view.asp Regards, Graham -- Dr Graham Wright Head, Microscopy Unit Institute of Medical Biology, A*STAR 8A Biomedical Grove, #06-06 Immunos, Singapore 138648 O: +65 6407 0167 E: [hidden email] W: http://www.a-star.edu.sg/imb/Tech-Platforms/IMB-Microscopy-Unit.aspx On 10 April 2015 at 03:51, Gary Laevsky <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > Did the various searches, but nothing seems particularly versatile. What > we have seen is a fixed hole in the center of the top of a chamber, with > little to no play, even with the sealing gasket. > > Does anybody have a favorite/s they could suggest? > > Thanks, as always. > > -- > Best, > > Gary Laevsky, Ph.D. > Confocal Imaging Facility Manager > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 > |
|
Smith, Benjamin E. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have used a Peltier stage from PhysiTemp with excellent success (TS-4SMP): http://www.physitemp.com/products/HeatingandCooling4Scopes/ Since we do everything from 8 hour live cell imaging at 37oC to Drosophila pupae live imaging for 48 hours at 25oC to live lamprey embryo imaging at 18oC, having a wide range of temperatures available is a must (this stage has a practical range of 0oC to 60oC, although in dry conditions it gets down to about -12oC). Also, because it is a Peltier device, it ramps to the proper temperature and stabilizes in about 30s (think PCR machines). We use tap water to remove waste heat, but they also have a water recirculator if you don't have access to a tap. Also, the dish holder can be removed to mount slides, which we have used to keep live tissue sections at 4oC to prevent necrosis while imaging. Originally we got the stage since we needed both heating and cooling, but we have found the temperature to be remarkably stable when measured at the sample (+/- 0.05oC), and we even use it for live imaging at room temp, as it will offset temperature fluctuations within the room, giving better consistency in our data. For more advanced needs, the PID controller can also be reprogrammed. Hope this helps, Ben Smith Benjamin E. Smith, Ph.D. Samuel Roberts Noble Microscopy Laboratory Research Scientist, Confocal Facility Manager University of Oklahoma Norman, OK 73019 E-mail: [hidden email] Voice 405-325-4391 FAX 405-325-7619 http://www.microscopy.ou.edu/ ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Graham Wright [[hidden email]] Sent: Tuesday, April 14, 2015 2:31 AM To: [hidden email] Subject: Re: Upright incubation solutions? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Gary, We had a look into this last year (although haven't yet committed) and had the Chamlide UM system suggested to us (no commercial interest): http://www.chamlide.com/product/product_view.asp Regards, Graham -- Dr Graham Wright Head, Microscopy Unit Institute of Medical Biology, A*STAR 8A Biomedical Grove, #06-06 Immunos, Singapore 138648 O: +65 6407 0167 E: [hidden email] W: http://www.a-star.edu.sg/imb/Tech-Platforms/IMB-Microscopy-Unit.aspx On 10 April 2015 at 03:51, Gary Laevsky <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > Did the various searches, but nothing seems particularly versatile. What > we have seen is a fixed hole in the center of the top of a chamber, with > little to no play, even with the sealing gasket. > > Does anybody have a favorite/s they could suggest? > > Thanks, as always. > > -- > Best, > > Gary Laevsky, Ph.D. > Confocal Imaging Facility Manager > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 > |
| Free forum by Nabble | Edit this page |