We are setting up a new lab section and we need to select a vacuum pump to hook up to our vacuum plumbing.  We have a large manifold system that ties multiple vacuum taps together into one central location.  We plan on using a single pump to provide for the whole lab.  My concern is in vacuum stability as users open and close the tap, as well as capacity.  Should I be looking for a pump with a strong vacuum draw (low Torr number), or large displacement (liters/minute)? 
				Thanks, Craig  | 
			
| 
					
		 | 
				
					
	
	 
		Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
 
				Craig, 
             The first thing you should be looking at is a big buffer 
tank to even out the fluctuations.  Having a lower Torr than you  
need will do very little for you - the difference in the total mass 
of air in a chamber at 0.1T vs 1T is tiny.  The displacement is 
the key thing - it must be able to handle everyone wanting it at 
once. 
                                                                       Guy 
Optical Imaging Techniques in Cell Biology 
by Guy Cox CRC Press / Taylor & Francis      http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net From: Confocal Microscopy List on behalf of Craig Brideau Sent: Tue 08/03/04 8:38 AM To: [hidden email] Subject: Vacuum pump advice? Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal We are setting up a new lab section and we need to select a vacuum pump to hook up to our vacuum plumbing.  We have a large manifold system that ties multiple vacuum taps together into one central location.  We plan on using a single pump to provide for the whole lab.  My concern is in vacuum stability as users open and close the tap, as well as capacity.  Should I be looking for a pump with a strong vacuum draw (low Torr number), or large displacement (liters/minute)? 
	
	Thanks, Craig  | 
			
| 
					
		 | 
				
					
	
	 
		Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
What can I use for the buffer tank?  Is there a particular style, or can I just stick any old pressure tank on the line and draw a vacuum reservoir on it? 
				Thanks, Craig On Tue, Mar 4, 2008 at 1:36 AM, Guy Cox <[hidden email]> wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  | 
			
| 
					
		 | 
				
					
	
	 
		Search the CONFOCAL archive at
 
				http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi! Use something intended for vacuum! Glass or metal will probably do, but what kind of chemicals might be drawn into the flask? Water can freeze at low pressure and collect. Make sure that the pump oil can be changed safely! Regards, Sophie ____________________________________________________ Sophie M. K. Brunet, Ph. D. Research Officer Optical Spectroscopy, Laser Systems and Applications Chemistry 112 sessional lecturer [hidden email] 306-966-1719 (office) 306-966-1702 (fax) ____________________________________________________ Saskatchewan Structural Sciences Centre University of Saskatchewan Thorvaldson Bldg. 110 Science Place Saskatoon, Sk S7N 5C9 ____________________________________________________ Quoting Craig Brideau <[hidden email]>: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > What can I use for the buffer tank? Is there a particular style, or can I > just stick any old pressure tank on the line and draw a vacuum reservoir on > it? > > Thanks, > > Craig > > > On Tue, Mar 4, 2008 at 1:36 AM, Guy Cox <[hidden email]> wrote: > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Craig, > > > > The first thing you should be looking at is a big buffer > > tank to even out the fluctuations. Having a lower Torr than you > > need will do very little for you - the difference in the total mass > > of air in a chamber at 0.1T vs 1T is tiny. The displacement is > > the key thing - it must be able to handle everyone wanting it at > > once. > > > > Guy > > > > Optical Imaging Techniques in Cell Biology > > by Guy Cox CRC Press / Taylor & Francis > > http://www.guycox.com/optical.htm > > ______________________________________________ > > Associate Professor Guy Cox, MA, DPhil(Oxon) > > Electron Microscope Unit, Madsen Building F09, > > University of Sydney, NSW 2006 > > ______________________________________________ > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > > Mobile 0413 281 861 > > ______________________________________________ > > > > > > > ------------------------------ > > *From:* Confocal Microscopy List on behalf of Craig Brideau > > *Sent:* Tue 08/03/04 8:38 AM > > *To:* [hidden email] > > *Subject:* Vacuum pump advice? > > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal We are setting up > > a new lab section and we need to select a vacuum pump to hook up to our > > vacuum plumbing. We have a large manifold system that ties multiple vacuum > > taps together into one central location. We plan on using a single pump to > > provide for the whole lab. My concern is in vacuum stability as users open > > and close the tap, as well as capacity. Should I be looking for a pump > with > > a strong vacuum draw (low Torr number), or large displacement > > (liters/minute)? > > > > Thanks, > > > > Craig > > >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Craig Brideau
			 
	
		Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
 
				Well, it only has to stand one atmosphere so 
the specification isn't onerous.  There  
are lots of options.  If you know anyone who is 
scrapping an old Philips EM400 electron  
microscope (now 30 years old) it has a particularly big 
and high-quality buffer tank.  I don't 
know the scale of your system so I can't really 
recommend a capacity but if you can 
get 20 litres (5 US gallons) I'd have thought it would 
do. 
                                                                                                       
Guy Optical Imaging Techniques in Cell Biology From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Thursday, 6 March 2008 9:30 AM To: [hidden email] Subject: Re: Vacuum pump advice? Thanks, Craig On Tue, Mar 4, 2008 at 1:36 AM, Guy Cox <[hidden email]> wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal No virus found in this incoming message. No virus found in this outgoing message.  | 
			