Shalin Mehta Shalin Mehta
Reply | Threaded
Open this post in threaded view
♦
♦
|

What methods are used to image morphology in high content screening systems?

Dear listers,
DIC and phase-contrast are used widely (albeit, less than fluorescence) for imaging cellular morphology . I am curious about what approaches are used to image morphology in the high content screening applications.
I have come across HCS systems on the web that use either DIC or phase-contrast.
Do some systems use hoffman modulation contrast? 

If you could list the system that you have and the type of contrast they employ for imaging the morphology, that will be helpful. Since high-content screening requires automated analysis, and these methods are not really quantitative, do people find it difficult to address some biological questions? My thesis is related to quantitative morphological imaging and I am trying to get the users' perspective on the topic.

Thanks very much for your inputs, 
Shalin
blog: shalin.wordpress.com

Bioimaging Lab, Block-E3A, #7-10                
Div of Bioengineering, NUS Singapore 117574  
website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What methods are used to image morphology in high content screening systems?

Hi Shalin,

BD Pathway 855 HCS ... brightfield


Phase contrast (halo), DIC (edges along the shear axis, nothing along the perpendicular axis), Hoffman (similar problem to DIC), are all problematic for clean morphometry. Darkfield is problematic is there is any junk in the lightpath that scatters light (see CytoViva product for interesting NA>1.2 condenser product). A 12-bit CCD camera has plenty of information in it for segmenting brightfield images.

I believe GE Healthcare licensed IATIA's quantitative phase map QPm algorithm for (some of?) their InCell HCS instruments ( http://www.iatia.com.au/products/lifeSciences/prodappGe.asp ). Whether anyone is using QPm - ask customers. QPm and similar algorithms (Michael Feld lab, see Holmes chapter in Pawkey, Kam/Agard/Sedat) compute the refractive index map in the image. Graham Dunn published something similar with DRIMAS (PMID: 2592444). These methods are quantitative.

By the way, IAITA ( http://www.iatia.com.au/technology/insideQpi.asp  ) emphasizes on their web site and literature the conversion of the QPm data to phase contrast, DIC or darkfield like images. This seems bizarre to me because the real value is in the quantitative map of R.I.  At least two papers computing QPm maps from the (non-confocal) transmitted light path of confocal microscopes have been published:

  • A simple method allowing DIC imaging in conjunction with confocal microscopy. S.H. Cody, S.D. Xiang, M.J. Layton, E. Handman, M.H.C. Lam, J.E. Layton, E.C. Nice, and J.K. Heath, Journal of Microscopy 217, 265-274 (2005).
  • Refractive Index Measurement in Viable Cells Using Quantitative Phase-Amplitude Microscopy and Confocal Microscopy. Claire L. Curl, , Catherine J. Bellair, Trudi Harris, Brendan E. Allman, Peter J. Harris, Alastair G. Stewart , Ann Roberts, Keith A. Nugent, and Lea M.D. Delbridge, Cytometry Part A, 65A, 88-92 (2005).
You can make arrangements to have test images processed by IATIA - send them an email.


Plan B; soluble GFP (if anyone can find a promoter that results in uniform expression in #GFP/um^3 in every cell in a population - good luck ... could be any color FP of course) or much simpler, a fluorescent dye that forms a covalent bond to most cell proteins, see Coates et al 1992, PMID: 1577856, for example (the Texas Red conjugate, not the phalloidin conjugate, in that paper).

Best wishes,

George
p.s. speaking of CytoViva - has anyone used a laser pointer as illuminator with a CytoViva?




At 11:40 AM 7/16/2010, you wrote:
Dear listers,
DIC and phase-contrast are used widely (albeit, less than fluorescence) for imaging cellular morphology . I am curious about what approaches are used to image morphology in the high content screening applications.
I have come across HCS systems on the web that use either DIC or phase-contrast.
Do some systems use hoffman modulation contrast? 

If you could list the system that you have and the type of contrast they employ for imaging the morphology, that will be helpful. Since high-content screening requires automated analysis, and these methods are not really quantitative, do people find it difficult to address some biological questions? My thesis is related to quantitative morphological imaging and I am trying to get the users' perspective on the topic.

Thanks very much for your inputs,
Shalin
blog: shalin.wordpress.com

Bioimaging Lab, Block-E3A, #7-10                
Div of Bioengineering, NUS Singapore 117574  
website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm




George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What methods are used to image morphology in high content screening systems?

In reply to this post by Shalin Mehta
Hi Shalin,

Two more items...

1. Interference reflection microscopy (IRM) and reflection interference contrast microscopy (RICM) can enable precise imaging of both cell-substratum adhesive sites (black = focal adhesions) as well as the outlines of cells (including ultrathin lamellipodia and filopodia). Easy to implement on a confocal microscope (no emission filter, though circularly polarized light or antiflex objective lens may help improve image quality). Same optical configuration provides the cell outline as you focus into the cell (provided R.I. or media is different than the plasma membrane and cell!. See PMID: 19816893, 19821476, 3900106, and especially 20013754 (if you do not have access to C.P.C., email the author for the article) for quantitative details.

2. efficient use of cells ... see http://www.nature.com/nmeth/journal/v7/n7/abs/nmeth.1462.html or at least the company's press release at http://www.cytoo.com/cytoo_newsroom.php  and be sure to crank up your audio and go to http://www.cellebrate.fr/

Enjoy,

George


 


At 11:40 AM 7/16/2010, you wrote:
Dear listers,
DIC and phase-contrast are used widely (albeit, less than fluorescence) for imaging cellular morphology . I am curious about what approaches are used to image morphology in the high content screening applications.
I have come across HCS systems on the web that use either DIC or phase-contrast.
Do some systems use hoffman modulation contrast? 

If you could list the system that you have and the type of contrast they employ for imaging the morphology, that will be helpful. Since high-content screening requires automated analysis, and these methods are not really quantitative, do people find it difficult to address some biological questions? My thesis is related to quantitative morphological imaging and I am trying to get the users' perspective on the topic.

Thanks very much for your inputs,
Shalin
blog: shalin.wordpress.com

Bioimaging Lab, Block-E3A, #7-10                
Div of Bioengineering, NUS Singapore 117574  
website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm



Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What methods are used to image morphology in high content screening systems?

In reply to this post by George McNamara

I was involved as an early evaluation site for the Iatia system (based on Keith Nugent’s algorithms) so I feel I should comment on this post.

 

The idea behind the system is something that every microscopist  is familiar with, that contrast is lowest at true focus.  A bit out of focus and phase effects enhance the contrast.   This is measurable, so by taking a through-focus series one can recover the phase at the image plane.  The primary output of the software  is therefore a phase map of the whole image.  If that is what one wants of course one needs to go no further.  But once one has this information it is easy to synthesize other imaging modalities such as phase contrast and DIC.  The really useful aspect of this is that one can then control how much of the amplitude image to feed in to the final output.  Optical DIC and phase contrast systems do not give you that control.

 

So why didn’t this system take the world by storm?  First off, most of us have phase and DIC anyway, and in fact it can be very hard to buy high-end lenses without phase plates.   So unless we need that full phase map it isn’t going to give us much for the purchase price.  Secondly, and this is the real killer, most people use phase and DIC to look at living cells and we can’t get them to sit still for a through-focus series, so it just doesn’t work. 

 

The bottom line?  A brilliant idea, and one which works, but still seeking the right niche in the market.

 

                                                                                 Guy

 

Optical Imaging Techniques in Cell Biology

by Guy Cox    CRC Press / Taylor & Francis

     http://www.guycox.com/optical.htm

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara
Sent: Sunday, 18 July 2010 11:39 PM
To: [hidden email]
Subject: Re: What methods are used to image morphology in high content screening systems?

 

Hi Shalin,

BD Pathway 855 HCS ... brightfield


Phase contrast (halo), DIC (edges along the shear axis, nothing along the perpendicular axis), Hoffman (similar problem to DIC), are all problematic for clean morphometry. Darkfield is problematic is there is any junk in the lightpath that scatters light (see CytoViva product for interesting NA>1.2 condenser product). A 12-bit CCD camera has plenty of information in it for segmenting brightfield images.

I believe GE Healthcare licensed IATIA's quantitative phase map QPm algorithm for (some of?) their InCell HCS instruments ( http://www.iatia.com.au/products/lifeSciences/prodappGe.asp ). Whether anyone is using QPm - ask customers. QPm and similar algorithms (Michael Feld lab, see Holmes chapter in Pawkey, Kam/Agard/Sedat) compute the refractive index map in the image. Graham Dunn published something similar with DRIMAS (PMID: 2592444). These methods are quantitative.

By the way, IAITA ( http://www.iatia.com.au/technology/insideQpi.asp  ) emphasizes on their web site and literature the conversion of the QPm data to phase contrast, DIC or darkfield like images. This seems bizarre to me because the real value is in the quantitative map of R.I.  At least two papers computing QPm maps from the (non-confocal) transmitted light path of confocal microscopes have been published:

  • A simple method allowing DIC imaging in conjunction with confocal microscopy. S.H. Cody, S.D. Xiang, M.J. Layton, E. Handman, M.H.C. Lam, J.E. Layton, E.C. Nice, and J.K. Heath, Journal of Microscopy 217, 265-274 (2005).
  • Refractive Index Measurement in Viable Cells Using Quantitative Phase-Amplitude Microscopy and Confocal Microscopy. Claire L. Curl, , Catherine J. Bellair, Trudi Harris, Brendan E. Allman, Peter J. Harris, Alastair G. Stewart , Ann Roberts, Keith A. Nugent, and Lea M.D. Delbridge, Cytometry Part A, 65A, 88-92 (2005).

You can make arrangements to have test images processed by IATIA - send them an email.


Plan B; soluble GFP (if anyone can find a promoter that results in uniform expression in #GFP/um^3 in every cell in a population - good luck ... could be any color FP of course) or much simpler, a fluorescent dye that forms a covalent bond to most cell proteins, see Coates et al 1992, PMID: 1577856, for example (the Texas Red conjugate, not the phalloidin conjugate, in that paper).

Best wishes,

George
p.s. speaking of CytoViva - has anyone used a laser pointer as illuminator with a CytoViva?




At 11:40 AM 7/16/2010, you wrote:

Dear listers,
DIC and phase-contrast are used widely (albeit, less than fluorescence) for imaging cellular morphology . I am curious about what approaches are used to image morphology in the high content screening applications.
I have come across HCS systems on the web that use either DIC or phase-contrast.
Do some systems use hoffman modulation contrast? 

If you could list the system that you have and the type of contrast they employ for imaging the morphology, that will be helpful. Since high-content screening requires automated analysis, and these methods are not really quantitative, do people find it difficult to address some biological questions? My thesis is related to quantitative morphological imaging and I am trying to get the users' perspective on the topic.

Thanks very much for your inputs,
Shalin
blog: shalin.wordpress.com

Bioimaging Lab, Block-E3A, #7-10                
Div of Bioengineering, NUS Singapore 117574  
website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm



No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.830 / Virus Database: 271.1.1/2914 - Release Date: 07/18/10 04:35:00

Pascal Lorentz-2 Pascal Lorentz-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Magnetic fields and microscopy

Dear list,

this is not exactly a confocal question but anyway related to it. I hope
to get some input from people who have some experience with magnetic
fields and the influence on microscopy but also on confocal scanners.
In our building we are planning to centralize our microscopy facility in
the second basement. Right now the architects are doing a feasibility
study.
Because the new rooms are located between the elevator machine room and
the high voltage current room and since the electric cables from the
whole house go through these rooms they now ask about magnetic fields.
Does somebody know if magnetic fields can influence microscopy (cells
for example), the microscopes itself, the lasers or scanners (galvos)?
If yes, is there a threshold that is acceptable?
I would be very thankful if I could get more information on that topic.

Best regards

Pascal

--
Pascal Lorentz
BioOptics Facility
Department of Biomedicine
University of Basel
Mattenstrasse 28
4058 Basel
Switzerland
Johannes Helm Johannes Helm
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Magnetic fields and microscopy

Good afternoon,

whether the electromagnetic fields will have any non-negligible influence
on the biological specimens certainly is a difficult problem.
Whether a reviewer refereeing a manuscript for publication would think
that the effects of the electromagnetic fields in your lab would cast any
doubts on the reliability of the presented observations and results most
probably is to a large degree dependent on the reviewer's "personal
opinion". There has been and still is a lot of writing about electro smog
caused by AC HV lines used to transport electric energy, by AC HV lines
for train traffic and, even more important, the effect of the electro
magnetic fields to which one is exposed when using cell phones.

Concerning the microscopes, lasers, detectors a.s.o.:

I cannot see any influence that an electromagnetic field of the strength
of one generated by the sources you mention would have on the microscope
itself.
The magnetic field necessary to alter the plane of polarization of
polarized light in a non negligble way, see "Faraday Effect", certainly is
much stronger than the fields you are going to experience. There should,
hence, not be any noticeable effect even of you plan to do quantitative
polarization microscopy.
Light detectors using HV, e.g. photomultiplier tubes, are usually mounted
in tubes made out of a material, which shields magnetic fields.

Best wishes for your work with and in the new lab.

Johannes




> Dear list,
>
> this is not exactly a confocal question but anyway related to it. I hope
> to get some input from people who have some experience with magnetic
> fields and the influence on microscopy but also on confocal scanners.
> In our building we are planning to centralize our microscopy facility in
> the second basement. Right now the architects are doing a feasibility
> study.
> Because the new rooms are located between the elevator machine room and
> the high voltage current room and since the electric cables from the
> whole house go through these rooms they now ask about magnetic fields.
> Does somebody know if magnetic fields can influence microscopy (cells
> for example), the microscopes itself, the lasers or scanners (galvos)?
> If yes, is there a threshold that is acceptable?
> I would be very thankful if I could get more information on that topic.
>
> Best regards
>
> Pascal
>
> --
> Pascal Lorentz
> BioOptics Facility
> Department of Biomedicine
> University of Basel
> Mattenstrasse 28
> 4058 Basel
> Switzerland
>


--
P. Johannes Helm

Voice: (+47) 228 51159 (office)
Fax: (+47) 228 51499 (office)
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Magnetic fields and microscopy

The electrical noise will probably not be good for your readings.
Unless you are willing to put the whole room in a Faraday cage I would
be concerned.  You will also probably want line filters as well; this
will filter the power coming out of your wall sockets, so any EM they
pick up from the nearby high voltage stuff will not make it into your
equipment.  Overall I'd say your planned location is less than ideal,
but with sufficient measures you could still make it work.  Good line
filters and shielding should keep out the nearby sources of electrical
interference.

Craig


On Tue, Jul 20, 2010 at 8:08 AM, Johannes Helm <[hidden email]> wrote:

> Good afternoon,
>
> whether the electromagnetic fields will have any non-negligible influence
> on the biological specimens certainly is a difficult problem.
> Whether a reviewer refereeing a manuscript for publication would think
> that the effects of the electromagnetic fields in your lab would cast any
> doubts on the reliability of the presented observations and results most
> probably is to a large degree dependent on the reviewer's "personal
> opinion". There has been and still is a lot of writing about electro smog
> caused by AC HV lines used to transport electric energy, by AC HV lines
> for train traffic and, even more important, the effect of the electro
> magnetic fields to which one is exposed when using cell phones.
>
> Concerning the microscopes, lasers, detectors a.s.o.:
>
> I cannot see any influence that an electromagnetic field of the strength
> of one generated by the sources you mention would have on the microscope
> itself.
> The magnetic field necessary to alter the plane of polarization of
> polarized light in a non negligble way, see "Faraday Effect", certainly is
> much stronger than the fields you are going to experience. There should,
> hence, not be any noticeable effect even of you plan to do quantitative
> polarization microscopy.
> Light detectors using HV, e.g. photomultiplier tubes, are usually mounted
> in tubes made out of a material, which shields magnetic fields.
>
> Best wishes for your work with and in the new lab.
>
> Johannes
>
>
>
>
>> Dear list,
>>
>> this is not exactly a confocal question but anyway related to it. I hope
>> to get some input from people who have some experience with magnetic
>> fields and the influence on microscopy but also on confocal scanners.
>> In our building we are planning to centralize our microscopy facility in
>> the second basement. Right now the architects are doing a feasibility
>> study.
>> Because the new rooms are located between the elevator machine room and
>> the high voltage current room and since the electric cables from the
>> whole house go through these rooms they now ask about magnetic fields.
>> Does somebody know if magnetic fields can influence microscopy (cells
>> for example), the microscopes itself, the lasers or scanners (galvos)?
>> If yes, is there a threshold that is acceptable?
>> I would be very thankful if I could get more information on that topic.
>>
>> Best regards
>>
>> Pascal
>>
>> --
>> Pascal Lorentz
>> BioOptics Facility
>> Department of Biomedicine
>> University of Basel
>> Mattenstrasse 28
>> 4058 Basel
>> Switzerland
>>
>
>
> --
> P. Johannes Helm
>
> Voice:  (+47) 228 51159 (office)
> Fax:    (+47) 228 51499 (office)
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Magnetic fields and microscopy

To clarify my comment; any wires in your system carrying signal will
couple like an antenna to the nearby sources of EM.  It's not so much
that the EM itself will influence your samples or light, its just that
you can expect to see a bunch of 60 Hz or other frequency electrical
noise on every wire/cable/connection in the lab if you don't have good
shielding.  A fun test is to get a long BNC cable, or better yet an
unshielded wire, plug it into an oscilloscope, and then see what you
pick up in the room as you move the cable around.  The frequencies and
magnitudes on the oscilloscope will give you some idea of how bad the
EM noise signals in the room are.

Craig


On Tue, Jul 20, 2010 at 4:47 PM, Craig Brideau <[hidden email]> wrote:

> The electrical noise will probably not be good for your readings.
> Unless you are willing to put the whole room in a Faraday cage I would
> be concerned.  You will also probably want line filters as well; this
> will filter the power coming out of your wall sockets, so any EM they
> pick up from the nearby high voltage stuff will not make it into your
> equipment.  Overall I'd say your planned location is less than ideal,
> but with sufficient measures you could still make it work.  Good line
> filters and shielding should keep out the nearby sources of electrical
> interference.
>
> Craig
>
>
> On Tue, Jul 20, 2010 at 8:08 AM, Johannes Helm <[hidden email]> wrote:
>> Good afternoon,
>>
>> whether the electromagnetic fields will have any non-negligible influence
>> on the biological specimens certainly is a difficult problem.
>> Whether a reviewer refereeing a manuscript for publication would think
>> that the effects of the electromagnetic fields in your lab would cast any
>> doubts on the reliability of the presented observations and results most
>> probably is to a large degree dependent on the reviewer's "personal
>> opinion". There has been and still is a lot of writing about electro smog
>> caused by AC HV lines used to transport electric energy, by AC HV lines
>> for train traffic and, even more important, the effect of the electro
>> magnetic fields to which one is exposed when using cell phones.
>>
>> Concerning the microscopes, lasers, detectors a.s.o.:
>>
>> I cannot see any influence that an electromagnetic field of the strength
>> of one generated by the sources you mention would have on the microscope
>> itself.
>> The magnetic field necessary to alter the plane of polarization of
>> polarized light in a non negligble way, see "Faraday Effect", certainly is
>> much stronger than the fields you are going to experience. There should,
>> hence, not be any noticeable effect even of you plan to do quantitative
>> polarization microscopy.
>> Light detectors using HV, e.g. photomultiplier tubes, are usually mounted
>> in tubes made out of a material, which shields magnetic fields.
>>
>> Best wishes for your work with and in the new lab.
>>
>> Johannes
>>
>>
>>
>>
>>> Dear list,
>>>
>>> this is not exactly a confocal question but anyway related to it. I hope
>>> to get some input from people who have some experience with magnetic
>>> fields and the influence on microscopy but also on confocal scanners.
>>> In our building we are planning to centralize our microscopy facility in
>>> the second basement. Right now the architects are doing a feasibility
>>> study.
>>> Because the new rooms are located between the elevator machine room and
>>> the high voltage current room and since the electric cables from the
>>> whole house go through these rooms they now ask about magnetic fields.
>>> Does somebody know if magnetic fields can influence microscopy (cells
>>> for example), the microscopes itself, the lasers or scanners (galvos)?
>>> If yes, is there a threshold that is acceptable?
>>> I would be very thankful if I could get more information on that topic.
>>>
>>> Best regards
>>>
>>> Pascal
>>>
>>> --
>>> Pascal Lorentz
>>> BioOptics Facility
>>> Department of Biomedicine
>>> University of Basel
>>> Mattenstrasse 28
>>> 4058 Basel
>>> Switzerland
>>>
>>
>>
>> --
>> P. Johannes Helm
>>
>> Voice:  (+47) 228 51159 (office)
>> Fax:    (+47) 228 51499 (office)
>>
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Magnetic fields and microscopy

PMTs need a trans-impedance amplifier to convert the current spikes
caused by the photon impacts into something instrumentation can read.
They can be very noise sensitive, especially if your ground lines
start picking up surrounding signals.

Craig



On Tue, Jul 20, 2010 at 5:37 PM, Donnelly, Tom <[hidden email]> wrote:

>  Don't know how this effects PMT's and galvo's but an CCD or EMCCD are great detectors of EM noise.
>
> TD
>
> -----Original Message-----
> From: Craig Brideau [mailto:[hidden email]]
> Sent: Tuesday, July 20, 2010 3:51 PM
> To: [hidden email]
> Subject: Re: Magnetic fields and microscopy
>
> To clarify my comment; any wires in your system carrying signal will couple like an antenna to the nearby sources of EM.  It's not so much that the EM itself will influence your samples or light, its just that you can expect to see a bunch of 60 Hz or other frequency electrical noise on every wire/cable/connection in the lab if you don't have good shielding.  A fun test is to get a long BNC cable, or better yet an unshielded wire, plug it into an oscilloscope, and then see what you pick up in the room as you move the cable around.  The frequencies and magnitudes on the oscilloscope will give you some idea of how bad the EM noise signals in the room are.
>
> Craig
>
>
> On Tue, Jul 20, 2010 at 4:47 PM, Craig Brideau <[hidden email]> wrote:
>> The electrical noise will probably not be good for your readings.
>> Unless you are willing to put the whole room in a Faraday cage I would
>> be concerned.  You will also probably want line filters as well; this
>> will filter the power coming out of your wall sockets, so any EM they
>> pick up from the nearby high voltage stuff will not make it into your
>> equipment.  Overall I'd say your planned location is less than ideal,
>> but with sufficient measures you could still make it work.  Good line
>> filters and shielding should keep out the nearby sources of electrical
>> interference.
>>
>> Craig
>>
>>
>> On Tue, Jul 20, 2010 at 8:08 AM, Johannes Helm <[hidden email]> wrote:
>>> Good afternoon,
>>>
>>> whether the electromagnetic fields will have any non-negligible
>>> influence on the biological specimens certainly is a difficult problem.
>>> Whether a reviewer refereeing a manuscript for publication would
>>> think that the effects of the electromagnetic fields in your lab
>>> would cast any doubts on the reliability of the presented
>>> observations and results most probably is to a large degree dependent
>>> on the reviewer's "personal opinion". There has been and still is a
>>> lot of writing about electro smog caused by AC HV lines used to
>>> transport electric energy, by AC HV lines for train traffic and, even
>>> more important, the effect of the electro magnetic fields to which one is exposed when using cell phones.
>>>
>>> Concerning the microscopes, lasers, detectors a.s.o.:
>>>
>>> I cannot see any influence that an electromagnetic field of the
>>> strength of one generated by the sources you mention would have on
>>> the microscope itself.
>>> The magnetic field necessary to alter the plane of polarization of
>>> polarized light in a non negligble way, see "Faraday Effect",
>>> certainly is much stronger than the fields you are going to
>>> experience. There should, hence, not be any noticeable effect even of
>>> you plan to do quantitative polarization microscopy.
>>> Light detectors using HV, e.g. photomultiplier tubes, are usually
>>> mounted in tubes made out of a material, which shields magnetic fields.
>>>
>>> Best wishes for your work with and in the new lab.
>>>
>>> Johannes
>>>
>>>
>>>
>>>
>>>> Dear list,
>>>>
>>>> this is not exactly a confocal question but anyway related to it. I
>>>> hope to get some input from people who have some experience with
>>>> magnetic fields and the influence on microscopy but also on confocal scanners.
>>>> In our building we are planning to centralize our microscopy
>>>> facility in the second basement. Right now the architects are doing
>>>> a feasibility study.
>>>> Because the new rooms are located between the elevator machine room
>>>> and the high voltage current room and since the electric cables from
>>>> the whole house go through these rooms they now ask about magnetic fields.
>>>> Does somebody know if magnetic fields can influence microscopy
>>>> (cells for example), the microscopes itself, the lasers or scanners (galvos)?
>>>> If yes, is there a threshold that is acceptable?
>>>> I would be very thankful if I could get more information on that topic.
>>>>
>>>> Best regards
>>>>
>>>> Pascal
>>>>
>>>> --
>>>> Pascal Lorentz
>>>> BioOptics Facility
>>>> Department of Biomedicine
>>>> University of Basel
>>>> Mattenstrasse 28
>>>> 4058 Basel
>>>> Switzerland
>>>>
>>>
>>>
>>> --
>>> P. Johannes Helm
>>>
>>> Voice:  (+47) 228 51159 (office)
>>> Fax:    (+47) 228 51499 (office)
>>>
>>
>
>
>
> This email message, together with any attachments, is for the sole use of the intended recipient(s) and is the confidential information of Applied Precision Inc. If you are not the intended recipient, your review, use, disclosure, copying or dissemination of this email message or its attachments, or the information contained therein, is strictly prohibited. If you are not the intended recipient or if you think this email was sent to you in error, please notify the sender by reply email and delete this message and its attachments, as well as all copies, from your system.
>
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What methods are used to image morphology in high content screening systems?

In reply to this post by Shalin Mehta
Hi again Shalin,

Check out the HD (high definition imaging pdf at http://essenbioscience.com/products.html
I came across this because IncuCyteHD was used by:


Glioma stem cell lines expanded in adherent culture have tumor-specific phenotypes and are suitable for chemical and genetic screens.

Pollard SM, Yoshikawa K, Clarke ID, Danovi D, Stricker S, Russell R, Bayani J, Head R, Lee M, Bernstein M, Squire JA, Smith A, Dirks P.

Cell Stem Cell. 2009 Jun 5;4(6):568-80.PMID: 19497285

The authors are to be congratulated for posting so many movies online at http://www.cell.com/cell-stem-cell/supplemental/S1934-5909(09)00149-0


Enjoy,

George
p.s. listserv - My core facility's web site underwent some changes recently affecting most of  the web links. For anyone looking for the Pubspectra dataset, you can go to http://www.sylvester.org/AICF/pubspectra.zip and click on the PubSpectra zip  link to download (the dataset is the xlsx file, the zip file has additional files that may be useful).




At 11:40 AM 7/16/2010, you wrote:
Dear listers,
DIC and phase-contrast are used widely (albeit, less than fluorescence) for imaging cellular morphology . I am curious about what approaches are used to image morphology in the high content screening applications.
I have come across HCS systems on the web that use either DIC or phase-contrast.
Do some systems use hoffman modulation contrast? 

If you could list the system that you have and the type of contrast they employ for imaging the morphology, that will be helpful. Since high-content screening requires automated analysis, and these methods are not really quantitative, do people find it difficult to address some biological questions? My thesis is related to quantitative morphological imaging and I am trying to get the users' perspective on the topic.

Thanks very much for your inputs,
Shalin
blog: shalin.wordpress.com

Bioimaging Lab, Block-E3A, #7-10                
Div of Bioengineering, NUS Singapore 117574  
website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm







George McNamara, Ph.D.
Image Core Manager
Analytical Imaging Core Facility
University of Miami, Miller School of Medicine
Miami, FL 33136
[hidden email]
[hidden email]
305-243-8436 office
http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx file is in the zip file)
http://home.earthlink.net/~geomcnamara

Shalin Mehta Shalin Mehta
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What methods are used to image morphology in high content screening systems?

Thank you George and Guy for very informative inputs.

I am *guessing* the reason IATIA incorporated computation of DIC and
phase-contrast images from quantitative RI is that our brains tend to
pick up sharp features from the images. The quantitative optical path
length appears 'smooth', which may not be quite appealing to the eye.
DIC and phase contrast highlight edge features, which are easier to
interpret by human.

best
shalin


On Sun, Jul 25, 2010 at 3:09 AM, George McNamara
<[hidden email]> wrote:

> Hi again Shalin,
>
> Check out the HD (high definition imaging pdf at
> http://essenbioscience.com/products.html
> I came across this because IncuCyteHD was used by:
>
>
> Glioma stem cell lines expanded in adherent culture have tumor-specific
> phenotypes and are suitable for chemical and genetic screens.
>
> Pollard SM, Yoshikawa K, Clarke ID, Danovi D, Stricker S, Russell R, Bayani
> J, Head R, Lee M, Bernstein M, Squire JA, Smith A, Dirks P.
>
> Cell Stem Cell. 2009 Jun 5;4(6):568-80.PMID: 19497285
>
> The authors are to be congratulated for posting so many movies online at
> http://www.cell.com/cell-stem-cell/supplemental/S1934-5909(09)00149-0
>
>
> Enjoy,
>
> George
> p.s. listserv - My core facility's web site underwent some changes recently
> affecting most of  the web links. For anyone looking for the Pubspectra
> dataset, you can go to http://www.sylvester.org/AICF/pubspectra.zip and
> click on the PubSpectra zip  link to download (the dataset is the xlsx file,
> the zip file has additional files that may be useful).
>
>
>
>
> At 11:40 AM 7/16/2010, you wrote:
>
> Dear listers,
> DIC and phase-contrast are used widely (albeit, less than fluorescence) for
> imaging cellular morphology . I am curious about what approaches are used to
> image morphology in the high content screening applications.
> I have come across HCS systems on the web that use either DIC or
> phase-contrast.
> Do some systems use hoffman modulation contrast?
>
> If you could list the system that you have and the type of contrast they
> employ for imaging the morphology, that will be helpful. Since high-content
> screening requires automated analysis, and these methods are not really
> quantitative, do people find it difficult to address some biological
> questions? My thesis is related to quantitative morphological imaging and I
> am trying to get the users' perspective on the topic.
>
> Thanks very much for your inputs,
> Shalin
> blog: shalin.wordpress.com
>
> Bioimaging Lab, Block-E3A, #7-10
> Div of Bioengineering, NUS Singapore 117574
> website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm
>
>
>
>
>
> George McNamara, Ph.D.
> Image Core Manager
> Analytical Imaging Core Facility
> University of Miami, Miller School of Medicine
> Miami, FL 33136
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://www.sylvester.org/AICF (Analytical Imaging Core Facility)
> http://www.sylvester.org/AICF/pubspectra.zip (the entire 2000+ spectra .xlsx
> file is in the zip file)
> http://home.earthlink.net/~geomcnamara
>