Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I need some help. We have a 1 year old Leica SP5 inverted scope and pretty much all of a sudden we started to get z-drift. Every time you do a time lapse, no matter what the z-step or time interval, the system does not start at the top of the sample, it moves down significantly and progressively. Anything I can do to fix this? I have called the service rep but would like to fix the problem ASAP. It is possible that I am the first person to use the system for this application since we upgraded the software and firmware. Nathalia -- Nathalia Glickman Holtzman Ph.D. Department of Biology SB D328 Queens College, CUNY 65-30 Kissena Boulevard Flushing, NY 11367 Office: 718-997-3678 Lab: 718-997-3517 Fax: 718-997-3445 |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Nathalia, I'm not sure if it's the same problem, but we have had a similar problem twice on the DM IRBE inverted body of our SP1, it turned out to be immersion oil that had dripped down the objectives and into the mechanism causing the turret to sink slowly - the loss of focus was visible over a short period of time using the eyepieces. Our field service engineer fixed it quickly both times (the first time was approximately a year after we got the scope, the second time was much much later, due mainly to better cleaning and care from the users once they were made aware of the risk). Cheers Ray ----- Original Message ----- From: "Nathalia Holtzman" <[hidden email]> To: <[hidden email]> Sent: Thursday, November 08, 2007 1:58 PM Subject: Z-drift > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi all, > I need some help. We have a 1 year old Leica SP5 inverted scope and pretty > much all of a sudden we started to get z-drift. Every time you do a time > lapse, no matter what the z-step or time interval, the system does not > start > at the top of the sample, it moves down significantly and progressively. > Anything I can do to fix this? > > I have called the service rep but would like to fix the problem ASAP. It > is > possible that I am the first person to use the system for this application > since we upgraded the software and firmware. > > Nathalia > > -- > Nathalia Glickman Holtzman Ph.D. > Department of Biology SB D328 > Queens College, CUNY > 65-30 Kissena Boulevard > Flushing, NY 11367 > > Office: 718-997-3678 > Lab: 718-997-3517 > Fax: 718-997-3445 |
In reply to this post by Nathalia Holtzman
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Nathalia, Not knowing anything about your setup or what you are looking at, my first guess would be temperature fluctuations. Try and let the microscope and stage warm up for a couple of hours, enclose as much as possible in some sort of chamber (a large bag may work in a pinch). One check- during z-drift, does the z-encoder indicate that the z-position is fixed? Another check- try setting upper and lower limits for a z-stack and (rapidly) move between those tow setpoints. Is there hysteresis present? Those can tell you if there is a hardware problem. Andy At 08:58 AM 11/8/2007, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi all, >I need some help. We have a 1 year old Leica SP5 inverted scope and pretty >much all of a sudden we started to get z-drift. Every time you do a time >lapse, no matter what the z-step or time interval, the system does not start >at the top of the sample, it moves down significantly and progressively. >Anything I can do to fix this? > >I have called the service rep but would like to fix the problem ASAP. It is >possible that I am the first person to use the system for this application >since we upgraded the software and firmware. > >Nathalia > >-- >Nathalia Glickman Holtzman Ph.D. >Department of Biology SB D328 >Queens College, CUNY >65-30 Kissena Boulevard >Flushing, NY 11367 > >Office: 718-997-3678 >Lab: 718-997-3517 >Fax: 718-997-3445 Andrew Resnick, Ph. D. Instructor Department of Physiology and Biophysics Case Western Reserve University 216-368-6899 (V) 216-368-4223 (F) |
In reply to this post by Nathalia Holtzman
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Nathalia, We have been experiencing the same thing on our SP2 AOBS system. This not a simple "drift" problem but a large sudden drop downward which only happens when attempting to use the time lapse setup. Like you we are seeing this after a software and firmware upgrade. We are still trying to identify the problem but suspect it is the software/firmware itself as it only happens using the time-lapse setup. We'll let you know what we discover, we have a very good Leica service engineer. Please let us know if you learn anything. Thanks Jon At 05:58 AM 11/8/2007, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi all, >I need some help. We have a 1 year old Leica SP5 inverted scope and pretty >much all of a sudden we started to get z-drift. Every time you do a time >lapse, no matter what the z-step or time interval, the system does not start >at the top of the sample, it moves down significantly and progressively. >Anything I can do to fix this? > >I have called the service rep but would like to fix the problem ASAP. It is >possible that I am the first person to use the system for this application >since we upgraded the software and firmware. > >Nathalia > >-- >Nathalia Glickman Holtzman Ph.D. >Department of Biology SB D328 >Queens College, CUNY >65-30 Kissena Boulevard >Flushing, NY 11367 > >Office: 718-997-3678 >Lab: 718-997-3517 >Fax: 718-997-3445 Jon Mulholland, Director Cell Sciences Imaging Facility Beckman Center, B050 Stanford University School of Medicine Stanford, CA 94305-5301 voice 650-725-7532 fax 650-725-4951 http://taltos.stanford.edu |
In reply to this post by Nathalia Holtzman
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Nathalia, We had a problem like this on our SP2 some years ago. This may not be that relevant because we have an upright scope with a galvo stage but it turned out that the galvo actually needed recalibrating. We had a Leica engineer recalibrate it and it has been fine since then. Cheers, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research UnitĀ School of Medical SciencesĀ Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.health.auckland.ac.nz/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Nathalia Holtzman Sent: 09 November 2007 02:59 To: [hidden email] Subject: Z-drift Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I need some help. We have a 1 year old Leica SP5 inverted scope and pretty much all of a sudden we started to get z-drift. Every time you do a time lapse, no matter what the z-step or time interval, the system does not start at the top of the sample, it moves down significantly and progressively. Anything I can do to fix this? I have called the service rep but would like to fix the problem ASAP. It is possible that I am the first person to use the system for this application since we upgraded the software and firmware. Nathalia -- Nathalia Glickman Holtzman Ph.D. Department of Biology SB D328 Queens College, CUNY 65-30 Kissena Boulevard Flushing, NY 11367 Office: 718-997-3678 Lab: 718-997-3517 Fax: 718-997-3445 |
In reply to this post by Nathalia Holtzman
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Nathalia, I would guess that 95% of the time, z-drift is caused by temperature instability in your room, incubator or both, rather than a mechanical fault with your microscope. This has been discussed many times previously on this list. You should search the archives for "z-drift" or "Focus Drift". Objective lenses are made of brass which will expand and contract with very slight changes of temperature. It is not a trivial matter to control the temperature of the room and incubator accurately enough to be able to perform time lapse microscopy. I have just spent two weeks playing with the air-conditioning system and incubator to achieve stability after the hospital switch from winter heating to summer cooling. To gain more understanding of your problem while you are waiting for a service call I suggest you look at the tutorial I have placed online at: www.ludwig.edu.au/confocal/drift If you follow this procedure you should be able to work out if the movement of the lens is always in: -A downward direction - Gravity causing drift may be the cause -An upward direction - Your lens is still warming up to the incubator temperature - Oscillating up and down - temperature instability caused by heating or cooling system in the room. -Or a microscope incubator system with either an on/off controller (get a controller with "fuzzy logic" or "PID control") -PID controller requires "Tuning" (read the manual). Temperature control is not trivial, you need to invest quite a bit of time to get it right. Cheers Stephen H. Cody Microscopy Manager Central Resource for Advanced Microscopy Ludwig Institute for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 Australia Tel: 61 3 9341 3155 Fax: 61 3 9341 3104 email: [hidden email] www.ludwig.edu.au/labs/confocal.html www.ludwig.edu.au/confocal Tip: Learn how to receive reminders about you microscope booking: www.ludwig.edu.au/confocal/Local/Booking_Hint.htm -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Nathalia Holtzman Sent: Friday, 9 November 2007 12:59 AM To: [hidden email] Subject: Z-drift Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I need some help. We have a 1 year old Leica SP5 inverted scope and pretty much all of a sudden we started to get z-drift. Every time you do a time lapse, no matter what the z-step or time interval, the system does not start at the top of the sample, it moves down significantly and progressively. Anything I can do to fix this? I have called the service rep but would like to fix the problem ASAP. It is possible that I am the first person to use the system for this application since we upgraded the software and firmware. Nathalia -- Nathalia Glickman Holtzman Ph.D. Department of Biology SB D328 Queens College, CUNY 65-30 Kissena Boulevard Flushing, NY 11367 Office: 718-997-3678 Lab: 718-997-3517 Fax: 718-997-3445 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research. |
Amanda M. Lawrence |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Dr. Cody, I tried to follow the links to your tutorial suggested in this email. The "home page" comes up but when I click on any of the menus at the top of the page (tutorials, image gallery, etc.), I get the message "this page cannot be found". As I am still new to confocal microscopy I am interested in your tutorials esp. the one on time-lapse and controlling z-drift, since time lapse is one thing our facility is going to have need of shortly. Thanks so much, Amanda Lawrence Electron Microscope Center Mississippi State University >>> Stephen Cody <[hidden email]> 11/8/2007 6:45 PM >>> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Nathalia, I would guess that 95% of the time, z-drift is caused by temperature instability in your room, incubator or both, rather than a mechanical fault with your microscope. This has been discussed many times previously on this list. You should search the archives for "z-drift" or "Focus Drift". Objective lenses are made of brass which will expand and contract with very slight changes of temperature. It is not a trivial matter to control the temperature of the room and incubator accurately enough to be able to perform time lapse microscopy. I have just spent two weeks playing with the air-conditioning system and incubator to achieve stability after the hospital switch from winter heating to summer cooling. To gain more understanding of your problem while you are waiting for a service call I suggest you look at the tutorial I have placed online at: www.ludwig.edu.au/confocal/drift If you follow this procedure you should be able to work out if the movement of the lens is always in: -A downward direction - Gravity causing drift may be the cause -An upward direction - Your lens is still warming up to the incubator temperature - Oscillating up and down - temperature instability caused by heating or cooling system in the room. -Or a microscope incubator system with either an on/off controller (get a controller with "fuzzy logic" or "PID control") -PID controller requires "Tuning" (read the manual). Temperature control is not trivial, you need to invest quite a bit of time to get it right. Cheers Stephen H. Cody Microscopy Manager Central Resource for Advanced Microscopy Ludwig Institute for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 Australia Tel: 61 3 9341 3155 Fax: 61 3 9341 3104 email: [hidden email] www.ludwig.edu.au/labs/confocal.html www.ludwig.edu.au/confocal Tip: Learn how to receive reminders about you microscope booking: www.ludwig.edu.au/confocal/Local/Booking_Hint.htm -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Nathalia Holtzman Sent: Friday, 9 November 2007 12:59 AM To: [hidden email] Subject: Z-drift Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi all, I need some help. We have a 1 year old Leica SP5 inverted scope and pretty much all of a sudden we started to get z-drift. Every time you do a time lapse, no matter what the z-step or time interval, the system does not start at the top of the sample, it moves down significantly and progressively. Anything I can do to fix this? I have called the service rep but would like to fix the problem ASAP. It is possible that I am the first person to use the system for this application since we upgraded the software and firmware. Nathalia -- Nathalia Glickman Holtzman Ph.D. Department of Biology SB D328 Queens College, CUNY 65-30 Kissena Boulevard Flushing, NY 11367 Office: 718-997-3678 Lab: 718-997-3517 Fax: 718-997-3445 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research. |