Zeiss 5-LIVE and Timing Issues

classic Classic list List threaded Threaded
♦
♦
6 messages Options
Holly L. AARON Holly L. AARON
Reply | Threaded
Open this post in threaded view
♦
♦
|

Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocal Community –

 

This question may be very specific to the Zeiss community and even more specific to the 5-LIVE users out there.

We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time intervals shorter than 100msec. By this I mean that if we set up a time-series in which an image should be taken every 500ms (let’s say the time required for the image is 30ms), it works well: an image is in fact taken every 500ms. However, if we then decide we want a shorter interval, 100ms or less (which should be fine given only 30ms for each frame), the images are taken at random times, some greater than 100ms, some less:

 

Image #

Time Expected

Time Actual

1

0

0

2

0.1

0.0987

3

0.2

0.1998

4

0.3

0.3009

5

0.4

0.4008

 

This gets worse when we go to shorter and shorter intervals, for example, 50ms:

Image #

Time Expected

Time [ s]

1

0

0

2

0.05

0.0209

3

0.1

0.1064

4

0.15

0.1273

5

0.2

0.2104

 

And even worse for shorter…

 

I am wondering: Does anyone else see this phenomenon? Have you been able to get around it? We are doing electrophysiology in concert with imaging and timing is crucial. So far we have not been able to set it up to trigger each image because that takes too long and is even less predictable. We would be very grateful for anyone who has found a work-around for this problem.

 

Thank you!

__________________

Holly L. Aaron

Molecular Imaging Center

Cancer Research Laboratory

University of California Berkeley

447 LSA #2751

Berkeley, CA  94720-2751

510.642.2901

510.642.5741 fax

[hidden email]

http://imaging.berkeley.edu

 

 

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Holly!  This may be dependent on the computer hardware running the `scope.  If it's Windows, then all bets are off for timing when you get down to the low hundreds or tens of ms.  Basically the OS can't reliably shuffle data around fast enough at those speeds; you need a real-time operating system for that.  In the case of most multitasking operating systems no single process can 'bogart' the CPU, and the time to switch between process threads can vary.  What this boils down to is that the computer will take its own sweet time handling the data coming in from your microscope depending on what else is going on with the computer and the time scale we are talking about.

Craig



On Fri, Jul 11, 2008 at 12:14 PM, Holly Aaron <[hidden email]> wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocal Community –

 

This question may be very specific to the Zeiss community and even more specific to the 5-LIVE users out there.

We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time intervals shorter than 100msec. By this I mean that if we set up a time-series in which an image should be taken every 500ms (let's say the time required for the image is 30ms), it works well: an image is in fact taken every 500ms. However, if we then decide we want a shorter interval, 100ms or less (which should be fine given only 30ms for each frame), the images are taken at random times, some greater than 100ms, some less:

 

Image #

Time Expected

Time Actual

1

0

0

2

0.1

0.0987

3

0.2

0.1998

4

0.3

0.3009

5

0.4

0.4008

 

This gets worse when we go to shorter and shorter intervals, for example, 50ms:

Image #

Time Expected

Time [ s]

1

0

0

2

0.05

0.0209

3

0.1

0.1064

4

0.15

0.1273

5

0.2

0.2104

 

And even worse for shorter…

 

I am wondering: Does anyone else see this phenomenon? Have you been able to get around it? We are doing electrophysiology in concert with imaging and timing is crucial. So far we have not been able to set it up to trigger each image because that takes too long and is even less predictable. We would be very grateful for anyone who has found a work-around for this problem.

 

Thank you!

__________________

Holly L. Aaron

Molecular Imaging Center

Cancer Research Laboratory

University of California Berkeley

447 LSA #2751

Berkeley, CA  94720-2751

510.642.2901

510.642.5741 fax

[hidden email]

http://imaging.berkeley.edu

 

 


Lambright, Geoffrey Lambright, Geoffrey
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Craig, and list,

 

So if it is the case of the computer hardware “shuffling” the data that is causing the imprecision in the data timing, does that mean that the timing of the actual image acquisition is precise and the variance that Holly sees comes from the computer’s ability to process and record the captured data? Any idea how one could check for that if that was the case?

 

Geoff

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Friday, July 11, 2008 1:03 PM
To: [hidden email]
Subject: Re: Zeiss 5-LIVE and Timing Issues

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Holly!  This may be dependent on the computer hardware running the `scope.  If it's Windows, then all bets are off for timing when you get down to the low hundreds or tens of ms.  Basically the OS can't reliably shuffle data around fast enough at those speeds; you need a real-time operating system for that.  In the case of most multitasking operating systems no single process can 'bogart' the CPU, and the time to switch between process threads can vary.  What this boils down to is that the computer will take its own sweet time handling the data coming in from your microscope depending on what else is going on with the computer and the time scale we are talking about.

Craig


On Fri, Jul 11, 2008 at 12:14 PM, Holly Aaron <[hidden email]> wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocal Community –

 

This question may be very specific to the Zeiss community and even more specific to the 5-LIVE users out there.

We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time intervals shorter than 100msec. By this I mean that if we set up a time-series in which an image should be taken every 500ms (let's say the time required for the image is 30ms), it works well: an image is in fact taken every 500ms. However, if we then decide we want a shorter interval, 100ms or less (which should be fine given only 30ms for each frame), the images are taken at random times, some greater than 100ms, some less:

 

Image #

Time Expected

Time Actual

1

0

0

2

0.1

0.0987

3

0.2

0.1998

4

0.3

0.3009

5

0.4

0.4008

 

This gets worse when we go to shorter and shorter intervals, for example, 50ms:

Image #

Time Expected

Time [ s]

1

0

0

2

0.05

0.0209

3

0.1

0.1064

4

0.15

0.1273

5

0.2

0.2104

 

And even worse for shorter…

 

I am wondering: Does anyone else see this phenomenon? Have you been able to get around it? We are doing electrophysiology in concert with imaging and timing is crucial. So far we have not been able to set it up to trigger each image because that takes too long and is even less predictable. We would be very grateful for anyone who has found a work-around for this problem.

 

Thank you!

__________________

Holly L. Aaron

Molecular Imaging Center

Cancer Research Laboratory

University of California Berkeley

447 LSA #2751

Berkeley, CA  94720-2751

510.642.2901

510.642.5741 fax

[hidden email]

http://imaging.berkeley.edu

 

 

 

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Try doing the same thing with a different microscope but similar image parameters; same bit depth per pixel, number of pixels, and acquisition times.  I'll bet you see similar performance issues.  We have some homebrew microscopes that I have worked with where we encountered operating system timing problems.  I actually had to offload a lot of timing stuff, like the pixel clock generation, to external hardware (pre-programmed timer/counter modules with their own hardware clocks) because windows cannot generate a reliable timing signal in the ms range due to the operating system's event handling limitations.  Again, you either need dedicated external hardware or a real-time OS to get reliable control over timing for small timing intervals, and Windows (and actually most desktop OSes) simply don't provide this.

Craig


On Fri, Jul 11, 2008 at 2:50 PM, Lambright, Geoffrey <[hidden email]> wrote:

Hi Craig, and list,

 

So if it is the case of the computer hardware "shuffling" the data that is causing the imprecision in the data timing, does that mean that the timing of the actual image acquisition is precise and the variance that Holly sees comes from the computer's ability to process and record the captured data? Any idea how one could check for that if that was the case?

 

Geoff

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Friday, July 11, 2008 1:03 PM
To: [hidden email]
Subject: Re: Zeiss 5-LIVE and Timing Issues

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Holly!  This may be dependent on the computer hardware running the `scope.  If it's Windows, then all bets are off for timing when you get down to the low hundreds or tens of ms.  Basically the OS can't reliably shuffle data around fast enough at those speeds; you need a real-time operating system for that.  In the case of most multitasking operating systems no single process can 'bogart' the CPU, and the time to switch between process threads can vary.  What this boils down to is that the computer will take its own sweet time handling the data coming in from your microscope depending on what else is going on with the computer and the time scale we are talking about.

Craig


On Fri, Jul 11, 2008 at 12:14 PM, Holly Aaron <[hidden email]> wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocal Community –

 

This question may be very specific to the Zeiss community and even more specific to the 5-LIVE users out there.

We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time intervals shorter than 100msec. By this I mean that if we set up a time-series in which an image should be taken every 500ms (let's say the time required for the image is 30ms), it works well: an image is in fact taken every 500ms. However, if we then decide we want a shorter interval, 100ms or less (which should be fine given only 30ms for each frame), the images are taken at random times, some greater than 100ms, some less:

 

Image #

Time Expected

Time Actual

1

0

0

2

0.1

0.0987

3

0.2

0.1998

4

0.3

0.3009

5

0.4

0.4008

 

This gets worse when we go to shorter and shorter intervals, for example, 50ms:

Image #

Time Expected

Time [ s]

1

0

0

2

0.05

0.0209

3

0.1

0.1064

4

0.15

0.1273

5

0.2

0.2104

 

And even worse for shorter…

 

I am wondering: Does anyone else see this phenomenon? Have you been able to get around it? We are doing electrophysiology in concert with imaging and timing is crucial. So far we have not been able to set it up to trigger each image because that takes too long and is even less predictable. We would be very grateful for anyone who has found a work-around for this problem.

 

Thank you!

__________________

Holly L. Aaron

Molecular Imaging Center

Cancer Research Laboratory

University of California Berkeley

447 LSA #2751

Berkeley, CA  94720-2751

510.642.2901

510.642.5741 fax

[hidden email]

http://imaging.berkeley.edu

 

 

 


Holly L. AARON Holly L. AARON
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi, Craig & Geoff & Everyone –

 

The 5-LIVE has a “real-time computer” or something it calls a real-time computer that the Windows box talks to. I would assume all the timing and handling is done with the real-time computer, but maybe it’s not.

 

The 5-LIVE is capable of 120 frames per second, but maybe not accurately? Even with a dedicated real-time computer? Or is there something not working correctly on our system?

 

Thanks for this great discussion!

-Holly

__________________

Holly L. Aaron

CRL Molecular Imaging Center

http://imaging.berkeley.edu


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Friday, July 11, 2008 1:56 PM
To: [hidden email]
Subject: Re: Zeiss 5-LIVE and Timing Issues

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Try doing the same thing with a different microscope but similar image parameters; same bit depth per pixel, number of pixels, and acquisition times.  I'll bet you see similar performance issues.  We have some homebrew microscopes that I have worked with where we encountered operating system timing problems.  I actually had to offload a lot of timing stuff, like the pixel clock generation, to external hardware (pre-programmed timer/counter modules with their own hardware clocks) because windows cannot generate a reliable timing signal in the ms range due to the operating system's event handling limitations.  Again, you either need dedicated external hardware or a real-time OS to get reliable control over timing for small timing intervals, and Windows (and actually most desktop OSes) simply don't provide this.

Craig

On Fri, Jul 11, 2008 at 2:50 PM, Lambright, Geoffrey <[hidden email]> wrote:

Hi Craig, and list,

 

So if it is the case of the computer hardware "shuffling" the data that is causing the imprecision in the data timing, does that mean that the timing of the actual image acquisition is precise and the variance that Holly sees comes from the computer's ability to process and record the captured data? Any idea how one could check for that if that was the case?

 

Geoff

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Friday, July 11, 2008 1:03 PM
To: [hidden email]
Subject: Re: Zeiss 5-LIVE and Timing Issues

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Holly!  This may be dependent on the computer hardware running the `scope.  If it's Windows, then all bets are off for timing when you get down to the low hundreds or tens of ms.  Basically the OS can't reliably shuffle data around fast enough at those speeds; you need a real-time operating system for that.  In the case of most multitasking operating systems no single process can 'bogart' the CPU, and the time to switch between process threads can vary.  What this boils down to is that the computer will take its own sweet time handling the data coming in from your microscope depending on what else is going on with the computer and the time scale we are talking about.

Craig

On Fri, Jul 11, 2008 at 12:14 PM, Holly Aaron <[hidden email]> wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocal Community –

 

This question may be very specific to the Zeiss community and even more specific to the 5-LIVE users out there.

We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time intervals shorter than 100msec. By this I mean that if we set up a time-series in which an image should be taken every 500ms (let's say the time required for the image is 30ms), it works well: an image is in fact taken every 500ms. However, if we then decide we want a shorter interval, 100ms or less (which should be fine given only 30ms for each frame), the images are taken at random times, some greater than 100ms, some less:

 

Image #

Time Expected

Time Actual

1

0

0

2

0.1

0.0987

3

0.2

0.1998

4

0.3

0.3009

5

0.4

0.4008

 

This gets worse when we go to shorter and shorter intervals, for example, 50ms:

Image #

Time Expected

Time [ s]

1

0

0

2

0.05

0.0209

3

0.1

0.1064

4

0.15

0.1273

5

0.2

0.2104

 

And even worse for shorter…

 

I am wondering: Does anyone else see this phenomenon? Have you been able to get around it? We are doing electrophysiology in concert with imaging and timing is crucial. So far we have not been able to set it up to trigger each image because that takes too long and is even less predictable. We would be very grateful for anyone who has found a work-around for this problem.

 

Thank you!

__________________

Holly L. Aaron

Molecular Imaging Center

Cancer Research Laboratory

University of California Berkeley

447 LSA #2751

Berkeley, CA  94720-2751

510.642.2901

510.642.5741 fax

[hidden email]

http://imaging.berkeley.edu

 

 

 

 

Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss 5-LIVE and Timing Issues

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Holly and list,

the scan head is controlled by an external computer, that is correct.
Using Nobody would use Windows for this task, also not for a point
scanner, if you see the time scale for the x/y scan mirrors. Time steps
for Windows and Mac OS are in the range of 60 ms, if I remember correct.

Unfortunately companies think that customers only buy systems that run
Windows. One won't get this problem with Linux (but maybe others). These
external "real-time computers" are usually controlled by Linux.

The faulty timing in the final image is added on the way from the external
computer to the Windows machine, since incoming data appears faster than
the OS can stamp it.

Now when it comes to triggering, as you plan it, Holly, I am not sure
which computer controls and handles this. In principle this needs to be
done directly via the RTC, but I recommend to get in touch with your local
Zeiss application specialist.

Michael


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi, Craig & Geoff & Everyone -
>
>
>
> The 5-LIVE has a "real-time computer" or something it calls a real-time
> computer that the Windows box talks to. I would assume all the timing and
> handling is done with the real-time computer, but maybe it's not.
>
>
>
> The 5-LIVE is capable of 120 frames per second, but maybe not accurately?
> Even with a dedicated real-time computer? Or is there something not
> working
> correctly on our system?
>
>
>
> Thanks for this great discussion!
>
> -Holly
>
> __________________
>
> Holly L. Aaron
>
> CRL Molecular Imaging Center
>
> http://imaging.berkeley.edu
>
>   _____
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Craig Brideau
> Sent: Friday, July 11, 2008 1:56 PM
> To: [hidden email]
> Subject: Re: Zeiss 5-LIVE and Timing Issues
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Try doing the same
> thing with a different microscope but similar image parameters; same bit
> depth per pixel, number of pixels, and acquisition times.  I'll bet you
> see
> similar performance issues.  We have some homebrew microscopes that I have
> worked with where we encountered operating system timing problems.  I
> actually had to offload a lot of timing stuff, like the pixel clock
> generation, to external hardware (pre-programmed timer/counter modules
> with
> their own hardware clocks) because windows cannot generate a reliable
> timing
> signal in the ms range due to the operating system's event handling
> limitations.  Again, you either need dedicated external hardware or a
> real-time OS to get reliable control over timing for small timing
> intervals,
> and Windows (and actually most desktop OSes) simply don't provide this.
>
> Craig
>
>
>
> On Fri, Jul 11, 2008 at 2:50 PM, Lambright, Geoffrey
> <[hidden email]> wrote:
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Craig, and list,
>
>
>
> So if it is the case of the computer hardware "shuffling" the data that is
> causing the imprecision in the data timing, does that mean that the timing
> of the actual image acquisition is precise and the variance that Holly
> sees
> comes from the computer's ability to process and record the captured data?
> Any idea how one could check for that if that was the case?
>
>
>
> Geoff
>
>
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Craig Brideau
> Sent: Friday, July 11, 2008 1:03 PM
> To: [hidden email]
> Subject: Re: Zeiss 5-LIVE and Timing Issues
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Holly!  This
> may
> be dependent on the computer hardware running the `scope.  If it's
> Windows,
> then all bets are off for timing when you get down to the low hundreds or
> tens of ms.  Basically the OS can't reliably shuffle data around fast
> enough
> at those speeds; you need a real-time operating system for that.  In the
> case of most multitasking operating systems no single process can 'bogart'
> the CPU, and the time to switch between process threads can vary.  What
> this
> boils down to is that the computer will take its own sweet time handling
> the
> data coming in from your microscope depending on what else is going on
> with
> the computer and the time scale we are talking about.
>
> Craig
>
>
>
> On Fri, Jul 11, 2008 at 12:14 PM, Holly Aaron <[hidden email]> wrote:
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Dear Confocal Community -
>
>
>
> This question may be very specific to the Zeiss community and even more
> specific to the 5-LIVE users out there.
>
> We find the 5-LIVE to be unreliable/unpredictable/unrepeatable in time
> intervals shorter than 100msec. By this I mean that if we set up a
> time-series in which an image should be taken every 500ms (let's say the
> time required for the image is 30ms), it works well: an image is in fact
> taken every 500ms. However, if we then decide we want a shorter interval,
> 100ms or less (which should be fine given only 30ms for each frame), the
> images are taken at random times, some greater than 100ms, some less:
>
>
>
>
> Image #
>
> Time Expected
>
> Time Actual
>
>
> 1
>
> 0
>
> 0
>
>
> 2
>
> 0.1
>
> 0.0987
>
>
> 3
>
> 0.2
>
> 0.1998
>
>
> 4
>
> 0.3
>
> 0.3009
>
>
> 5
>
> 0.4
>
> 0.4008
>
>
>
> This gets worse when we go to shorter and shorter intervals, for example,
> 50ms:
>
>
> Image #
>
> Time Expected
>
> Time [ s]
>
>
> 1
>
> 0
>
> 0
>
>
> 2
>
> 0.05
>
> 0.0209
>
>
> 3
>
> 0.1
>
> 0.1064
>
>
> 4
>
> 0.15
>
> 0.1273
>
>
> 5
>
> 0.2
>
> 0.2104
>
>
>
> And even worse for shorter.
>
>
>
> I am wondering: Does anyone else see this phenomenon? Have you been able
> to
> get around it? We are doing electrophysiology in concert with imaging and
> timing is crucial. So far we have not been able to set it up to trigger
> each
> image because that takes too long and is even less predictable. We would
> be
> very grateful for anyone who has found a work-around for this problem.
>
>
>
> Thank you!
>
> __________________
>
> Holly L. Aaron
>
> Molecular Imaging Center
>
> Cancer Research Laboratory
>
> University of California Berkeley
>
> 447 LSA #2751
>
> Berkeley, CA  94720-2751
>
> 510.642.2901
>
> 510.642.5741 fax
>
> [hidden email]
>
> http://imaging.berkeley.edu