Login  Register

Zeiss Axiocam Mrc5 with Windows 7

classic Classic list List threaded Threaded
2 messages Options Options
Embed post
Permalink
Cselenyak Attila-2 Cselenyak Attila-2
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Zeiss Axiocam Mrc5 with Windows 7

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Listers,

this is not a confocal question but you might know the solution for
this problem. I tried to use Axiocam Mrc5 camera with Windows 7 but
the program is not finding the camera. Is there a special driver for
Win7 or can I use the drivers for XP and Vista?

Thanks,

Attila


--
Cselenyák Attila

Semmelweis Egyetem
Klinikai Kísérleti Kutató- és Humán Élettani Intézet
1094, Budapest, Tűzoltó utca 37-47.
Telefon: +36-20-6660342
Mobil: +36-70-3874931
Web: www.tissueengineering.hu

--------

Attila Cselenyak

Semmelweis University
Basic Medical Science Center
Institute Of Human Physiology And Clinical Experimental Research
H-1094, Budapest
Tűzoltó utca 37-47.
Phone: +36-20-6660342
Mobile: +36-70-3874931
Web: www.tissueengineering.hu
Viereck, Friedhelm Viereck, Friedhelm
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Zeiss Axiocam Mrc5 with Windows 7

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Attila,

the current version of AxioVision (4.8.2) provides AxioCam drivers for
Windows 7 x32.
You can get a DVD with the current version from your local Carl Zeiss
MicroImaging representative.

Best regards,
Friedhelm Viereck

___________________________

Carl Zeiss MicroImaging GmbH
Geschäftsbereich Biosciences / Business Unit Biosciences
Imaging Software München
Produktmanager / Product Manager

F r i e d h e l m   V i e r e c k
Dipl.-Inform. Med.

Kistlerhofstr. 75, 81379 München
Phone: +49 89 90 90 00 309
Fax:       +49 89 90 90 00 555
mailto:[hidden email]
http://www.zeiss.de/mikro

___________________________

Carl Zeiss MicroImaging GmbH
Vorsitzender des Aufsichtsrates: Dr. Michael Kaschke
Geschäftsführung: Dr. Ulrich Simon, Wilhelm Nörthemann, Dr. Bernhard Ohnesorge
Sitz der Gesellschaft: 07740 Jena, Deutschland
Amtsgericht Jena, HRB 210.536, USt-IdNr: DE 814 503 774