Zeiss C-mount with Hamamatsu Fusion camera

classic Classic list List threaded Threaded
♦
♦
2 messages Options
Milan Esner Milan Esner
Reply | Threaded
Open this post in threaded view
♦
♦
|

Zeiss C-mount with Hamamatsu Fusion camera

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello,

recently we upgraded our Zeiss AxioImager.Z2 with new Hamamatsu ORCA Fusion camera and used standard 1x C-mount top adapter to mount the camera. Surprisingly we see dark shadows in the image corners. We identified that shadows are from C-mount adapter - we measured the inner diameter of C-mount and it is less than 22 mm. Camera has sensor which should fit into 22 mm FOV.
I wanted to ask you if you have similar problems with this camera on other microscopes? And is C-mount inner diameter different in different brands of microscopes?

Thanks a lot. Best regards,

Milan


--
Milan Esner
Head of Celular Imaging Core Facility CELLIM
CEITEC Masaryk University
Kamenice 5, building A2
625 00 Brno, Czech Republic

Phone: +420 54949 6133
E-mail: [hidden email]
weber weber
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss C-mount with Hamamatsu Fusion camera

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all,

I had have the same problem. So i do mi own adaptator. You have four threads on the front ring.
Use them to design your adaptator. I bought all components in Thorlabs.

I can send you an image

Regards