|
Michelle Peckham |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Just wondering if anyone has had either of these two systems demonstrated? Or has even bought one? And what their experiences with these systems is – do they really give good 'super-resolution' results? |
|
Dan Metcalf |
|
|
I've been using an in-house built dSTORM microscope and so I don't have any direct experience with commercial super-resolution microscopes. Here are some thoughts though:
dSTORM (also called ground state depletion microscopy by Stefan Hell and Leica). This will get you up to approximately 20 nm resolution. You can use Alexa, Cy and Atto dyes so long as they're matched up with the relevant lasers and filters on the microscopes. The key for us to getting good resolution is to have a nicely labelled sample with nice bright structures and low background. Having all the fluorophores in a similar plane of focus also seems to help. For both of these reasons you'll see that these techniques are combined with TIRF or a 'partial' TIRF. The only real down-side for me so far is that you need to acquire multiple frames over several minutes. This means it's better suited to a fixed sample and a microscope that is in a thermally stable and vibration-free environment. The fluorophore blinking behaviour is achieved with a combination of reducing buffer and a high laser power. This is in constrast to nSTORM which has coupled activator-receptor fluorophores with normal laser powers. I'd recommend thinking about dSTORM as well as nSTORM. I believe Zeiss will be launching a dSTORM system soon and Leica are starting to demonstrate their version. |
|
Schoenle, Andreas |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** In my humble opinion, your statement should read *GSDIM*[+] (also called dSTORM). In the paper introducing the name 'dSTORM', Heilemann et al.[3] do not point out that (unpaired!) Cy5 molecules had been used a year earlier by Bock et al. [2] for a stochastic single molecule switching type of nanoscopy. In simple words, they coined the acronym dSTORM for something that had been published already.[#] GSDIM [4] introduces the mechanism of pumping molecules out of the ground state to an arbitrary dark state. Explicitely, it states that this concepts should be applicable to virtually any dye under the right conditions. In their second paper[5], published after the GSDIM paper, Heilemann et al. then analyze in detail the switching mechanism of this concept but decided to ignore the acronym GSDIM. Rather they rebranded their acronym 'dSTORM' (which was originally coined just for Cy5 without Cy3) to other 'conventional' dyes such that it also contained the GSDIM method, which was shown already earlier by Fölling. Moreover, Fölling et al. pointed out the generality of the ground state depletion switching mechanism. [1] Rust et al. (2006) Nature Meth. (http://www.nature.com/nmeth/journal/v3/n10/full/nmeth929.html) [2] Bock et al. (2007) Appl. Phys. B (http://www.nanoscopy.de/publications/pdf/Appl._Phys._B_88_161-165.pdf) [3] Heilemann et al. (2008) Angew. Chem. (http://onlinelibrary.wiley.com/doi/10.1002/ange.200802376/full) [4] Fölling et al. (2008) Nature Meth. (http://www.nanoscopy.de/publications/pdf/Nature_Meth._5_943-945.pdf) [5] Heilemann et al. (2009) Angew. Chem. (http://onlinelibrary.wiley.com/doi/10.1002/ange.200902073/full) [+] *GSD* refers to the general mechanism of switching a dye by ground state depletion, which had been introduced as early as 1995 for STED-type nanoscopies. This very same mechanism of switching a dye to a metastable dark state is then used in *GSDIM* (aka dSTORM) to stochastically switch single molecules. [#] The fluorophores (Cy5/Alexa647) and off-switch used in what was originally termed dSTORM are identical to those in the original STORM paper[1], which BTW also states that the concept is applicable to any on-switchable dye. The possibility of directly switching on was both known from earlier publications and, more importantly, demonstrated in the context of nanoscopy in [2], where two-color single molecule switching nanoscopy using the fluorescent protein rsFastLime and Cy5 is presented. -- Andreas Schoenle, Dr. Department of NanoBiophotonics Max-Planck-Institute for Biophysical Chemistry D-37070 Gottingen, Germany phone:+49 551 201 2611 fax:+49 551 201 2505 http://www.mpibpc.gwdg.de/abteilungen/200/ mailto:[hidden email] PGP-Key: http://www.mpibpc.gwdg.de/abteilungen/200/personals/aschoen.asc -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dan Metcalf Sent: Thursday, June 23, 2011 3:49 PM To: [hidden email] Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I've been using an in-house built dSTORM microscope and so I don't have any direct experience with commercial super-resolution microscopes. Here are some thoughts though: *dSTORM * (also called ground state depletion microscopy by Stefan Hell and Leica). This will get you up to approximately 20 nm resolution. You can use Alexa, Cy and Atto dyes so long as they're matched up with the relevant lasers and filters on the microscopes. The key for us to getting good resolution is to have a nicely labelled sample with nice bright structures and low background. Having all the fluorophores in a similar plane of focus also seems to help. For both of these reasons you'll see that these techniques are combined with TIRF or a 'partial' TIRF. The only real down-side for me so far is that you need to acquire multiple frames over several minutes. This means it's better suited to a fixed sample and a microscope that is in a thermally stable and vibration-free environment. The fluorophore blinking behaviour is achieved with a combination of reducing buffer and a high laser power. This is in constrast to nSTORM which has coupled activator-receptor fluorophores with normal laser powers. I'd recommend thinking about dSTORM as well as nSTORM. I believe Zeiss will be launching a dSTORM system soon and Leica are starting to demonstrate their version. -- View this message in context: http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-system- and-Nikon-N-storm-system-experiences-tp6503765p6508337.html Sent from the Confocal Microscopy List mailing list archive at Nabble.com. |
|
Jose Moranosa |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** He asks about experiences with available commercial instruments. This information would be useful for all. Please, share your experiences if you have them. * *Also, really there are any differences between GSDIM/STORM/dSTORM/PALM/FPALM/PALMIRA/iPALM/etc? At the heart, all these techniques are the same thing, yes? Please tell me what distinguishes GSDIM/STORM/PALM. To me, it looks Leica/Nikon/Zeiss have claimed the same thing with different names. Ciao, Jose* From: *"Schoenle, Andreas" <[hidden email]> > *Date: *June 25, 2011 10:47:06 AM EDT > *To: *[hidden email] > *Subject: **Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences?* > *Reply-To: *Confocal Microscopy List <[hidden email]> > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In my humble opinion, your statement should read *GSDIM*[+] (also called > dSTORM). > > In the paper introducing the name 'dSTORM', Heilemann et al.[3] do not > point > out that (unpaired!) Cy5 molecules had been used a year earlier by Bock et > al. [2] for a stochastic single molecule switching type of nanoscopy. In > simple words, they coined the acronym dSTORM for something that had been > published already.[#] > GSDIM [4] introduces the mechanism of pumping molecules out of the ground > state to an arbitrary dark state. Explicitely, it states that this concepts > should be applicable to virtually any dye under the right conditions. In > their second paper[5], published after the GSDIM paper, Heilemann et al. > then > analyze in detail the switching mechanism of this concept but decided to > ignore the acronym GSDIM. Rather they rebranded their acronym 'dSTORM' > (which > was originally coined just for Cy5 without Cy3) to other 'conventional' > dyes > such that it also contained the GSDIM method, which was shown already > earlier > by Fölling. > Moreover, Fölling et al. pointed out the generality of the ground state > depletion switching mechanism. > > [1] Rust et al. (2006) Nature Meth. > (http://www.nature.com/nmeth/journal/v3/n10/full/nmeth929.html) > [2] Bock et al. (2007) Appl. Phys. B > (http://www.nanoscopy.de/publications/pdf/Appl._Phys._B_88_161-165.pdf) > [3] Heilemann et al. (2008) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200802376/full) > [4] Fölling et al. (2008) Nature Meth. > (http://www.nanoscopy.de/publications/pdf/Nature_Meth._5_943-945.pdf) > [5] Heilemann et al. (2009) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200902073/full) > > [+] > *GSD* refers to the general mechanism of switching a dye by ground state > depletion, which had been introduced as early as 1995 for STED-type > nanoscopies. This very same mechanism of switching a dye to a metastable > dark > state is then used in *GSDIM* (aka dSTORM) to stochastically switch single > molecules. > [#] > The fluorophores (Cy5/Alexa647) and off-switch used in what was originally > termed dSTORM are identical to those in the original STORM paper[1], which > BTW also states that the concept is applicable to any on-switchable dye. > The > possibility of directly switching on was both known from earlier > publications > and, more importantly, demonstrated in the context of nanoscopy in [2], > where > two-color single molecule switching nanoscopy using the fluorescent protein > rsFastLime and Cy5 is presented. > > -- > Andreas Schoenle, Dr. > Department of NanoBiophotonics > Max-Planck-Institute for Biophysical Chemistry D-37070 Gottingen, Germany > > phone:+49 551 201 2611 > fax:+49 551 201 2505 > http://www.mpibpc.gwdg.de/abteilungen/200/ > mailto:[hidden email] > > PGP-Key: http://www.mpibpc.gwdg.de/abteilungen/200/personals/aschoen.asc > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Dan Metcalf > Sent: Thursday, June 23, 2011 3:49 PM > To: [hidden email] > Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I've been using an in-house built dSTORM microscope and so I don't have any > direct experience with commercial super-resolution microscopes. Here are > some > thoughts though: > > *dSTORM * (also called ground state depletion microscopy by Stefan Hell and > Leica). This will get you up to approximately 20 nm resolution. You can use > Alexa, Cy and Atto dyes so long as they're matched up with the relevant > lasers and filters on the microscopes. The key for us to getting good > resolution is to have a nicely labelled sample with nice bright structures > and low background. Having all the fluorophores in a similar plane of focus > also seems to help. For both of these reasons you'll see that these > techniques are combined with TIRF or a 'partial' TIRF. The only real > down-side for me so far is that you need to acquire multiple frames over > several minutes. This means it's better suited to a fixed sample and a > microscope that is in a thermally stable and vibration-free environment. > > The fluorophore blinking behaviour is achieved with a combination of > reducing > buffer and a high laser power. This is in constrast to nSTORM which has > coupled activator-receptor fluorophores with normal laser powers. I'd > recommend thinking about dSTORM as well as nSTORM. > > I believe Zeiss will be launching a dSTORM system soon and Leica are > starting > to demonstrate their version. > > > -- > View this message in context: > > http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-system- > and-Nikon-N-storm-system-experiences-tp6503765p6508337.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. > > > |
|
David Baddeley |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The short (and somewhat simplified) answer is that anything with PALM in it uses fluorescent proteins and/or caged compounds, whereas the rest use different subsets of organic dyes which must first be shelved into a dark state. The equipment and data analysis algorithms required are largely identical and the proliferation of names has more to do with turf wars, marketing, and intellectual property circumvention than anything else. This acronym proliferation has not been aided by the fact that many of the key discoveries have happened more or less simultaneously in a number of different labs (PALM, STORM, and fPALM were all in press at the same time. With GSDIM, dSTORM and RPM [Baddeley et Al, 2009], it was even closer, with all papers having been submitted to the same journal at the same time - unfortunately the editors chose to only publish the GSDIM paper and the other two had to be resubmitted elsewhere, resulting in somewhat delayed publication). Whilst I empathise with Andreas's frustration, I believe that there is a lot of 'prior art' in the field and that one also could construct a similar argument for the use of other acronyms such as RPM or SPDM (eg. Reymann et Al 2008 - first use of non carboxycyanine organic dyes for localisation microscopy, Kaufman 1971, Wick 1975, Rundquist 1976 and others - observation of 'reversible photobleaching' in ensemble populations, Song et Al 1996 and others - spectroscopic characterisation of the mechanisms behind reversible photobleaching and the effect of MEA - notably they came up with the same long lived dark states which have more recently been described by Heilemann et Al). In the end, however, this leads to a whole pile of confusion for the end user and I believe that dSTORM is not a particularly bad acronym, due to the relatively minimal embellishment over, and clear linkage to traditional STORM. Both GSD and SPDM have the disadvantage of having previously been associated with different techniques: a STED-like point scanning approach and a spectrally separated position measurement method respectively, not helping the confusion. GSD also has the disadvantage of being strongly linked to the triplet as the long lived dark state - although the original paper does raise the possibility of other dark states. Typical triplet lifetimes are usually orders of magnitude too short to constitute the long lived dark state (particularly in the presence of triplet quenchers such as MEA) and a partially reduced radical state seems much more likely. Cheers, David ----- Original Message ---- From: Jose Moranosa <[hidden email]> To: [hidden email] Sent: Sun, 26 June, 2011 8:24:19 AM Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** He asks about experiences with available commercial instruments. This information would be useful for all. Please, share your experiences if you have them. * *Also, really there are any differences between GSDIM/STORM/dSTORM/PALM/FPALM/PALMIRA/iPALM/etc? At the heart, all these techniques are the same thing, yes? Please tell me what distinguishes GSDIM/STORM/PALM. To me, it looks Leica/Nikon/Zeiss have claimed the same thing with different names. Ciao, Jose* From: *"Schoenle, Andreas" <[hidden email]> > *Date: *June 25, 2011 10:47:06 AM EDT > *To: *[hidden email] > *Subject: **Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences?* > *Reply-To: *Confocal Microscopy List <[hidden email]> > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In my humble opinion, your statement should read *GSDIM*[+] (also called > dSTORM). > > In the paper introducing the name 'dSTORM', Heilemann et al.[3] do not > point > out that (unpaired!) Cy5 molecules had been used a year earlier by Bock et > al. [2] for a stochastic single molecule switching type of nanoscopy. In > simple words, they coined the acronym dSTORM for something that had been > published already.[#] > GSDIM [4] introduces the mechanism of pumping molecules out of the ground > state to an arbitrary dark state. Explicitely, it states that this concepts > should be applicable to virtually any dye under the right conditions. In > their second paper[5], published after the GSDIM paper, Heilemann et al. > then > analyze in detail the switching mechanism of this concept but decided to > ignore the acronym GSDIM. Rather they rebranded their acronym 'dSTORM' > (which > was originally coined just for Cy5 without Cy3) to other 'conventional' > dyes > such that it also contained the GSDIM method, which was shown already > earlier > by Fölling. > Moreover, Fölling et al. pointed out the generality of the ground state > depletion switching mechanism. > > [1] Rust et al. (2006) Nature Meth. > (http://www.nature.com/nmeth/journal/v3/n10/full/nmeth929.html) > [2] Bock et al. (2007) Appl. Phys. B > (http://www.nanoscopy.de/publications/pdf/Appl._Phys._B_88_161-165.pdf) > [3] Heilemann et al. (2008) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200802376/full) > [4] Fölling et al. (2008) Nature Meth. > (http://www.nanoscopy.de/publications/pdf/Nature_Meth._5_943-945.pdf) > [5] Heilemann et al. (2009) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200902073/full) > > [+] > *GSD* refers to the general mechanism of switching a dye by ground state > depletion, which had been introduced as early as 1995 for STED-type > nanoscopies. This very same mechanism of switching a dye to a metastable > dark > state is then used in *GSDIM* (aka dSTORM) to stochastically switch single > molecules. > [#] > The fluorophores (Cy5/Alexa647) and off-switch used in what was originally > termed dSTORM are identical to those in the original STORM paper[1], which > BTW also states that the concept is applicable to any on-switchable dye. > The > possibility of directly switching on was both known from earlier > publications > and, more importantly, demonstrated in the context of nanoscopy in [2], > where > two-color single molecule switching nanoscopy using the fluorescent protein > rsFastLime and Cy5 is presented. > > -- > Andreas Schoenle, Dr. > Department of NanoBiophotonics > Max-Planck-Institute for Biophysical Chemistry D-37070 Gottingen, Germany > > phone:+49 551 201 2611 > fax:+49 551 201 2505 > http://www.mpibpc.gwdg.de/abteilungen/200/ > mailto:[hidden email] > > PGP-Key: http://www.mpibpc.gwdg.de/abteilungen/200/personals/aschoen.asc > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Dan Metcalf > Sent: Thursday, June 23, 2011 3:49 PM > To: [hidden email] > Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I've been using an in-house built dSTORM microscope and so I don't have any > direct experience with commercial super-resolution microscopes. Here are > some > thoughts though: > > *dSTORM * (also called ground state depletion microscopy by Stefan Hell and > Leica). This will get you up to approximately 20 nm resolution. You can use > Alexa, Cy and Atto dyes so long as they're matched up with the relevant > lasers and filters on the microscopes. The key for us to getting good > resolution is to have a nicely labelled sample with nice bright structures > and low background. Having all the fluorophores in a similar plane of focus > also seems to help. For both of these reasons you'll see that these > techniques are combined with TIRF or a 'partial' TIRF. The only real > down-side for me so far is that you need to acquire multiple frames over > several minutes. This means it's better suited to a fixed sample and a > microscope that is in a thermally stable and vibration-free environment. > > The fluorophore blinking behaviour is achieved with a combination of > reducing > buffer and a high laser power. This is in constrast to nSTORM which has > coupled activator-receptor fluorophores with normal laser powers. I'd > recommend thinking about dSTORM as well as nSTORM. > > I believe Zeiss will be launching a dSTORM system soon and Leica are > starting > to demonstrate their version. > > > -- > View this message in context: > > http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-system- > and-Nikon-N-storm-system-experiences-tp6503765p6508337.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. > > > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, well, I was reduced to coining this acronym: Arbitrary Conventions for Refining Original Names for Your Microscope. OK, more seriously, I don't have any axe to grind in this one. My only connection was being a referee for one of the papers David mentions (and no, I didn't reject it and force it to be delayed!). However, the term STORM (Stochastic Optical Resolution Microscopy) does actually cover all the methods described here, so to my mind inventing a new name has no merit other than staking a claim. Hence, to me, dSTORM seems sensible, and acknowledges what has been done before. GSDim and RPM are unnecessary coinages. They are all varieties of the pre-existing technique STORM - in fact they are all essentially the same variety of that technique. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Baddeley Sent: Sunday, 26 June 2011 11:35 AM To: [hidden email] Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The short (and somewhat simplified) answer is that anything with PALM in it uses fluorescent proteins and/or caged compounds, whereas the rest use different subsets of organic dyes which must first be shelved into a dark state. The equipment and data analysis algorithms required are largely identical and the proliferation of names has more to do with turf wars, marketing, and intellectual property circumvention than anything else. This acronym proliferation has not been aided by the fact that many of the key discoveries have happened more or less simultaneously in a number of different labs (PALM, STORM, and fPALM were all in press at the same time. With GSDIM, dSTORM and RPM [Baddeley et Al, 2009], it was even closer, with all papers having been submitted to the same journal at the same time - unfortunately the editors chose to only publish the GSDIM paper and the other two had to be resubmitted elsewhere, resulting in somewhat delayed publication). Whilst I empathise with Andreas's frustration, I believe that there is a lot of 'prior art' in the field and that one also could construct a similar argument for the use of other acronyms such as RPM or SPDM (eg. Reymann et Al 2008 - first use of non carboxycyanine organic dyes for localisation microscopy, Kaufman 1971, Wick 1975, Rundquist 1976 and others - observation of 'reversible photobleaching' in ensemble populations, Song et Al 1996 and others - spectroscopic characterisation of the mechanisms behind reversible photobleaching and the effect of MEA - notably they came up with the same long lived dark states which have more recently been described by Heilemann et Al). In the end, however, this leads to a whole pile of confusion for the end user and I believe that dSTORM is not a particularly bad acronym, due to the relatively minimal embellishment over, and clear linkage to traditional STORM. Both GSD and SPDM have the disadvantage of having previously been associated with different techniques: a STED-like point scanning approach and a spectrally separated position measurement method respectively, not helping the confusion. GSD also has the disadvantage of being strongly linked to the triplet as the long lived dark state - although the original paper does raise the possibility of other dark states. Typical triplet lifetimes are usually orders of magnitude too short to constitute the long lived dark state (particularly in the presence of triplet quenchers such as MEA) and a partially reduced radical state seems much more likely. Cheers, David ----- Original Message ---- From: Jose Moranosa <[hidden email]> To: [hidden email] Sent: Sun, 26 June, 2011 8:24:19 AM Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** He asks about experiences with available commercial instruments. This information would be useful for all. Please, share your experiences if you have them. * *Also, really there are any differences between GSDIM/STORM/dSTORM/PALM/FPALM/PALMIRA/iPALM/etc? At the heart, all these techniques are the same thing, yes? Please tell me what distinguishes GSDIM/STORM/PALM. To me, it looks Leica/Nikon/Zeiss have claimed the same thing with different names. Ciao, Jose* From: *"Schoenle, Andreas" <[hidden email]> > *Date: *June 25, 2011 10:47:06 AM EDT > *To: *[hidden email] > *Subject: **Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences?* > *Reply-To: *Confocal Microscopy List <[hidden email]> > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In my humble opinion, your statement should read *GSDIM*[+] (also called > dSTORM). > > In the paper introducing the name 'dSTORM', Heilemann et al.[3] do not > point > out that (unpaired!) Cy5 molecules had been used a year earlier by Bock et > al. [2] for a stochastic single molecule switching type of nanoscopy. In > simple words, they coined the acronym dSTORM for something that had been > published already.[#] > GSDIM [4] introduces the mechanism of pumping molecules out of the ground > state to an arbitrary dark state. Explicitely, it states that this concepts > should be applicable to virtually any dye under the right conditions. In > their second paper[5], published after the GSDIM paper, Heilemann et al. > then > analyze in detail the switching mechanism of this concept but decided to > ignore the acronym GSDIM. Rather they rebranded their acronym 'dSTORM' > (which > was originally coined just for Cy5 without Cy3) to other 'conventional' > dyes > such that it also contained the GSDIM method, which was shown already > earlier > by Fölling. > Moreover, Fölling et al. pointed out the generality of the ground state > depletion switching mechanism. > > [1] Rust et al. (2006) Nature Meth. > (http://www.nature.com/nmeth/journal/v3/n10/full/nmeth929.html) > [2] Bock et al. (2007) Appl. Phys. B > (http://www.nanoscopy.de/publications/pdf/Appl._Phys._B_88_161-165.pdf) > [3] Heilemann et al. (2008) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200802376/full) > [4] Fölling et al. (2008) Nature Meth. > (http://www.nanoscopy.de/publications/pdf/Nature_Meth._5_943-945.pdf) > [5] Heilemann et al. (2009) Angew. Chem. > (http://onlinelibrary.wiley.com/doi/10.1002/ange.200902073/full) > > [+] > *GSD* refers to the general mechanism of switching a dye by ground state > depletion, which had been introduced as early as 1995 for STED-type > nanoscopies. This very same mechanism of switching a dye to a metastable > dark > state is then used in *GSDIM* (aka dSTORM) to stochastically switch single > molecules. > [#] > The fluorophores (Cy5/Alexa647) and off-switch used in what was originally > termed dSTORM are identical to those in the original STORM paper[1], which > BTW also states that the concept is applicable to any on-switchable dye. > The > possibility of directly switching on was both known from earlier > publications > and, more importantly, demonstrated in the context of nanoscopy in [2], > where > two-color single molecule switching nanoscopy using the fluorescent protein > rsFastLime and Cy5 is presented. > > -- > Andreas Schoenle, Dr. > Department of NanoBiophotonics > Max-Planck-Institute for Biophysical Chemistry D-37070 Gottingen, Germany > > phone:+49 551 201 2611 > fax:+49 551 201 2505 > http://www.mpibpc.gwdg.de/abteilungen/200/ > mailto:[hidden email] > > PGP-Key: http://www.mpibpc.gwdg.de/abteilungen/200/personals/aschoen.asc > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Dan Metcalf > Sent: Thursday, June 23, 2011 3:49 PM > To: [hidden email] > Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - > experiences? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I've been using an in-house built dSTORM microscope and so I don't have any > direct experience with commercial super-resolution microscopes. Here are > some > thoughts though: > > *dSTORM * (also called ground state depletion microscopy by Stefan Hell and > Leica). This will get you up to approximately 20 nm resolution. You can use > Alexa, Cy and Atto dyes so long as they're matched up with the relevant > lasers and filters on the microscopes. The key for us to getting good > resolution is to have a nicely labelled sample with nice bright structures > and low background. Having all the fluorophores in a similar plane of focus > also seems to help. For both of these reasons you'll see that these > techniques are combined with TIRF or a 'partial' TIRF. The only real > down-side for me so far is that you need to acquire multiple frames over > several minutes. This means it's better suited to a fixed sample and a > microscope that is in a thermally stable and vibration-free environment. > > The fluorophore blinking behaviour is achieved with a combination of > reducing > buffer and a high laser power. This is in constrast to nSTORM which has > coupled activator-receptor fluorophores with normal laser powers. I'd > recommend thinking about dSTORM as well as nSTORM. > > I believe Zeiss will be launching a dSTORM system soon and Leica are > starting > to demonstrate their version. > > > -- > View this message in context: > > http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-system- > and-Nikon-N-storm-system-experiences-tp6503765p6508337.html > Sent from the Confocal Microscopy List mailing list archive at Nabble.com. > > > |
|
James Pawley |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** STORM in a teacup? As I may have noted to this list before, the technique of centroid-finding as a method of improving the resolution of an optical system (but admittedly not of combining this with a system for turning point emitters on and off) goes back to a least to the 1970s: An earlier digital system for locating and storing the centroid of a single-photoelectron event in an electronic image acquisition system is described in Herrmann, Krahl Rust and Ulrichs, Optik, 44,(1976)-4, pp 393-412 (see figure 7) Herrmann, Krahl and Rust, Ultramicroscopy, 3 (1978) pp 227-235. Herrmann, and Krahl, J. Microscopy, 127-1 (1982) pp 17-28 and many other places The system was for viewing very low dose images in an electron microscope and used an intensified vidicon camera coupled to a phosphor conversion screen. Each electron in the electron microscope made a flash on the screen and this was then amplified and detected by the intensified vidicon. Using this procedure, the spot produced by each microscope electron was quite large (4-5 tv lines wide) and consequently the resulting image was very blurred. To overcome this, the authors built (!) a 256x256 video-rate digital image memory and then developed software (or more likely hardware) that ran in real time to calculate the centroid of each flash, slowly accumulating an image in the digital memory in which each flash was recorded as a single unit event located where the centroid of the original flash had been. Of course, in this case, the source of the blurring to be removed was the low resolution of the intensified vidicon, not an imperfection in the electron microscope itself. However, if you consider the microscope+vidicon as a single functional unit, the parallel is quite close: it was made to "see and record" an electron landing place; the newer LM analogs is made to "see and record" the location of a fluorescent molecule. I saw the Rust system operate in the early 1980s. It was set to accumulate in an exponential (running?) average manner (stored pixel value added to new value and the sum divided by a constant to create new stored value) suitable for averaging about 120 frames (4 second decay). One could watch the image of a single biological molecule first begin to appear out of the background and then turn into a featureless blob as radiation damage took its toll. This general system was later commercialized by Hamamatsu in a system that they called PIAS (Photon-counting image acquisition system, See: http://www.sao.ru/drabek/CCDP/Hamamatsu/CAMERAS/PiasE.htm for updated version.). Some PIAS system simply gated the single-photoelectron flashes to a standard height but a more advanced system was also offered in which a computer again computed the centroid of each flash and stored a single count at its location. I have a catalog describing the system dated May 1985. Like many productive new developments, PALM and STORM and the other variant are actually improvements or recombinations of earlier techniques although often the new "inventors" were unaware of the earlier work. You can make the case that the confocal was "invented" at least 7 times, and Guy makes a good case for a much longer list. Priority is important. But it is not all-important. Putting the last piece in the puzzle may produce something that works, but this doesn't change the fact that without many of the earlier pieces, the puzzle would still remain incomplete. All that stated, my problem wit all of these techniques is how they display the data. Usually they place a blob into the image memory at the location of the centroid and the narrowness and brightness of this blob is often related to the accuracy of the fit fit of the "Gaussian" to the detected blob. But this says nothing about the statistics of the dye molecules themselves. How many were actually detected? Why are some shown brighter and smaller when the main variable may simply be how many times these they emitted before being bleached (or exactly where they are located in the evanescent field) rather than the significance of the labeled molecule to the function of the cell. I worry that the proclivity of our visual system to look for "edges" and then interpret these collections of blob as though they were some sort of "tapioca" pudding may lead us draw unsupported interpretations from these images. We may have solved "The Diffraction" problem but "The Statistics" are still unclear because it has stopped being the (nice, understandable Poisson) statistics related to the difficulty in counting a very small number of quantum events and become the statistics that determine which individual fluorophors have been recorded and what factors in their local microenviroment might have been important in this selection, as well as the (Poisson-like?) errors associated with observing very small numbers of molecules. Regards, Jim Pawley -- > >Yes, well, I was reduced to coining this >acronym: Arbitrary Conventions for Refining >Original Names for Your Microscope. > >OK, more seriously, I don't have any axe to >grind in this one. My only connection was being >a referee for one of the papers David mentions >(and no, I didn't reject it and force it to be >delayed!). However, the term STORM (Stochastic >Optical Resolution Microscopy) does actually >cover all the methods described here, so to my >mind inventing a new name has no merit other >than staking a claim. Hence, to me, dSTORM >seems sensible, and acknowledges what has been >done before. GSDim and RPM are unnecessary >coinages. They are all varieties of the >pre-existing technique STORM - in fact they are >all essentially the same variety of that >technique. > > Guy > >Optical Imaging Techniques in Cell Biology >by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm >______________________________________________ >Associate Professor Guy Cox, MA, DPhil(Oxon) >Australian Centre for Microscopy & Microanalysis, >Madsen Building F09, University of Sydney, NSW 2006 > >Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 >______________________________________________ > http://www.guycox.net > > > >-----Original Message----- >From: Confocal Microscopy List >[mailto:[hidden email]] On >Behalf Of David Baddeley >Sent: Sunday, 26 June 2011 11:35 AM >To: [hidden email] >Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >The short (and somewhat simplified) answer is >that anything with PALM in it uses >fluorescent proteins and/or caged compounds, whereas the rest use different >subsets of organic dyes which must first be shelved into a dark state. The >equipment and data analysis algorithms required are largely identical and the >proliferation of names has more to do with turf wars, marketing, and >intellectual property circumvention than anything else. > >This acronym proliferation has not been aided by the fact that many of the key >discoveries have happened more or less simultaneously in a number of different >labs (PALM, STORM, and fPALM were all in press at the same time. With GSDIM, >dSTORM and RPM [Baddeley et Al, 2009], it was even closer, with all papers >having been submitted to the same journal at the same time - unfortunately the >editors chose to only publish the GSDIM paper and the other two had to be >resubmitted elsewhere, resulting in somewhat delayed publication). > >Whilst I empathise with Andreas's frustration, I >believe that there is a lot of >'prior art' in the field and that one also could construct a similar argument >for the use of other acronyms such as RPM or SPDM (eg. Reymann et Al 2008 - >first use of non carboxycyanine organic dyes for localisation microscopy, >Kaufman 1971, Wick 1975, Rundquist 1976 and >others - observation of 'reversible >photobleaching' in ensemble populations, Song et Al 1996 and others - >spectroscopic characterisation of the mechanisms behind reversible >photobleaching and the effect of MEA - notably they came up with the same long >lived dark states which have more recently been described by Heilemann et Al). >In the end, however, this leads to a whole pile of confusion for the end user >and I believe that dSTORM is not a particularly bad acronym, due to the >relatively minimal embellishment over, and clear linkage to traditional STORM. > >Both GSD and SPDM have the disadvantage of having previously >been associated with different techniques: a STED-like point scanning approach >and a spectrally separated position measurement >method respectively, not helping >the confusion. GSD also has the disadvantage of being strongly linked to the >triplet as the long lived dark state - although the original paper does raise >the possibility of other dark states. Typical triplet lifetimes are usually >orders of magnitude too short to constitute the long lived dark state >(particularly in the presence of triplet >quenchers such as MEA) and a partially >reduced radical state seems much more likely. > >Cheers, >David > > >----- Original Message ---- >From: Jose Moranosa <[hidden email]> >To: [hidden email] >Sent: Sun, 26 June, 2011 8:24:19 AM >Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - experiences? > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >He asks about experiences with available commercial instruments. This >information would be useful for all. Please, share your experiences if you >have them. >* >*Also, really there are any differences between >GSDIM/STORM/dSTORM/PALM/FPALM/PALMIRA/iPALM/etc? At the heart, all these >techniques are the same thing, yes? Please tell me what distinguishes >GSDIM/STORM/PALM. To me, it looks Leica/Nikon/Zeiss have claimed the same >thing with different names. > >Ciao, > >Jose* > > >From: *"Schoenle, Andreas" <[hidden email]> > >> *Date: *June 25, 2011 10:47:06 AM EDT >> *To: *[hidden email] >> *Subject: **Re: Zeiss Elyra PALM system and Nikon N-storm system - >> experiences?* >> *Reply-To: *Confocal Microscopy List <[hidden email]> >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> In my humble opinion, your statement should read *GSDIM*[+] (also called >> dSTORM). >> >> In the paper introducing the name 'dSTORM', Heilemann et al.[3] do not >> point >> out that (unpaired!) Cy5 molecules had been used a year earlier by Bock et >> al. [2] for a stochastic single molecule switching type of nanoscopy. In >> simple words, they coined the acronym dSTORM for something that had been >> published already.[#] >> GSDIM [4] introduces the mechanism of pumping molecules out of the ground >> state to an arbitrary dark state. Explicitely, it states that this concepts >> should be applicable to virtually any dye under the right conditions. In >> their second paper[5], published after the GSDIM paper, Heilemann et al. >> then >> analyze in detail the switching mechanism of this concept but decided to >> ignore the acronym GSDIM. Rather they rebranded their acronym 'dSTORM' >> (which >> was originally coined just for Cy5 without Cy3) to other 'conventional' >> dyes >> such that it also contained the GSDIM method, which was shown already >> earlier >> by Fölling. >> Moreover, Fölling et al. pointed out the generality of the ground state >> depletion switching mechanism. >> >> [1] Rust et al. (2006) Nature Meth. >> (http://www.nature.com/nmeth/journal/v3/n10/full/nmeth929.html) >> [2] Bock et al. (2007) Appl. Phys. B >> (http://www.nanoscopy.de/publications/pdf/Appl._Phys._B_88_161-165.pdf) > > [3] Heilemann et al. (2008) Angew. Chem. >> (http://onlinelibrary.wiley.com/doi/10.1002/ange.200802376/full) >> [4] Fölling et al. (2008) Nature Meth. >> (http://www.nanoscopy.de/publications/pdf/Nature_Meth._5_943-945.pdf) >> [5] Heilemann et al. (2009) Angew. Chem. >> (http://onlinelibrary.wiley.com/doi/10.1002/ange.200902073/full) >> >> [+] >> *GSD* refers to the general mechanism of switching a dye by ground state >> depletion, which had been introduced as early as 1995 for STED-type >> nanoscopies. This very same mechanism of switching a dye to a metastable > > dark >> state is then used in *GSDIM* (aka dSTORM) to stochastically switch single >> molecules. >> [#] >> The fluorophores (Cy5/Alexa647) and off-switch used in what was originally >> termed dSTORM are identical to those in the original STORM paper[1], which >> BTW also states that the concept is applicable to any on-switchable dye. >> The >> possibility of directly switching on was both known from earlier >> publications >> and, more importantly, demonstrated in the context of nanoscopy in [2], >> where >> two-color single molecule switching nanoscopy using the fluorescent protein >> rsFastLime and Cy5 is presented. >> >> -- >> Andreas Schoenle, Dr. >> Department of NanoBiophotonics >> Max-Planck-Institute for Biophysical Chemistry D-37070 Gottingen, Germany >> >> phone:+49 551 201 2611 >> fax:+49 551 201 2505 >> http://www.mpibpc.gwdg.de/abteilungen/200/ >> mailto:[hidden email] >> >> PGP-Key: http://www.mpibpc.gwdg.de/abteilungen/200/personals/aschoen.asc >> >> >> -----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] >> On >> Behalf Of Dan Metcalf >> Sent: Thursday, June 23, 2011 3:49 PM >> To: [hidden email] >> Subject: Re: Zeiss Elyra PALM system and Nikon N-storm system - >> experiences? >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I've been using an in-house built dSTORM microscope and so I don't have any >> direct experience with commercial super-resolution microscopes. Here are >> some >> thoughts though: >> >> *dSTORM * (also called ground state depletion microscopy by Stefan Hell and >> Leica). This will get you up to approximately 20 nm resolution. You can use >> Alexa, Cy and Atto dyes so long as they're matched up with the relevant >> lasers and filters on the microscopes. The key for us to getting good >> resolution is to have a nicely labelled sample with nice bright structures >> and low background. Having all the fluorophores in a similar plane of focus >> also seems to help. For both of these reasons you'll see that these >> techniques are combined with TIRF or a 'partial' TIRF. The only real >> down-side for me so far is that you need to acquire multiple frames over >> several minutes. This means it's better suited to a fixed sample and a >> microscope that is in a thermally stable and vibration-free environment. >> >> The fluorophore blinking behaviour is achieved with a combination of >> reducing >> buffer and a high laser power. This is in constrast to nSTORM which has >> coupled activator-receptor fluorophores with normal laser powers. I'd >> recommend thinking about dSTORM as well as nSTORM. >> >> I believe Zeiss will be launching a dSTORM system soon and Leica are >> starting >> to demonstrate their version. >> >> >> -- >> View this message in context: >> >> >>http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-system- >> and-Nikon-N-storm-system-experiences-tp6503765p6508337.html >> Sent from the Confocal Microscopy List mailing list archive at Nabble.com. >> >> >> -- **************************************** Prof. James B. Pawley, Phone: 604-822-7801 3D Microscopy of Living Cells: Summer Course CELL: 778-919-3176 Info at: http://www.3dcourse.ubc.ca "If it isn't diffraction, it is statistics":Microscopist's complaint, Anon. |
|
Mark Bates |
|
|
hi all,
Thanks Jim for your synopsis of the related work in the EM field - I wasn't aware of that. Just to touch on a couple of the points raised, I agree with Jim regarding the statistics of switchable molecules - this can result in the STORM image becoming non-linear with respect to how the intensity in the image corresponds to the local concentration of fluorophores on the sample. This all depends on the nature of the switchable fluorophore of course, whether it activates once and then bleaches, or if it's reversibly switchable, etc. In principle it would be very useful to treat these images as quantitative maps of fluorophore position and concentration, and this is where one needs to be careful in considering the switching statistics of the fluorophore. Regarding the name debate, PALM, STORM, FPALM etc all refer to precisely the same concept. PALM should not be associated strictly with proteins, nor should STORM be associated with one type of fluorescent probe. We made it clear in our original paper that the concept is applicable to any switchable fluorophore. I don't see the need for the acronyms which have been subsequently coined, except where there is new science involved (such as GSDIM, which extends the concept to any photophysical dark state). best, Mark Bates, Ph.D. Max Planck Institute for Biophysical Chemistry Göttingen 37077 Germany |
|
Csúcs Gábor |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Leaving the discussion aside, I've seen both systems demonstrated and recently took a look also at the Leica system. For me the main difference (if I neglect all the licensing/patent issues) was that in the GSDIM (Leica) they use considerably stronger lasers ( around 600 mW). Greetings Gabor Light Microscopy Centre ETH Zurich, Schafmattstrasse 18 CH-8093, Zurich Switzerland Phone: +41 44 633 6221 On 6/27/11 2:48 PM, "Mark Bates" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >hi all, > >Thanks Jim for your synopsis of the related work in the EM field - I >wasn't >aware of that. Just to touch on a couple of the points raised, I agree >with >Jim regarding the statistics of switchable molecules - this can result in >the STORM image becoming non-linear with respect to how the intensity in >the >image corresponds to the local concentration of fluorophores on the >sample. >This all depends on the nature of the switchable fluorophore of course, >whether it activates once and then bleaches, or if it's reversibly >switchable, etc. In principle it would be very useful to treat these >images >as quantitative maps of fluorophore position and concentration, and this >is >where one needs to be careful in considering the switching statistics of >the >fluorophore. > >Regarding the name debate, PALM, STORM, FPALM etc all refer to precisely >the >same concept. PALM should not be associated strictly with proteins, nor >should STORM be associated with one type of fluorescent probe. We made it >clear in our original paper that the concept is applicable to any >switchable >fluorophore. I don't see the need for the acronyms which have been >subsequently coined, except where there is new science involved (such as >GSDIM, which extends the concept to any photophysical dark state). > >best, >Mark Bates, Ph.D. >Max Planck Institute for Biophysical Chemistry >Göttingen 37077 >Germany > >-- >View this message in context: >http://confocal-microscopy-list.588098.n2.nabble.com/Zeiss-Elyra-PALM-syst >em-and-Nikon-N-storm-system-experiences-tp6503765p6520262.html >Sent from the Confocal Microscopy List mailing list archive at Nabble.com. |
|
Clements, Ian |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Please do not respond directly to this post. Send inquires to Heidi Rudi ([hidden email]). Bio Engineer Service A position of Bio Engineer is part of the Life Sciences Service Team. Under limited supervision, provides installation, training and advanced technical and applications support for API end customers and distributors. Major Responsibilities: * Perform complex field service installations, upgrades and repairs with up to 75% time on the road * Provide on-site technical training * Analyze, resolve and document technical/logistical issues encountered in field ensuring maximum customer satisfaction * Create and maintain service calls records. Maintain Service Knowledge Database * Prepare required activity reports. Secondary Functions: When working from the home office, expected to: * Develop Service training / technical documentation * Provide assistance to the Helpdesk TSE (Technical Service Engineer) handling hotline calls (phone & e-mail) * Provide technical training for distributors and end users using Internet based teleconference tools * Provide assistance for traveling FSEs (Field Service Engineers) * Additional tasks as defined by Department Manager. Specific Job Skills: * Exceptional interpersonal skills * Excellent communication skills both written and oral * Train users on hardware and software included in Applied Precision's products * Ability to effectively analyze and solve technical problems * First-class customer service skills with the ability to handle and diffuse difficult customers and situations * Practical implementation of the time management skills to identify priorities and organize daily workload * Ability to gather data, compile and prepare reports. Education and/or Experience: * Bachelor's degree in Biomedical Engineering or equivalent experience * A minimum 3+ years of Cell or Molecular Biology or related field * Experience with microscopic life science imaging systems is preferred * Ability to work with complex electromechanical / optical devices * Working knowledge of DOS / Windows and Linux OS * Very good understanding of computer hardware / software and networking * Prior experience in Field Service environments * Proficiency with MS Office software tools. Job Conditions: * Operating from API Headquarters - Issaquah, WA * Position requires limited weekend and holiday coverage. This email message, together with any attachments, is for the sole use of the intended recipient(s) and is the confidential information of Applied Precision Inc. If you are not the intended recipient, your review, use, disclosure, copying or dissemination of this email message or its attachments, or the information contained therein, is strictly prohibited. If you are not the intended recipient or if you think this email was sent to you in error, please notify the sender by reply email and delete this message and its attachments, as well as all copies, from your system. |
| Free forum by Nabble | Edit this page |