Hello,
We are having some problems with our Zeiss LSM 510
META argon laser (488 et al). Basically it's getting old and due a replacement,
it's no longer producing the power it should and we're also getting frequent
error messages:
Object: COM3
Read time out Object: COM3 Read time out Object: AOTF Driver IsHwAccessible failed which means we need to restart part, or
all, of the system. Zeiss have told us that we should replace the laser and
power-supply and the problem will be resolved. We will be replacing the laser as
soon as possible, but I wondered what the advice from the list is
regards whether or not we need to replace the power-supply too? It adds quite a
lot to the price. So, from your experiences, should the power-supply be replaced
at the same time as the laser? If we don't could our communication problems
persist?
Thanks in advance for your
help.
Graham
-----
Dr Graham Wright Microscopy & Imaging Facility Temasek Life Sciences Laboratory
1 Research Link National University of Singapore Singapore 117604 |
|
Dale Callaham |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal This is just a general comment about the laser power supply electronics. To keep sizes small, modern power supply components are worked fairly hard; most use switchmode power conversion technology and the capacitors run a very high ripple current. All electronics have a lifetime that is dependent on the conditions of use and these components are probably run at good design parameters but still operate at elevated temps and are probably "tired". With aging, the Equivalent Series Resistance of the capcitors increases and the ripple voltage increases. What I am saying then is that it may operate the laser if you keep the old one, but the performance is likely dropping off or may do so in the near future. From the company perspective it is more likely to give good performance and happy customers if they are cycled at the same time. I can't comment on the error messages. I would trust Zeiss since they know these messages. Our Zeiss service has been very good with getting the correct parts first time based on the system messages and symptoms. I have no association with any commercial interests. Dale Callaham Graham Wright wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > Hello, > We are having some problems with our Zeiss LSM 510 META argon laser (488 > et al). Basically it's getting old and due a replacement, it's no longer > producing the power it should and we're also getting frequent error > messages: > /Object: COM3 > Read time out > Object: COM3 > Read time out > Object: AOTF Driver > IsHwAccessible failed/ > which means we need to restart part, or all, of the system. Zeiss have > told us that we should replace the laser and power-supply and the > problem will be resolved. We will be replacing the laser as soon as > possible, but I wondered what the advice from the list is regards > whether or not we need to replace the power-supply too? It adds quite a > lot to the price. So, from your experiences, should the power-supply be > replaced at the same time as the laser? If we don't could > our communication problems persist? > Thanks in advance for your help. > Graham > ----- > Dr Graham Wright > Microscopy & Imaging Facility > > Temasek Life Sciences Laboratory > 1 Research Link > National University of Singapore > Singapore 117604 > > P: +65 6872 8406 > E: [hidden email] <mailto:[hidden email]> > W: http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm |
|
|
In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
I'd try to have it refurbished; we just did that with an old, non-functional Ar laser, including an overhaul of the power supply, replacement of the laser tube, and increase of power output, for significantly less than the price of a new one. The work was done by Dynamic Laser in Salt Lake City, Utah. No commercial interest of course, just a happy customer. Let me know if you'd like further details.
Zoltan
On Feb 12, 2008 2:28 AM, Graham Wright <[hidden email]> wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -- -- Zoltan Cseresnyes Facility manager, Imaging Suite Dept. of Zoology University of Cambridge Downing Street, Cambridge CB2 3EJ UK Tel.: (++44) (0)1223 769282 Fax.: (++44) (0)1223 336676 |
|
|
In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi Graham, Hope this was helpful, Fred On 11-Feb-08, at 7:28 PM, Graham Wright wrote:
Fred D. Mast Department of Cell Biology Medical Sciences Building Room 5-14 University of Alberta Edmonton, Alberta, T6G 2H7 Canada Tel: 1-780-492-7407 |
|
Gert van Cappellen |
|
|
In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal In the past we had some laser replacements without power supply or replaced power supply without laser and it often tuned out that we needed a second replacement a few weeks later. Nowadays the laser and power supply are replaced together leading to less down time and less service visits. The error message doesn't seem to indicate a problem with the laser, but with your AOTF. Good luck, Gert van Cappellen Graham Wright schreef: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > Hello, > We are having some problems with our Zeiss LSM 510 META argon laser > (488 et al). Basically it's getting old and due a replacement, it's no > longer producing the power it should and we're also getting frequent > error messages: > /Object: COM3 > Read time out > Object: COM3 > Read time out > Object: AOTF Driver > IsHwAccessible failed/ > which means we need to restart part, or all, of the system. Zeiss have > told us that we should replace the laser and power-supply and the > problem will be resolved. We will be replacing the laser as soon as > possible, but I wondered what the advice from the list is regards > whether or not we need to replace the power-supply too? It adds quite > a lot to the price. So, from your experiences, should the power-supply > be replaced at the same time as the laser? If we don't could > our communication problems persist? > Thanks in advance for your help. > Graham > ----- > Dr Graham Wright > Microscopy & Imaging Facility > > Temasek Life Sciences Laboratory > 1 Research Link > National University of Singapore > Singapore 117604 > > P: +65 6872 8406 > E: [hidden email] <mailto:[hidden email]> > W: http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm |
|
Frank-Peter Schoew |
|
|
In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello, the error message about the COM port is not related to the Argon laser head or the laser power supply. It seems more a problem with the AOTF driver in the system. We ll get in contact with our Service team in Singapore to solve the problem immidiately. Best regards Frank-Peter Schoew Carl Zeiss MicroImaging GmbH Technical Support and Training BioSciences F r a n k - P e t e r S c h o e w Phone: +49 3641 64-3488 Fax: +49 3641 64-3144 mailto:[hidden email] http://www.zeiss.de/micro Carl Zeiss MicroImaging GmbH, Carl-Zeiss-Promenade 10, 07745 Jena Vorsitzender des Aufsichtsrates: Dr. Michael Kaschke Geschäftsführung: Dr. Ulrich Simon, Michael Mathias, Wilhelm Nörthemann Sitz der Gesellschaft: 07740 Jena, Deutschland Amtsgericht Jena, HRB 210.536, USt-IdNr: DE814503774 |