Dear all,
we've got an upright microscope with a 20x water dipping lens that we use for Ca imaging in tissues. Much like a patch clamp experiment we insert a perfusion pipette between the lens and the tissue to administer drugs etc. The solutions are all oxygenated CO2/bicarbonate buffered. Unfortunately it happens (too often) that an bubble (air/O2/CO2?) gets trapped under the lens, which obviously perturbs the image to an extent that measuring is no longer possible. I'm sure some of you have experienced similar problems. Any thought on how to solve this (in a simple way)? Best regards, Pieter |
![]() |
Dear Pieter
For our acute cerebellar thick slices we used a custom-designed chamber with so-called bubble traps. That is, the tubing with the oxygenated perfusion terminated within a small circular trough at the side of the chamber. Bubbles escaped from this trough while the perfusate streamed through a slit below the liquid surface into the main chamber where the slice was placed. I am sure there are commercial versions available. Cheers Franco Pieter Vanden Berghe wrote: > Dear all, > > we've got an upright microscope with a 20x water dipping lens that we > use for Ca imaging in tissues. Much like a patch clamp experiment we > insert a perfusion pipette between the lens and the tissue to administer > drugs etc. The solutions are all oxygenated CO2/bicarbonate buffered. > Unfortunately it happens (too often) that an bubble (air/O2/CO2?) gets > trapped under the lens, which obviously perturbs the image to an extent > that measuring is no longer possible. > > I'm sure some of you have experienced similar problems. Any thought on > how to solve this (in a simple way)? > > Best regards, > > Pieter > > > -- ___ ___ _______ / / / / / ____/ / / / / / /___ / / / / /___ / / /___ / / ____/ / /______/ /__/ /______/ Dr. Franco Del Principe ------------------------------------------------------------ LIFE IMAGING SERVICES GmbH Fon +41 61 711 6461 Efringerstrasse 79 Fax +41 61 711 6462 CH-4057 Basel, Switzerland Mob +41 79 672 4694 E-mail [hidden email] Web http://www.lis.ch |
![]() |
fyi
At 07:01 AM 9/2/2009, you wrote: >Dear Pieter > >For our acute cerebellar thick slices we used a custom-designed chamber >with so-called bubble traps. That is, the tubing with the oxygenated >perfusion terminated within a small circular trough at the side of the >chamber. Bubbles escaped from this trough while the perfusate streamed >through a slit below the liquid surface into the main chamber where the >slice was placed. I am sure there are commercial versions available. > >Cheers > >Franco > >Pieter Vanden Berghe wrote: >>Dear all, >>we've got an upright microscope with a 20x water dipping lens that we use >>for Ca imaging in tissues. Much like a patch clamp experiment we insert a >>perfusion pipette between the lens and the tissue to administer drugs >>etc. The solutions are all oxygenated CO2/bicarbonate buffered. >>Unfortunately it happens (too often) that an bubble (air/O2/CO2?) gets >>trapped under the lens, which obviously perturbs the image to an extent >>that measuring is no longer possible. >>I'm sure some of you have experienced similar problems. Any thought on >>how to solve this (in a simple way)? >>Best regards, >>Pieter >> > >-- > ___ ___ _______ > / / / / / ____/ > / / / / / /___ > / / / / /___ / > / /___ / / ____/ / > /______/ /__/ /______/ Dr. Franco Del Principe >------------------------------------------------------------ >LIFE IMAGING SERVICES GmbH Fon +41 61 711 6461 >Efringerstrasse 79 Fax +41 61 711 6462 >CH-4057 Basel, Switzerland Mob +41 79 672 4694 >E-mail [hidden email] Web http://www.lis.ch |
Free forum by Nabble | Edit this page |