To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, Would be interested to hear your views on air-conditioning requirements for a confocal microscope rooms. We have central airconditioning with temperature sensors in each microsocpe room but this set up does not keep the temperature sufficiently stable. We are thinking of installing a dedicated air-con unit as used in computer rooms (CRAC). This should maintain the temperature and control humidity but prices for such units are very high. Best regards, Anya Dr Anya Salih Confocal Bio-Imaging Facility Hawkesbury Campus, Bld S8 University of Western Sydney |
|
Gary G. Li-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Dr. Anya Salih, Talk with the vendors of your confocal microscopes to get suggestions, if you have not done so. Regards Gary G. Li, PhD Johns Hopkins School of Medicine Baltimore, MD 21224 On Mon, Dec 20, 2010 at 9:26 PM, Anya Salih <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, Would be interested to hear your views on air-conditioning requirements > for a confocal microscope rooms. > > We have central airconditioning with temperature sensors in each microsocpe > room but this set up does not keep the temperature sufficiently stable. We > are thinking of installing a dedicated air-con unit as used in computer > rooms (CRAC). This should maintain the temperature and control humidity but > prices for such units are very high. > > Best regards, > Anya > > Dr Anya Salih > Confocal Bio-Imaging Facility > Hawkesbury Campus, Bld S8 > University of Western Sydney > |
|
|
In reply to this post by Anya Salih
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Anya, We found that there's a benefit to using a large room (with a large plenum above the acoustical tiles) subdivided into sections using blackout curtains. The large size of the airspace evens out temperature differences. The only important thing to ensure is that the supply registers are NOT above your confocals. This type of setup also promotes a more social microscope experience but that may not be great for everyone. Another option we are using is a modular cleanroom. We're using a biobubble (see: http://www.biobubble.com/biobubble_ppt/img127.html) to house 2 microscopy setups. Airflow and temperature are stabilized. Cheers, - Damir On 12/20/2010 6:26 PM, Anya Salih wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, Would be interested to hear your views on air-conditioning requirements for a confocal microscope rooms. > > We have central airconditioning with temperature sensors in each microsocpe room but this set up does not keep the temperature sufficiently stable. We are thinking of installing a dedicated air-con unit as used in computer rooms (CRAC). This should maintain the temperature and control humidity but prices for such units are very high. > > Best regards, > Anya > > Dr Anya Salih > Confocal Bio-Imaging Facility > Hawkesbury Campus, Bld S8 > University of Western Sydney -- Damir Sudar - Staff Scientist and Deputy for Technology Lawrence Berkeley Laboratory / Life Sciences Division One Cyclotron Road, MS 977R225A, Berkeley, CA 94720, USA T: 510/486-5346 - F: 510/486-5586 - E: [hidden email] WWW: http://www.lbl.gov/lifesciences/labs/sudar_lab.html |
|
Sylvie Le Guyader-2 |
|
|
In reply to this post by Anya Salih
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Anya We had the same problem with our argon and 2 P lasers warming up the room (3 other microscopes) and the air con not being able to stabilize the temperature. Installing above the lasers hoods (no fan, just large metal funnels) that channel the heat to the main ventilation output worked well for us and was cheap. Additionally, you can also hang a curtain around the lasers, but you must make sure that you have the proper ventilation/hood on the laser side of the curtain. This has the advantage that the curtain can be pushed to the side for servicing. It often requires moving the lasers lightly away from the microscope and grouping them together which is not always possible. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet 14183 Huddinge Sweden office: 08 608 9240 from 17.12 +46 (0) 8 5248 1107 new number! LCI room: 08 608 9248 from 17.12 +46 (0) 8 5248 1172 new number! mobile: +46 (0) 73 733 5008 > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Anya Salih > Sent: 21 December 2010 03:26 > To: [hidden email] > Subject: air conditioning requirements > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, Would be interested to hear your views on air-conditioning requirements for a > confocal microscope rooms. > > We have central airconditioning with temperature sensors in each microsocpe room > but this set up does not keep the temperature sufficiently stable. We are thinking > of installing a dedicated air-con unit as used in computer rooms (CRAC). This > should maintain the temperature and control humidity but prices for such units are > very high. > > Best regards, > Anya > > Dr Anya Salih > Confocal Bio-Imaging Facility > Hawkesbury Campus, Bld S8 > University of Western Sydney |
| Free forum by Nabble | Edit this page |