![]() |
Hi, When you do colocalization measurement (in image pro for example) between let say the red and a green channel the co-localization coefficient tells you how much red and green are co-localized. But as much as I understood, it doesn't give you information on the fact that for example red is always co-localized with green but green is not always colocalized with red ? What kind of analysis would I need to compare this between samples ? Thank you in advance, JP - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - Jean-Pierre CLAMME, PhD Senior Scientist Nitto Denko Technical 501 Via Del Monte Oceanside, CA 92058 E-mail: [hidden email] Phone: +760.435.7065 |
![]() |
- Actually, many colocalization programs, including the imageJ plugin "Colocalization finder" will tell you the percent of each channel that is colocalized. see for instance:http://rsb.info.nih.gov/ij/plugins/colocalization-finder.html
-- Julio Vazquez Fred Hutchinson Cancer Research Center Seattle, WA 98109-1024 http://www.fhcrc.org/ On Feb 4, 2009, at 5:05 PM, Jean-Pierre CLAMME wrote:
|
![]() |
In reply to this post by Jean-Pierre CLAMME-2
> But as much as I understood,
> it doesn't give you information on the fact that for example red is always > co-localized with green but green is not always colocalized with red ? What > kind of analysis would I need to compare this between samples ? Mander's coefficients are intended to give this kind of information, see e.g. http://support.svi.nl/wiki/ColocalizationCoefficients (coefficients interpretation at the end of the article). Cheers, jose. |
![]() ![]() |
Yuri Gaidoukevitch |
![]() |
In reply to this post by Jean-Pierre CLAMME-2
Jean-Pierre, Image-Pro provides this information in m1/m2 and M1/M2 coefficients. You can find detailed information in http://www.mediacy.com/pdfs/colocfluorprobes.pdf Regards, Yuri Gaidoukevitch Senior Software Engineer Media Cybernetics Frambozenweg 139 2321KA Leiden The Netherlands tel +31-71-5730639 mobile +31-621580163 fax +31-71-5730640 |
![]() |
In reply to this post by Julio Vazquez
As Jose and Julio point out the data is
readily available in the form you are looking for. A wider question is what you take measurements
to mean. These coefficients need to be qualified by
the area each fluorophore occupies in the ROI: if each occurred in 50% of the
pixels, then chance alone would produce an overlap of 25% (uninteresting), while
overlaps of say 10% or 40% would be more interesting – a non random
mechanism. If say they are both water soluble they
might well appear throughout the cytoplasm and show 100% colocalization (using
some variant of overlap). Also bear in mind that these coefficients
are very sensitive to the threshold used to decide if a fluorophore is
present/absent in individual pixels. A more interesting measure is correlation –
does the variation in the intensity of one fluorophore match that of the
second. This shows whether there is some underlying linkage between the distribution
of the fluorophores. The variation would be caused by inhomegeneity in the
volume of distribution. A correlation measurement for our two fluorophores that
simply appear in the cytoplasm will be very low. In practice use both overlap and
correlation when considering colocalization and, as with all measurements,
always consider exactly what has been measured and how it might be misleading. Jeremy Dr F451a Cell Biologi Wenner-Gren
Inst. The Arhenius
Lab S-106 91 From: - Actually, many colocalization programs, including the imageJ plugin
"Colocalization finder" will tell you the percent of each channel
that is colocalized. see for instance: http://rsb.info.nih.gov/ij/plugins/colocalization-finder.html -- Julio Vazquez http://www.fhcrc.org/
On Feb 4, 2009, at 5:05 PM, Jean-Pierre CLAMME wrote:
|
![]() ![]() |
Daniel James White |
![]() |
In reply to this post by Jean-Pierre CLAMME-2
Dear JP,
You need to measure thresholded Manders coefficients (1 for red , 1 for green!) search this list archives for my previous long posts on this topic. in short, use ImageJ or easier BioImageXD (both free) to calculate threshold using the methods of costes, then calculate the Manders coefficents 1 and 2 then do the costes statistical significance test. Publish: thresholds, manders coefficents, statistical significance, 2D histogram . scatter plots with straight line fit and thresholds. You need to read 2 papers at least Costes et al and Manders et al. I will send them to you in a moment cheers Dan Begin forwarded message: > > Date: Wed, 4 Feb 2009 17:05:30 -0800 > From: Jean-Pierre CLAMME <[hidden email]> > Subject: colocalization question > > This is a multipart message in MIME format. > --=_alternative 0005FF9888257554_= > Content-Type: text/plain; charset="US-ASCII" > > Hi, > > When you do colocalization measurement (in image pro for example) > between > let say the red and a green channel the co-localization coefficient > tells > you how much red and green are co-localized. But as much as I > understood, > it doesn't give you information on the fact that for example red is > always > co-localized with green but green is not always colocalized with red ? > What kind of analysis would I need to compare this between samples ? > > Thank you in advance, > > JP > > > - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - > Jean-Pierre CLAMME, PhD > Senior Scientist > Nitto Denko Technical > 501 Via Del Monte > Oceanside, CA 92058 > E-mail: [hidden email] > Phone: +760.435.7065 Dr. Daniel James White BSc. (Hons.) PhD Senior Microscopist / Image Processing and Analysis Light Microscopy Facility Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstrasse 108 01307 DRESDEN Germany New Mobile Number!!! +49 (0)15114966933 (German Mobile) +49 (0)351 210 2627 (Work phone at MPI-CBG) +49 (0)351 210 1078 (Fax MPI-CBG LMF) http://www.bioimagexd.net http://www.chalkie.org.uk [hidden email] ( [hidden email] ) |
![]() ![]() |
B. Prabhakar Pandian |
![]() |
In reply to this post by Julio Vazquez
Hello,
Does anybody have advice on what substrates (Gelatin/Lysine/Collagen/Etc.) is best for growing adherent cells on glass slides. Is there a study where cell differentiation was compared on various substrates. Any help would be great. Thanks, -Prabhakar |
![]() ![]() |
Jean-Pierre CLAMME-2 |
![]() |
In reply to this post by Daniel James White
Thank you all very much. Your inputs helped a lot ! JP - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - Jean-Pierre CLAMME, PhD Senior Scientist Nitto Denko Technical 501 Via Del Monte Oceanside, CA 92058 E-mail: [hidden email] Phone: +760.435.7065 Confocal Microscopy List <[hidden email]> wrote on 02/05/2009 03:00:41 AM: > Dear JP, > You need to measure thresholded Manders coefficients (1 for red , 1 > for green!) > search this list archives for my previous long posts on this topic. > > in short, use ImageJ or easier BioImageXD (both free) > to calculate threshold using the methods of costes, > then calculate the Manders coefficents 1 and 2 > then do the costes statistical significance test. > Publish: > thresholds, manders coefficents, statistical significance, > 2D histogram . scatter plots with straight line fit and thresholds. > You need to read 2 papers at least > Costes et al > and Manders et al. > I will send them to you in a moment > cheers > Dan > > Begin forwarded message: > > > > Date: Wed, 4 Feb 2009 17:05:30 -0800 > > From: Jean-Pierre CLAMME <[hidden email]> > > Subject: colocalization question > > > > This is a multipart message in MIME format. > > --=_alternative 0005FF9888257554_= > > Content-Type: text/plain; charset="US-ASCII" > > > > Hi, > > > > When you do colocalization measurement (in image pro for example) > > between > > let say the red and a green channel the co-localization coefficient > > tells > > you how much red and green are co-localized. But as much as I > > understood, > > it doesn't give you information on the fact that for example red is > > always > > co-localized with green but green is not always colocalized with red ? > > What kind of analysis would I need to compare this between samples ? > > > > Thank you in advance, > > > > JP > > > > > > - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - > > Jean-Pierre CLAMME, PhD > > Senior Scientist > > Nitto Denko Technical > > 501 Via Del Monte > > Oceanside, CA 92058 > > E-mail: [hidden email] > > Phone: +760.435.7065 > Dr. Daniel James White BSc. (Hons.) PhD > Senior Microscopist / Image Processing and Analysis > Light Microscopy Facility > Max Planck Institute of Molecular Cell Biology and Genetics > Pfotenhauerstrasse 108 > 01307 DRESDEN > Germany > > New Mobile Number!!! > +49 (0)15114966933 (German Mobile) > +49 (0)351 210 2627 (Work phone at MPI-CBG) > +49 (0)351 210 1078 (Fax MPI-CBG LMF) > http://www.bioimagexd.net > http://www.chalkie.org.uk > [hidden email] > ( [hidden email] ) |
![]() ![]() |
Anchall ............ |
![]() |
In reply to this post by B. Prabhakar Pandian
Transfection microarray of human mesenchymal stem cells and on-chip siRNA gene knockdown
On Thu, Feb 5, 2009 at 5:56 PM, Anchall ............ <[hidden email]> wrote: This study they have tested PLL ( poly L-Lysine),fibronectin and gelatin.. |
![]() ![]() |
Bowman, Doug |
![]() |
Not specifically a confocal question: Any opinions on commercially available
media/drug perfusion system for multi-well plate live-cell imaging? We are looking for a system that would
allow for media exchange / compound addition during a timelapse experiment
without having to remove the plate from the environmental enclosure. I am
familiar with the Wafergen system and Fluxion BioFlux 200 products – but
only from a web search. Any opinions on either system or other
companies would be great.
This e-mail, including any attachments, is a confidential business communication, and may contain information that is confidential, proprietary and/or privileged. This e-mail is intended only for the individual(s) to whom it is addressed, and may not be saved, copied, printed, disclosed or used by anyone else. If you are not the(an) intended recipient, please immediately delete this e-mail from your computer system and notify the sender. Thank you. |
![]() ![]() |
Kurt Thorn |
![]() |
There is a system similar to the BioFlux from CellASIC
(http://www.cellasic.com/). I've worked with both the BioFlux and the CellASIC and both work well and are easy to use. Kurt Bowman, Doug wrote: > > Not specifically a confocal question: > > Any opinions on commercially available media/drug perfusion system for > multi-well plate live-cell imaging? > > We are looking for a system that would allow for media exchange / > compound addition during a timelapse experiment without having to > remove the plate from the environmental enclosure. I am familiar with > the Wafergen system and Fluxion BioFlux 200 products – but only from a > web search. > > Any opinions on either system or other companies would be great. > > > Thanks, > Doug > > > > This e-mail, including any attachments, is a confidential business communication, and may contain information that is confidential, proprietary and/or privileged. This e-mail is intended only for the individual(s) to whom it is addressed, and may not be saved, copied, printed, disclosed or used by anyone else. If you are not the(an) intended recipient, please immediately delete this e-mail from your computer system and notify the sender. Thank you. > -- Kurt Thorn, PhD Director, Nikon Imaging Center University of California San Francisco UCSF MC 2140 Genentech Hall Room S252 600 16th St. San Francisco, CA 94158-2517 http://nic.ucsf.edu phone 415.514.9709 fax 415.514.4300 |
![]() |
In reply to this post by Jean-Pierre CLAMME-2
Howdy,
Sorry to chip in a bit late to this discussion but one thing you need to always
remember with colocalisation of fluorecent images, that is resolution. Sorry to
those that already know all this but is is an important point to emphasise when
talking about colocalisation.
Resolution is governed by NA and can be calculated using
0.61 x emission wavlength/NA (for widfield) and 0.4 x emision wavelength/NA (for
confocal). There are a few variations of this formula, and while they give
theoretical best resolutions, the resolution acheived in practice is usually
always worse. Also this is only for lateral resolution, axial resolution is
worse by a facotr of about 3.
So for a 60x 1.4NA objective on a confocal your best
lateral resolution for say GFP is about 150nm (220nm for widefield). That means
that your proteins could be 200nm apart and still "colocalise". 200nm (2000
angstroms) is a rather large distance when you are talking about protein
interactions. For example a human herpes virus is aorund 200nm across, there are
a lot of proteins in that small particle, but not all of them are directly
interating with each other.
Normally when people are doing colocalisation they have
immunoprecipitation or other such data to show the proteins they are interested
in interact, couple this with the fluorecent colocalisation and it is much more
convincing. The other choice is FRET imaging, as FRET requires proteins to be
1-10nm appart for it to occur. So even though you can't resolve the protein with
a confocal you can indirectly measure its interaction.
If you want an accurate image to show proteins interacting
you have to use immuno EM as EM can give you the resolution you
need.
Sorry for the rant :)
Cam
Cameron J.
Nowell From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jean-Pierre CLAMME Sent: Friday, 6 February 2009 3:57 AM To: [hidden email] Subject: Re: colocalization question Thank you all very much. Your inputs helped a lot ! JP - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - Jean-Pierre CLAMME, PhD Senior Scientist Nitto Denko Technical 501 Via Del Monte Oceanside, CA 92058 E-mail: [hidden email] Phone: +760.435.7065 Confocal Microscopy List <[hidden email]> wrote on 02/05/2009 03:00:41 AM: > Dear JP, > You need to measure thresholded Manders coefficients (1 for red , 1 > for green!) > search this list archives for my previous long posts on this topic. > > in short, use ImageJ or easier BioImageXD (both free) > to calculate threshold using the methods of costes, > then calculate the Manders coefficents 1 and 2 > then do the costes statistical significance test. > Publish: > thresholds, manders coefficents, statistical significance, > 2D histogram . scatter plots with straight line fit and thresholds. > You need to read 2 papers at least > Costes et al > and Manders et al. > I will send them to you in a moment > cheers > Dan > > Begin forwarded message: > > > > Date: Wed, 4 Feb 2009 17:05:30 -0800 > > From: Jean-Pierre CLAMME <[hidden email]> > > Subject: colocalization question > > > > This is a multipart message in MIME format. > > --=_alternative 0005FF9888257554_= > > Content-Type: text/plain; charset="US-ASCII" > > > > Hi, > > > > When you do colocalization measurement (in image pro for example) > > between > > let say the red and a green channel the co-localization coefficient > > tells > > you how much red and green are co-localized. But as much as I > > understood, > > it doesn't give you information on the fact that for example red is > > always > > co-localized with green but green is not always colocalized with red ? > > What kind of analysis would I need to compare this between samples ? > > > > Thank you in advance, > > > > JP > > > > > > - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - > > Jean-Pierre CLAMME, PhD > > Senior Scientist > > Nitto Denko Technical > > 501 Via Del Monte > > Oceanside, CA 92058 > > E-mail: [hidden email] > > Phone: +760.435.7065 > Dr. Daniel James White BSc. (Hons.) PhD > Senior Microscopist / Image Processing and Analysis > Light Microscopy Facility > Max Planck Institute of Molecular Cell Biology and Genetics > Pfotenhauerstrasse 108 > 01307 DRESDEN > Germany > > New Mobile Number!!! > +49 (0)15114966933 (German Mobile) > +49 (0)351 210 2627 (Work phone at MPI-CBG) > +49 (0)351 210 1078 (Fax MPI-CBG LMF) > http://www.bioimagexd.net > http://www.chalkie.org.uk > [hidden email] > ( [hidden email] ) This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. |
![]() |
Cameron Nowell wrote:
> That means that your proteins could be 200nm apart and still "colocalise". Some images to illustrate that can be found at http://support.svi.nl/wiki/BlurAndNoiseAffectColocalization Maybe Dr. Nowell wants to quote that email in that wiki article! |
![]() |
In reply to this post by Jean-Pierre CLAMME-2
Indeed, this kind of information is expressed by M1 and M2 (what the
community named the "Manders' coefficients"). However, be aware that you should be very careful with your conclusions from your measurements. Like all co-localization coefficients these coefficients are very sensitive for off-set and cross-talk. A high offset gives a high background and background correlates and may co-localize with backgound and this will strongly bias your coefficient (especially in 3D data-sets since there are lot of background pixels in 3D-images). Cross-talk is undistinguishable from co-localisation and should therefore be reduced during scanning (sequential scanning) of removed by post-processing. So you should allways measure your cross-talk in control experiments (single colour labelling) with the same microscope settings (laser power, pinhole, PMT voltage, etc) as your real experiment. Another point of attention is the threshold that has to be set in the Manders' coefficients. This may be a very critical setting, so be careful and do your control experiments. The numbers that come out should be compared with these control experiments. And this counts for all colocalisation coefficients. You should compare the numbers with the numbers of control experiments. There are two kind of experiments that you may do. 1) Biological control experiments: e.g. Do another pair of stainings of which you know that they should colocalize and do a pair of stainings of which you know that they cannot colocalize at all. Then measure your relative "zero" and "one". 2) Image processing experiment (only possible to measure your "zero"). Take two dual-colour images of look-alike cells. Take the red component of the first and measure the colocalisation with the green component of the second image. Repeat this for some more combinations of images and see what the colocalisation coefficents are of two uncorrelated patterns. It is clear that you apply the same post-processing (cross-talk reduction, noise reduction, off-set reduction, deconvolution, etc) and colocalisation measurement (thresholds, etc). So, my take home message is: It is easy to get numbers from an image. The interpretation of these numbers needs a lot of attention. Good luck and succes with your research, Erik Manders |
![]() |
Hi,
I would appreciate any input on the Leica Tandem Scanner confocal microscope (pros and cons). Many thanks ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara |
![]() |
Hello, what are we talking about here? I've used the Leica STED
(and I'm seriously impressed with it) but I didn't know Leica made a tandem scanner. Please clarify! I'm presenting results from the Leica STED at NZCM next week in Rotorua and in April at FOM in Krakow. I do realise both are a bit remote from Brazil. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Friday, 6 February 2009 9:37 PM To: [hidden email] Subject: Input on the Leica STED Hi, I would appreciate any input on the Leica Tandem Scanner confocal microscope (pros and cons). Many thanks ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM |
![]() ![]() |
simon walker (BI) |
![]() |
Presumably we're talking about the SP5 with conventional and resonant scanners and adapted for STED..?
-----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: 06 February 2009 11:41 To: [hidden email] Subject: Re: Input on the Leica STED Hello, what are we talking about here? I've used the Leica STED (and I'm seriously impressed with it) but I didn't know Leica made a tandem scanner. Please clarify! I'm presenting results from the Leica STED at NZCM next week in Rotorua and in April at FOM in Krakow. I do realise both are a bit remote from Brazil. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Friday, 6 February 2009 9:37 PM To: [hidden email] Subject: Input on the Leica STED Hi, I would appreciate any input on the Leica Tandem Scanner confocal microscope (pros and cons). Many thanks ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM |
![]() |
In reply to this post by Guy Cox
Hi, tandem scanner means you have a slider with both the galvano scanner
and a resonant scanner. The position of the slider (which scanner will be used) can be choosen during system initation. So kind of like the Nikon A1R, just that you cannot use both in parallel with the Leica SP5. But this whole thing is independet from STED. So I don't know what we are talking about :). Michael > Hello, what are we talking about here? I've used the Leica STED > (and I'm seriously impressed with it) but I didn't know Leica made > a tandem scanner. Please clarify! I'm presenting results from > the Leica STED at NZCM next week in Rotorua and in April at FOM in > Krakow. I do realise both are a bit remote from Brazil. > > > Guy > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Electron Microscope Unit, Madsen Building F09, > University of Sydney, NSW 2006 > ______________________________________________ > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Renato Mortara > Sent: Friday, 6 February 2009 9:37 PM > To: [hidden email] > Subject: Input on the Leica STED > > Hi, > > I would appreciate any input on the Leica Tandem Scanner confocal > microscope (pros and cons). > > Many thanks ! > > Renato > > Renato A. Mortara > Parasitology Division > UNIFESP - Escola Paulista de Medicina > Rua Botucatu, 862, 6th floor > São Paulo, SP > 04023-062 > Brazil > Phone: 55 11 5579-8306 > Fax: 55 11 5571-1095 > email: [hidden email] > home page: www.ecb.epm.br/~ramortara > > > No virus found in this incoming message. > Checked by AVG. > Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: > 4/02/2009 4:35 PM > > > No virus found in this outgoing message. > Checked by AVG. > Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: > 4/02/2009 4:35 PM |
![]() |
In reply to this post by simon walker (BI)
Well, tandem scanning was the term coined by Mojimir Petran for
his spinning-disk confocal system. Ever since then there have been companies offering similar systems. Leica used to offer such a system (based on the BD CARV) but have since discontinued it. But it is something very different in all respects (including both price and performance) from the Leica STED which is a super- resolution system offering 85nm resolution. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker (BI) Sent: Friday, 6 February 2009 10:46 PM To: [hidden email] Subject: Re: Input on the Leica STED Presumably we're talking about the SP5 with conventional and resonant scanners and adapted for STED..? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: 06 February 2009 11:41 To: [hidden email] Subject: Re: Input on the Leica STED Hello, what are we talking about here? I've used the Leica STED (and I'm seriously impressed with it) but I didn't know Leica made a tandem scanner. Please clarify! I'm presenting results from the Leica STED at NZCM next week in Rotorua and in April at FOM in Krakow. I do realise both are a bit remote from Brazil. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Friday, 6 February 2009 9:37 PM To: [hidden email] Subject: Input on the Leica STED Hi, I would appreciate any input on the Leica Tandem Scanner confocal microscope (pros and cons). Many thanks ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM |
![]() ![]() |
Renato A. Mortara |
![]() |
In reply to this post by simon walker (BI)
Hope it is slightly clearer...
Leica is offering the STED system that that can be an option to their TCS SP5 confocal. Besides, they also offer a tandem scanner for the same instrument, presumably to achieve higher acquisitions speed. Thanks again ! Renato -----Mensagem original----- De: Confocal Microscopy List [mailto:[hidden email]] Em nome de simon walker (BI) Enviada em: sexta-feira, 6 de fevereiro de 2009 09:46 Para: [hidden email] Assunto: Re: Input on the Leica STED Presumably we're talking about the SP5 with conventional and resonant scanners and adapted for STED..? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: 06 February 2009 11:41 To: [hidden email] Subject: Re: Input on the Leica STED Hello, what are we talking about here? I've used the Leica STED (and I'm seriously impressed with it) but I didn't know Leica made a tandem scanner. Please clarify! I'm presenting results from the Leica STED at NZCM next week in Rotorua and in April at FOM in Krakow. I do realise both are a bit remote from Brazil. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Friday, 6 February 2009 9:37 PM To: [hidden email] Subject: Input on the Leica STED Hi, I would appreciate any input on the Leica Tandem Scanner confocal microscope (pros and cons). Many thanks ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.18/1935 - Release Date: 4/02/2009 4:35 PM |
Free forum by Nabble | Edit this page |