I would like to perform simple FRET measurements (with a three-filter method) to detect interaction between two proteins in living cells. Most of the people use a CFP/YFP couple (or variants like mCerrulean/mCitrine), but I would prefer to use an eGFP/mCherry couple. Do you know if this pair would be good for FRET ? There was a paper showing that it could be the case (
http://www.ncbi.nlm.nih.gov/sites/entrez?Db=pubmed&Cmd=ShowDetailView&TermToSearch=16941642
) but there was also a discussion on this list saying that mCherry behaves poorly as a FRET acceptor, due to its amino and carboxy terminus sequences (
http://listserv.buffalo.edu/cgi-bin/wa?A2=ind0707&L=confocal&D=0&P=16580).
So what is your advice, should I try to see eGFP/mCherry FRET or directly order a three-filer CFP/YFP set and start to switch the FPs on my constructs ? Another question : do you have a good positive control for FRET that works well in hippocampal neuronal cultures ? Thank you for your advices, Christophe Leterrier Postdoc INSERM UMR641 Neurobiology of ionic channels IFR Jean Roche - Mediterranee University Marseille, France |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Christophe, We recently published a paper (PMID: 17925859) on red FRET acceptors: http://dx.doi.org/10.1371/journal.pone.0001011 We concluded that YFP is preffered as donor over GFP when mCherry is the acceptor, based on R0 values (the YFP-mCherry construct is characterized in the above-mentioned paper and is available as a positive control) In your filter-cube approach, you are essentially measuring sensitized emission. So you would benefit from an acceptor with a high quantum yield. (This is why CFP-YFP is such a nice FRET pair). If you want to stick to GFP as a donor, you're probably better of using tagRFP or mStrawberry as an acceptor. Regards, Joachim. Dr. Ir. Joachim Goedhart Section Molecular Cytology Swammerdam Institute for Life Sciences University of Amsterdam Kruislaan 316 NL-1098 SM Amsterdam The Netherlands Tel: +31(0)20 525 7774 Fax: +31(0)20 525 6271 http://wwwmc.bio.uva.nl/~joachim |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Christophe, Adding to Joachim comments, you should also consider that red-shifted donor-acceptor pairs show wider tails of excitation and emission. Thus, if your are thinking on sensitized emission to measure FRET, be aware that cross-excitation (acceptor) and bleed-through (donor) values are higher than those of CFP-YFP. For a discussion have a look at Fluorescent protein FRET: the good, the bad and the ugly.(2007) Piston, DW and Kremers, GJ. Trends Biochem Sci. 32(9):407-14. Regards, At 09:22 25/01/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear Christophe, > >We recently published a paper (PMID: 17925859) on red FRET acceptors: >http://dx.doi.org/10.1371/journal.pone.0001011 >We concluded that YFP is preffered as donor over >GFP when mCherry is the acceptor, based on R0 >values (the YFP-mCherry construct is >characterized in the above-mentioned paper and is available >as a positive control) >In your filter-cube approach, you are >essentially measuring sensitized emission. So you would >benefit from an acceptor with a high quantum >yield. (This is why CFP-YFP is such a nice FRET pair). >If you want to stick to GFP as a donor, you're >probably better of using tagRFP or mStrawberry as an >acceptor. > >Regards, >Joachim. > >Dr. Ir. Joachim Goedhart >Section Molecular Cytology >Swammerdam Institute for Life Sciences >University of Amsterdam >Kruislaan 316 >NL-1098 SM Amsterdam >The Netherlands >Tel: +31(0)20 525 7774 >Fax: +31(0)20 525 6271 >http://wwwmc.bio.uva.nl/~joachim F Javier Diez-Guerra, PhD Profesor Titular Centro de Biologia Molecular Severo Ochoa C/ Nicolás Cabrera, 1 Universidad Autónoma Ctra Colmenar Viejo Km 15 Cantoblanco, 28049 Madrid SPAIN phone: +34 91 196 4612 e-mail: [hidden email] |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Thank you for these usefull advices and references. Do anyone know
where (web, article...) I can find primers and hints on how to design
fusion proteins to optimize the chance to observe FRET ?
Christophe Leterrier On Jan 25, 2008 10:05 AM, F Javier Diez Guerra <[hidden email]> wrote: Christophe, |