field service for Coherent and SP NIR lasers

classic Classic list List threaded Threaded
♦
♦
3 messages Options
Gavin Rumbaugh Gavin Rumbaugh
Reply | Threaded
Open this post in threaded view
♦
♦
|

field service for Coherent and SP NIR lasers

Hello,

This message is in response to the suggestion that Spectra Physics NIR pulsed lasers for multi-photon microscopy nearly always require the user to send the laser cavity back to the factory for service. In my experience, this is not the case.

I have used Coherent and SP lasers (Chameleon Series and Mai Tai Series) for more than seven years now. Routine service to each laser can be performed in the field. Actually, Coherent lasers are more "closed" than similar product from SP.  This is why Coherent has a laser replacement policy, because most work requires the product to shipped to their factory in Scotland (as was explained to me by a Coherent engineer). In contrast, SP Mai Tai is more modular in its design, allowing major service and even technological upgrades in the field. In my case, SP was able to  upgrade the Mai Tai HP to a DeepSee version in the field. This upgrade took a few hours, and my system was only down for an afternoon. I have now been informed by SP that its is very likely that my laser can be further field-upgraded to the EHP version. Thus, because of SP's modular design, service is less of an intrusion and the user has an upgrade path. This upgradability is not necessary available on similar Coherent products, which is a plus for SP, as pulse-NIR laser technology evolves rapidly.

I have no bias toward either company. I currently own each laser and am completely happy both the Vision II and the DeepSee. I only wish to provide an honest comparison of both companies; from sales to use to service. The companies are both excellent and provide top-notch sales and service support. The technology that each offer is similar (i would give the edge to SP now, but this could change in six months). One major difference between the companies that i have seen is in their sales structure. It appears that Coherent uses more of a sales rep structure, while SP uses field-sales engineers. The former being salesman with knowledge of lasers and the later being engineers (or PhDs ) with knowledge of sales. Personally, i would rather work with a scientist or engineer.

Any one with similar or different experiences?

All the best, Gavin


Gavin Rumbaugh, PhD
Assistant Professor
Department of Neuroscience, #3C2
The Scripps Research Institute
Jupiter, FL    33458
205-306-7096
[hidden email]
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: field service for Coherent and SP NIR lasers

Hello,
i just want to add that Spectra Physics (UK) also offered me to get the laser under a service contract after it was out of a contract for a while. They checked the laser (working hours, different parameters) and then offered the service contract similar to what George reported from Coherent. So far i had  only good experiences with Spectra Physics, the local service engineer always responds quickly if there is an issue. So far all problems could be quickly solved over the phone. We also got the deepsee upgrade in the field.

best wishes

Andreas



-----Original Message-----
From: Gavin Rumbaugh <[hidden email]>
To: [hidden email]
Sent: Mon, 12 Jul 2010 14:33
Subject: field service for Coherent and SP NIR lasers

Hello,

This message is in response to the suggestion that Spectra Physics NIR pulsed
lasers for multi-photon microscopy nearly always require the user to send the
laser cavity back to the factory for service. In my experience, this is not the
case.

I have used Coherent and SP lasers (Chameleon Series and Mai Tai Series) for
more than seven years now. Routine service to each laser can be performed in the
field. Actually, Coherent lasers are more "closed" than similar product from SP.
This is why Coherent has a laser replacement policy, because most work requires
the product to shipped to their factory in Scotland (as was explained to me by a
Coherent engineer). In contrast, SP Mai Tai is more modular in its design,
allowing major service and even technological upgrades in the field. In my case,
SP was able to upgrade the Mai Tai HP to a DeepSee version in the field. This
upgrade took a few hours, and my system was only down for an afternoon. I have
now been informed by SP that its is very likely that my laser can be further
field-upgraded to the EHP version. Thus, because of SP's modular design, service
is less of an intrusion and the user has an upgrade path. This upgradability is
not necessary available on similar Coherent products, which is a plus for SP, as
pulse-NIR laser technology evolves rapidly.

I have no bias toward either company. I currently own each laser and am
completely happy both the Vision II and the DeepSee. I only wish to provide an
honest comparison of both companies; from sales to use to service. The companies
are both excellent and provide top-notch sales and service support. The
technology that each offer is similar (i would give the edge to SP now, but this
could change in six months). One major difference between the companies that i
have seen is in their sales structure. It appears that Coherent uses more of a
sales rep structure, while SP uses field-sales engineers. The former being
salesman with knowledge of lasers and the later being engineers (or PhDs ) with
knowledge of sales. Personally, i would rather work with a scientist or
engineer.

Any one with similar or different experiences?

All the best, Gavin


Gavin Rumbaugh, PhD
Assistant Professor
Department of Neuroscience, #3C2
The Scripps Research Institute
Jupiter, FL 33458
205-306-7096
[hidden email]=
Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: field service for Coherent and SP NIR lasers

Hi, I can add that my experience was identical to George’s and that Coherent replaced our Chameleon in a manner that resulted in our 2P capabilities being down for just a few hours.

Chameleon support has been excellent.

 

I would not be too surprised to hear that Spectra Physics matches this service.

 

Cheers,

 

Brian D Armstrong PhD

Light Microscopy Core Manager

Beckman Research Institute

City of Hope

Dept of Neuroscience

1450 E Duarte Rd

Duarte, CA 91010

626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andreas Bruckbauer
Sent: Tuesday, July 13, 2010 2:28 AM
To: [hidden email]
Subject: Re: field service for Coherent and SP NIR lasers

 

Hello,
i just want to add that Spectra Physics (UK) also offered me to get the laser under a service contract after it was out of a contract for a while. They checked the laser (working hours, different parameters) and then offered the service contract similar to what George reported from Coherent. So far i had  only good experiences with Spectra Physics, the local service engineer always responds quickly if there is an issue. So far all problems could be quickly solved over the phone. We also got the deepsee upgrade in the field.


best wishes

Andreas

 

 

 

-----Original Message-----
From: Gavin Rumbaugh <[hidden email]>
To: [hidden email]
Sent: Mon, 12 Jul 2010 14:33
Subject: field service for Coherent and SP NIR lasers

Hello,

 


 
This message is in response to the suggestion that Spectra Physics NIR pulsed 

 
lasers for multi-photon microscopy nearly always require the user to send the 

 
laser cavity back to the factory for service. In my experience, this is not the 

 
case. 

 


 
I have used Coherent and SP lasers (Chameleon Series and Mai Tai Series) for 

 
more than seven years now. Routine service to each laser can be performed in the 

 
field. Actually, Coherent lasers are more "closed" than similar product from SP.  

 
This is why Coherent has a laser replacement policy, because most work requires 

 
the product to shipped to their factory in Scotland (as was explained to me by a 

 
Coherent engineer). In contrast, SP Mai Tai is more modular in its design, 

 
allowing major service and even technological upgrades in the field. In my case, 

 
SP was able to  upgrade the Mai Tai HP to a DeepSee version in the field. This 

 
upgrade took a few hours, and my system was only down for an afternoon. I have 

 
now been informed by SP that its is very likely that my laser can be further 

 
field-upgraded to the EHP version. Thus, because of SP's modular design, service 

 
is less of an intrusion and the user has an upgrade path. This upgradability is 

 
not necessary available on similar Coherent products, which is a plus for SP, as 

 
pulse-NIR laser technology evolves rapidly.

 


 
I have no bias toward either company. I currently own each laser and am 

 
completely happy both the Vision II and the DeepSee. I only wish to provide an 

 
honest comparison of both companies; from sales to use to service. The companies 

 
are both excellent and provide top-notch sales and service support. The 

 
technology that each offer is similar (i would give the edge to SP now, but this 

 
could change in six months). One major difference between the companies that i 

 
have seen is in their sales structure. It appears that Coherent uses more of a 

 
sales rep structure, while SP uses field-sales engineers. The former being 

 
salesman with knowledge of lasers and the later being engineers (or PhDs ) with 

 
knowledge of sales. Personally, i would rather work with a scientist or 

 
engineer.

 


 
Any one with similar or different experiences?

 


 
All the best, Gavin 

 


 


 
Gavin Rumbaugh, PhD

 
Assistant Professor

 
Department of Neuroscience, #3C2

 
The Scripps Research Institute

 
Jupiter, FL    33458

 
205-306-7096

 
[hidden email]=

 


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.
---------------------------------------------------------------------