fluorescent Ab counterstain for chicken somites (lectins?)

classic Classic list List threaded Threaded
♦
♦
9 messages Options
matthewmakes matthewmakes
Reply | Threaded
Open this post in threaded view
♦
♦
|

fluorescent Ab counterstain for chicken somites (lectins?)

hi.

i'm interested in picking up a quick, simple and fairly affordable fluorescent counterstain for viewing cell morphology in antibody labeled sections of chicken embryos.  from what i can put together, some sort of lectin would likely do the job, but i'm having a hard time finding material with lectin labels used in the chicken embryo.  i've seen some papers with nice wheat germ agglutinin labels in similar staged somites of mice.  i primarily need this as an Ab counterstain, but if i could get something that would work live as well, that would be great.  (i've got a collection of membrane labels for live imaging, but more options are always nice.)

any advice would be greatly appreciated.

thank you,  matthew


See how Windows Mobile brings your life together—at home, work, or on the go. See Now
Carol Heckman Carol Heckman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: fluorescent Ab counterstain for chicken somites (lectins?)

Matthew-
I don't seen any reason why the wheat germ agglutinin labels wouldn't work.  The cell membrane protein structure might be highly conserved between mouse and chicken.
Carol
________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Matthew Smith [[hidden email]]
Sent: Tuesday, September 30, 2008 9:12 PM
To: [hidden email]
Subject: fluorescent Ab counterstain for chicken somites (lectins?)

hi.

i'm interested in picking up a quick, simple and fairly affordable fluorescent counterstain for viewing cell morphology in antibody labeled sections of chicken embryos.  from what i can put together, some sort of lectin would likely do the job, but i'm having a hard time finding material with lectin labels used in the chicken embryo.  i've seen some papers with nice wheat germ agglutinin labels in similar staged somites of mice.  i primarily need this as an Ab counterstain, but if i could get something that would work live as well, that would be great.  (i've got a collection of membrane labels for live imaging, but more options are always nice.)

any advice would be greatly appreciated.

thank you,  matthew

________________________________
See how Windows Mobile brings your life together—at home, work, or on the go. See Now<http://clk.atdmt.com/MRT/go/msnnkwxp1020093182mrt/direct/01/>
Koo, Lily (NIH/NIAID) [E] Koo, Lily (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

cell membrane image

Dear all,

When we imaging erythrocytes labeled with a membrane dye on a SP5
confocal, we noticed that in all images, the top and bottom of the cell
membrane appeared much dimmer than the left and right sides of it.  Is
there a particular reason for this observation?  Is there a difference
in the resolution of the x-scan vs. y-scan?

Thanks much.

Lily
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cell membrane image

Interesting observation. Assuming the PSF is not assymetric in x-y,
could this be due to the interaction of the orientation of the dye
molecules and the polarization of the laser?

Regards Mark Cannell


Koo, Lily (NIH/NIAID) [F] wrote:

> Dear all,
>
> When we imaging erythrocytes labeled with a membrane dye on a SP5
> confocal, we noticed that in all images, the top and bottom of the cell
> membrane appeared much dimmer than the left and right sides of it.  Is
> there a particular reason for this observation?  Is there a difference
> in the resolution of the x-scan vs. y-scan?
>
> Thanks much.
>
> Lily
>  
Goodhouse, Joseph G. Goodhouse, Joseph G.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cell membrane image

In reply to this post by Koo, Lily (NIH/NIAID) [E]
        We see they same thing in dye labeled lipid vesicles.  It has to
do with the dipole orientation with respect to the laser polarization.
Taking 2 images using a quarter wave plate with one should show you
equal presence.  You may also see this effect if you use the beam
rotator on the SP5


Joe Goodhouse
Confocal Core Lab Manager
Dept. of Molecular Biology
Princeton University
609-258-5432

Visit us at http://www.molbio1.princeton.edu/facility/confocal/   


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Koo, Lily (NIH/NIAID) [F]
Sent: Thursday, October 02, 2008 4:40 PM
To: [hidden email]
Subject: cell membrane image

Dear all,

When we imaging erythrocytes labeled with a membrane dye on a SP5
confocal, we noticed that in all images, the top and bottom of the cell
membrane appeared much dimmer than the left and right sides of it.  Is
there a particular reason for this observation?  Is there a difference
in the resolution of the x-scan vs. y-scan?

Thanks much.

Lily
John Oreopoulos John Oreopoulos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cell membrane image

There have been two replies now on this potentially being a light polarization and dye photoselection/dichroism interaction. There are ways to verify this. 
First, what membrane probe are you using? Is it one of the carbocyanine dyes like DiI? Dan Axelrod showed many years ago that probes like this orient a particular way in the membrane and that this effect can be exploited several ways (See, Axelrod Biophysical Journal, 1979). Try rotating your sample 90 degrees on the microscope stage and see if the cells are still lighting up the same way. If so, then you can be sure that photoselection is occurring.

You can check the degree and direction of polarization of your microscope laser source by placing film polarizer (polaroid film) direction on the stage and rotating it. Remove the objective for this measurement and set the microscope to do a point scan in the middle of the scan field. If when you rotate the film with its reference direction initially pointing North-South to East-West and the laser light coming through disappears, then you can be sure that your laser beam is strongly polarized in the NS direction (and therefore photoselection would occur in the top-bottom direction of your image if the dye has a dipole that orients perpendicular to the membrane, something like DPH or Laurdan. For DiI or DiO, it would be the opposite - left and right side would preferentially light up - since the transition dipoles for this dye tend to align parallel to the membrane). 

Joe's suggestion of inserting a quarter-wave plate into the excitation beam and turning it 45 degrees relative to the laser beam polarization direction should get rid of this problem since by doing so you circularize the polarization of the laser beam, thus equally exciting all directions that are perpendicular to the laser beam direction of travel (ie: anywhere in the xy plane).

This posting by Lily brings up the point that this effect is present all the time in all confocal microscopes - at least the ones that use laser sources (which are usually strongly polarized) and polarization-maintaining optical fibers to deliver the laser beams into the scan head. If your probes are oriented in your sample, you might unintentionally be selecting parts of your sample to light up by fluorescence.


John Oreopoulos, BSc,
PhD Candidate
University of Toronto
Institute For Biomaterials and Biomedical Engineering
Centre For Studies in Molecular Imaging

Tel: W:416-946-5022


On 2-Oct-08, at 5:14 PM, Goodhouse, Joseph G. wrote:

We see they same thing in dye labeled lipid vesicles.  It has to
do with the dipole orientation with respect to the laser polarization.
Taking 2 images using a quarter wave plate with one should show you
equal presence.  You may also see this effect if you use the beam
rotator on the SP5 


Joe Goodhouse
Confocal Core Lab Manager
Dept. of Molecular Biology
Princeton University
609-258-5432



-----Original Message-----
From: Confocal Microscopy List [[hidden email]]
On Behalf Of Koo, Lily (NIH/NIAID) [F]
Sent: Thursday, October 02, 2008 4:40 PM
Subject: cell membrane image

Dear all,

When we imaging erythrocytes labeled with a membrane dye on a SP5
confocal, we noticed that in all images, the top and bottom of the cell
membrane appeared much dimmer than the left and right sides of it.  Is
there a particular reason for this observation?  Is there a difference
in the resolution of the x-scan vs. y-scan?

Thanks much.

Lily





Koo, Lily (NIH/NIAID) [E] Koo, Lily (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cell membrane image

Sincere thanks to John and Joe’s replies.  The Polaroid film idea is neat, I will definitely try it.  Thanks for the quarter-wave plate idea.

 

Lily

 


From: John Oreopoulos [mailto:[hidden email]]
Sent: Thursday, October 02, 2008 5:38 PM
To: [hidden email]
Subject: Re: cell membrane image

 

There have been two replies now on this potentially being a light polarization and dye photoselection/dichroism interaction. There are ways to verify this. 

First, what membrane probe are you using? Is it one of the carbocyanine dyes like DiI? Dan Axelrod showed many years ago that probes like this orient a particular way in the membrane and that this effect can be exploited several ways (See, Axelrod Biophysical Journal, 1979). Try rotating your sample 90 degrees on the microscope stage and see if the cells are still lighting up the same way. If so, then you can be sure that photoselection is occurring.

 

You can check the degree and direction of polarization of your microscope laser source by placing film polarizer (polaroid film) direction on the stage and rotating it. Remove the objective for this measurement and set the microscope to do a point scan in the middle of the scan field. If when you rotate the film with its reference direction initially pointing North-South to East-West and the laser light coming through disappears, then you can be sure that your laser beam is strongly polarized in the NS direction (and therefore photoselection would occur in the top-bottom direction of your image if the dye has a dipole that orients perpendicular to the membrane, something like DPH or Laurdan. For DiI or DiO, it would be the opposite - left and right side would preferentially light up - since the transition dipoles for this dye tend to align parallel to the membrane). 

 

Joe's suggestion of inserting a quarter-wave plate into the excitation beam and turning it 45 degrees relative to the laser beam polarization direction should get rid of this problem since by doing so you circularize the polarization of the laser beam, thus equally exciting all directions that are perpendicular to the laser beam direction of travel (ie: anywhere in the xy plane).

 

This posting by Lily brings up the point that this effect is present all the time in all confocal microscopes - at least the ones that use laser sources (which are usually strongly polarized) and polarization-maintaining optical fibers to deliver the laser beams into the scan head. If your probes are oriented in your sample, you might unintentionally be selecting parts of your sample to light up by fluorescence.

 

 

John Oreopoulos, BSc,

PhD Candidate

University of Toronto

Institute For Biomaterials and Biomedical Engineering

Centre For Studies in Molecular Imaging

 

Tel: W:416-946-5022

 

 

On 2-Oct-08, at 5:14 PM, Goodhouse, Joseph G. wrote:



            We see they same thing in dye labeled lipid vesicles.  It has to

do with the dipole orientation with respect to the laser polarization.

Taking 2 images using a quarter wave plate with one should show you

equal presence.  You may also see this effect if you use the beam

rotator on the SP5 

 

 

Joe Goodhouse

Confocal Core Lab Manager

Dept. of Molecular Biology

Princeton University

609-258-5432

 

 

 

-----Original Message-----

From: Confocal Microscopy List [[hidden email]]

On Behalf Of Koo, Lily (NIH/NIAID) [F]

Sent: Thursday, October 02, 2008 4:40 PM

Subject: cell membrane image

 

Dear all,

 

When we imaging erythrocytes labeled with a membrane dye on a SP5

confocal, we noticed that in all images, the top and bottom of the cell

membrane appeared much dimmer than the left and right sides of it.  Is

there a particular reason for this observation?  Is there a difference

in the resolution of the x-scan vs. y-scan?

 

Thanks much.

 

Lily

 

 




 

Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: fluorescent Ab counterstain for chicken somites (lectins?)

In reply to this post by Carol Heckman
Hi Matthew,
As a point of clarification, lectins will bind specific sequences of sugars
on glycoproteins.  As such, identical proteins with different glyco-residues
will have different lectin binding properties.  Given that, you may need to
do what I had to do years ago, and that was sample a bunch of different
lectins to see what works.  When  lectin chemistry was new, one could find
samplers with small amounts of several lectins to test.  Now, I don't see
any.  Sigma Chemical has many lectins, though I don't know how many have
been fluorescently tagged.  Molecular Probes is another source of labelled
lectins.
Good luck,
Carl

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "Carol Heckman" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, October 02, 2008 12:22 PM
Subject: Re: fluorescent Ab counterstain for chicken somites (lectins?)


Matthew-
I don't seen any reason why the wheat germ agglutinin labels wouldn't work.
The cell membrane protein structure might be highly conserved between mouse
and chicken.
Carol
________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf
Of Matthew Smith [[hidden email]]
Sent: Tuesday, September 30, 2008 9:12 PM
To: [hidden email]
Subject: fluorescent Ab counterstain for chicken somites (lectins?)

hi.

i'm interested in picking up a quick, simple and fairly affordable
fluorescent counterstain for viewing cell morphology in antibody labeled
sections of chicken embryos.  from what i can put together, some sort of
lectin would likely do the job, but i'm having a hard time finding material
with lectin labels used in the chicken embryo.  i've seen some papers with
nice wheat germ agglutinin labels in similar staged somites of mice.  i
primarily need this as an Ab counterstain, but if i could get something that
would work live as well, that would be great.  (i've got a collection of
membrane labels for live imaging, but more options are always nice.)

any advice would be greatly appreciated.

thank you,  matthew

________________________________
See how Windows Mobile brings your life together—at home, work, or on the
go. See Now<http://clk.atdmt.com/MRT/go/msnnkwxp1020093182mrt/direct/01/>
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: fluorescent Ab counterstain for chicken somites (lectins?)

Vector Labs sells lectin kits.

-Esteban


On Fri, Oct 3, 2008 at 1:36 PM, Carl Boswell <[hidden email]> wrote:
Hi Matthew,
As a point of clarification, lectins will bind specific sequences of sugars on glycoproteins.  As such, identical proteins with different glyco-residues will have different lectin binding properties.  Given that, you may need to do what I had to do years ago, and that was sample a bunch of different lectins to see what works.  When  lectin chemistry was new, one could find samplers with small amounts of several lectins to test.  Now, I don't see any.  Sigma Chemical has many lectins, though I don't know how many have been fluorescently tagged.  Molecular Probes is another source of labelled lectins.
Good luck,
Carl

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message ----- From: "Carol Heckman" <[hidden email]> Sent: Thursday, October 02, 2008 12:22 PM
Subject: Re: fluorescent Ab counterstain for chicken somites (lectins?)



Matthew-
I don't seen any reason why the wheat germ agglutinin labels wouldn't work. The cell membrane protein structure might be highly conserved between mouse and chicken.
Carol
________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Matthew Smith [[hidden email]]
Sent: Tuesday, September 30, 2008 9:12 PM
To: [hidden email]
Subject: fluorescent Ab counterstain for chicken somites (lectins?)

hi.

i'm interested in picking up a quick, simple and fairly affordable fluorescent counterstain for viewing cell morphology in antibody labeled sections of chicken embryos.  from what i can put together, some sort of lectin would likely do the job, but i'm having a hard time finding material with lectin labels used in the chicken embryo.  i've seen some papers with nice wheat germ agglutinin labels in similar staged somites of mice.  i primarily need this as an Ab counterstain, but if i could get something that would work live as well, that would be great.  (i've got a collection of membrane labels for live imaging, but more options are always nice.)

any advice would be greatly appreciated.

thank you,  matthew

________________________________
See how Windows Mobile brings your life together—at home, work, or on the go. See Now<http://clk.atdmt.com/MRT/go/msnnkwxp1020093182mrt/direct/01/>



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax