Hi there,

I was looking at multiple vendors' websites for an optical fiber wavelength
combiner (two or three colors), or what is called fused fiber WDM. I have
noticed that most of them are designed for low to medium power level
applications, e.g. up to 300 mW. I was wondering if anyone here has
experience on using these devices, e.g. what is the maximum power it can
take, or is there any high power version of it that can take up to 1 W of
CW in the visible. I am thinking of using it in a STED microscopy setup but
I think 300 mW power limit might be at the lower side for this application.

Thank you very much in advance,
Lu

-----------------------------------------------------
Lu Yan
Nanostructured Fibers and Nonlinear Optics Laboratory
Electrical and Computer Engineering
Boston University
8 St. Mary St., Boston, MA, 02215
617.353.0286 (office) | 617.358.5917 (lab)
[hidden email]
-----------------------------------------------------
Michael Giacomelli Michael Giacomelli
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Lu,

I'm assuming this is single mode?  Putting 1 watt into a few square
microns can be fairly tricky, as the power density is gigantic.  What
wavelength are you using?  I've seen fiber faces blown apart by ~150
mW of 800 nm (femtosecond) ti:saph light.

My bet is that the coupler itself won't be much more vulnerable to
damage than the fiber its made of, but that you might have trouble
just getting that kind of power (depending on wavelength) into a fiber
without ablating the connectors or (depending on defects) your
splices.


Mike

On Fri, Jan 22, 2016 at 3:36 PM, Yan, Lu <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi there,
>
> I was looking at multiple vendors' websites for an optical fiber wavelength
> combiner (two or three colors), or what is called fused fiber WDM. I have
> noticed that most of them are designed for low to medium power level
> applications, e.g. up to 300 mW. I was wondering if anyone here has
> experience on using these devices, e.g. what is the maximum power it can
> take, or is there any high power version of it that can take up to 1 W of
> CW in the visible. I am thinking of using it in a STED microscopy setup but
> I think 300 mW power limit might be at the lower side for this application.
>
> Thank you very much in advance,
> Lu
>
> -----------------------------------------------------
> Lu Yan
> Nanostructured Fibers and Nonlinear Optics Laboratory
> Electrical and Computer Engineering
> Boston University
> 8 St. Mary St., Boston, MA, 02215
> 617.353.0286 (office) | 617.358.5917 (lab)
> [hidden email]
> -----------------------------------------------------
Lu Yan Lu Yan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike,

Yes. It will be single moded. We have coupled few hundreds of mW of Ti:
sapphire laser at around 800 nm (CW though) into a single mode fiber
regularly without burning the fiber facet (we used bare fiber so that if
the fiber burnt, we cleave it and redo the coupling). I am not quite sure
what will be the case for pulse laser. I guess the damaging mechanism
sometimes varies from case to case. For the WDM, I think it might be the
fused coupler that may be vulnerable to damage but I could be completely
wrong about it. Will try to talk to some vendors tomorrow to figure it out.

Thanks,
Lu

-----------------------------------------------------
Lu Yan
Nanostructured Fibers and Nonlinear Optics Laboratory
Electrical and Computer Engineering
Boston University
8 St. Mary St., Boston, MA, 02215
617.353.0286 (office) | 617.358.5917 (lab)
[hidden email]
-----------------------------------------------------

On Fri, Jan 22, 2016 at 4:05 PM, Michael Giacomelli <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Lu,
>
> I'm assuming this is single mode?  Putting 1 watt into a few square
> microns can be fairly tricky, as the power density is gigantic.  What
> wavelength are you using?  I've seen fiber faces blown apart by ~150
> mW of 800 nm (femtosecond) ti:saph light.
>
> My bet is that the coupler itself won't be much more vulnerable to
> damage than the fiber its made of, but that you might have trouble
> just getting that kind of power (depending on wavelength) into a fiber
> without ablating the connectors or (depending on defects) your
> splices.
>
>
> Mike
>
> On Fri, Jan 22, 2016 at 3:36 PM, Yan, Lu <[hidden email]> wrote:
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi there,
> >
> > I was looking at multiple vendors' websites for an optical fiber
> wavelength
> > combiner (two or three colors), or what is called fused fiber WDM. I have
> > noticed that most of them are designed for low to medium power level
> > applications, e.g. up to 300 mW. I was wondering if anyone here has
> > experience on using these devices, e.g. what is the maximum power it can
> > take, or is there any high power version of it that can take up to 1 W of
> > CW in the visible. I am thinking of using it in a STED microscopy setup
> but
> > I think 300 mW power limit might be at the lower side for this
> application.
> >
> > Thank you very much in advance,
> > Lu
> >
> > -----------------------------------------------------
> > Lu Yan
> > Nanostructured Fibers and Nonlinear Optics Laboratory
> > Electrical and Computer Engineering
> > Boston University
> > 8 St. Mary St., Boston, MA, 02215
> > 617.353.0286 (office) | 617.358.5917 (lab)
> > [hidden email]
> > -----------------------------------------------------
>
Julio MATEOS_LANGERAK Julio MATEOS_LANGERAK
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi,

I know qioptiq has SM fibers for coupling up to 500mW in the visible range. They might have more options offline

Best, Julio

Sent from my iPhone

> On 23 Jan 2016, at 00:32, Yan, Lu <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mike,
>
> Yes. It will be single moded. We have coupled few hundreds of mW of Ti:
> sapphire laser at around 800 nm (CW though) into a single mode fiber
> regularly without burning the fiber facet (we used bare fiber so that if
> the fiber burnt, we cleave it and redo the coupling). I am not quite sure
> what will be the case for pulse laser. I guess the damaging mechanism
> sometimes varies from case to case. For the WDM, I think it might be the
> fused coupler that may be vulnerable to damage but I could be completely
> wrong about it. Will try to talk to some vendors tomorrow to figure it out.
>
> Thanks,
> Lu
>
> -----------------------------------------------------
> Lu Yan
> Nanostructured Fibers and Nonlinear Optics Laboratory
> Electrical and Computer Engineering
> Boston University
> 8 St. Mary St., Boston, MA, 02215
> 617.353.0286 (office) | 617.358.5917 (lab)
> [hidden email]
> -----------------------------------------------------
>
>> On Fri, Jan 22, 2016 at 4:05 PM, Michael Giacomelli <[hidden email]> wrote:
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hi Lu,
>>
>> I'm assuming this is single mode?  Putting 1 watt into a few square
>> microns can be fairly tricky, as the power density is gigantic.  What
>> wavelength are you using?  I've seen fiber faces blown apart by ~150
>> mW of 800 nm (femtosecond) ti:saph light.
>>
>> My bet is that the coupler itself won't be much more vulnerable to
>> damage than the fiber its made of, but that you might have trouble
>> just getting that kind of power (depending on wavelength) into a fiber
>> without ablating the connectors or (depending on defects) your
>> splices.
>>
>>
>> Mike
>>
>>> On Fri, Jan 22, 2016 at 3:36 PM, Yan, Lu <[hidden email]> wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your
>> posting.
>>> *****
>>>
>>> Hi there,
>>>
>>> I was looking at multiple vendors' websites for an optical fiber
>> wavelength
>>> combiner (two or three colors), or what is called fused fiber WDM. I have
>>> noticed that most of them are designed for low to medium power level
>>> applications, e.g. up to 300 mW. I was wondering if anyone here has
>>> experience on using these devices, e.g. what is the maximum power it can
>>> take, or is there any high power version of it that can take up to 1 W of
>>> CW in the visible. I am thinking of using it in a STED microscopy setup
>> but
>>> I think 300 mW power limit might be at the lower side for this
>> application.
>>>
>>> Thank you very much in advance,
>>> Lu
>>>
>>> -----------------------------------------------------
>>> Lu Yan
>>> Nanostructured Fibers and Nonlinear Optics Laboratory
>>> Electrical and Computer Engineering
>>> Boston University
>>> 8 St. Mary St., Boston, MA, 02215
>>> 617.353.0286 (office) | 617.358.5917 (lab)
>>> [hidden email]
>>> -----------------------------------------------------
>>
Lu Yan Lu Yan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Julio,

Thanks for the information. I will try to reach them.

Cheers,
Lu
On 2016年1月23日 周六 at 下午12:57 Julio MATEOS_LANGERAK <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
>
> I know qioptiq has SM fibers for coupling up to 500mW in the visible
> range. They might have more options offline
>
> Best, Julio
>
> Sent from my iPhone
>
> > On 23 Jan 2016, at 00:32, Yan, Lu <[hidden email]> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Mike,
> >
> > Yes. It will be single moded. We have coupled few hundreds of mW of Ti:
> > sapphire laser at around 800 nm (CW though) into a single mode fiber
> > regularly without burning the fiber facet (we used bare fiber so that if
> > the fiber burnt, we cleave it and redo the coupling). I am not quite sure
> > what will be the case for pulse laser. I guess the damaging mechanism
> > sometimes varies from case to case. For the WDM, I think it might be the
> > fused coupler that may be vulnerable to damage but I could be completely
> > wrong about it. Will try to talk to some vendors tomorrow to figure it
> out.
> >
> > Thanks,
> > Lu
> >
> > -----------------------------------------------------
> > Lu Yan
> > Nanostructured Fibers and Nonlinear Optics Laboratory
> > Electrical and Computer Engineering
> > Boston University
> > 8 St. Mary St., Boston, MA, 02215
> > 617.353.0286 (office) | 617.358.5917 (lab)
> > [hidden email]
> > -----------------------------------------------------
> >
> >> On Fri, Jan 22, 2016 at 4:05 PM, Michael Giacomelli <[hidden email]>
> wrote:
> >>
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> Post images on http://www.imgur.com and include the link in your
> posting.
> >> *****
> >>
> >> Hi Lu,
> >>
> >> I'm assuming this is single mode?  Putting 1 watt into a few square
> >> microns can be fairly tricky, as the power density is gigantic.  What
> >> wavelength are you using?  I've seen fiber faces blown apart by ~150
> >> mW of 800 nm (femtosecond) ti:saph light.
> >>
> >> My bet is that the coupler itself won't be much more vulnerable to
> >> damage than the fiber its made of, but that you might have trouble
> >> just getting that kind of power (depending on wavelength) into a fiber
> >> without ablating the connectors or (depending on defects) your
> >> splices.
> >>
> >>
> >> Mike
> >>
> >>> On Fri, Jan 22, 2016 at 3:36 PM, Yan, Lu <[hidden email]> wrote:
> >>> *****
> >>> To join, leave or search the confocal microscopy listserv, go to:
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> Post images on http://www.imgur.com and include the link in your
> >> posting.
> >>> *****
> >>>
> >>> Hi there,
> >>>
> >>> I was looking at multiple vendors' websites for an optical fiber
> >> wavelength
> >>> combiner (two or three colors), or what is called fused fiber WDM. I
> have
> >>> noticed that most of them are designed for low to medium power level
> >>> applications, e.g. up to 300 mW. I was wondering if anyone here has
> >>> experience on using these devices, e.g. what is the maximum power it
> can
> >>> take, or is there any high power version of it that can take up to 1 W
> of
> >>> CW in the visible. I am thinking of using it in a STED microscopy setup
> >> but
> >>> I think 300 mW power limit might be at the lower side for this
> >> application.
> >>>
> >>> Thank you very much in advance,
> >>> Lu
> >>>
> >>> -----------------------------------------------------
> >>> Lu Yan
> >>> Nanostructured Fibers and Nonlinear Optics Laboratory
> >>> Electrical and Computer Engineering
> >>> Boston University
> >>> 8 St. Mary St., Boston, MA, 02215
> >>> 617.353.0286 (office) | 617.358.5917 (lab)
> >>> [hidden email]
> >>> -----------------------------------------------------
> >>
>
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

One more thing, when the fibre survives, you might disintegrate your sample...

Best wishes

Andreas

-----Original Message-----
From: "Yan, Lu" <[hidden email]>
Sent: ‎23/‎01/‎2016 20:34
To: "[hidden email]" <[hidden email]>
Subject: Re: high power fiber wavelength combiner or wdm

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Julio,

Thanks for the information. I will try to reach them.

Cheers,
Lu
On 2016年1月23日 周六 at 下午12:57 Julio MATEOS_LANGERAK <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
>
> I know qioptiq has SM fibers for coupling up to 500mW in the visible
> range. They might have more options offline
>
> Best, Julio
>
> Sent from my iPhone
>
> > On 23 Jan 2016, at 00:32, Yan, Lu <[hidden email]> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Mike,
> >
> > Yes. It will be single moded. We have coupled few hundreds of mW of Ti:
> > sapphire laser at around 800 nm (CW though) into a single mode fiber
> > regularly without burning the fiber facet (we used bare fiber so that if
> > the fiber burnt, we cleave it and redo the coupling). I am not quite sure
> > what will be the case for pulse laser. I guess the damaging mechanism
> > sometimes varies from case to case. For the WDM, I think it might be the
> > fused coupler that may be vulnerable to damage but I could be completely
> > wrong about it. Will try to talk to some vendors tomorrow to figure it
> out.
> >
> > Thanks,
> > Lu
> >
> > -----------------------------------------------------
> > Lu Yan
> > Nanostructured Fibers and Nonlinear Optics Laboratory
> > Electrical and Computer Engineering
> > Boston University
> > 8 St. Mary St., Boston, MA, 02215
> > 617.353.0286 (office) | 617.358.5917 (lab)
> > [hidden email]
> > -----------------------------------------------------
> >
> >> On Fri, Jan 22, 2016 at 4:05 PM, Michael Giacomelli <[hidden email]>
> wrote:
> >>
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> Post images on http://www.imgur.com and include the link in your
> posting.
> >> *****
> >>
> >> Hi Lu,
> >>
> >> I'm assuming this is single mode?  Putting 1 watt into a few square
> >> microns can be fairly tricky, as the power density is gigantic.  What
> >> wavelength are you using?  I've seen fiber faces blown apart by ~150
> >> mW of 800 nm (femtosecond) ti:saph light.
> >>
> >> My bet is that the coupler itself won't be much more vulnerable to
> >> damage than the fiber its made of, but that you might have trouble
> >> just getting that kind of power (depending on wavelength) into a fiber
> >> without ablating the connectors or (depending on defects) your
> >> splices.
> >>
> >>
> >> Mike
> >>
> >>> On Fri, Jan 22, 2016 at 3:36 PM, Yan, Lu <[hidden email]> wrote:
> >>> *****
> >>> To join, leave or search the confocal microscopy listserv, go to:
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> Post images on http://www.imgur.com and include the link in your
> >> posting.
> >>> *****
> >>>
> >>> Hi there,
> >>>
> >>> I was looking at multiple vendors' websites for an optical fiber
> >> wavelength
> >>> combiner (two or three colors), or what is called fused fiber WDM. I
> have
> >>> noticed that most of them are designed for low to medium power level
> >>> applications, e.g. up to 300 mW. I was wondering if anyone here has
> >>> experience on using these devices, e.g. what is the maximum power it
> can
> >>> take, or is there any high power version of it that can take up to 1 W
> of
> >>> CW in the visible. I am thinking of using it in a STED microscopy setup
> >> but
> >>> I think 300 mW power limit might be at the lower side for this
> >> application.
> >>>
> >>> Thank you very much in advance,
> >>> Lu
> >>>
> >>> -----------------------------------------------------
> >>> Lu Yan
> >>> Nanostructured Fibers and Nonlinear Optics Laboratory
> >>> Electrical and Computer Engineering
> >>> Boston University
> >>> 8 St. Mary St., Boston, MA, 02215
> >>> 617.353.0286 (office) | 617.358.5917 (lab)
> >>> [hidden email]
> >>> -----------------------------------------------------
> >>
>