Needham, Maurice |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Everyone, We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses. Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem. In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue. Thanks for your help, Maurice |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Maurice, Olympus objectives have a RMS thread, while Nikon uses M25. You can get adapters at Thorlabs (no commercial interest). Also the parfocal distance is different which might result in a problem with your z-drive. This is a good summary: http://www.microscopyu.com/articles/optics/cfintro.html (no commercial interest). Not sure what you mean with beam expander. Maybe you mean tube lens? The tube lens focal length for Nikon is 200 mm while Olympus uses 180 mm. But that should not be a big problem. Both companies do their optical corrections in the objective and not the tube lens, I believe. So mixing Leica and Zeiss with the "Japanese" manufacturers might be a bigger problem. Maybe someone on the list can confirm? Hope that helps, Jens > On Nov 26, 2014, at 1:23 PM, Needham, Maurice <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses. Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem. > > In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue. > > Thanks for your help, > > Maurice |
|
Michael Giacomelli |
|
|
In reply to this post by Needham, Maurice
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Maurice, Usually if both objectives are sold separately from the microscope system, you can interchange them. If they are sold only as part of a system, they often depend on customized tube lenses to get best performance (although sometimes they still work well). However, besides different thread lengths, be aware that different objectives, even with the same magnification, may have very different back focal plane locations. Before you swap the objectives, request the back focal plane location for the old objective and the new from the manufacturers and figure out if you can mechanically make it work in your system (that is, the new objective won't have its back focal plane such that its threading would be in the middle of your dichroic). Assuming you can make it work, thorlabs is great for adapters. Mike On Wed, Nov 26, 2014 at 1:23 PM, Needham, Maurice <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > We have an Olympus two-photon system that we would like to use a Nikon 16X > lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam > expander on the Olympus will not compensate for this lens but wondered if > anyone has experience with “fooling” the software to allow compensation > when using non-Olympus lenses. Also, we would appreciate comments on the > utility of this lens for two-photon imaging as we also have a custom system > that won’t have the beam expander problem. > > In addition, does anyone know of a supplier or accurate method of > fabricating adaptors to allow use of objective lens between different > manufacturer's scopes. Different objective screw thread is the main issue. > > Thanks for your help, > > Maurice > |
|
Craig Brideau |
|
|
In reply to this post by Jens
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Actually the larger, newer Olympus lenses have switched to M25 thread as well. In this case it depends on what your nosepiece is threaded for. RMS is fairly small and limits the back aperture of objectives pretty drastically, so the move has been towards M25, and even M32 on big bore objectives like the 16X/0.8 and 25X/1.1 sold by Nikon. The back aperture of these lenses is on the order of 18mm or so! Craig On Wed, Nov 26, 2014 at 11:47 AM, Jens-Bernhard Bosse <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Maurice, > > Olympus objectives have a RMS thread, while Nikon uses M25. You can get > adapters at Thorlabs (no commercial interest). > Also the parfocal distance is different which might result in a problem > with your z-drive. > This is a good summary: > http://www.microscopyu.com/articles/optics/cfintro.html (no commercial > interest). > Not sure what you mean with beam expander. Maybe you mean tube lens? > The tube lens focal length for Nikon is 200 mm while Olympus uses 180 mm. > But that should not be a big problem. > Both companies do their optical corrections in the objective and not the > tube lens, I believe. So mixing Leica and Zeiss with the "Japanese" > manufacturers might be a bigger problem. Maybe someone on the list can > confirm? > > Hope that helps, > > Jens > > > > On Nov 26, 2014, at 1:23 PM, Needham, Maurice <[hidden email]> > wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Everyone, > > > > We have an Olympus two-photon system that we would like to use a Nikon > 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam > expander on the Olympus will not compensate for this lens but wondered if > anyone has experience with “fooling” the software to allow compensation > when using non-Olympus lenses. Also, we would appreciate comments on the > utility of this lens for two-photon imaging as we also have a custom system > that won’t have the beam expander problem. > > > > In addition, does anyone know of a supplier or accurate method of > fabricating adaptors to allow use of objective lens between different > manufacturer's scopes. Different objective screw thread is the main issue. > > > > Thanks for your help, > > > > Maurice > |
|
Andreas Bruckbauer |
|
|
In reply to this post by Jens
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The beam expander on the newer Olympus multiphotons is just that, it is motorized and makes sure the back aperture is filled when using different objectives. You can just define a new objective with the right magnification and NA to 'fool' it. This way it is also possible to underfill existing objectives. Best wishes Andreas -----Original Message----- From: "Jens-Bernhard Bosse" <[hidden email]> Sent: 26/11/2014 18:51 To: "[hidden email]" <[hidden email]> Subject: Re: inter-manufacturer objective compatibility ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Maurice, Olympus objectives have a RMS thread, while Nikon uses M25. You can get adapters at Thorlabs (no commercial interest). Also the parfocal distance is different which might result in a problem with your z-drive. This is a good summary: http://www.microscopyu.com/articles/optics/cfintro.html (no commercial interest). Not sure what you mean with beam expander. Maybe you mean tube lens? The tube lens focal length for Nikon is 200 mm while Olympus uses 180 mm. But that should not be a big problem. Both companies do their optical corrections in the objective and not the tube lens, I believe. So mixing Leica and Zeiss with the "Japanese" manufacturers might be a bigger problem. Maybe someone on the list can confirm? Hope that helps, Jens > On Nov 26, 2014, at 1:23 PM, Needham, Maurice <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses. Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem. > > In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue. > > Thanks for your help, > > Maurice |
|
Zdenek Svindrych-2 |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes, Nikon (200 mm tube lens) 16x/0.8 has to have BFP diameter of 20 mm. So with small Olympus thread you won't be able to use the whole NA of your big Nikon lens... You may also need to rise the stage if the parfocal distance of the Nikon lens is much larger than that of the Olympus lenses (i.e. the Nikon lens is much longer). But if you manage to fit it mechanically, it should work. Best, zdenek ---------- Původní zpráva ---------- Od: Craig Brideau <[hidden email]> Komu: [hidden email] Datum: 26. 11. 2014 14:08:33 Předmět: Re: inter-manufacturer objective compatibility "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Actually the larger, newer Olympus lenses have switched to M25 thread as well. In this case it depends on what your nosepiece is threaded for. RMS is fairly small and limits the back aperture of objectives pretty drastically, so the move has been towards M25, and even M32 on big bore objectives like the 16X/0.8 and 25X/1.1 sold by Nikon. The back aperture of these lenses is on the order of 18mm or so! Craig On Wed, Nov 26, 2014 at 11:47 AM, Jens-Bernhard Bosse <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Maurice, > > Olympus objectives have a RMS thread, while Nikon uses M25. You can get > adapters at Thorlabs (no commercial interest). > Also the parfocal distance is different which might result in a problem > with your z-drive. > This is a good summary: > http://www.microscopyu.com/articles/optics/cfintro.html (no commercial > interest). > Not sure what you mean with beam expander. Maybe you mean tube lens? > The tube lens focal length for Nikon is 200 mm while Olympus uses 180 mm. > But that should not be a big problem. > Both companies do their optical corrections in the objective and not the > tube lens, I believe. So mixing Leica and Zeiss with the "Japanese" > manufacturers might be a bigger problem. Maybe someone on the list can > confirm? > > Hope that helps, > > Jens > > > > On Nov 26, 2014, at 1:23 PM, Needham, Maurice <[hidden email]> > wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Everyone, > > > > We have an Olympus two-photon system that we would like to use a Nikon > 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam > expander on the Olympus will not compensate for this lens but wondered if > anyone has experience with “fooling” the software to allow compensation > when using non-Olympus lenses. Also, we would appreciate comments on the > utility of this lens for two-photon imaging as we also have a custom > that won’t have the beam expander problem. > > > > In addition, does anyone know of a supplier or accurate method of > fabricating adaptors to allow use of objective lens between different > manufacturer's scopes. Different objective screw thread is the main issue. > > > > Thanks for your help, > > > > Maurice >" |
|
Steffen Dietzel |
|
|
In reply to this post by Needham, Maurice
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Maurice, we have a LaVision Biotec multi-photon which is built around an Olympus stand. A few years ago bought additional objectives, to supplement the (very good) 20x 0.95 Objective from Olympus for specific tasks. After testing several ones we ended up buying one from Leica (40x, 1.1, better resolution, but flat top) and the 16x Nikon (steeper angle, could access places the others couldn't). Putting all theory aside, they both work nicely for their specific purpose. Steffen Am 26.11.2014 19:23, schrieb Needham, Maurice: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses. Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem. > > In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue. > > Thanks for your help, > > Maurice > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
|
Needham, Maurice |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Everyone, Thank you so much for all your responses. I have some homework to do but i'm sure I will get it sorted due to all your helpful comments. This is my first post to the Microscopy list and i'm very impressed with the response. Its comforting to have access to such a wealth of knowledge on Imaging. Take care, with thanks, Maurice ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Steffen Dietzel [[hidden email]] Sent: Thursday, November 27, 2014 5:54 AM To: [hidden email] Subject: Re: inter-manufacturer objective compatibility ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Maurice, we have a LaVision Biotec multi-photon which is built around an Olympus stand. A few years ago bought additional objectives, to supplement the (very good) 20x 0.95 Objective from Olympus for specific tasks. After testing several ones we ended up buying one from Leica (40x, 1.1, better resolution, but flat top) and the 16x Nikon (steeper angle, could access places the others couldn't). Putting all theory aside, they both work nicely for their specific purpose. Steffen Am 26.11.2014 19:23, schrieb Needham, Maurice: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Everyone, > > We have an Olympus two-photon system that we would like to use a Nikon 16X lens (CFI LWD 16XW, NA 0.8, WD 3mm) with. We realize that the beam expander on the Olympus will not compensate for this lens but wondered if anyone has experience with “fooling” the software to allow compensation when using non-Olympus lenses. Also, we would appreciate comments on the utility of this lens for two-photon imaging as we also have a custom system that won’t have the beam expander problem. > > In addition, does anyone know of a supplier or accurate method of fabricating adaptors to allow use of objective lens between different manufacturer's scopes. Different objective screw thread is the main issue. > > Thanks for your help, > > Maurice > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |
|
Stanislav Vitha-2 |
|
|
In reply to this post by Needham, Maurice
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** On Wed, 26 Nov 2014 19:21:50 +0000, Andreas Bruckbauer <[hidden email]> wrote: >The beam expander on the newer Olympus multiphotons is just that, it is motorized and makes sure the back aperture is filled when using different objectives. You can just define a new objective with the right magnification and NA to 'fool' it. This way it is also possible to underfill existing objectives. >Best wishes >Andreas At least on our single-photon Olympus FV1000 (software v. 3.1), there is a plain text file somewhere in the Fluoview installation folder that contains the list of objectives. You can edit the file to add definition of another objective. If I remember correctly, it contains information like magnification, NA, immersion medium, working distance, back aperture diameter, .. I did that when testing a Zeiss 25x/0.8 multiimmersion objective on the Olympus (worked, but as expected had a pretty significant chromatic aberration). I do not remember the name of the definition file. Stan Vitha Texas A&M University Microscopy and Imaging Center |