![]() ![]() |
Aryeh Weiss |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
We are observing what appear to be interference fringes form reflected light in our Leica SP5, with our 10x objective. Example images can be
seen using this link: Basically, when the 405nm excitation is used alone, we do not see the fringes. When the other lines (ie the "visible" lines, as opposed to what Leica calls UV, which is really V) are used, there are fringes even in the channel which is set to pass wavelengths shorter than the shortest excitation in use (in our case 488nm). So, in brief, the "visible" laser lines all produce interference fringes in all channels. The fringe spacing changes with excitation wavelength (as expected), but not with scan speed. They are not apparent when working with 40x oil. However, I note that with the oil objective, there is much less reflection from the coverslip . We do not have other air objectives with which to test the system. The fringes appear with various samples. It is quite possible that this has always been there, but no one noticed because confocal at 10x is not done very often. If anyone on the list can tell us more about these fringes, we would appreciate it. Thanks in advance. --aryeh -- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
mmodel |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hello Aryeh,
You are doing fluorescence or reflection? If reflection then the fringes are almost unavoidable. For example, light reflecting from the front and from the back of the coverslip should interfere because of the long coherence length of the laser
Mike From: Confocal Microscopy List <[hidden email]> on behalf of Aryeh Weiss <[hidden email]>
Sent: Sunday, November 20, 2016 7:25 AM To: [hidden email] Subject: interference fringes in SP5 scans at 10x ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
We are observing what appear to be interference fringes form reflected light in our Leica SP5, with our 10x objective. Example images can be seen using this link: Basically, when the 405nm excitation is used alone, we do not see the fringes. When the other lines (ie the "visible" lines, as opposed to what Leica calls UV, which is really V) are used, there are fringes even in the channel which is set to pass wavelengths shorter than the shortest excitation in use (in our case 488nm). So, in brief, the "visible" laser lines all produce interference fringes in all channels. The fringe spacing changes with excitation wavelength (as expected), but not with scan speed. They are not apparent when working with 40x oil. However, I note that with the oil objective, there is much less reflection from the coverslip . We do not have other air objectives with which to test the system. The fringes appear with various samples. It is quite possible that this has always been there, but no one noticed because confocal at 10x is not done very often. If anyone on the list can tell us more about these fringes, we would appreciate it. Thanks in advance. --aryeh -- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
Aryeh Weiss |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Thank you for your reply. This is
fluorescence, not reflectance, imaging. The passbands are set to
reject the excitation wavelengths.
--aryeh On 20/11/2016 16:18, MODEL, MICHAEL wrote:
... [show rest of quote]
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
mmodel |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Fluorescent light cannot produce interference. So this has to be reflection. It could be due to something with filters or from reflection from something inside. Have you try to remove the sample? Defocus the sample? Try single line scanning? From: Aryeh Weiss <[hidden email]> on behalf of Aryeh Weiss <[hidden email]>
Sent: Sunday, November 20, 2016 9:49 AM To: MODEL, MICHAEL; [hidden email] Subject: Re: interference fringes in SP5 scans at 10x Thank you for your reply. This is fluorescence, not reflectance, imaging. The passbands are set to reject the excitation wavelengths.
--aryeh On 20/11/2016 16:18, MODEL, MICHAEL wrote:
... [show rest of quote]
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
mcammer |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Reflection. We especially had this problem with the far-red lines with the SP2 and SP5 AOBS systems. If you have one of the spectral systems, move the detection bounds further away from the laser lines. Also, is there a polarizing filter that can be put in/out to help block laser reflection? Don't recall if Leica had this.
_________________________________________ Michael Cammer, Optical Microscopy Specialist http://ocs.med.nyu.edu/microscopy
http://microscopynotes.com/
Cell: (914) 309-3270
From: Confocal Microscopy List [[hidden email]] on behalf of MODEL, MICHAEL [[hidden email]]
Sent: Sunday, November 20, 2016 10:08 AM To: [hidden email] Subject: Re: interference fringes in SP5 scans at 10x ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Fluorescent light cannot produce interference. So this has to be reflection. It could be due to something with filters or from reflection from something inside. Have you try to remove the sample? Defocus the sample? Try single line scanning? From: Aryeh Weiss <[hidden email]> on behalf of Aryeh Weiss <[hidden email]>
Sent: Sunday, November 20, 2016 9:49 AM To: MODEL, MICHAEL; [hidden email] Subject: Re: interference fringes in SP5 scans at 10x Thank you for your reply. This is fluorescence, not reflectance, imaging. The passbands are set to reject the excitation wavelengths.
--aryeh On 20/11/2016 16:18, MODEL, MICHAEL wrote:
... [show rest of quote]
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051
|
![]() ![]() |
manoj mathew |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Dear Aryeh, As mentioned already fringes are almost always due to coherent excitation light from the laser/s entering the detectors. Fluorescence is non-choherent and cannot possibly give off fringes. On SP5, essentially you have three barriers preventing the excitation light from entering the detectors: 1. AOBS 2. Notch filters 3. Spectral Sliders This might be problem with the AOBS settings. The 405nm on Leica SP5 is not routed through the AOBS but through a separate primary dichroic mirror. This might explain why you don't see fringes with the 405nm. Check the following AOBS settings: 1. Check if it is in reflection or fluorescence mode. You need to select the fluorescence mode 2. Check if appropriate notch filter (to cut of excitation light) is inserted. Also check your emission spectral slider settings: If they are not properly set (sufficiently away from the excitation wavelengths) it could allow excitation light into the detectors. However even if the spectral sliders are set properly there could be issues if the spectral calibration of the system is off. Spectral calibration is done through the service software and needs to be done periodically along with pinhole alignment etc. by the service engineer. Best Regards, Manoj -- Manoj Mathew, PhD Facility Incharge Central Imaging and Flow Cytometry Facility National Centre for Biological Sciences GKVK Post, Bellary Road Bangalore-560065, India Ph. +91 8067176277 Mob. 09886789049 CIFF Webpage: https://www.ncbs.res.in/ Linkedin:https://in.linkedin. On Sun, Nov 20, 2016 at 8:38 PM, MODEL, MICHAEL <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/w ... [show rest of quote] |
![]() ![]() |
Aryeh Weiss |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Thanks to the many members of the list
that have replied to my post.
As many have pointed out, fluorescence is not coherent, and the fringes must be from reflections. One suggestion has been to check the AOBS alignment, and we will try to do that. A misaligned AOBS might be passing some of the reflected excitation light. We made sure that the spectral bands were not close to the excitation lines, so it is unlikely to to direct leakage of the excitation. I was not aware that the SP5 has notch filters to block the excitation light. I thought that the AOBS and knife edges were sufficient. So we have something to check (the AOBS), and we will report on whether this was the problem. --aryeh On 20/11/2016 20:08, Manoj Mathew wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** ... [show rest of quote]
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
Zdenek Svindrych-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Aryeh Weiss
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Aryeh,
we've seen something similar when one of our AOBS drivers failed. The microscope still worked with bright samples, but the reflections were quite strong. You can try to switch between fluorescence and reflection mode, obviously, the reflections should be much stronger in the 'reflection' mode. But it was the other way round in our case.... Or the AOBS may need some tuning (just like AOTFs do). I don't know whether there are any notch filters, we have the 'X' configuration (with white light laser), so no notch filters here. The microscope also comes with a diagnostic software that lets you diagnose and tune AOBS (among other things), but there is not much documentation available. Your local Leica rep should be able to help you. Good luck! Best, zdenek --
Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) University of Virginia, Charlottesville, VA http://www.kcci.virginia.edu/ tel: 434-982-4869 Annual FRET Workshop: http://kcci.virginia.edu/workshop-2017 ---------- Původní zpráva ----------
... [show rest of quote] |
![]() ![]() |
Aryeh Weiss |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Thank you for your detailed reply.
Looks like a consensus has formed to check the AOBS and AOTF tuning, so we will do that. It occurs to me that one way to check if it is leakage through the AOBS and AOTF is to set the passbands so that there are no laser lines that overlap the passband. For us, that could be 520-550, since we have a multiline argon-ion, 561, 594 and 633 lasers. In that case, I would expect to see the fringes disappear or be very much weaker. However, we also see the fringes in the far red channel, where there are no laser lines. Maybe a mistuned AOBS also affects the direction (angle) of the reflected light? Best regards, --aryeh On 20/11/2016 21:48, [hidden email] wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Aryeh, ... [show rest of quote]
-- Aryeh Weiss Faculty of Engineering Bar Ilan University Ramat Gan 52900 Israel Ph: 972-3-5317638 FAX: 972-3-7384051 |
![]() ![]() |
manoj mathew |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Clarification to my previous post:
We have a Leica SP5 with STED attachment and it comes with notch filters. I now believe they are probably specific to the STED configuration and not a standard option. If the notch filters are installed they are accessible not through the AOBS settings (sorry my bad in the previous post) but through an icon called Analyzer Filter below the AOBS icon on the configuration window. AOBS and spectral sliders alone should be able to block out all excitation light without the need for the notch filters. They are probably more useful when the STED laser is ON. Best, Manoj -- Manoj Mathew, PhD Facility Incharge Central Imaging and Flow Cytometry Facility National Centre for Biological Sciences GKVK Post, Bellary Road Bangalore-560065, India Ph. +91 8067176277 Mob. 09886789049 CIFF Webpage: https://www.ncbs.res.in/research-facilities/ciff Linkedin:https://in.linkedin.com/in/manojvmathew/ On Mon, Nov 21, 2016 at 8:52 AM, Aryeh Weiss <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ ... [show rest of quote] |
![]() ![]() |
Sathya Srinivasan |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Aryeh Weiss
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Aryeh, There was a discussion about this issue during 2011:
Seems that discussion also concluded AOBS to be the problem: Good luck. Sathya Srinivasan ONPRC On Sun, Nov 20, 2016 at 4:25 AM, Aryeh Weiss <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ ... [show rest of quote] |
![]() ![]() |
Rosemary.White |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Thanks everyone for this useful discussion. I was always convinced this was reflection, and now will check more closely when (not if…) we see this again.
cheers,
Rosemary
Dr Rosemary White
CSIRO Black Mountain
GPO Box 1700
Canberra, ACT 2601
Australia
T 61 2 6246 5475
E [hidden email]
From: Confocal Microscopy List <[hidden email]> on behalf of Sathya Srinivasan <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]> Date: Wednesday, 23 November 2016 9:08 am To: "[hidden email]" <[hidden email]> Subject: Re: interference fringes in SP5 scans at 10x ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Hi Aryeh,
There was a discussion about this issue during 2011:
Seems that discussion also concluded AOBS to be the problem:
Good luck.
Sathya Srinivasan
ONPRC
On Sun, Nov 20, 2016 at 4:25 AM, Aryeh Weiss
<[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ ... [show rest of quote]
|
![]() ![]() |
Steffen Dietzel |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Aryeh Weiss
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Aryeh, in the SP8, additional notch filters are used only to block out
2-photon lasers or STED depletion lasers, but not for the normal
excitation. I would imagine it is the same for the SP5. Concerning the interference phenomenon, this may be a stupid
question, but could interference or another involved process
somehow change the wavelength of the excitation light from a
single wavelength peak to a low intensity broader profile? When I
see this phenomenon, I have the impression it is strongest close
to the excitation wavelength, on both sides of it. By excitation
with a pulsed laser we can gate it out if we dismiss the first 0.3
ns after the pulse, so it is clearly not fluorescence, as was
already said by others. Steffen Am 20.11.2016 um 19:34 schrieb Aryeh
Weiss:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** ... [show rest of quote]
-- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Biomedical Center (BMC) Head of the Core Facility Bioimaging Großhaderner Straße 9 D-82152 Planegg-Martinsried Germany http://www.bioimaging.bmc.med.uni-muenchen.de |
Free forum by Nabble | Edit this page |