Login  Register

lateral drifting

classic Classic list List threaded Threaded
7 messages Options Options
Embed post
Permalink
Chen, Jianbo (NIH/NCI) [E] Chen, Jianbo (NIH/NCI) [E]
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

lateral drifting

Dear everyone

I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks

 

Jianbo Chen

[hidden email]

301-846-1841

 

Axel Kurt Preuss Axel Kurt Preuss
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

Dear Jianbao
we needed details. What kind of stage, what sample etc. Which software controls the stage? Which kind of environmental chamber
Three hours is short, there should not be any problem. I would check with fluorescent beads or scratches on the coverglass.
 Often its a controller problem.



________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Chen, Jianbo (NIH/NCI) [E] [[hidden email]]
Sent: 20 February 2010 05:02
To: [hidden email]
Subject: lateral drifting

Dear everyone
I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks

Jianbo Chen
[hidden email]
301-846-1841


Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.
Shalin Mehta Shalin Mehta
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

In reply to this post by Chen, Jianbo (NIH/NCI) [E]
Hi Jianbo,
We experienced that with our Ludl stage. It was because of the calibration of joystick at startup. We tried two things to get rid of the problem and then the drifts reduced to minimal (a pixel or so in random direction):
1) Lock the joystick in central position during and after sometime of starting the controller and software. Once we mark the positions, we put the joystick away from any source of vibration (e.g. heater).
2) Ask the software to 'go back' to the position (even though imaging only one position) before acquiring an image.

Best
Shalin
On Sat, Feb 20, 2010 at 5:02 AM, Chen, Jianbo (NIH/NCI) [E] <[hidden email]> wrote:

Dear everyone

I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks

 

Jianbo Chen

[hidden email]

301-846-1841

 


mahogny mahogny
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

In reply to this post by Chen, Jianbo (NIH/NCI) [E]
Chen, Jianbo (NIH/NCI) [E] wrote:
> Dear everyone
> I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks
>  
we've had the same problem and it is unavoidable. in addition, we wore
out the gear box (expensive!) by excessive movement at least once, so
avoid using it for this. a piezo stage on top of the motorized stage
solved the problem.

an additional note there, we have noticed that the piezo also can drift
a very small way in the beginning but not if you constantly apply a
force to have it in the pushed down state. e.g. just leave the old
sample there until you replace it.

/Johan

> Jianbo Chen
> [hidden email]
> 301-846-1841
>
>
>  
mahogny mahogny
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

Johan Henriksson wrote:

> Chen, Jianbo (NIH/NCI) [E] wrote:
>> Dear everyone
>> I have some lateral drifting problem with a new microscope during
>> time lapse experiment. The drifting is usually around 10 pixel for a
>> 512 image during 3 hours. The microscope is set on a floating table.
>> The company people think it is unavoidable to have some drifting for
>> anything that is controlled by gear (motorized stage). Is that true
>> or just their excuse? I would appreciate any insight from people
>> having experiences in time-lapse experiment. Is it common or
>> unavoidable to have some drifting? Thanks
>>  
> we've had the same problem and it is unavoidable. in addition, we wore
> out the gear box (expensive!) by excessive movement at least once, so
> avoid using it for this. a piezo stage on top of the motorized stage
> solved the problem.
>
> an additional note there, we have noticed that the piezo also can
> drift a very small way in the beginning but not if you constantly
> apply a force to have it in the pushed down state. e.g. just leave the
> old sample there until you replace it.
noticed the keyword "lateral", I assume it means horizontal. the
joystick theory in the other post makes more sense, our problem was
vertical drift.

/Johan


>
> /Johan
>
>> Jianbo Chen
>> [hidden email]
>> 301-846-1841
>>
>>
>>  
Ralf Zenke Ralf Zenke
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

In reply to this post by Chen, Jianbo (NIH/NCI) [E]
Dear Jianbo,

When you talk of drift there are several effects that could occur. If you leave the stage at one position for your 3 hour experiment (no point visiting etc.) - do you see a change in the coordinates of the stage? The software should show the xyz coordinates somewhere. If coordinates change without your interaction this could mean that you have one of these rubbish joysticks. I also have one of these. It's a faulty design - you should just ask for a new one.
If you notice the drift in an experiment with point visiting you should compare the drift (pixel size of image x number of drift pixels) with the specification on the stage (accuracy/repeatability). Good stages with linear encoders have an accuracy/repeatability of around 1µm. But there are also stages with 8µm and more.

Hope that helps.

Cheers,

Ralf

Ralf Zenke

Max Planck Institute of Biochemistry
Core Facility
Am Klopferspitz 18
DE-82152 Martinsried near Munich
GERMANY
Phone: (+49) (89) 8578 3798
Fax: (+49) (89) 8578 2847
www.biochem.mpg.de

Am Freitag, den 19.02.2010, 16:02 -0500 schrieb Chen, Jianbo (NIH/NCI) [E]:
Dear everyone

I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks

 

Jianbo Chen

[hidden email]

301-846-1841

 


Emmanuel Gustin Emmanuel Gustin
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: lateral drifting

 

If you could translate that pixel distance to a physical distance in micron, it would be easier to judge. But a drift of a few micron is indeed unavoidable using a stage that works with gears, because the teeth of the gears cannot be sized to match each other exactly. There has to be a small size differential to leave a gap, which is filled by a thin film of lubricant. So even if the position of the driving gear is fixed, there can be a little motion in the stage as the pressure between the surfaces causes the oil to flow. (Usually this problem affects the focus more than the stage.)

 

Best Regards,

 

Emmanuel

 

 

--
 Emmanuel Gustin,    Tel. (+32) 15 46 1586,    e-mail: [hidden email]

 

 

 

Am Freitag, den 19.02.2010, 16:02 -0500 schrieb Chen, Jianbo (NIH/NCI) [E]:

Dear everyone

I have some lateral drifting problem with a new microscope during time lapse experiment. The drifting is usually around 10 pixel for a 512 image during 3 hours. The microscope is set on a floating table. The company people think it is unavoidable to have some drifting for anything that is controlled by gear (motorized stage). Is that true or just their excuse? I would appreciate any insight from people having experiences in time-lapse experiment. Is it common or unavoidable to have some drifting? Thanks

 

Jianbo Chen

[hidden email]

301-846-1841