Thomas Horn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello all, we are currently using DiD to label a structure on a micro fluidics device. However, the cells that we need to visualize do contain mCherry. Problem is that the absorption/excitation curve of DiD has a long tail "on the left" so that 561 excites it very well. As the DiD fluorescence is very strong we do get a lot of Signal in the mCherry channel when exciting with a 561nm laser - even if we use a small band pass on the lower emission spectrum of mCherry. The DiD signal is very strong versus a weak mCherry so that spectral unmixing is problematic. To cut a long story short: can someone recommend a lipophilic dye that is good for 647 excitation but is not excited by 561nm. Kind regards, Thomas Dr. Thomas Horn, The Single Cell Unit, U1.46 Department of Biosystems Science and Engineering (D-BSSE) Swiss Federal Institute of Technology Zurich (ETH) Mattenstrasse 26 CH 4048 Basel Switzerland Phone: +41 61 387 3373 mail: [hidden email] |
|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Thomas, Since the lipophilic probes like DiI, DiO, and DiD usually absorb and emit light strongly compared to mCherry (as in your case), perhaps you could switch over to the longer wavelength probe DiR (no commercial interest): https://products.invitrogen.com/ivgn/product/D12731?ICID=search-product This probe has a much lower absorption at 647 nm, but I bet you'll still be able to get plenty of signal to form an image with the same laser line. The emission is further out in the red around 780 nm, so you'll need a different emission filter (your existing filter might work if it's a long-pass), but the bleedthrough into your mCherry channel should be much more reduced since the absorption at 561 nm is less than 5%. You didn't mention what type of imaging system you're using. If it's a laser scanning confocal with a PMT, you might need to play with the settings a bit. If it's a camera-based system (like a spinning disk confocal), you should have no problems since most CCDs can still detect out in that wavelength range. Don't bother trying to look for the probe in the eyepieces, though! Cheers, John Oreopoulos Research Assistant Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca On 2012-10-22, at 10:46 AM, Horn Thomas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello all, > we are currently using DiD to label a structure on a micro fluidics device. However, the cells that we need to visualize do contain mCherry. Problem is that the absorption/excitation curve of DiD has a long tail "on the left" so that 561 excites it very well. As the DiD fluorescence is very strong we do get a lot of Signal in the mCherry channel when exciting with a 561nm laser - even if we use a small band pass on the lower emission spectrum of mCherry. The DiD signal is very strong versus a weak mCherry so that spectral unmixing is problematic. > > To cut a long story short: can someone recommend a lipophilic dye that is good for 647 excitation but is not excited by 561nm. > > Kind regards, > Thomas > > > > > Dr. Thomas Horn, > The Single Cell Unit, U1.46 > Department of Biosystems Science and Engineering (D-BSSE) > Swiss Federal Institute of Technology Zurich (ETH) > Mattenstrasse 26 > CH 4048 Basel > Switzerland > Phone: +41 61 387 3373 > mail: [hidden email] |
|
George McNamara |
|
|
In reply to this post by Thomas Horn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** DiR abs max 750 nm ... about 20% of max at 633 nm em max 780 nm ... FW10% is 740 to 850 nm Not all detectors count photons in the near infrared. Not all filters are blocked in the NIR. http://www.invitrogen.com/site/us/en/home/support/Research-Tools/Fluorescence-SpectraViewer.html (The U arizona spectra Viewer has bad trace for DiR emission) note some "alternation" (I think they should have written "alteration") of spectra of DiI and DiD by confocal laser illumination was reported in: Michal Majkowski, Pawel Póda, Julita Kulbacka, Jolanta Saczko,and Aleksander F. Sikorski Alternation of Fluorescent Spectra of Membrane Markers DiI C18(3) and DiI C18(5) Evoked by Laser Illumination J Histochem Cytochem October 2012 60: 789-791, first published on July 21, 2012 doi:10.1369/0022155412457572 http://jhc.sagepub.com/content/current Personally, I would not worry about the effect(s). George On 10/22/2012 10:46 AM, Horn Thomas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello all, > we are currently using DiD to label a structure on a micro fluidics device. However, the cells that we need to visualize do contain mCherry. Problem is that the absorption/excitation curve of DiD has a long tail "on the left" so that 561 excites it very well. As the DiD fluorescence is very strong we do get a lot of Signal in the mCherry channel when exciting with a 561nm laser - even if we use a small band pass on the lower emission spectrum of mCherry. The DiD signal is very strong versus a weak mCherry so that spectral unmixing is problematic. > > To cut a long story short: can someone recommend a lipophilic dye that is good for 647 excitation but is not excited by 561nm. > > Kind regards, > Thomas > > > > > Dr. Thomas Horn, > The Single Cell Unit, U1.46 > Department of Biosystems Science and Engineering (D-BSSE) > Swiss Federal Institute of Technology Zurich (ETH) > Mattenstrasse 26 > CH 4048 Basel > Switzerland > Phone: +41 61 387 3373 > mail: [hidden email] > > |
| Free forum by Nabble | Edit this page |