Hi David and Listserv,

David wrote: "Photon counting mode on EMCCDs  works by making the em
gain very high, such that the read noise is much less than 1 photon per
pixel."

But read noise is measured with no light, so no photons. Just electrons.
For example, the Photometrics Evolve datasheet:

Read RMS noise (e- rms @ Gain State 3)
10 MHz EM Port            45 electrons
   5 MHz EM Port            32 electrons
   5 MHz non-EM port     12 electrons
1.25 MHz non-EM port    6 electrons

Dark current 0.001 electrons/pixel/sec

Background events
0.0045              events/pixel/sec (10 MHz, 1000x EM gain) standard
operation
not detectable    events/pixel/sec (10 MHz, 1000x EM gain) BERT operation

    (footnote 4) Dark current -- this is measured in a traditional
    manner (as with all CCD cameras) by taking a long integration to
    obtain a signal. An average measurement is taken over the CCD area
    (excluding blemishes). It should be noted that dark current can vary
    significantly between different CCDs and the numbers here are typical.

    (footnote 5) Background events -- as EMCCD cameras are actually
    capable of detecting single photons the real detection limit of
    these cameras is set by the number of dark background events. These
    can arise from two things, dark current (which is thermal generation
    of an electron and is a temperature dependent phenomenon) and also
    clock induced charge (CIC) electrons (also called spurious charge).
    Each can lead to the generation of non-photon derived electrons
    which are multiplied through the electron-multiplication register
    generating random high
    value pixels which are above the read noise.
    These background events are measured by taking 30 ms exposure at
    10MHz speed with 1000X EM Gain applied and counting the number of
    random high value pixels which are at a single event threshold above
    the modal value of the image histogram. This number is expressed as
    a probability of an event per pixel. The number can vary from frame
    to frame and sensor to sensor however a typical value is provided.


Someone else posted an equation about scientific CMOS camera(s) with "EM
gain" in the equation ... sCMOS does not have EM gain, so that equation
would either result in zero or infinity, which would be silly.

Hamamatsu FLASH4.0 read noise (from "Changing the Game" pdf) is less
than 1.3 electrons (at 100 fps).


Best wishes,

George


On 1/25/2012 8:47 PM, David Baddeley wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Photon counting mode on EMCCDs  works by making the em gain very high, such that the read noise is much less than 1 photon per pixel. If the photon flux is also very low (much less than 1 photon/pixel/frame) you can threshold each frame and classify each pixel as either having seen a photon or not, and thereby remove the multiplicative noise effects. This only works for very low photon counts as are observed in some aspects of astronomy. In PALM/STORM the photon fluxes are much higher (typically several 100-1000 /pixel/frame) and we can't do photon counting per se. As to EMCCD vs CCD vs SCMOS, they all work with a slight theoretical advantage to SCMOS followed closely by EMCCD. In most practical cases the background is sufficiently high that a small amount of read noise can be tolerated (even as much as the ~5 e- you get from a good conventional CCD @ ~10hz readout - although you'll often want to record faster than 10 hz at which point standard ccds
>   become less useful). The important factors for PALM/STORM are thus effective quantum efficiency (remembering that the electron multiplying step effectively halves the QE), and readout speed.  I remember hearing that Leica was using a sCMOS device in their system (they've definitely published a proof of concept with one), and when we trialled a CMOS sensor it seemed to work just as well as our EMCCD cameras.
>
> As to orientation effects, I was unable to see any difference between linearly and circularly polarised excitation when imaging using antibody conjugated dyes - this suggests that they (at least) are on a flexible linker and free to rotate rapidly. Can't really speculate as to how well this extends to other labelling methods.
>
> I'd also suggest investigating other visualisation methods, as the original method of drawing Gaussians has a number of potential limitations, some of which, such as the tendency to inspire more confidence in the data than might be warranted, have already been pointed out (what hasn't been mentioned yet is the fact that it sacrifices resolution if you do happen to have a high density of localisations).  Try a number of methods and see how they work with your data (and ideally also with simulated data of known objects). At the very least do a simple 2D histogram as well as the Gaussian method - this will allow you to directly relate pixel intensity to the number of events within that pixel. If you want to go further, our paper (Baddeley, Cannell, Soeller, 'Visualization of Localization Microscopy Data', Microscopy and Microanalysis 2010) might give you some ideas.
>
> Best wishes,
> David
>
>
> ________________________________
>   From: Guy Cox<[hidden email]>
> To: [hidden email]
> Sent: Wednesday, 25 January 2012 11:20 AM
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I just assumed that photon-counting technology would be used for this but I guess I was wrong.  I also thought astronomers used CCD and CMOS sensors for photon counting.  I guess it depends on how many electrons one photon produces in the potential well.  Jim probably knows the answer to that.
>
>                                 Guy
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor&  Francis
>      http://www.guycox.com/optical.htm
> ______________________________________________
> Guy Cox, MA, DPhil(Oxon), Honorary Associate,
> Australian Centre for Microscopy&  Microanalysis,
> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>               Mobile 0413 281 861
> ______________________________________________
>        http://www.guycox.net
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
> Sent: Wednesday, 25 January 2012 2:30 AM
> To: [hidden email]
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Essentially yes.
>
> Guy's reply would seem to apply to photon-counting with PMTs when run in photon-counting-mode, not to cameras. Guy, please correct me if I am wrong.
>
> Christian
>
> On 25/01/2012, at 4:18 AM, Roger Phillips wrote:
>
>    
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Thanks, Christian,
>> So the sCMOS photon count would come from the formula
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>> with EM-gain set to 1 and no 'additional noise introduced by the EM gain process'?
>>
>> In Guy's reply, he said 'signals below a certain threshold are regarded as noise, and discarded, and signals above a higher threshold are regarded as 'pile-up' and also discarded.'  This methods seems to account for the 'low threshold' but not for the possible 'pile-up'?
>> Roger
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
>> Sent: 24 January 2012 14:44
>> To: [hidden email]
>> Subject: Re: localization precision in PALM/STORM
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> You basically want to convert the AD counts in your image into photon numbers. That requires several bits of info about the camera. We have Andor cameras and I can look up the required values from the data/test sheet that comes with the individual camera (these change from cam to cam). You need to know the
>> - electrons per count
>> - absolute EM gain (with the quantEM you might have to measure/calibrate this)
>> - A/D offset (this can be measured in dark frames with no light impinging on the cam, i.e. shutter closed)
>>
>> The formula is then something like
>>
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>>
>> Some more recent camera software packages may have functions to make this conversion for you. I am not sure about photometrics cams/SDKs.
>>
>> Due the additional noise introduced by the EM gain process you should divide the resulting photon-numbers by 2 before looking up values from the Thompson et al. formula. There are papers on EM CCDs that explain this. I also seem to recall that some more recent papers have a subtle correction to the Thompson formula.
>>
>> Hope this helps,
>>
>> Christian
>>
>> On 25/01/2012, at 2:08 AM, Christophe Leterrier wrote:
>>
>>      
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Mark,
>>>
>>> I don't think the quantEM has a built-in photon calibration function, in contrast to the newer Evolve camera, also from Photometrics. The spec sheet for the quantEM is available here :
>>> http://www.photometrics.com/products/datasheets/qem512sc.pdf
>>>
>>> Do I have to calibrate it myself or is Photometrics supposed to provide a photon/intensity calibration curve? I don't want exact experimental values for my precise camera, just a reasonable estimate to derive a theoretical "best" value for localization accuracy.
>>>
>>> Christophe
>>>
>>>
>>> Le mardi 24 janvier 2012 à 10:58, Mark Cannell a écrit :
>>>
>>>        
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> Doesn't the quantEM have a photon calibration function? The background noise should be estimated from the variance of the background (extracted from image regions when/where flashes were not detected...). You can also calibrate the camera with weak sources to double check the manufactures stated read-out calibration.
>>>>
>>>> Hope this helps
>>>>
>>>> Mark
>>>>
>>>>
>>>> On 24/01/2012, at 9:26 AM, Christophe Leterrier wrote:
>>>>
>>>>          
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Hi,
>>>>>
>>>>> Not strictly a confocal question, but I'm pretty sure this list is the best place to get thorough and insightful answers.
>>>>>
>>>>> I have made 2D STORM (stochastic optical reconstruction microscopy) acquisitions and processing and I end up with a table of XY localized fluorophores together with the integrated intensity of the localized diffraction-limited spot.
>>>>>
>>>>> I'd like to plot each fluorophore as a gaussian with a width corresponding to the localization precision, similar to what was done in Bates et al. Science 2007. According to equation (17) in Thompson, Larson&  Webb Biophys J. 2002 (http://goo.gl/5GIXM), this precision depends on the number of photons collected, the width of the diffraction-limited spot, the size of the camera pixel, and the background noise.
>>>>>
>>>>> So my question is : How do I get the number of photons from the intensity level of an image? I'm using a Photometrics 512*512 QuantEM camera. What is the background noise and how do I estimate it? Then using these values in the Thompson et al. equation, I can get a theoretical spot intensity / localization precision calibration curve that I could use for the gaussian-based reconstruction.
>>>>>
>>>>> Thanks for your help,
>>>>>
>>>>> --
>>>>> Christophe Leterrier
>>>>> Researcher
>>>>> Axonal Domains Architecture Team
>>>>> CRN2M CNRS UMR 7286
>>>>> Aix Marseille University, France
>>>>>
>>>>>            
>>>>
>>>>
>>>>          
>>      
>    


--


George McNamara, PhD
Analytical Imaging Core Facility
University of Miami
Sebastian Rhode Sebastian Rhode
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: localization precision in PALM/STORM

In reply to this post by lechristophe
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi,

Do you think it is an appropiate way to determine a thing like stage drift
imaging 100nm beads while recording a timelapse.
I usually fit the beads positions with a 2d Gauss and due to the very high
SNR, it should be possible to localize the peak with nm precsision. And
assuming that you have fixed beads, this will give you a good idea about
your drift.

Do you guys think, thisway of determine the drift is fine or how would you
do this?

Cheers,

Sebi
David Baddeley David Baddeley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: localization precision in PALM/STORM

In reply to this post by George McNamara
I guess I should have said 'photon equivalents' - the scena
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi George,

I guess I should have said 'photon equivalents' - the scenario I had in mind is when the read noise is, e.g., 45 electrons and the EM gain is such that each photon produces ~1000 electrons (ie much more than the read noise). You can then set a threshold at, for example, 500 electrons and get a yes/no image as to whether a photon hit.

cheers,
David


________________________________
 From: George McNamara <[hidden email]>
To: [hidden email]
Sent: Friday, 27 January 2012 1:02 AM
Subject: Re: localization precision in PALM/STORM
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi David and Listserv,

David wrote: "Photon counting mode on EMCCDs  works by making the em
gain very high, such that the read noise is much less than 1 photon per
pixel."

But read noise is measured with no light, so no photons. Just electrons.
For example, the Photometrics Evolve datasheet:

Read RMS noise (e- rms @ Gain State 3)
10 MHz EM Port            45 electrons
   5 MHz EM Port            32 electrons
   5 MHz non-EM port     12 electrons
1.25 MHz non-EM port    6 electrons

Dark current 0.001 electrons/pixel/sec

Background events
0.0045              events/pixel/sec (10 MHz, 1000x EM gain) standard
operation
not detectable    events/pixel/sec (10 MHz, 1000x EM gain) BERT operation

    (footnote 4) Dark current -- this is measured in a traditional
    manner (as with all CCD cameras) by taking a long integration to
    obtain a signal. An average measurement is taken over the CCD area
    (excluding blemishes). It should be noted that dark current can vary
    significantly between different CCDs and the numbers here are typical.

    (footnote 5) Background events -- as EMCCD cameras are actually
    capable of detecting single photons the real detection limit of
    these cameras is set by the number of dark background events. These
    can arise from two things, dark current (which is thermal generation
    of an electron and is a temperature dependent phenomenon) and also
    clock induced charge (CIC) electrons (also called spurious charge).
    Each can lead to the generation of non-photon derived electrons
    which are multiplied through the electron-multiplication register
    generating random high
    value pixels which are above the read noise.
    These background events are measured by taking 30 ms exposure at
    10MHz speed with 1000X EM Gain applied and counting the number of
    random high value pixels which are at a single event threshold above
    the modal value of the image histogram. This number is expressed as
    a probability of an event per pixel. The number can vary from frame
    to frame and sensor to sensor however a typical value is provided.


Someone else posted an equation about scientific CMOS camera(s) with "EM
gain" in the equation ... sCMOS does not have EM gain, so that equation
would either result in zero or infinity, which would be silly.

Hamamatsu FLASH4.0 read noise (from "Changing the Game" pdf) is less
than 1.3 electrons (at 100 fps).


Best wishes,

George


On 1/25/2012 8:47 PM, David Baddeley wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Photon counting mode on EMCCDs  works by making the em gain very high, such that the read noise is much less than 1 photon per pixel. If the photon flux is also very low (much less than 1 photon/pixel/frame) you can threshold each frame and classify each pixel as either having seen a photon or not, and thereby remove the multiplicative noise effects. This only works for very low photon counts as are observed in some aspects of astronomy. In PALM/STORM the photon fluxes are much higher (typically several 100-1000 /pixel/frame) and we can't do photon counting per se. As to EMCCD vs CCD vs SCMOS, they all work with a slight theoretical advantage to SCMOS followed closely by EMCCD. In most practical cases the background is sufficiently high that a small amount of read noise can be tolerated (even as much as the ~5 e- you get from a good conventional CCD @ ~10hz readout - although you'll often want to record faster than 10 hz at which point standard ccds
>   become less useful). The important factors for PALM/STORM are thus effective quantum efficiency (remembering that the electron multiplying step effectively halves the QE), and readout speed.  I remember hearing that Leica was using a sCMOS device in their system (they've definitely published a proof of concept with one), and when we trialled a CMOS sensor it seemed to work just as well as our EMCCD cameras.
>
> As to orientation effects, I was unable to see any difference between linearly and circularly polarised excitation when imaging using antibody conjugated dyes - this suggests that they (at least) are on a flexible linker and free to rotate rapidly. Can't really speculate as to how well this extends to other labelling methods.
>
> I'd also suggest investigating other visualisation methods, as the original method of drawing Gaussians has a number of potential limitations, some of which, such as the tendency to inspire more confidence in the data than might be warranted, have already been pointed out (what hasn't been mentioned yet is the fact that it sacrifices resolution if you do happen to have a high density of localisations).  Try a number of methods and see how they work with your data (and ideally also with simulated data of known objects). At the very least do a simple 2D histogram as well as the Gaussian method - this will allow you to directly relate pixel intensity to the number of events within that pixel. If you want to go further, our paper (Baddeley, Cannell, Soeller, 'Visualization of Localization Microscopy Data', Microscopy and Microanalysis 2010) might give you some ideas.
>
> Best wishes,
> David
>
>
> ________________________________
>   From: Guy Cox<[hidden email]>
> To: [hidden email]
> Sent: Wednesday, 25 January 2012 11:20 AM
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I just assumed that photon-counting technology would be used for this but I guess I was wrong.  I also thought astronomers used CCD and CMOS sensors for photon counting.  I guess it depends on how many electrons one photon produces in the potential well.  Jim probably knows the answer to that.
>
>                                 Guy
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor&  Francis
>      http://www.guycox.com/optical.htm
> ______________________________________________
> Guy Cox, MA, DPhil(Oxon), Honorary Associate,
> Australian Centre for Microscopy&  Microanalysis,
> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>               Mobile 0413 281 861
> ______________________________________________
>        http://www.guycox.net
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
> Sent: Wednesday, 25 January 2012 2:30 AM
> To: [hidden email]
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Essentially yes.
>
> Guy's reply would seem to apply to photon-counting with PMTs when run in photon-counting-mode, not to cameras. Guy, please correct me if I am wrong.
>
> Christian
>
> On 25/01/2012, at 4:18 AM, Roger Phillips wrote:
>
>   
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Thanks, Christian,
>> So the sCMOS photon count would come from the formula
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>> with EM-gain set to 1 and no 'additional noise introduced by the EM gain process'?
>>
>> In Guy's reply, he said 'signals below a certain threshold are regarded as noise, and discarded, and signals above a higher threshold are regarded as 'pile-up' and also discarded.'  This methods seems to account for the 'low threshold' but not for the possible 'pile-up'?
>> Roger
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
>> Sent: 24 January 2012 14:44
>> To: [hidden email]
>> Subject: Re: localization precision in PALM/STORM
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> You basically want to convert the AD counts in your image into photon numbers. That requires several bits of info about the camera. We have Andor cameras and I can look up the required values from the data/test sheet that comes with the individual camera (these change from cam to cam). You need to know the
>> - electrons per count
>> - absolute EM gain (with the quantEM you might have to measure/calibrate this)
>> - A/D offset (this can be measured in dark frames with no light impinging on the cam, i.e. shutter closed)
>>
>> The formula is then something like
>>
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>>
>> Some more recent camera software packages may have functions to make this conversion for you. I am not sure about photometrics cams/SDKs.
>>
>> Due the additional noise introduced by the EM gain process you should divide the resulting photon-numbers by 2 before looking up values from the Thompson et al. formula. There are papers on EM CCDs that explain this. I also seem to recall that some more recent papers have a subtle correction to the Thompson formula.
>>
>> Hope this helps,
>>
>> Christian
>>
>> On 25/01/2012, at 2:08 AM, Christophe Leterrier wrote:
>>
>>     
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Mark,
>>>
>>> I don't think the quantEM has a built-in photon calibration function, in contrast to the newer Evolve camera, also from Photometrics. The spec sheet for the quantEM is available here :
>>> http://www.photometrics.com/products/datasheets/qem512sc.pdf
>>>
>>> Do I have to calibrate it myself or is Photometrics supposed to provide a photon/intensity calibration curve? I don't want exact experimental values for my precise camera, just a reasonable estimate to derive a theoretical "best" value for localization accuracy.
>>>
>>> Christophe
>>>
>>>
>>> Le mardi 24 janvier 2012 à 10:58, Mark Cannell a écrit :
>>>
>>>       
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> Doesn't the quantEM have a photon calibration function? The background noise should be estimated from the variance of the background (extracted from image regions when/where flashes were not detected...). You can also calibrate the camera with weak sources to double check the manufactures stated read-out calibration.
>>>>
>>>> Hope this helps
>>>>
>>>> Mark
>>>>
>>>>
>>>> On 24/01/2012, at 9:26 AM, Christophe Leterrier wrote:
>>>>
>>>>         
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Hi,
>>>>>
>>>>> Not strictly a confocal question, but I'm pretty sure this list is the best place to get thorough and insightful answers.
>>>>>
>>>>> I have made 2D STORM (stochastic optical reconstruction microscopy) acquisitions and processing and I end up with a table of XY localized fluorophores together with the integrated intensity of the localized diffraction-limited spot.
>>>>>
>>>>> I'd like to plot each fluorophore as a gaussian with a width corresponding to the localization precision, similar to what was done in Bates et al. Science 2007. According to equation (17) in Thompson, Larson&  Webb Biophys J. 2002 (http://goo.gl/5GIXM), this precision depends on the number of photons collected, the width of the diffraction-limited spot, the size of the camera pixel, and the background noise.
>>>>>
>>>>> So my question is : How do I get the number of photons from the intensity level of an image? I'm using a Photometrics 512*512 QuantEM camera. What is the background noise and how do I estimate it? Then using these values in the Thompson et al. equation, I can get a theoretical spot intensity / localization precision calibration curve that I could use for the gaussian-based reconstruction.
>>>>>
>>>>> Thanks for your help,
>>>>>
>>>>> --
>>>>> Christophe Leterrier
>>>>> Researcher
>>>>> Axonal Domains Architecture Team
>>>>> CRN2M CNRS UMR 7286
>>>>> Aix Marseille University, France
>>>>>
>>>>>           
>>>>
>>>>
>>>>         
>>     
>   


--


George McNamara, PhD
Analytical Imaging Core Facility
University of Miami
Periasamy, Ammasi (ap3t) Periasamy, Ammasi (ap3t)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: localization precision in PALM/STORM

In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hamamatsu ORCA-Flash4.0 provides incredible speed. I was testing the camera in the SPIE meeting two days ago.
2048x2048, 100 frames/sec. If the size is reduced the frames/sec increases. For 2048x256, 800 frames/sec


Dr. Ammasi Periasamy
Professor & Center Director
Keck Center for Cellular Imaging (KCCI)
Biology, Gilmer Hall (064), 485 McCormick Rd
University of Virginia
Charlottesville, VA 22904
(Campus Mail - P.O. Box 400328)
Voice: 434-243-7602 (Office); 982-4869 (lab)
Fax:434-982-5210; Email:[hidden email]
http://www.kcci.virginia.edu/contact/peri.php
************************
11th Annual Workshop on FRET Microscopy, March 5-10, 2012
http://www.kcci.virginia.edu/workshop/workshop2012/
*************************


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara
Sent: Thursday, January 26, 2012 7:03 AM
To: [hidden email]
Subject: Re: localization precision in PALM/STORM

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi David and Listserv,

David wrote: "Photon counting mode on EMCCDs  works by making the em
gain very high, such that the read noise is much less than 1 photon per
pixel."

But read noise is measured with no light, so no photons. Just electrons.
For example, the Photometrics Evolve datasheet:

Read RMS noise (e- rms @ Gain State 3)
10 MHz EM Port            45 electrons
   5 MHz EM Port            32 electrons
   5 MHz non-EM port     12 electrons
1.25 MHz non-EM port    6 electrons

Dark current 0.001 electrons/pixel/sec

Background events
0.0045              events/pixel/sec (10 MHz, 1000x EM gain) standard
operation
not detectable    events/pixel/sec (10 MHz, 1000x EM gain) BERT operation

    (footnote 4) Dark current -- this is measured in a traditional
    manner (as with all CCD cameras) by taking a long integration to
    obtain a signal. An average measurement is taken over the CCD area
    (excluding blemishes). It should be noted that dark current can vary
    significantly between different CCDs and the numbers here are typical.

    (footnote 5) Background events -- as EMCCD cameras are actually
    capable of detecting single photons the real detection limit of
    these cameras is set by the number of dark background events. These
    can arise from two things, dark current (which is thermal generation
    of an electron and is a temperature dependent phenomenon) and also
    clock induced charge (CIC) electrons (also called spurious charge).
    Each can lead to the generation of non-photon derived electrons
    which are multiplied through the electron-multiplication register
    generating random high
    value pixels which are above the read noise.
    These background events are measured by taking 30 ms exposure at
    10MHz speed with 1000X EM Gain applied and counting the number of
    random high value pixels which are at a single event threshold above
    the modal value of the image histogram. This number is expressed as
    a probability of an event per pixel. The number can vary from frame
    to frame and sensor to sensor however a typical value is provided.


Someone else posted an equation about scientific CMOS camera(s) with "EM
gain" in the equation ... sCMOS does not have EM gain, so that equation
would either result in zero or infinity, which would be silly.

Hamamatsu FLASH4.0 read noise (from "Changing the Game" pdf) is less
than 1.3 electrons (at 100 fps).


Best wishes,

George


On 1/25/2012 8:47 PM, David Baddeley wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Photon counting mode on EMCCDs  works by making the em gain very high, such that the read noise is much less than 1 photon per pixel. If the photon flux is also very low (much less than 1 photon/pixel/frame) you can threshold each frame and classify each pixel as either having seen a photon or not, and thereby remove the multiplicative noise effects. This only works for very low photon counts as are observed in some aspects of astronomy. In PALM/STORM the photon fluxes are much higher (typically several 100-1000 /pixel/frame) and we can't do photon counting per se. As to EMCCD vs CCD vs SCMOS, they all work with a slight theoretical advantage to SCMOS followed closely by EMCCD. In most practical cases the background is sufficiently high that a small amount of read noise can be tolerated (even as much as the ~5 e- you get from a good conventional CCD @ ~10hz readout - although you'll often want to record faster than 10 hz at which point standard ccds
>   become less useful). The important factors for PALM/STORM are thus effective quantum efficiency (remembering that the electron multiplying step effectively halves the QE), and readout speed.  I remember hearing that Leica was using a sCMOS device in their system (they've definitely published a proof of concept with one), and when we trialled a CMOS sensor it seemed to work just as well as our EMCCD cameras.
>
> As to orientation effects, I was unable to see any difference between linearly and circularly polarised excitation when imaging using antibody conjugated dyes - this suggests that they (at least) are on a flexible linker and free to rotate rapidly. Can't really speculate as to how well this extends to other labelling methods.
>
> I'd also suggest investigating other visualisation methods, as the original method of drawing Gaussians has a number of potential limitations, some of which, such as the tendency to inspire more confidence in the data than might be warranted, have already been pointed out (what hasn't been mentioned yet is the fact that it sacrifices resolution if you do happen to have a high density of localisations).  Try a number of methods and see how they work with your data (and ideally also with simulated data of known objects). At the very least do a simple 2D histogram as well as the Gaussian method - this will allow you to directly relate pixel intensity to the number of events within that pixel. If you want to go further, our paper (Baddeley, Cannell, Soeller, 'Visualization of Localization Microscopy Data', Microscopy and Microanalysis 2010) might give you some ideas.
>
> Best wishes,
> David
>
>
> ________________________________
>   From: Guy Cox<[hidden email]>
> To: [hidden email]
> Sent: Wednesday, 25 January 2012 11:20 AM
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I just assumed that photon-counting technology would be used for this but I guess I was wrong.  I also thought astronomers used CCD and CMOS sensors for photon counting.  I guess it depends on how many electrons one photon produces in the potential well.  Jim probably knows the answer to that.
>
>                                 Guy
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor&  Francis
>      http://www.guycox.com/optical.htm
> ______________________________________________
> Guy Cox, MA, DPhil(Oxon), Honorary Associate,
> Australian Centre for Microscopy&  Microanalysis,
> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>               Mobile 0413 281 861
> ______________________________________________
>        http://www.guycox.net
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
> Sent: Wednesday, 25 January 2012 2:30 AM
> To: [hidden email]
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Essentially yes.
>
> Guy's reply would seem to apply to photon-counting with PMTs when run in photon-counting-mode, not to cameras. Guy, please correct me if I am wrong.
>
> Christian
>
> On 25/01/2012, at 4:18 AM, Roger Phillips wrote:
>
>    
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Thanks, Christian,
>> So the sCMOS photon count would come from the formula
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>> with EM-gain set to 1 and no 'additional noise introduced by the EM gain process'?
>>
>> In Guy's reply, he said 'signals below a certain threshold are regarded as noise, and discarded, and signals above a higher threshold are regarded as 'pile-up' and also discarded.'  This methods seems to account for the 'low threshold' but not for the possible 'pile-up'?
>> Roger
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
>> Sent: 24 January 2012 14:44
>> To: [hidden email]
>> Subject: Re: localization precision in PALM/STORM
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> You basically want to convert the AD counts in your image into photon numbers. That requires several bits of info about the camera. We have Andor cameras and I can look up the required values from the data/test sheet that comes with the individual camera (these change from cam to cam). You need to know the
>> - electrons per count
>> - absolute EM gain (with the quantEM you might have to measure/calibrate this)
>> - A/D offset (this can be measured in dark frames with no light impinging on the cam, i.e. shutter closed)
>>
>> The formula is then something like
>>
>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>>
>> Some more recent camera software packages may have functions to make this conversion for you. I am not sure about photometrics cams/SDKs.
>>
>> Due the additional noise introduced by the EM gain process you should divide the resulting photon-numbers by 2 before looking up values from the Thompson et al. formula. There are papers on EM CCDs that explain this. I also seem to recall that some more recent papers have a subtle correction to the Thompson formula.
>>
>> Hope this helps,
>>
>> Christian
>>
>> On 25/01/2012, at 2:08 AM, Christophe Leterrier wrote:
>>
>>      
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Mark,
>>>
>>> I don't think the quantEM has a built-in photon calibration function, in contrast to the newer Evolve camera, also from Photometrics. The spec sheet for the quantEM is available here :
>>> http://www.photometrics.com/products/datasheets/qem512sc.pdf
>>>
>>> Do I have to calibrate it myself or is Photometrics supposed to provide a photon/intensity calibration curve? I don't want exact experimental values for my precise camera, just a reasonable estimate to derive a theoretical "best" value for localization accuracy.
>>>
>>> Christophe
>>>
>>>
>>> Le mardi 24 janvier 2012 à 10:58, Mark Cannell a écrit :
>>>
>>>        
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> Doesn't the quantEM have a photon calibration function? The background noise should be estimated from the variance of the background (extracted from image regions when/where flashes were not detected...). You can also calibrate the camera with weak sources to double check the manufactures stated read-out calibration.
>>>>
>>>> Hope this helps
>>>>
>>>> Mark
>>>>
>>>>
>>>> On 24/01/2012, at 9:26 AM, Christophe Leterrier wrote:
>>>>
>>>>          
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Hi,
>>>>>
>>>>> Not strictly a confocal question, but I'm pretty sure this list is the best place to get thorough and insightful answers.
>>>>>
>>>>> I have made 2D STORM (stochastic optical reconstruction microscopy) acquisitions and processing and I end up with a table of XY localized fluorophores together with the integrated intensity of the localized diffraction-limited spot.
>>>>>
>>>>> I'd like to plot each fluorophore as a gaussian with a width corresponding to the localization precision, similar to what was done in Bates et al. Science 2007. According to equation (17) in Thompson, Larson&  Webb Biophys J. 2002 (http://goo.gl/5GIXM), this precision depends on the number of photons collected, the width of the diffraction-limited spot, the size of the camera pixel, and the background noise.
>>>>>
>>>>> So my question is : How do I get the number of photons from the intensity level of an image? I'm using a Photometrics 512*512 QuantEM camera. What is the background noise and how do I estimate it? Then using these values in the Thompson et al. equation, I can get a theoretical spot intensity / localization precision calibration curve that I could use for the gaussian-based reconstruction.
>>>>>
>>>>> Thanks for your help,
>>>>>
>>>>> --
>>>>> Christophe Leterrier
>>>>> Researcher
>>>>> Axonal Domains Architecture Team
>>>>> CRN2M CNRS UMR 7286
>>>>> Aix Marseille University, France
>>>>>
>>>>>            
>>>>
>>>>
>>>>          
>>      
>    


--


George McNamara, PhD
Analytical Imaging Core Facility
University of Miami
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: localization precision in PALM/STORM

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

CMOS is intrinsically faster due to each pixel having a readout transistior -it's not a CCD! Of course nothing is for free and the inter pixel gain /offset variations need to be accounted for. I assume they have some memory somewhere of darks and flats -either in the camera or in software? Old CMOS cameras used to bloom at long integration times due to light coming from the readout electronics at the side of the chip. Does that still happen?

Regards Mark

On 28/01/2012, at 1:45 AM, Periasamy, Ammasi (ap3t) wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hamamatsu ORCA-Flash4.0 provides incredible speed. I was testing the camera in the SPIE meeting two days ago.
> 2048x2048, 100 frames/sec. If the size is reduced the frames/sec increases. For 2048x256, 800 frames/sec
>
>
> Dr. Ammasi Periasamy
> Professor & Center Director
> Keck Center for Cellular Imaging (KCCI)
> Biology, Gilmer Hall (064), 485 McCormick Rd
> University of Virginia
> Charlottesville, VA 22904
> (Campus Mail - P.O. Box 400328)
> Voice: 434-243-7602 (Office); 982-4869 (lab)
> Fax:434-982-5210; Email:[hidden email]
> http://www.kcci.virginia.edu/contact/peri.php
> ************************
> 11th Annual Workshop on FRET Microscopy, March 5-10, 2012
> http://www.kcci.virginia.edu/workshop/workshop2012/
> *************************
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara
> Sent: Thursday, January 26, 2012 7:03 AM
> To: [hidden email]
> Subject: Re: localization precision in PALM/STORM
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi David and Listserv,
>
> David wrote: "Photon counting mode on EMCCDs  works by making the em
> gain very high, such that the read noise is much less than 1 photon per
> pixel."
>
> But read noise is measured with no light, so no photons. Just electrons.
> For example, the Photometrics Evolve datasheet:
>
> Read RMS noise (e- rms @ Gain State 3)
> 10 MHz EM Port            45 electrons
>   5 MHz EM Port            32 electrons
>   5 MHz non-EM port     12 electrons
> 1.25 MHz non-EM port    6 electrons
>
> Dark current 0.001 electrons/pixel/sec
>
> Background events
> 0.0045              events/pixel/sec (10 MHz, 1000x EM gain) standard
> operation
> not detectable    events/pixel/sec (10 MHz, 1000x EM gain) BERT operation
>
>    (footnote 4) Dark current -- this is measured in a traditional
>    manner (as with all CCD cameras) by taking a long integration to
>    obtain a signal. An average measurement is taken over the CCD area
>    (excluding blemishes). It should be noted that dark current can vary
>    significantly between different CCDs and the numbers here are typical.
>
>    (footnote 5) Background events -- as EMCCD cameras are actually
>    capable of detecting single photons the real detection limit of
>    these cameras is set by the number of dark background events. These
>    can arise from two things, dark current (which is thermal generation
>    of an electron and is a temperature dependent phenomenon) and also
>    clock induced charge (CIC) electrons (also called spurious charge).
>    Each can lead to the generation of non-photon derived electrons
>    which are multiplied through the electron-multiplication register
>    generating random high
>    value pixels which are above the read noise.
>    These background events are measured by taking 30 ms exposure at
>    10MHz speed with 1000X EM Gain applied and counting the number of
>    random high value pixels which are at a single event threshold above
>    the modal value of the image histogram. This number is expressed as
>    a probability of an event per pixel. The number can vary from frame
>    to frame and sensor to sensor however a typical value is provided.
>
>
> Someone else posted an equation about scientific CMOS camera(s) with "EM
> gain" in the equation ... sCMOS does not have EM gain, so that equation
> would either result in zero or infinity, which would be silly.
>
> Hamamatsu FLASH4.0 read noise (from "Changing the Game" pdf) is less
> than 1.3 electrons (at 100 fps).
>
>
> Best wishes,
>
> George
>
>
> On 1/25/2012 8:47 PM, David Baddeley wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Photon counting mode on EMCCDs  works by making the em gain very high, such that the read noise is much less than 1 photon per pixel. If the photon flux is also very low (much less than 1 photon/pixel/frame) you can threshold each frame and classify each pixel as either having seen a photon or not, and thereby remove the multiplicative noise effects. This only works for very low photon counts as are observed in some aspects of astronomy. In PALM/STORM the photon fluxes are much higher (typically several 100-1000 /pixel/frame) and we can't do photon counting per se. As to EMCCD vs CCD vs SCMOS, they all work with a slight theoretical advantage to SCMOS followed closely by EMCCD. In most practical cases the background is sufficiently high that a small amount of read noise can be tolerated (even as much as the ~5 e- you get from a good conventional CCD @ ~10hz readout - although you'll often want to record faster than 10 hz at which point standard ccds
>>  become less useful). The important factors for PALM/STORM are thus effective quantum efficiency (remembering that the electron multiplying step effectively halves the QE), and readout speed.  I remember hearing that Leica was using a sCMOS device in their system (they've definitely published a proof of concept with one), and when we trialled a CMOS sensor it seemed to work just as well as our EMCCD cameras.
>>
>> As to orientation effects, I was unable to see any difference between linearly and circularly polarised excitation when imaging using antibody conjugated dyes - this suggests that they (at least) are on a flexible linker and free to rotate rapidly. Can't really speculate as to how well this extends to other labelling methods.
>>
>> I'd also suggest investigating other visualisation methods, as the original method of drawing Gaussians has a number of potential limitations, some of which, such as the tendency to inspire more confidence in the data than might be warranted, have already been pointed out (what hasn't been mentioned yet is the fact that it sacrifices resolution if you do happen to have a high density of localisations).  Try a number of methods and see how they work with your data (and ideally also with simulated data of known objects). At the very least do a simple 2D histogram as well as the Gaussian method - this will allow you to directly relate pixel intensity to the number of events within that pixel. If you want to go further, our paper (Baddeley, Cannell, Soeller, 'Visualization of Localization Microscopy Data', Microscopy and Microanalysis 2010) might give you some ideas.
>>
>> Best wishes,
>> David
>>
>>
>> ________________________________
>>  From: Guy Cox<[hidden email]>
>> To: [hidden email]
>> Sent: Wednesday, 25 January 2012 11:20 AM
>> Subject: Re: localization precision in PALM/STORM
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> I just assumed that photon-counting technology would be used for this but I guess I was wrong.  I also thought astronomers used CCD and CMOS sensors for photon counting.  I guess it depends on how many electrons one photon produces in the potential well.  Jim probably knows the answer to that.
>>
>>                                Guy
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor&  Francis
>>     http://www.guycox.com/optical.htm
>> ______________________________________________
>> Guy Cox, MA, DPhil(Oxon), Honorary Associate,
>> Australian Centre for Microscopy&  Microanalysis,
>> Madsen Building F09, University of Sydney, NSW 2006
>>
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>              Mobile 0413 281 861
>> ______________________________________________
>>       http://www.guycox.net
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
>> Sent: Wednesday, 25 January 2012 2:30 AM
>> To: [hidden email]
>> Subject: Re: localization precision in PALM/STORM
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Essentially yes.
>>
>> Guy's reply would seem to apply to photon-counting with PMTs when run in photon-counting-mode, not to cameras. Guy, please correct me if I am wrong.
>>
>> Christian
>>
>> On 25/01/2012, at 4:18 AM, Roger Phillips wrote:
>>
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Thanks, Christian,
>>> So the sCMOS photon count would come from the formula
>>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>>> with EM-gain set to 1 and no 'additional noise introduced by the EM gain process'?
>>>
>>> In Guy's reply, he said 'signals below a certain threshold are regarded as noise, and discarded, and signals above a higher threshold are regarded as 'pile-up' and also discarded.'  This methods seems to account for the 'low threshold' but not for the possible 'pile-up'?
>>> Roger
>>>
>>> -----Original Message-----
>>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Soeller
>>> Sent: 24 January 2012 14:44
>>> To: [hidden email]
>>> Subject: Re: localization precision in PALM/STORM
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> You basically want to convert the AD counts in your image into photon numbers. That requires several bits of info about the camera. We have Andor cameras and I can look up the required values from the data/test sheet that comes with the individual camera (these change from cam to cam). You need to know the
>>> - electrons per count
>>> - absolute EM gain (with the quantEM you might have to measure/calibrate this)
>>> - A/D offset (this can be measured in dark frames with no light impinging on the cam, i.e. shutter closed)
>>>
>>> The formula is then something like
>>>
>>> photons = (counts - A/D offset)*electrons-per-count/EM-gain
>>>
>>> Some more recent camera software packages may have functions to make this conversion for you. I am not sure about photometrics cams/SDKs.
>>>
>>> Due the additional noise introduced by the EM gain process you should divide the resulting photon-numbers by 2 before looking up values from the Thompson et al. formula. There are papers on EM CCDs that explain this. I also seem to recall that some more recent papers have a subtle correction to the Thompson formula.
>>>
>>> Hope this helps,
>>>
>>> Christian
>>>
>>> On 25/01/2012, at 2:08 AM, Christophe Leterrier wrote:
>>>
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> Hi Mark,
>>>>
>>>> I don't think the quantEM has a built-in photon calibration function, in contrast to the newer Evolve camera, also from Photometrics. The spec sheet for the quantEM is available here :
>>>> http://www.photometrics.com/products/datasheets/qem512sc.pdf
>>>>
>>>> Do I have to calibrate it myself or is Photometrics supposed to provide a photon/intensity calibration curve? I don't want exact experimental values for my precise camera, just a reasonable estimate to derive a theoretical "best" value for localization accuracy.
>>>>
>>>> Christophe
>>>>
>>>>
>>>> Le mardi 24 janvier 2012 à 10:58, Mark Cannell a écrit :
>>>>
>>>>
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Doesn't the quantEM have a photon calibration function? The background noise should be estimated from the variance of the background (extracted from image regions when/where flashes were not detected...). You can also calibrate the camera with weak sources to double check the manufactures stated read-out calibration.
>>>>>
>>>>> Hope this helps
>>>>>
>>>>> Mark
>>>>>
>>>>>
>>>>> On 24/01/2012, at 9:26 AM, Christophe Leterrier wrote:
>>>>>
>>>>>
>>>>>> *****
>>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>>> *****
>>>>>>
>>>>>> Hi,
>>>>>>
>>>>>> Not strictly a confocal question, but I'm pretty sure this list is the best place to get thorough and insightful answers.
>>>>>>
>>>>>> I have made 2D STORM (stochastic optical reconstruction microscopy) acquisitions and processing and I end up with a table of XY localized fluorophores together with the integrated intensity of the localized diffraction-limited spot.
>>>>>>
>>>>>> I'd like to plot each fluorophore as a gaussian with a width corresponding to the localization precision, similar to what was done in Bates et al. Science 2007. According to equation (17) in Thompson, Larson&  Webb Biophys J. 2002 (http://goo.gl/5GIXM), this precision depends on the number of photons collected, the width of the diffraction-limited spot, the size of the camera pixel, and the background noise.
>>>>>>
>>>>>> So my question is : How do I get the number of photons from the intensity level of an image? I'm using a Photometrics 512*512 QuantEM camera. What is the background noise and how do I estimate it? Then using these values in the Thompson et al. equation, I can get a theoretical spot intensity / localization precision calibration curve that I could use for the gaussian-based reconstruction.
>>>>>>
>>>>>> Thanks for your help,
>>>>>>
>>>>>> --
>>>>>> Christophe Leterrier
>>>>>> Researcher
>>>>>> Axonal Domains Architecture Team
>>>>>> CRN2M CNRS UMR 7286
>>>>>> Aix Marseille University, France
>>>>>>
>>>>>>
>>>>>
>>>>>
>>>>>
>>>
>>
>
>
> --
>
>
> George McNamara, PhD
> Analytical Imaging Core Facility
> University of Miami
12