Sylvie Le Guyader |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all One of our users needs to culture a piece of tissue on some sort of membrane. I am wondering if anyone knows where we could find such a membrane. The requirements are: - It should be compatible with tissue culture (not toxic) - It should float (so that the top of the tissue is in the air and the bottom touches the medium) - It should be somehow rigid (so that it doesn't curl up) - It should not have any high edges like transwell inserts (because we need to image it on an upright system with a short working distance objective) Would anyone know of something we can use? thanks Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Unit Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 5248 1107 LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> |
WILDE Geraint |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I my past research days I grew organotypic hippocampal cultures which required the precise conditions you list below, and also used them so that I could image the cultures (LWD objectives at the time on an inverted scope). I used these. http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Inserts,MM_NF-C10504?bd=1 It seems they are still manufactured even after all these years. Best Geraint Product Manager for Microscopy Systems Andor Technology Part of the Oxford Instruments Group -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: 09 September 2015 08:09 To: [hidden email] Subject: [CONFOCALMICROSCOPY] membrane for tissue culture ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all One of our users needs to culture a piece of tissue on some sort of membrane. I am wondering if anyone knows where we could find such a membrane. The requirements are: - It should be compatible with tissue culture (not toxic) - It should float (so that the top of the tissue is in the air and the bottom touches the medium) - It should be somehow rigid (so that it doesn't curl up) - It should not have any high edges like transwell inserts (because we need to image it on an upright system with a short working distance objective) Would anyone know of something we can use? thanks Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Unit Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 5248 1107 LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> This message has been scanned for malware by SurfControl plc. www.surfcontrol.com |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I'm interested to have the same performances for the cell cultures I use. Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. Best regards, Dario DARIO DONNARUMMA POSTDOC Laboratoire Charles Coulomb (L2C) Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 46 37 [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> UNIVERSITÉ DE MONTPELLIER CC 0026 - Place Eugène Bataillon 34095 Montpellier Cedex 5 WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I my past research days I grew organotypic hippocampal cultures which > required the precise conditions you list below, and also used them so that > I could image the cultures (LWD objectives at the time on an inverted > scope). I used these. > http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Inserts,MM_NF-C10504?bd=1 > It seems they are still manufactured even after all these years. > > Best > > Geraint > > Product Manager for Microscopy Systems > > Andor Technology > Part of the Oxford Instruments Group > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Sylvie Le Guyader > Sent: 09 September 2015 08:09 > To: [hidden email] > Subject: [CONFOCALMICROSCOPY] membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of > membrane. I am wondering if anyone knows where we could find such a > membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air > and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts > (because we need to image it on an upright system with a short working > distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > > This message has been scanned for malware by SurfControl plc. > www.surfcontrol.com > |
WILDE Geraint |
Sorry, I should probably have added that the transwell I referred to, if you pick the appropriate size (e.g. for a 6-well plate) would allow access for a shorter working distance objective on an upright microscope.
-----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dar Do Sent: 09 September 2015 11:12 To: [hidden email] Subject: Re: [CONFOCALMICROSCOPY] membrane for tissue culture ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I'm interested to have the same performances for the cell cultures I use. Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. Best regards, Dario DARIO DONNARUMMA POSTDOC Laboratoire Charles Coulomb (L2C) Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 46 37 [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> UNIVERSITÉ DE MONTPELLIER CC 0026 - Place Eugène Bataillon 34095 Montpellier Cedex 5 WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I my past research days I grew organotypic hippocampal cultures which > required the precise conditions you list below, and also used them so > that I could image the cultures (LWD objectives at the time on an > inverted scope). I used these. > http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Ins > erts,MM_NF-C10504?bd=1 It seems they are still manufactured even after > all these years. > > Best > > Geraint > > Product Manager for Microscopy Systems > > Andor Technology > Part of the Oxford Instruments Group > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Sylvie Le Guyader > Sent: 09 September 2015 08:09 > To: [hidden email] > Subject: [CONFOCALMICROSCOPY] membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of > membrane. I am wondering if anyone knows where we could find such a > membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air > and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts > (because we need to image it on an upright system with a short working > distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > > This message has been scanned for malware by SurfControl plc. > www.surfcontrol.com > |
Sylvie Le Guyader |
Hi Geraint
The problem with the transwells you suggest is that on an upright microscope, especially with a short WD objective, the objective will bump into the wall of the well before you reach the focus plane. Or did I misunderstand? We use a Zeiss 20x water dipping lens so it is quite wide. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Unit Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 5248 1107 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of WILDE Geraint Sent: den 9 september 2015 12:22 To: [hidden email] Subject: Re: membrane for tissue culture Sorry, I should probably have added that the transwell I referred to, if you pick the appropriate size (e.g. for a 6-well plate) would allow access for a shorter working distance objective on an upright microscope. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dar Do Sent: 09 September 2015 11:12 To: [hidden email] Subject: Re: [CONFOCALMICROSCOPY] membrane for tissue culture ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I'm interested to have the same performances for the cell cultures I use. Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. Best regards, Dario DARIO DONNARUMMA POSTDOC Laboratoire Charles Coulomb (L2C) Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 46 37 [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> UNIVERSITÉ DE MONTPELLIER CC 0026 - Place Eugène Bataillon 34095 Montpellier Cedex 5 WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I my past research days I grew organotypic hippocampal cultures which > required the precise conditions you list below, and also used them so > that I could image the cultures (LWD objectives at the time on an > inverted scope). I used these. > http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Ins > erts,MM_NF-C10504?bd=1 It seems they are still manufactured even after > all these years. > > Best > > Geraint > > Product Manager for Microscopy Systems > > Andor Technology > Part of the Oxford Instruments Group > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Sylvie Le Guyader > Sent: 09 September 2015 08:09 > To: [hidden email] > Subject: [CONFOCALMICROSCOPY] membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of > membrane. I am wondering if anyone knows where we could find such a > membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air > and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts > (because we need to image it on an upright system with a short working > distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > > This message has been scanned for malware by SurfControl plc. > www.surfcontrol.com > |
Sylvie Le Guyader |
In reply to this post by Dar Do
Thanks to everyone for the very informative answers. These are really nice suggestions.
Jean-Marie, do you have any experience with the SimPore membranes? Are they compatible with cell culture? And do you know if it is possible to remove the reservoirs from the Ibidi slides to allow more space for the objective? Dario: interesting product! Thanks. There is no info on the Sigma website about the high of the walls around the membrane after it has been popped out of the insert. I said in my answer to Geraint that we use the 20x dipping lens but that was a mistake. We use a 20x dry lens NA 0.6 WD 1.7 mm. Quite a lot of WD but considering that the insert is 12mm in diameter, the walls still look too high to me. do you have any idea how high they are? Thanks everyone! :) Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Unit Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 5248 1107 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dar Do Sent: den 9 september 2015 12:12 To: [hidden email] Subject: Re: membrane for tissue culture ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I'm interested to have the same performances for the cell cultures I use. Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. Best regards, Dario DARIO DONNARUMMA POSTDOC Laboratoire Charles Coulomb (L2C) Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 46 37 [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> UNIVERSITÉ DE MONTPELLIER CC 0026 - Place Eugène Bataillon 34095 Montpellier Cedex 5 WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I my past research days I grew organotypic hippocampal cultures which > required the precise conditions you list below, and also used them so > that I could image the cultures (LWD objectives at the time on an > inverted scope). I used these. > http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Ins > erts,MM_NF-C10504?bd=1 It seems they are still manufactured even after > all these years. > > Best > > Geraint > > Product Manager for Microscopy Systems > > Andor Technology > Part of the Oxford Instruments Group > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Sylvie Le Guyader > Sent: 09 September 2015 08:09 > To: [hidden email] > Subject: [CONFOCALMICROSCOPY] membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of > membrane. I am wondering if anyone knows where we could find such a > membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air > and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts > (because we need to image it on an upright system with a short working > distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > > This message has been scanned for malware by SurfControl plc. > www.surfcontrol.com > |
Giovanna Expósito Romero |
In reply to this post by Sylvie Le Guyader
Dear Sylvie,
You could try with MILLICELL-CM 0.4um, PICM0RG50 with 35mm dishes. They have 30mm diameter and just 5mm height. https://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Insert%2C-30%C2%A0mm%2C-hydrophilic-PTFE%2C-0.4%C2%A0%C2%B5m,MM_NF-PICM0RG50#documentation We normally use this membranes on an inverted microscope with long working distance objectives and 35mm dishes (MaTtek) for brain slice culture. I think they should work for your application also. Best regards, Giovanna Giovanna Expósito Servicio de Imagen Instituto de Neurociencias Telf: +34 965919256 Avd. Ramón y Cajal s/n Campus de San Juan 03550 Sant Joan d´Alacant (Alicante) Spain -----Mensaje original----- De: Confocal Microscopy List [mailto:[hidden email]] En nombre de Sylvie Le Guyader Enviado el: miércoles, 09 de septiembre de 2015 12:30 Para: [hidden email] Asunto: Re: membrane for tissue culture Hi Geraint The problem with the transwells you suggest is that on an upright microscope, especially with a short WD objective, the objective will bump into the wall of the well before you reach the focus plane. Or did I misunderstand? We use a Zeiss 20x water dipping lens so it is quite wide. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Unit Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 5248 1107 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of WILDE Geraint Sent: den 9 september 2015 12:22 To: [hidden email] Subject: Re: membrane for tissue culture Sorry, I should probably have added that the transwell I referred to, if you pick the appropriate size (e.g. for a 6-well plate) would allow access for a shorter working distance objective on an upright microscope. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dar Do Sent: 09 September 2015 11:12 To: [hidden email] Subject: Re: [CONFOCALMICROSCOPY] membrane for tissue culture ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, I'm interested to have the same performances for the cell cultures I use. Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. Best regards, Dario DARIO DONNARUMMA POSTDOC Laboratoire Charles Coulomb (L2C) Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 46 37 [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> UNIVERSITÉ DE MONTPELLIER CC 0026 - Place Eugène Bataillon 34095 Montpellier Cedex 5 WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I my past research days I grew organotypic hippocampal cultures which > required the precise conditions you list below, and also used them so > that I could image the cultures (LWD objectives at the time on an > inverted scope). I used these. > http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Ins > erts,MM_NF-C10504?bd=1 It seems they are still manufactured even after > all these years. > > Best > > Geraint > > Product Manager for Microscopy Systems > > Andor Technology > Part of the Oxford Instruments Group > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Sylvie Le Guyader > Sent: 09 September 2015 08:09 > To: [hidden email] > Subject: [CONFOCALMICROSCOPY] membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of > membrane. I am wondering if anyone knows where we could find such a > membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air > and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts > (because we need to image it on an upright system with a short working > distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > > This message has been scanned for malware by SurfControl plc. > www.surfcontrol.com > |
George McNamara |
In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Read Wakabayashi et al 2007 and use their method of growing cells on the underside of the transwell, and imaging that: http://www.ncbi.nlm.nih.gov/pubmed/17308935 Histochem Cell Biol. 2007 May;127(5):463-72. Epub 2007 Feb 17. Four-dimensional imaging of filter-grown polarized epithelial cells. Wakabayashi Y(1), Chua J, Larkin JM, Lippincott-Schwartz J, Arias IM. Author information: (1)Cell Biology and Metabolism Branch, National Institute of Child Health and Human Development, National Institutes of Health, Bethesda, MD 20892, USA. [hidden email] Understanding how epithelial cells generate and maintain polarity and function requires live cell imaging. In order for cells to become fully polarized, it is necessary to grow them on a permeable membrane filter; however, the translucent filter obstructs the microscope light path required for quantitative live cell imaging. Alternatively, the membrane filter may be excised but this eliminates selective access to apical and basolateral surfaces. Conversely, epithelial cells cultured directly on glass exhibit different phenotypes and functions from filter grown cells. Here, we describe a new method for culturing polarized epithelial cells on a Transwell filter insert that allows superior live cell imaging with spatial and temporal image resolution previously unachievable using conventional methods. Cells were cultured on the underside of a filter support. Epithelial cells grown in this inverted configuration exhibit a fully polarized architecture, including the presence of functional tight junctions. This new culturing system permits four-dimensional (three spatial dimension over time) imaging of endosome and Golgi apparatus dynamics, and permits selective manipulation of the apical and basolateral surfaces. This new technique has wide applicability for visualization and manipulation of polarized epithelial cells. PMID: 17308935 [PubMed - indexed for MEDLINE] By now they may have published more such papers, and other groups may have published and cited their work. Enjoy, George p.s. you do not have to stand on your heads while imaging. On 9/9/2015 2:08 AM, Sylvie Le Guyader wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > > > One of our users needs to culture a piece of tissue on some sort of membrane. I am wondering if anyone knows where we could find such a membrane. > > > > The requirements are: > > - It should be compatible with tissue culture (not toxic) > > - It should float (so that the top of the tissue is in the air and the bottom touches the medium) > > - It should be somehow rigid (so that it doesn't curl up) > > - It should not have any high edges like transwell inserts (because we need to image it on an upright system with a short working distance objective) > > > > Would anyone know of something we can use? > > > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/42 |
Glen MacDonald-2 |
In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Sylvie, If you use the flexible membrane inserts, they can be removed from the polystyrene wall with a scalpel or small scissors. Then place in a petri dish for imaging with an upright system, or in a large coverslip culture chamber to image on an inverted system. The membrane can be returned to culture by placing onto another transwell insert. Contamination is a concern, but we had reasonable success with brain slices. Another option is to culture the inserts upside down. Regards, Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Cellular Morphology Core Center on Human Development and Disability Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] On Sep 9, 2015, at 3:39 AM, Sylvie Le Guyader <[hidden email]> wrote: > Thanks to everyone for the very informative answers. These are really nice suggestions. > Jean-Marie, do you have any experience with the SimPore membranes? Are they compatible with cell culture? > And do you know if it is possible to remove the reservoirs from the Ibidi slides to allow more space for the objective? > > Dario: interesting product! Thanks. There is no info on the Sigma website about the high of the walls around the membrane after it has been popped out of the insert. I said in my answer to Geraint that we use the 20x dipping lens but that was a mistake. We use a 20x dry lens NA 0.6 WD 1.7 mm. Quite a lot of WD but considering that the insert is 12mm in diameter, the walls still look too high to me. do you have any idea how high they are? > > Thanks everyone! :) > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Unit Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7, > Novum, G lift, floor 6 > 14157 Huddinge > Sweden > mobile: +46 (0) 73 733 5008 > office: +46 (0) 8 5248 1107 > LCI website > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dar Do > Sent: den 9 september 2015 12:12 > To: [hidden email] > Subject: Re: membrane for tissue culture > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Sylvie, > > I'm interested to have the same performances for the cell cultures I use. > Right now we are using Multiwell with Polycarbonate Membrane Transwell® Inserts with a long range objective, but we would like to work with a short range objective also. So we are thinking about to use Snapwell™ Inserts ( http://www.sigmaaldrich.com/catalog/product/sigma/cls3407?lang=fr®ion=FR), > that give you the possibility to detach the membrane and place it on a glass slide. My doubt are about contamination and sollicitation of the cells, so we never try this solution so far, but I think we will in the near future. > > Best regards, > > Dario > > DARIO DONNARUMMA > > POSTDOC > Laboratoire Charles Coulomb (L2C) > > Tél : +33 (0) 4 67 14 38 60 <+33+(0)+4+67+14+38+60> | Fax : +33 (0) 4 67 14 > 46 37 > > [hidden email] <[hidden email]> <http://www.umontpellier.fr/> <http://www.umontpellier.fr/> > > UNIVERSITÉ DE MONTPELLIER > > CC 0026 - Place Eugène Bataillon > 34095 Montpellier Cedex 5 > > WWW.UMONTPELLIER.FR <http://www.umontpellier.fr/> > > 2015-09-09 10:05 GMT+02:00 WILDE Geraint <[hidden email]>: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear Sylvie, >> >> I my past research days I grew organotypic hippocampal cultures which >> required the precise conditions you list below, and also used them so >> that I could image the cultures (LWD objectives at the time on an >> inverted scope). I used these. >> http://www.merckmillipore.com/GB/en/product/Millicell-Cell-Culture-Ins >> erts,MM_NF-C10504?bd=1 It seems they are still manufactured even after >> all these years. >> >> Best >> >> Geraint >> >> Product Manager for Microscopy Systems >> >> Andor Technology >> Part of the Oxford Instruments Group >> >> -----Original Message----- >> From: Confocal Microscopy List >> [mailto:[hidden email]] >> On Behalf Of Sylvie Le Guyader >> Sent: 09 September 2015 08:09 >> To: [hidden email] >> Subject: [CONFOCALMICROSCOPY] membrane for tissue culture >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear all >> >> >> >> One of our users needs to culture a piece of tissue on some sort of >> membrane. I am wondering if anyone knows where we could find such a >> membrane. >> >> >> >> The requirements are: >> >> - It should be compatible with tissue culture (not toxic) >> >> - It should float (so that the top of the tissue is in the air >> and the bottom touches the medium) >> >> - It should be somehow rigid (so that it doesn't curl up) >> >> - It should not have any high edges like transwell inserts >> (because we need to image it on an upright system with a short working >> distance objective) >> >> >> >> Would anyone know of something we can use? >> >> >> >> thanks >> >> Med vänlig hälsning / Best regards >> >> Sylvie >> >> @@@@@@@@@@@@@@@@@@@@@@@@ >> Sylvie Le Guyader, PhD >> Live Cell Imaging Unit Manager >> Karolinska Institutet- Bionut Dpt >> Hälsovägen 7, >> Novum, G lift, floor 6 >> 14157 Huddinge >> Sweden >> mobile: +46 (0) 73 733 5008 >> office: +46 (0) 8 5248 1107 >> LCI website<http://ki.se/en/bionut/welcome-to-the-lci-unit> >> >> >> This message has been scanned for malware by SurfControl plc. >> www.surfcontrol.com >> |
Free forum by Nabble | Edit this page |