Axel Kurt Preuss |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear List Would anybody know a vendor who offers Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm 0.5 x 3 inch 12.7 x 76.19 mm I found following article I cant get a hold of, but Corning assured me they don't have these slides. Does anybody have that article? A large format membrane-based x-ray mask for microfluidic chip ... by L Wang - 2006 - Cited by 2 - Related articles Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm were... Figure 3(a) is a silicon nitride x-ray mask with a window ... iopscience.iop.org/0960-1317/16/2/026/pdf/jmm6_2_026.pdf Thanks, Cheers Axel ------------------ Axel K Preuss, PhD Central Imaging Facility -- A*STAR Institute of Molecular and Cell Biology IMCB - 6-19B -- Singapore - Republic of Singapore 138673 Tel +65 6586.9711 Fax +65 6779.1117 mobile +65 9271.5622 Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you. |
|
Julio Vazquez |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Axel, I can't get the article unless I pay $ 30.00, but I'm wondering whether this is just a typo. 1" x 3" x 1 mm (or 25 mm x 75 mm x 1 mm) would be a standard microscope slide. This is also the format typically used for micro arrays. Does it look from the figure that their slide is half the width of a normal slide? Does it need to be?.... -- Julio Vazquez Fred Hutchinson Cancer Research Center 1100 Fairview Ave. N., mailstop DE-512 Seattle, WA 98109-1024 http://www.fhcrc.org On Dec 10, 2010, at 2:37 AM, Axel Kurt Preuss wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear List > > Would anybody know a vendor who offers > > Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch > and a thickness of 1 mm > 0.5 x 3 inch > 12.7 x 76.19 mm > > I found following article I cant get a hold of, but Corning assured > me they don't have these slides. Does anybody have that article? > > > A large format membrane-based x-ray mask for microfluidic chip ... > by L Wang - 2006 - Cited by 2 - Related articles > Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch > and a thickness of 1 mm were... Figure 3(a) is a silicon nitride x- > ray mask with a window ... > iopscience.iop.org/0960-1317/16/2/026/pdf/jmm6_2_026.pdf > > > Thanks, > Cheers > > Axel > ------------------ > Axel K Preuss, PhD Central Imaging Facility -- A*STAR Institute > of Molecular and Cell Biology IMCB - 6-19B -- Singapore - > Republic of Singapore 138673 Tel +65 6586.9711 Fax +65 > 6779.1117 mobile +65 9271.5622 > > > > Note: This message may contain confidential information. If this > Email/Fax has been sent to you by mistake, please notify the sender > and delete it immediately. Thank you. |
|
Julio Vazquez |
|
|
In reply to this post by Axel Kurt Preuss
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Of course, I suppose you can also have them cut by a local glass or machine shop. A Google search for "custom microscope slide" gave me a few hits, such as this company that creates custom microfluidics slides: http://www.translume.com/slides.htm and this site lists slide/glass manufacturers: http://www.thomasnet.com/products/microscope-slides-75850602-1.html -- Julio Vazquez Fred Hutchinson Cancer Research Center Seattle, WA 98109-1024 http://www.fhcrc.org/ On Dec 10, 2010, at 2:37 AM, Axel Kurt Preuss wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear List > > Would anybody know a vendor who offers > > Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch > and a thickness of 1 mm > 0.5 x 3 inch > 12.7 x 76.19 mm > > I found following article I cant get a hold of, but Corning assured > me they don't have these slides. Does anybody have that article? > > > A large format membrane-based x-ray mask for microfluidic chip ... > by L Wang - 2006 - Cited by 2 - Related articles > Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch > and a thickness of 1 mm were... Figure 3(a) is a silicon nitride x- > ray mask with a window ... > iopscience.iop.org/0960-1317/16/2/026/pdf/jmm6_2_026.pdf > > > Thanks, > Cheers > > Axel > ------------------ > Axel K Preuss, PhD Central Imaging Facility -- A*STAR Institute > of Molecular and Cell Biology IMCB - 6-19B -- Singapore - > Republic of Singapore 138673 Tel +65 6586.9711 Fax +65 > 6779.1117 mobile +65 9271.5622 > > > > Note: This message may contain confidential information. If this > Email/Fax has been sent to you by mistake, please notify the sender > and delete it immediately. Thank you. |
|
|
In reply to this post by Axel Kurt Preuss
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Are you sure it's not just a typo in the paper? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Axel Kurt Preuss Sent: Friday, 10 December 2010 9:37 PM To: [hidden email] Subject: microscope glass slides 0.5 inch × 3 inch and a thicknes s of 1 mm ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear List Would anybody know a vendor who offers Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm 0.5 x 3 inch 12.7 x 76.19 mm I found following article I cant get a hold of, but Corning assured me they don't have these slides. Does anybody have that article? A large format membrane-based x-ray mask for microfluidic chip ... by L Wang - 2006 - Cited by 2 - Related articles Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm were... Figure 3(a) is a silicon nitride x-ray mask with a window ... iopscience.iop.org/0960-1317/16/2/026/pdf/jmm6_2_026.pdf Thanks, Cheers Axel ------------------ Axel K Preuss, PhD Central Imaging Facility -- A*STAR Institute of Molecular and Cell Biology IMCB - 6-19B -- Singapore - Republic of Singapore 138673 Tel +65 6586.9711 Fax +65 6779.1117 mobile +65 9271.5622 Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you. |
|
Ignatius, Mike-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** This is a contact for someone who can create this size glass slide as a custom if it isn't a typo. [hidden email]. No commercial interest. Mike Ignatius Molecular Probes/Life Ttechnologies Inc. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Friday, December 10, 2010 8:38 PM To: [hidden email] Subject: Re: microscope glass slides 0.5 inch × 3 inch and a thicknes s of 1 mm ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Are you sure it's not just a typo in the paper? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Axel Kurt Preuss Sent: Friday, 10 December 2010 9:37 PM To: [hidden email] Subject: microscope glass slides 0.5 inch × 3 inch and a thicknes s of 1 mm ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear List Would anybody know a vendor who offers Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm 0.5 x 3 inch 12.7 x 76.19 mm I found following article I cant get a hold of, but Corning assured me they don't have these slides. Does anybody have that article? A large format membrane-based x-ray mask for microfluidic chip ... by L Wang - 2006 - Cited by 2 - Related articles Glass slides (Corning Inc. [8]) with a size of 0.5 inch × 3 inch and a thickness of 1 mm were... Figure 3(a) is a silicon nitride x-ray mask with a window ... iopscience.iop.org/0960-1317/16/2/026/pdf/jmm6_2_026.pdf Thanks, Cheers Axel ------------------ Axel K Preuss, PhD Central Imaging Facility -- A*STAR Institute of Molecular and Cell Biology IMCB - 6-19B -- Singapore - Republic of Singapore 138673 Tel +65 6586.9711 Fax +65 6779.1117 mobile +65 9271.5622 Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you. |