Dear list friends,

How do I calculate the divergence of fluorescence from a
diffraction-limited spot in an infinity focus system with
increasing path length?
Thanks for help,
Roger

Dr Roger Guy Phillips
Centre for Advanced Microscopy,
University of Sussex
School of Life Sciences
John Maynard Smith Building
Falmer, Brighton & Hove
BN1 9QG
United Kingdom

phone:44 (0)1273 877585
fax: 44 (0)1273 678433
email: [hidden email]
room:2C9 (ext 7585)/lab 4C2 (ext 2734)  
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

(no subject)

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Roger,

        Not sure about the question.  To the best
of my understanding (and I'm not a physicist), within
the 'infinity space' the rays from a spot will not
diverge but will remain parallel.  They will
diverge from the optic axis depending on the
distance of the spot from the centre of the field.
(That's why the 'infinity space' is actually very
finite.)

                                        Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Roger Phillips
Sent: Thursday, 24 January 2008 10:54 PM
To: [hidden email]
Subject:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear list friends,

How do I calculate the divergence of fluorescence from a diffraction-limited spot in an infinity focus system with increasing path length?
Thanks for help,
Roger

Dr Roger Guy Phillips
Centre for Advanced Microscopy,
University of Sussex
School of Life Sciences
John Maynard Smith Building
Falmer, Brighton & Hove
BN1 9QG
United Kingdom

phone:44 (0)1273 877585
fax: 44 (0)1273 678433
email: [hidden email]
room:2C9 (ext 7585)/lab 4C2 (ext 2734)  

No virus found in this incoming message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.10/1240 - Release Date: 23/01/2008 5:47 PM
 

No virus found in this outgoing message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.10/1240 - Release Date: 23/01/2008 5:47 PM
 
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

(no subject)

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi

Yes, I think that is right, the finite field size limits the collimation angle to field size/focal length.

Cheers Mark


-----Original Message-----
From: Confocal Microscopy List on behalf of Guy Cox
Sent: Fri 1/25/2008 1:09 AM
To: [hidden email]
Subject:
 
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Roger,

        Not sure about the question.  To the best
of my understanding (and I'm not a physicist), within
the 'infinity space' the rays from a spot will not
diverge but will remain parallel.  They will
diverge from the optic axis depending on the
distance of the spot from the centre of the field.
(That's why the 'infinity space' is actually very
finite.)

                                        Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Roger Phillips
Sent: Thursday, 24 January 2008 10:54 PM
To: [hidden email]
Subject:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear list friends,

How do I calculate the divergence of fluorescence from a diffraction-limited spot in an infinity focus system with increasing path length?
Thanks for help,
Roger

Dr Roger Guy Phillips
Centre for Advanced Microscopy,
University of Sussex
School of Life Sciences
John Maynard Smith Building
Falmer, Brighton & Hove
BN1 9QG
United Kingdom

phone:44 (0)1273 877585
fax: 44 (0)1273 678433
email: [hidden email]
room:2C9 (ext 7585)/lab 4C2 (ext 2734)  

No virus found in this incoming message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.10/1240 - Release Date: 23/01/2008 5:47 PM
 

No virus found in this outgoing message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.10/1240 - Release Date: 23/01/2008 5:47 PM