						yuansheng sun | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Listers, I have a question about filling the one-photon laser to the objective back aperture in a confocal. There is a paper (as shown below) which suggests that underfill will yield a better Guassian excitation volume for FCS measurements. Underfill will also give more power. However, would it be tricky for the alignment when you multiple lasers and dicrhoics? Thanks. The paper - Focal Volume Optics and Experimental Artifacts in Confocal Fluorescence Correlation Spectroscopy, Samuel T. Hess and Watt W. Webb, BJ 83: 2300-2317, 2002. Sheng  | 
			
 
	
	 | 
						James Pawley | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear Listers, > >I have a question about filling the one-photon laser to the objective back >aperture in a confocal. There is a paper (as shown below) which suggests >that underfill will yield a better Guassian excitation volume for FCS >measurements. Underfill will also give more power. However, would it be >tricky for the alignment when you multiple lasers and dicrhoics? Thanks. > >The paper - Focal Volume Optics and Experimental Artifacts in Confocal >Fluorescence Correlation Spectroscopy, Samuel T. Hess and Watt W. Webb, BJ >83: 2300-2317, 2002. > >Sheng Dear Sheng, It is true that a Gaussian distribution of light in the BFP will produce a Gaussian distribution of light in the image plane. Of course, as a Gaussian distrubution goes to infinity, no matter how you choose the "size" of the central spot to be in the BFP, you will always cut off some of the Gaussian and so not quite have a Gaussian in the image plane. But the full-width-at-half-maximum of the Gaussian in the image plane will be larger than the FWHM of the almost-Airy spot that you would get by overfilling the BFP (The BFP intensity would have to be completely uniform to be a true Airy. This is only approached by massive overfilling.). Of course, the AIry Disk has some light (about 20%) in the outer rings and these will reduce contrast but as the intensity (photons/µm2) is only about 1% of that in the central peak, they don't excite many photons. In terms of getting more light through the objective, of course light that hits glass may be transmitted while that which hits metal will not be. However, we seldom have the problem of not enough light. Indeed, even 1mw in a 0.5 µm FWHM spot it close to singlet-state saturation with most dies and you should try to stay at least 10 (preferably 100x) below that. I would over fill about 2x. Best, JP -- *************************************************************************** Prof. James B. Pawley, Ph. 608-238-3953 21. N. Prospect Ave. Madison, WI 53726 USA [hidden email] 3D Microscopy of Living Cells Course, June 9-21, 2012, UBC, Vancouver Canada Info: http://www.3dcourse.ubc.ca/ Application deadline 3/16/2012 "If it ain't diffraction, it must be statistics." Anon. 11/16/12  | 
			
 
	
	 | 
						Sudipta Maiti | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Also, for FCS, the easy way to get rid of the artifacts caused by overfilling is to use a confical pinhole that is small. The benefit is higher resolution, lower background and the ability to perform FCS at higher concentrations. Sudipta Fri, 16 Dec 2011 09:51:52 -0600, James Pawley wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > >***** > >To join, leave or search the confocal microscopy listserv, go to: > >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >***** > > > >Dear Listers, > > > >I have a question about filling the one-photon laser to the objective back > >aperture in a confocal. There is a paper (as shown below) which suggests > >that underfill will yield a better Guassian excitation volume for FCS > >measurements. Underfill will also give more power. However, would it be > >tricky for the alignment when you multiple lasers and dicrhoics? Thanks. > > > >The paper - Focal Volume Optics and Experimental Artifacts in Confocal > >Fluorescence Correlation Spectroscopy, Samuel T. Hess and Watt W. Webb, BJ > >83: 2300-2317, 2002. > > > >Sheng > > Dear Sheng, > > It is true that a Gaussian distribution of light > in the BFP will produce a Gaussian distribution > of light in the image plane. > > Of course, as a Gaussian distrubution goes to > infinity, no matter how you choose the "size" of > the central spot to be in the BFP, you will > always cut off some of the Gaussian and so not > quite have a Gaussian in the image plane. > > But the full-width-at-half-maximum of the > Gaussian in the image plane will be larger than > the FWHM of the almost-Airy spot that you would > get by overfilling the BFP (The BFP intensity > would have to be completely uniform to be a true > Airy. This is only approached by massive > overfilling.). > > Of course, the AIry Disk has some light (about > 20%) in the outer rings and these will reduce > contrast but as the intensity (photons/µm2) is > only about 1% of that in the central peak, they > don't excite many photons. > > In terms of getting more light through the > objective, of course light that hits glass may be > transmitted while that which hits metal will not > be. However, we seldom have the problem of not > enough light. Indeed, even 1mw in a 0.5 µm FWHM > spot it close to singlet-state saturation with > most dies and you should try to stay at least 10 > (preferably 100x) below that. > > I would over fill about 2x. > > Best, > > JP > -- > *************************************************************************** > Prof. James B. Pawley, > Ph. 608-238-3953 > 21. N. Prospect Ave. Madison, WI 53726 USA > [hidden email] > 3D Microscopy of Living Cells Course, June 9-21, 2012, UBC, > Vancouver Canada Info: http://www.3dcourse.ubc.ca/ Application > deadline 3/16/2012 "If it ain't diffraction, it must be > statistics." Anon. 11/16/12 Dr. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.weebly.com  | 
			
 
	
	 | 
						yuansheng sun | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear JP and Sudipta, Thanks a lot for your suggestions. Due to the cut-off at BFP, the excitation volume is not quite Guassian in the one-photon case, as JP pointed out. Most of FCS data fittings assume a Gaussian excitation volume characterized by w0 and z0 (e-2). The paper I mentioned earlier suggested that underfilling could improve on this. But certainly there are problems with underfilling, e.g. uneven illumination. Thus, I posted the question whether underfilling would be a practical way of getting a better Gaussian excitation volume in a single-photon confocal. I feel that Sudipta's small pinhole with overfilling is a better practical solution. Sheng On Fri, Dec 16, 2011 at 1:03 PM, Sudipta Maiti <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Also, for FCS, the easy way to get rid of the artifacts caused by > overfilling is to use a confical > pinhole that is small. The benefit is higher resolution, lower background > and the ability to perform > FCS at higher concentrations. > Sudipta > Fri, 16 Dec 2011 09:51:52 -0600, James Pawley wrote > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > >***** > > >To join, leave or search the confocal microscopy listserv, go to: > > >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > >***** > > > > > >Dear Listers, > > > > > >I have a question about filling the one-photon laser to the objective > back > > >aperture in a confocal. There is a paper (as shown below) which > suggests > > >that underfill will yield a better Guassian excitation volume for FCS > > >measurements. Underfill will also give more power. However, would it > be > > >tricky for the alignment when you multiple lasers and dicrhoics? > Thanks. > > > > > >The paper - Focal Volume Optics and Experimental Artifacts in Confocal > > >Fluorescence Correlation Spectroscopy, Samuel T. Hess and Watt W. Webb, > BJ > > >83: 2300-2317, 2002. > > > > > >Sheng > > > > Dear Sheng, > > > > It is true that a Gaussian distribution of light > > in the BFP will produce a Gaussian distribution > > of light in the image plane. > > > > Of course, as a Gaussian distrubution goes to > > infinity, no matter how you choose the "size" of > > the central spot to be in the BFP, you will > > always cut off some of the Gaussian and so not > > quite have a Gaussian in the image plane. > > > > But the full-width-at-half-maximum of the > > Gaussian in the image plane will be larger than > > the FWHM of the almost-Airy spot that you would > > get by overfilling the BFP (The BFP intensity > > would have to be completely uniform to be a true > > Airy. This is only approached by massive > > overfilling.). > > > > Of course, the AIry Disk has some light (about > > 20%) in the outer rings and these will reduce > > contrast but as the intensity (photons/痠2) is > > only about 1% of that in the central peak, they > > don't excite many photons. > > > > In terms of getting more light through the > > objective, of course light that hits glass may be > > transmitted while that which hits metal will not > > be. However, we seldom have the problem of not > > enough light. Indeed, even 1mw in a 0.5 痠 FWHM > > spot it close to singlet-state saturation with > > most dies and you should try to stay at least 10 > > (preferably 100x) below that. > > > > I would over fill about 2x. > > > > Best, > > > > JP > > -- > > > *************************************************************************** > > Prof. James B. Pawley, > > Ph. 608-238-3953 > > 21. N. Prospect Ave. Madison, WI 53726 USA > > [hidden email] > > 3D Microscopy of Living Cells Course, June 9-21, 2012, UBC, > > Vancouver Canada Info: http://www.3dcourse.ubc.ca/ Application > > deadline 3/16/2012 "If it ain't diffraction, it must be > > statistics." Anon. 11/16/12 > > > Dr. Sudipta Maiti > Dept. of Chemical Sciences > Tata Institute of Fundamental Research > Homi Bhabha Raod, Colaba, Mumbai 400005 > Ph. 91-22-2278-2716 / 2539 > Fax: 91-22-2280-4610 > alternate e-mail: [hidden email] > url: biophotonics.weebly.com >  | 
			
| Free forum by Nabble | Edit this page |