http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, I know this is a very basic question but I am having hard time to find the answer. How can I calculate the thickness of the optical section I am aquiring with different pinhole size? So waht is its relation with objective resolution? Thanks in advance Valeria |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Valeria I don't think there's a simple formula, but in the 2nd edition of Confocal Handbook there is a chapter by Wilson that talks about the effect of pinhole on resolution. Mike Michael Model, Ph.D. Confocal Microscopy Core Dpt. Biological Sciences Kent State University Kent, OH 44242 tel. 330-672-2874 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Valeria Berno Sent: Tuesday, July 08, 2008 7:47 AM To: [hidden email] Subject: pinhole-thickness Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, I know this is a very basic question but I am having hard time to find the answer. How can I calculate the thickness of the optical section I am aquiring with different pinhole size? So waht is its relation with objective resolution? Thanks in advance Valeria |
|
|
In reply to this post by Valeria Berno
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Valeria I would say that the size of your optical section is entirely dependant on the NA of your objective. What happens when you open your pinhole is that you let photons coming from "outside" contribute to image formation. Of course, you also get back difracted light coming from your focal plane that you otherwise you would have rejected. The end effect is more a blurring than a "thickening". the simplest solution is to get a infra red pulsed laser! No need of pinhole with multiphoton! Cheers Eric Eric Scarfone, PhD, CNRS, Center for Hearing and communication Research Department of Clinical Neuroscience Karolinska Institutet Postal Address: CFH, M1:02 Karolinska Hospital, SE-171 76 Stockholm, Sweden Work: +46 (0)8-517 79343, Cell: +46 (0)70 888 2352 Fax: +46 (0)8-301876 email: [hidden email] http://www.ki.se/cfh/ ----- Original Message ----- From: Valeria Berno <[hidden email]> Date: Tuesday, July 8, 2008 1:47 pm Subject: pinhole-thickness To: [hidden email] > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Dear all, > > I know this is a very basic question but I am having hard time to > find the > answer. > > How can I calculate the thickness of the optical section I am > with different pinhole size? > > So waht is its relation with objective resolution? > > Thanks in advance > > Valeria > |
|
|
In reply to this post by mmodel
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
-
Valeria, There is a Zeiss publication where they give the formula that calculates the thickness of the optical sections as a function of pinhole diameter, wavelength, refractive index, and numerical aperture. As Michael said, it is a rather complex formula. The more you open the pinhole, the more out of focus contribution you will get from farther away from the focal plane, so it's a somewhat arbitrary definition... it all depends where you set the threshold for what you "see" and what you don't "see" (typically, expressed in units of FWMH or full-width at max height). In any event, if you send me an email, I can reply with an attachment of the Zeiss paper (I did not get your original posting, so I don't have your personal address). As a side note, the Zeiss LSM 510 acquisition software will tell you what your optical thickness is for any combination of objective, wavelength and pinhole aperture. This is used to for instance make sure you are collecting images of similar "thickness" in different channels by adjusting the pinhole aperture, for instance for colocalization studies. The publication is titled: Microscopy from Carl Zeiss. Principles. Confocal laser scanning microscopy. You may be able to Google it too...
-- Julio Vazquez Fred Hutchinson Cancer Research Center 1100 Fairview Ave. N., mailstop DE-512 Seattle, WA 98109-1024 On Jul 8, 2008, at 7:02 AM, MODEL, MICHAEL wrote:
|
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
The publication link is:
On Jul 8, 2008, at 1:57 PM, Julio Vazquez wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal |