![]() ![]() |
Boswell, Carl A - (cboswell) |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
We have been having a chronic problem with the image quality (S/N,
resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
Julian Smith III |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps.
JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
Guy Cox |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
I can't speak for your particular problem, but there are problems with using a diode laser and an AOTF. Diode lasers typically have a bit of wavelength spread, so with the narrow cut-off of an
AOTF you are likely to lose some power. Diodes can also drift in wavelength - not usually enough to matter with a dichroic, but enough to bother an AOTF. I suggest you check your 405 laser with a power meter and spectrum analyser to see what it is actually delivering. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Smith, III, Julian P.S Sent: Saturday, 28 February 2009 10:57 AM To: [hidden email] Subject: Re: poor dapi images on LSM 510 Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps. JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 No virus found in this incoming message. Checked by AVG. Version: 7.5.557 / Virus Database: 270.11.4/1976 - Release Date: 27/02/2009 1:27 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.557 / Virus Database: 270.11.4/1976 - Release Date: 27/02/2009 1:27 PM |
![]() ![]() |
O'Toole, P |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Boswell, Carl A - (cboswell)
You may also want to check that the optic fibre is not being knocked. The
405 output on the 510 is sensitive to the optic fibre being moved close to the back of the laser module. This could account for the decay over time if the laser output remains high. Best Pete ----- Original Message ----- From: "Carl Boswell" <[hidden email]> To: <[hidden email]> Sent: Friday, February 27, 2009 11:19 PM Subject: poor dapi images on LSM 510 > We have been having a chronic problem with the image quality (S/N, > resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality > seems to decay over time (months) and pinhole/collimator tuning doesn't > make things right. We have had the AOM replaced once but the problem > eventually returned. We now have to use 100% 405 laser, rather than 10% > to get anything close to good images. Is there something about the AOM > stability in this light path, or alignment issues with the laser that > others have found to result in a slow decay in usefulness of this channel? > > Thanks for your thoughts. > Carl > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 ... [show rest of quote]
|
![]() ![]() |
Keith Morris |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Julian Smith III
Yep we had the same problem with our DAPI laser last year [3years old]. Kept
adjusting the collimator as illumination was uneven. DAPI faint etc.., initially with duff samples, then with nuclei that dazzle under the mercury lamp. Got the engineer out under the maintenance contract and he spent two days trying to work out what was wrong [laser was reading very low]. Never quite said what he thought it was. Came back and replaced the 405nm laser, AOTF, and probably some related stuff [lot of boxes], and from then on the 405nm laser images have been fine [10% power]. The 405nm laser failure was very gradual, so for a month or so didn't know quite what was going on. Seemed very poor image quality was more noticeable than any lack of laser power. Had a related 405nm problem with the DAPI nuclei sometimes seeming to be in a different Z plain to the FITC/TRITC z slice [could see a nuclei 'hole' shadow but the DAPI was out of focus elsewhere. Seemed to come & go. I was new to the post and the Zeiss 510, and hadn't noticed this sort of thing before with other Leica/Bio-Rad confocals. Couldn't find a way to Z correct the DAPI channel in LSM software [isn't one]. Had a chat with our Zeiss confocal rep, and he said I expect you have a Plan Neofluar 40x and a Plan Apochromat 63x [which we do]. The cheaper Plan Neofluar will cause this problems apparently, while if you see it on the Plan Apochromat call in the engineer. Hadn't twigged it was only being seen on the 40x, and probably always had high power Apochromats before. So now we tend to look at the nucleus and nucleoli with the 63x Apochromat. Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Smith, III, Julian P.S Sent: 27 February 2009 23:57 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps. JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
simon walker (BI) |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In my experience the major cause of loss of signal from 405 excitation
is the pinhole drifting out of alignment. If it's significantly out of alignment it can also cause the apparent large shift in focal plane you describe. This is a much more significant effect than the one you see when you image eg. 1 um beads and measure channel registration using plan apos and neofluar objectives. I know Carl said in his original posting that he had tried tuning the pinhole, but I would always look at this first before worrying about AOTFs etc. Simon -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Keith Morris Sent: 02 March 2009 10:10 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Yep we had the same problem with our DAPI laser last year [3years old]. Kept adjusting the collimator as illumination was uneven. DAPI faint etc.., initially with duff samples, then with nuclei that dazzle under the mercury lamp. Got the engineer out under the maintenance contract and he spent two days trying to work out what was wrong [laser was reading very low]. Never quite said what he thought it was. Came back and replaced the 405nm laser, AOTF, and probably some related stuff [lot of boxes], and from then on the 405nm laser images have been fine [10% power]. The 405nm laser failure was very gradual, so for a month or so didn't know quite what was going on. Seemed very poor image quality was more noticeable than any lack of laser power. Had a related 405nm problem with the DAPI nuclei sometimes seeming to be in a different Z plain to the FITC/TRITC z slice [could see a nuclei 'hole' shadow but the DAPI was out of focus elsewhere. Seemed to come & go. I was new to the post and the Zeiss 510, and hadn't noticed this sort of thing before with other Leica/Bio-Rad confocals. Couldn't find a way to Z correct the DAPI channel in LSM software [isn't one]. Had a chat with our Zeiss confocal rep, and he said I expect you have a Plan Neofluar 40x and a Plan Apochromat 63x [which we do]. The cheaper Plan Neofluar will cause this problems apparently, while if you see it on the Plan Apochromat call in the engineer. Hadn't twigged it was only being seen on the 40x, and probably always had high power Apochromats before. So now we tend to look at the nucleus and nucleoli with the 63x Apochromat. Keith ------------------------------------------------------------------------ --- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Smith, III, Julian P.S Sent: 27 February 2009 23:57 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps. JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
Rietdorf, Jens |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Similar problems over here, take a look at the beam profile of the
laser. It seems that can change if the laser ages. We had some success with adjusting the fiber coupling at the laser coupling side, though we have to do it from time to time. Best, jens -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker (BI) Sent: Monday, March 02, 2009 11:38 AM To: [hidden email] Subject: Re: poor dapi images on LSM 510 In my experience the major cause of loss of signal from 405 excitation is the pinhole drifting out of alignment. If it's significantly out of alignment it can also cause the apparent large shift in focal plane you describe. This is a much more significant effect than the one you see when you image eg. 1 um beads and measure channel registration using plan apos and neofluar objectives. I know Carl said in his original posting that he had tried tuning the pinhole, but I would always look at this first before worrying about AOTFs etc. Simon -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Keith Morris Sent: 02 March 2009 10:10 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Yep we had the same problem with our DAPI laser last year [3years old]. Kept adjusting the collimator as illumination was uneven. DAPI faint etc.., initially with duff samples, then with nuclei that dazzle under the mercury lamp. Got the engineer out under the maintenance contract and he spent two days trying to work out what was wrong [laser was reading very low]. Never quite said what he thought it was. Came back and replaced the 405nm laser, AOTF, and probably some related stuff [lot of boxes], and from then on the 405nm laser images have been fine [10% power]. The 405nm laser failure was very gradual, so for a month or so didn't know quite what was going on. Seemed very poor image quality was more noticeable than any lack of laser power. Had a related 405nm problem with the DAPI nuclei sometimes seeming to be in a different Z plain to the FITC/TRITC z slice [could see a nuclei 'hole' shadow but the DAPI was out of focus elsewhere. Seemed to come & go. I was new to the post and the Zeiss 510, and hadn't noticed this sort of thing before with other Leica/Bio-Rad confocals. Couldn't find a way to Z correct the DAPI channel in LSM software [isn't one]. Had a chat with our Zeiss confocal rep, and he said I expect you have a Plan Neofluar 40x and a Plan Apochromat 63x [which we do]. The cheaper Plan Neofluar will cause this problems apparently, while if you see it on the Plan Apochromat call in the engineer. Hadn't twigged it was only being seen on the 40x, and probably always had high power Apochromats before. So now we tend to look at the nucleus and nucleoli with the 63x Apochromat. Keith ------------------------------------------------------------------------ --- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Smith, III, Julian P.S Sent: 27 February 2009 23:57 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps. JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
Keith Morris |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by simon walker (BI)
I notice the new Zeiss 710 Meta only has one pinhole, for pretty much this
reason I suspect. With our failing 405nm laser & associated AOTF optics [whatever it was that was failing], pinhole adjustment made no difference whatsoever [we do that regularly anyway]. We did find we were adjusting the collimator a heck of a lot [despite the engineer saying you rarely have to adjust it]. In the end the DAPI signal became more fuzz and noise than signal, and increasing laser power made no difference [just brighter fuzz] - by then the engineer was already involved. Even the Zeiss engineer took two or three stabs to fix it [first time he thought it was OK, we didn't]. Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker (BI) Sent: 02 March 2009 10:38 To: [hidden email] Subject: Re: poor dapi images on LSM 510 In my experience the major cause of loss of signal from 405 excitation is the pinhole drifting out of alignment. If it's significantly out of alignment it can also cause the apparent large shift in focal plane you describe. This is a much more significant effect than the one you see when you image eg. 1 um beads and measure channel registration using plan apos and neofluar objectives. I know Carl said in his original posting that he had tried tuning the pinhole, but I would always look at this first before worrying about AOTFs etc. Simon -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Keith Morris Sent: 02 March 2009 10:10 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Yep we had the same problem with our DAPI laser last year [3years old]. Kept adjusting the collimator as illumination was uneven. DAPI faint etc.., initially with duff samples, then with nuclei that dazzle under the mercury lamp. Got the engineer out under the maintenance contract and he spent two days trying to work out what was wrong [laser was reading very low]. Never quite said what he thought it was. Came back and replaced the 405nm laser, AOTF, and probably some related stuff [lot of boxes], and from then on the 405nm laser images have been fine [10% power]. The 405nm laser failure was very gradual, so for a month or so didn't know quite what was going on. Seemed very poor image quality was more noticeable than any lack of laser power. Had a related 405nm problem with the DAPI nuclei sometimes seeming to be in a different Z plain to the FITC/TRITC z slice [could see a nuclei 'hole' shadow but the DAPI was out of focus elsewhere. Seemed to come & go. I was new to the post and the Zeiss 510, and hadn't noticed this sort of thing before with other Leica/Bio-Rad confocals. Couldn't find a way to Z correct the DAPI channel in LSM software [isn't one]. Had a chat with our Zeiss confocal rep, and he said I expect you have a Plan Neofluar 40x and a Plan Apochromat 63x [which we do]. The cheaper Plan Neofluar will cause this problems apparently, while if you see it on the Plan Apochromat call in the engineer. Hadn't twigged it was only being seen on the 40x, and probably always had high power Apochromats before. So now we tend to look at the nucleus and nucleoli with the 63x Apochromat. Keith ------------------------------------------------------------------------ --- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/cytogenetics/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Smith, III, Julian P.S Sent: 27 February 2009 23:57 To: [hidden email] Subject: Re: poor dapi images on LSM 510 Sudden death of the 405? We lost ours at 560+/- user hours (Olympus FV1000; laser replaced under warranty). New 405 is running at 5% on the same preps. JSIII -----Original Message----- From: Confocal Microscopy List on behalf of Carl Boswell Sent: Fri 2/27/2009 6:19 PM To: [hidden email] Subject: poor dapi images on LSM 510 We have been having a chronic problem with the image quality (S/N, resolution) of DAPI stained nuclei using a Zeiss 510 Meta. The quality seems to decay over time (months) and pinhole/collimator tuning doesn't make things right. We have had the AOM replaced once but the problem eventually returned. We now have to use 100% 405 laser, rather than 10% to get anything close to good images. Is there something about the AOM stability in this light path, or alignment issues with the laser that others have found to result in a slow decay in usefulness of this channel? Thanks for your thoughts. Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
![]() ![]() |
MICROSCOPIA CONFOCAL y CCD |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Good afternoon. I am planning to use QUASAR 670,
from Biosearch Technologies, for peptide labeling. I chose this molecule because its spectral properties and prize. I would like to know if anyone has experience with this fluorophore. Is it a quite photo-stable compound?, would you recommend it as a far-red fluorophore?. Does anyone know problems regarding this compound?. Thank you very much for your time. At 14:31 02/03/2009, you wrote: >[hidden email] SERVICIO de MICROSCOPIA CONFOCAL y CCD Mª Teresa Seisdedos Domínguez Fernando González Camacho [hidden email] Centro Investigaciones Biológicas (CIB) CSIC C/Ramiro de Maeztu, 9 28040 Madrid Phone: + 34-91 8373112 ext.4401 Fax: + 34-91 536 04 32 |
Free forum by Nabble | Edit this page |