|
Cromey, Douglas W - (dcromey) |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Is there a better or preferred "blue" fluorochrome? Mostly this is a widefield fluorescence question (using a DAPI cube), but there is the possibility for using a confocal with a 405nm laser line. I'm working with a group that is trying to work around a frustrating autofluorescence issue in liver tissue. What we ended up doing on an earlier project was not using any secondaries that fluoresced in the green channel (no fitc/AF488/GFP) and instead used the green channel on our widefield microscope as a control, ratioing this image so we could subtract it out of the other channels to remove some of the broad autofluorescent background. This worked nicely. In the published images we showed nuclei (blue) and two specific stains using AF568 and AF647, with reduced background autofluorescence. The PI now wants to have labelling with three specific fluorochromes, and I am advocating using the green channel as the generic "everything lights up" channel for context, so we can hopefully skip the Hoechst 3342 or DAPI. Since we won't be using DNA dyes, is there a good blue fluorochrome that can be used as a secondary for specific labelling of a marker found on an intracellular organelle (other than the nucleus)? Thanks, Doug ------------------------------------------------------------------------------------------ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: LSN 463 email: [hidden email]<mailto:[hidden email]> voice: 520-626-2824 fax: 520-626-2097 http://microscopy.arizona.edu/learn/microscopy-imaging-resources-www Home of: "Microscopy and Imaging Resources on the WWW" UA Microscopy Alliance - http://microscopy.arizona.edu<http://microscopy.arizona.edu/> |
|
Alfonso Schmidt |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Douglas: We use BV421 ( Biolegend or BD) or AF405, both works very good in our confocal ( FV1200) exciting with laser 405 and collecting emission with a BD 430-450. Alfonso Schmidt Hugh Green Cytometry Core Staff Scientist P: +64 4 499 6914 x 868 E: [hidden email] W: www.malaghan.org.nz A: PO Box 7060, Wellington 6242, New Zealand > On Aug 21, 2018, at 11:06 AM, Cromey, Douglas W - (dcromey) <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Is there a better or preferred "blue" fluorochrome? Mostly this is a widefield fluorescence question (using a DAPI cube), but there is the possibility for using a confocal with a 405nm laser line. > > I'm working with a group that is trying to work around a frustrating autofluorescence issue in liver tissue. What we ended up doing on an earlier project was not using any secondaries that fluoresced in the green channel (no fitc/AF488/GFP) and instead used the green channel on our widefield microscope as a control, ratioing this image so we could subtract it out of the other channels to remove some of the broad autofluorescent background. This worked nicely. In the published images we showed nuclei (blue) and two specific stains using AF568 and AF647, with reduced background autofluorescence. > > The PI now wants to have labelling with three specific fluorochromes, and I am advocating using the green channel as the generic "everything lights up" channel for context, so we can hopefully skip the Hoechst 3342 or DAPI. Since we won't be using DNA dyes, is there a good blue fluorochrome that can be used as a secondary for specific labelling of a marker found on an intracellular organelle (other than the nucleus)? > > Thanks, > Doug > > ------------------------------------------------------------------------------------------ > Douglas W. Cromey, M.S. - Associate Scientific Investigator > Dept. of Cellular & Molecular Medicine, University of Arizona > 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA > > office: LSN 463 email: [hidden email]<mailto:[hidden email]> > voice: 520-626-2824 fax: 520-626-2097 > > http://microscopy.arizona.edu/learn/microscopy-imaging-resources-www > Home of: "Microscopy and Imaging Resources on the WWW" > > UA Microscopy Alliance - http://microscopy.arizona.edu<http://microscopy.arizona.edu/> The information contained in this email and/or attachmentsmay contain confidential material.If it is not intended for you, please do not read, distribute or copy it or any attachments. Please notify the sender by return email and delete the original message and any attachments.Thank you. |
|
Moulding, Dale |
|
|
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Doug, I have a user that has dropped Dapi and swears by Alexa405. Used on an LSM880 it gives great results, even in wholemounts in PBS (no antifade). I didn't believe it until I watched a few scans. Best used for your more abundant antigens but works surprisingly well. good luck Dale http://www.ucl.ac.uk/ich/core-scientific-facilities-centres/confocal-microscopy ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Cromey, Douglas W - (dcromey) <[hidden email]> Sent: 21 August 2018 00:06:28 To: [hidden email] Subject: recommendation for a "blue" secondary fluorochrome ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy LISTSERV 16.0 - CONFOCALMICROSCOPY List at LISTS.UMN.EDU<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> lists.umn.edu [hidden email]: listserv archives. confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Is there a better or preferred "blue" fluorochrome? Mostly this is a widefield fluorescence question (using a DAPI cube), but there is the possibility for using a confocal with a 405nm laser line. I'm working with a group that is trying to work around a frustrating autofluorescence issue in liver tissue. What we ended up doing on an earlier project was not using any secondaries that fluoresced in the green channel (no fitc/AF488/GFP) and instead used the green channel on our widefield microscope as a control, ratioing this image so we could subtract it out of the other channels to remove some of the broad autofluorescent background. This worked nicely. In the published images we showed nuclei (blue) and two specific stains using AF568 and AF647, with reduced background autofluorescence. The PI now wants to have labelling with three specific fluorochromes, and I am advocating using the green channel as the generic "everything lights up" channel for context, so we can hopefully skip the Hoechst 3342 or DAPI. Since we won't be using DNA dyes, is there a good blue fluorochrome that can be used as a secondary for specific labelling of a marker found on an intracellular organelle (other than the nucleus)? Thanks, Doug ------------------------------------------------------------------------------------------ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: LSN 463 email: [hidden email]<mailto:[hidden email]> voice: 520-626-2824 fax: 520-626-2097 http://microscopy.arizona.edu/learn/microscopy-imaging-resources-www Home of: "Microscopy and Imaging Resources on the WWW" UA Microscopy Alliance - http://microscopy.arizona.edu<http://microscopy.arizona.edu/> |
|
Steffen Dietzel |
|
|
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Doug, I have heard the Brilliant Violett dyes are good. Haven't tried them myself for the lack of time, but I would be interested in results. Steffen Am 21.08.2018 um 01:06 schrieb Cromey, Douglas W - (dcromey): > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Is there a better or preferred "blue" fluorochrome? Mostly this is a widefield fluorescence question (using a DAPI cube), but there is the possibility for using a confocal with a 405nm laser line. > > I'm working with a group that is trying to work around a frustrating autofluorescence issue in liver tissue. What we ended up doing on an earlier project was not using any secondaries that fluoresced in the green channel (no fitc/AF488/GFP) and instead used the green channel on our widefield microscope as a control, ratioing this image so we could subtract it out of the other channels to remove some of the broad autofluorescent background. This worked nicely. In the published images we showed nuclei (blue) and two specific stains using AF568 and AF647, with reduced background autofluorescence. > > The PI now wants to have labelling with three specific fluorochromes, and I am advocating using the green channel as the generic "everything lights up" channel for context, so we can hopefully skip the Hoechst 3342 or DAPI. Since we won't be using DNA dyes, is there a good blue fluorochrome that can be used as a secondary for specific labelling of a marker found on an intracellular organelle (other than the nucleus)? > > Thanks, > Doug > > ------------------------------------------------------------------------------------------ > Douglas W. Cromey, M.S. - Associate Scientific Investigator > Dept. of Cellular & Molecular Medicine, University of Arizona > 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA > > office: LSN 463 email: [hidden email]<mailto:[hidden email]> > voice: 520-626-2824 fax: 520-626-2097 > > http://microscopy.arizona.edu/learn/microscopy-imaging-resources-www > Home of: "Microscopy and Imaging Resources on the WWW" > > UA Microscopy Alliance - http://microscopy.arizona.edu<http://microscopy.arizona.edu/> > ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Biomedical Center (BMC) Head of the Core Facility Bioimaging Großhaderner Straße 9 D-82152 Planegg-Martinsried Germany http://www.bioimaging.bmc.med.uni-muenchen.de |
|
Dmitry Ushakov-3 |
|
|
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** eFluor450 works really well in confocal with 405 nm line. |
|
Alison J. North |
|
|
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Doug, I may have missed this - lots of replies coming through and I'm busy at a course - but I've had the most success with DyLight 405, as long as you use it with either Prolong Diamond (not Gold in this case) or homemade N-propyl gallate antifade. Alexa 405 is also stable in either Prolong Gold or Prolong Diamond but it's much weaker than DyLight 405 in my hands. I've tested a lot of different blue dyes and the choice of antifade seems to be particularly important with them. Just my own findings, which I keep meaning to publish somehow but have never had time... All the best, Alison Sent from my Verizon Wireless 4G LTE DROID "Cromey, Douglas W - (dcromey)" <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFAg&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=xgcrADgSzrbikXnjMzUfjiEtJwrFqXVxEvVU1jH90WU&s=FfHbH9CPyauZN4dFOoqA7Wcpxq6n3crUWJqOM_IWxTg&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFAg&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=xgcrADgSzrbikXnjMzUfjiEtJwrFqXVxEvVU1jH90WU&s=ZqkVFz_JFlrPqs5CYlXJ9SS_FsBaIK8C0RUxmwll5tY&e= and include the link in your posting. ***** Is there a better or preferred "blue" fluorochrome? Mostly this is a widefield fluorescence question (using a DAPI cube), but there is the possibility for using a confocal with a 405nm laser line. I'm working with a group that is trying to work around a frustrating autofluorescence issue in liver tissue. What we ended up doing on an earlier project was not using any secondaries that fluoresced in the green channel (no fitc/AF488/GFP) and instead used the green channel on our widefield microscope as a control, ratioing this image so we could subtract it out of the other channels to remove some of the broad autofluorescent background. This worked nicely. In the published images we showed nuclei (blue) and two specific stains using AF568 and AF647, with reduced background autofluorescence. The PI now wants to have labelling with three specific fluorochromes, and I am advocating using the green channel as the generic "everything lights up" channel for context, so we can hopefully skip the Hoechst 3342 or DAPI. Since we won't be using DNA dyes, is there a good blue fluorochrome that can be used as a secondary for specific labelling of a marker found on an intracellular organelle (other than the nucleus)? Thanks, Doug ------------------------------------------------------------------------------------------ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: LSN 463 email: [hidden email]<mailto:[hidden email]> voice: 520-626-2824 fax: 520-626-2097 https://urldefense.proofpoint.com/v2/url?u=http-3A__microscopy.arizona.edu_learn_microscopy-2Dimaging-2Dresources-2Dwww&d=DwIFAg&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=xgcrADgSzrbikXnjMzUfjiEtJwrFqXVxEvVU1jH90WU&s=fC_N7W0MPQFCuJmpaW6JulMEVMGFBiOjUt8HD78YPI4&e= Home of: "Microscopy and Imaging Resources on the WWW" UA Microscopy Alliance - https://urldefense.proofpoint.com/v2/url?u=http-3A__microscopy.arizona.edu&d=DwIFAg&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=xgcrADgSzrbikXnjMzUfjiEtJwrFqXVxEvVU1jH90WU&s=WlmovRJxBvMhPmiH0ftNpZ8rLaK1OVseNxDEbjbhhXs&e=<https://urldefense.proofpoint.com/v2/url?u=http-3A__microscopy.arizona.edu_&d=DwIFAg&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=xgcrADgSzrbikXnjMzUfjiEtJwrFqXVxEvVU1jH90WU&s=GkN24mDZJchUI6AMARk8_6xAmXBW3RWAR_psdASFaWs&e=> |
| Free forum by Nabble | Edit this page |