To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, everyone: I'm looking for a 63X objective for live imaging and its resolution is my major concern. Zeiss recommands Objective LCI Plan-Neofluar 63x/1.3 Imm Corr DIC M27, saying it has NA = 1.3 in both water and glycerine. Does it really have NA=1.3 even in water? Your reply is highly appreciated. Xiao ******************* Xiao Liu School of Life Sciences Tsing-Hua University Beijing, China ******************* |
|
Pascal Weber |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** What is your goal ! Do you really nead glycerine if yes tehre just one objective. This one recommanded by Zeiss. But you juste have only 170 micron working distance. I worked with this one on a live mouse brain stained with GFP . And we obtained 700 microns penetration on a two photon microscope. Objectif W Plan-Apochromat 63x/1,0 M27. |
|
|
In reply to this post by Xiao Liu
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The only way this would be possible is if it were stopped down to reduce its actual aperture in glycerine. This is technically possible but (IMHO) unlikely. Glycerol has a refractive index of 1.4722 at 590nm (the closest I can find to the official 550nm). NA 1.3 therefore corresponds to an acceptance half-angle of 62 degrees. Water at 550nm has a refractive index of 1.33. This same half-angle would be an NA of 1.17. Looking at it the other way round, to achieve NA 1.3 in water requires an acceptance angle of 77 degrees, which while not impossible is pushing the envelope. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Xiao Liu Sent: Friday, 16 December 2011 11:09 AM To: [hidden email] Subject: resolution of Zeiss Water and Glycerine 63X lens ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, everyone: I'm looking for a 63X objective for live imaging and its resolution is my major concern. Zeiss recommands Objective LCI Plan-Neofluar 63x/1.3 Imm Corr DIC M27, saying it has NA = 1.3 in both water and glycerine. Does it really have NA=1.3 even in water? Your reply is highly appreciated. Xiao ******************* Xiao Liu School of Life Sciences Tsing-Hua University Beijing, China ******************* |
|
James Pawley |
|
|
In reply to this post by Xiao Liu
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > > >Hi, everyone: > >I'm looking for a 63X objective for live imaging and its resolution >is my major concern. Zeiss recommands Objective LCI Plan-Neofluar >63x/1.3 Imm Corr DIC M27, saying it has NA = 1.3 in both water and >glycerine. Does it really have NA=1.3 even in water? > >Your reply is highly appreciated. >Xiao > >******************* >Xiao Liu >School of Life Sciences >Tsing-Hua University >Beijing, China >******************* > Well, it may have such an NA in water, but most of the rays approaching the flat surfaces (two sides of the cover slip and the front element of the objective) will be reflected when transmitting these surface, so will not contribute much to the image. Keep in mind that, when actually tested, using the Methods in Chapter 11 of the Handbook, Juskaitis found that even high quality NA 1.4 objectives only worked "properly" up to 1.35 (The high-NA ray were indeed propagated by the lenses but had the wrong phase so actually did not contribute to image sharpness). In general, when viewing living specimens, "resolution" (that slippery concept that can only be considered in terms of more or less contrast) is more likely degraded by aberrations brought on by wrong immersion media or by the RI variations of the specimen itself than by the NA marked on the side of the lens. Cheers, Jim Pawley -- *************************************************************************** Prof. James B. Pawley, Ph. 608-238-3953 21. N. Prospect Ave. Madison, WI 53726 USA [hidden email] 3D Microscopy of Living Cells Course, June 9-21, 2012, UBC, Vancouver Canada Info: http://www.3dcourse.ubc.ca/ Application deadline 3/16/2012 "If it ain't diffraction, it must be statistics." Anon. 11/16/12 |
| Free forum by Nabble | Edit this page |