searching for a Leica power supply

classic Classic list List threaded Threaded
♦
♦
5 messages Options
Philipp Tripal Philipp Tripal
Reply | Threaded
Open this post in threaded view
♦
♦
|

searching for a Leica power supply

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear colleagues

I´m searching for a Leica power supply to run an ancient fluorescent
microscope:

*Power supply 12V 100W, *Lamp power, stabilized, AC-input range: 90-250V
AC / 50...60Hz , DC output voltage: Adjustable from 2.5V up to 12V,
ambient temperature range: +5-60°C, protected for short out and open
output connector

If you have such a power supply and you don´t need it anymore, I´d
appreciate your reply.


Sincerely yours,

Philipp Tripal


--
Philipp Tripal
Optical Imaging Centre Erlangen

Hartmannstr. 14
91052 Erlangen, Germany
+49-9131-85-64309 (Office)
+49-9131-85-64301 (Secretary)
+49-9131-85-64302 (Fax)

www.oice.uni-erlangen.de
Peter Call Peter Call
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: searching for a Leica power supply

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Commercial response *

Hi Phillipp

If unsuccessful you could try an LED transmitted source..

Best Wishes
 
Peter
 
Peter Call
CoolLED UK ltd
Field Sales Manager – Central Europe and Australia
Email: [hidden email]
Mobile +44 (0)7702 212183
Tel:  +44 1264 323040
Fax: +44 1264 723897

Happy New Year from everyone at CoolLED


Click here for information on the CoolLED product range

This message and any attachments are strictly confidential and intended solely for the addressee. Any unauthorized use or disclosure, in whole or in part, is prohibited. E-mails are subject to possible alteration. CoolLED Ltd and the sender decline any liability if this message and/or any attachments have been altered, changed or falsified. We have taken steps to ensure that this email and attachments are free from any virus, but it remains your responsibility to ensure that viruses do not adversely affect you. If you are not the intended recipient of this message, please delete it and notify the sender immediately.
Trading Office:  CoolLED Limited, Westmarch Business Centre, River Way, Andover, SP10 1NS, UK
CoolLED Ltd is a limited company registered in England. Registered number: 7654232. Registered office: 52c Borough High Street, London, SE1 1XN



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philipp Tripal
Sent: 01 February 2017 08:43
To: [hidden email]
Subject: searching for a Leica power supply

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear colleagues

I´m searching for a Leica power supply to run an ancient fluorescent
microscope:

*Power supply 12V 100W, *Lamp power, stabilized, AC-input range: 90-250V AC / 50...60Hz , DC output voltage: Adjustable from 2.5V up to 12V, ambient temperature range: +5-60°C, protected for short out and open output connector

If you have such a power supply and you don´t need it anymore, I´d appreciate your reply.


Sincerely yours,

Philipp Tripal


--
Philipp Tripal
Optical Imaging Centre Erlangen

Hartmannstr. 14
91052 Erlangen, Germany
+49-9131-85-64309 (Office)
+49-9131-85-64301 (Secretary)
+49-9131-85-64302 (Fax)

www.oice.uni-erlangen.de

G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: searching for a Leica power supply

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

You may be looking for used/donated but in case you're not, I bought a new
power supply exactly as you describe direct from Leica not that long ago,
so it should still be available from Leica. I think it cost me US$600-700.

-Esteban


On Feb 1, 2017 1:09 AM, "Peter Call" <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Commercial response *

Hi Phillipp

If unsuccessful you could try an LED transmitted source..

Best Wishes

Peter

Peter Call
CoolLED UK ltd
Field Sales Manager – Central Europe and Australia
Email: [hidden email]
Mobile +44 (0)7702 212183
Tel:  +44 1264 323040
Fax: +44 1264 723897

Happy New Year from everyone at CoolLED


Click here for information on the CoolLED product range

This message and any attachments are strictly confidential and intended
solely for the addressee. Any unauthorized use or disclosure, in whole or
in part, is prohibited. E-mails are subject to possible alteration. CoolLED
Ltd and the sender decline any liability if this message and/or any
attachments have been altered, changed or falsified. We have taken steps to
ensure that this email and attachments are free from any virus, but it
remains your responsibility to ensure that viruses do not adversely affect
you. If you are not the intended recipient of this message, please delete
it and notify the sender immediately.
Trading Office:  CoolLED Limited, Westmarch Business Centre, River Way,
Andover, SP10 1NS, UK
CoolLED Ltd is a limited company registered in England. Registered number:
7654232. Registered office: 52c Borough High Street, London, SE1 1XN



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Philipp Tripal
Sent: 01 February 2017 08:43
To: [hidden email]
Subject: searching for a Leica power supply

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear colleagues

I´m searching for a Leica power supply to run an ancient fluorescent
microscope:

*Power supply 12V 100W, *Lamp power, stabilized, AC-input range: 90-250V AC
/ 50...60Hz , DC output voltage: Adjustable from 2.5V up to 12V, ambient
temperature range: +5-60°C, protected for short out and open output
connector

If you have such a power supply and you don´t need it anymore, I´d
appreciate your reply.


Sincerely yours,

Philipp Tripal


--
Philipp Tripal
Optical Imaging Centre Erlangen

Hartmannstr. 14
91052 Erlangen, Germany
+49-9131-85-64309 (Office)
+49-9131-85-64301 (Secretary)
+49-9131-85-64302 (Fax)

www.oice.uni-erlangen.de
Zdenek Svindrych-2 Zdenek Svindrych-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: searching for a Leica power supply

In reply to this post by Philipp Tripal
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Philipp,
maybe some regulated laboratory power supply would be cheaper than "Leica
solution".

Try to google (ebay, amazon) for 'STP3010' to get an idea (cheap Chinese
product).

Check the input line voltage, and check that you can regulate current and
voltage independently from zero up to your required values (12V, 8.3A in
your case). That should be OK for driving an incandescent lamp...

Best, zdenek



---------- Původní zpráva ----------
Od: Philipp Tripal <[hidden email]>
Komu: [hidden email]
Datum: 1. 2. 2017 3:48:43
Předmět: searching for a Leica power supply

"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy 
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear colleagues

I´m searching for a Leica power supply to run an ancient fluorescent
microscope:

*Power supply 12V 100W, *Lamp power, stabilized, AC-input range: 90-250V
AC / 50...60Hz , DC output voltage: Adjustable from 2.5V up to 12V,
ambient temperature range: +5-60°C, protected for short out and open
output connector

If you have such a power supply and you don´t need it anymore, I´d
appreciate your reply.


Sincerely yours,

Philipp Tripal


--
Philipp Tripal
Optical Imaging Centre Erlangen

Hartmannstr. 14
91052 Erlangen, Germany
+49-9131-85-64309 (Office)
+49-9131-85-64301 (Secretary)
+49-9131-85-64302 (Fax)

www.oice.uni-erlangen.de
"
Stanislav Vitha-2 Stanislav Vitha-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: searching for a Leica power supply

In reply to this post by Philipp Tripal
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

If I understand correctly this is for the transmitted light/brightfield  incandescent lamp.

A cheap regulated lab power supply with the right output range (0-12V) works fine.
I have several Leica Ortholux I scopes (the black ones from 1960's) where I replaced the original Leica plug with two "banana" plugs (is that a right term?) that go into a 0-12V lab power supply.

Another option is to look for car/RV light dimmers, but you would probably have to feed it with 12 V DC.

Similarly, a cheap (~$3 off Ebay)  dimmable 0-12V, 8A dimmer switch should work if you can supply a DC power to it, e.g., from an old  laptop computer power adapter.
 
https://www.amazon.com/LEDwholesalers-LED-Strip-Lights-Controller/dp/B003L4KKF2/ref=pd_sim_60_2?_encoding=UTF8&psc=1&refRID=4YTD01ZPNVBFTPV2E0H5 


Best Regards,

Stan Vitha

Microscopy and Imaging Center
Texas A&M University