seeking "reliable literature" on correct imaging

classic Classic list List threaded Threaded
♦
♦
31 messages Options
12
Cammer, Michael Cammer, Michael
Reply | Threaded
Open this post in threaded view
♦
♦
|

seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Gulpreet Kaur Gulpreet Kaur
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello Michael,

I’m a recent graduate student from University of Wisconsin Madison and I will be using a lot of imaging techniques in he future. Can you send me a copy of the PowerPoint too?

Thanks,
Gul


Gulpreet Kaur
Research Assistant - Lakkaraju Lab
Department of Ophthalmology and Visual Sciences
Program in Cellular and Molecular Biology
University of Wisconsin - Madison


> On Nov 16, 2018, at 11:00 AM, MODEL, MICHAEL <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Michael,
>
> We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.
>
> Mike Model
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
> Sent: Friday, November 16, 2018 11:50 AM
> To: [hidden email]
> Subject: seeking "reliable literature" on correct imaging
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Increasingly we have found that microscope users do not understand proper imaging.
>
> A few examples:
> Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
> Exposures are changed in fluorescent micrographs to make each one "look good".
> 16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
> JPG compression is great because images can be emailed.
> People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
> Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.
>
> A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."
>
> I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.
>
> So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
> Looking forward to suggestions of easy to read definitive texts!
>
> Thank you.
>
> Cheers-
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
> =================================
VERMEREN Matthieu VERMEREN Matthieu
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Michael,

I've been reading Pete Bankhead's excellent "Analyzing fluorescence microscopy images with ImageJ"  (see https://petebankhead.gitbooks.io/imagej-intro/content/).  I think the first part of the course covers what you describe.  At least it's a good start.

Sincerely,

Matthieu


--
The University of Edinburgh is a charitable body, registered in
Scotland, with registration number SC005336.


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, 16 November, 2018 16:50
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by mmodel
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike;
        Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
        Your PowerPoint should hold more credibility that I obviously have....
        Best;
Kathy



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
Sent: Friday, November 16, 2018 9:00 AM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Lemasters, John J. Lemasters, John J.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike,

I would like a copy of the PowerPoint as well. This is a very important issue that Michael Cammer raises. Trainees (and even more senior folks) continue to make these mistakes even as you emphatically tell them not to.

Thanks again,

John

--
John J. Lemasters, MD, PhD
Professor and GlaxoSmithKline Distinguished Endowed Chair
Director, Center for Cell Death, Injury & Regeneration
Departments of Drug Discovery & Biomedical Sciences and Biochemistry & Molecular Biology
Medical University of South Carolina
DD504 Drug Discovery Building
70 President Street, MSC 139
Charleston, SC 29425

Office: 843-876-2360
Lab: 843-876-2354
Fax: 843-876-2353
Email: [hidden email]
http://academicdepartments.musc.edu/ccdir

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kathryn Spencer
Sent: Friday, November 16, 2018 12:32 PM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

CAUTION: External

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike;
        Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
        Your PowerPoint should hold more credibility that I obviously have....
        Best;
Kathy



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
Sent: Friday, November 16, 2018 9:00 AM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================




-------------------------------------------------------------------------
This message was secured via TLS by MUSC.
Julio Vazquez-2 Julio Vazquez-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by kspencer007
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

For the basics, I used the book by John Russ (The Image Procesing Handbook), but this is quite expensive for students to purchase, and overkill, although the introductory chapters would be useful. I also found the following articles useful:

Douglas Cromey: Digital Imaging Ethics (Google Digital Imaging Ethics), and references  in there.
Rossner and Yamada: What’s in a picture? The temptation of image manipulation J Cell Biology 166: 11-15

These papers will give some basic info about digital images and also point to the dangerous boundary between poor image acquisition/processing practices and data falsification.

--

Julio Vazquez, PhD
Director of Cellular Imaging
(206) 667-1215
(206) 667-4205

Fred Hutchinson Cancer Research Center
1100 Fairview Ave.N., Mailstop DE-512
Seattle, WA 98109



On Nov 16, 2018, at 9:31 AM, Kathryn Spencer <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=5B3zF6g2x4sYF2f3fSIusceTEzZCgVhT_jwlWOoqYw0&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=7OeCl1ldMz036QeWxyEIq7Fd9FzFkyes_PqYy5R-PDY&e= and include the link in your posting.
*****

Hi Mike;
Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
Your PowerPoint should hold more credibility that I obviously have....
Best;
Kathy



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
Sent: Friday, November 16, 2018 9:00 AM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=5B3zF6g2x4sYF2f3fSIusceTEzZCgVhT_jwlWOoqYw0&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=7OeCl1ldMz036QeWxyEIq7Fd9FzFkyes_PqYy5R-PDY&e= and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=5B3zF6g2x4sYF2f3fSIusceTEzZCgVhT_jwlWOoqYw0&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=eRAMFD45gAfqt84VtBcfhQ&r=ttNYkfRwLB_uJJmzFu8cIdRdnibIA0UXGXDLv4MFuvU&m=P-6UuvOKGqAi_cIaRLp3TgvtgmGUOG7tRli3BkF00dA&s=7OeCl1ldMz036QeWxyEIq7Fd9FzFkyes_PqYy5R-PDY&e= and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================

John Oreopoulos John Oreopoulos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Where to begin... Well, here are the most modern books I can think of  
off the top of my head, but there are other good ones out there. I  
have an entire library of "basic microscopy" articles collected over  
the years which I can share with you offline as well. But in terms of  
a total/coherent narrative...

Start with this paper by Pawley. This will put the fear of changing  
exposure levels arbitrarily into anyone:

https://www.future-science.com/doi/10.2144/00285bt01


Here are some good modern books on microscopy imaging:

https://www.elsevier.com/books/digital-microscopy/sluder/978-0-12-407761-4

https://www.sciencedirect.com/bookseries/methods-in-cell-biology/vol/123

https://onlinelibrary.wiley.com/doi/book/10.1002/9781118382905

https://www.springer.com/gp/book/9780306484681

https://www.springer.com/gp/book/9780387259215


And in terms of digital imaging ethics, EVERYONE should read Cromey's  
article on this topic:

https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4114110/


John Oreopoulos
Staff Scientist
Andor Technology

PLEASE NOTE OUR NEW ADDRESS:

250 West Beaver Creek Rd., Unit 1
Richmond Hill, Ontario Canada L4B 1C7
Tel: (905) 326-5040 ext. 412
Email: [hidden email]
WWW site: www.andor.com

Quoting "Cammer, Michael" <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Increasingly we have found that microscope users do not understand  
> proper imaging.
>
> A few examples:
> Confocal images are saturated and have the offset set to negative  
> numbers intentionally "to reduce background".
> Exposures are changed in fluorescent micrographs to make each one  
> "look good".
> 16 bit raw data with no saturation or clipping are set to 8 bits  
> with a lot of saturated and clipped pixels at the extremes and these  
>  are what are considered raw data.
> JPG compression is great because images can be emailed.
> People assume (and get angry when this doesn't work) that computers  
> can automatically segment and count anything regardless of  
> experimental design, image quality, etc.
> Even if they have saved everything as TIF with no metadata and  
> forgot what magnifications they used.
>
> A student who recognizes these problems and wants to learn the right  
>  way emailed me yesterday asking for "any reliable literature you  
> recommend if I'd like to read more about correct imaging."
>
> I did Google searches and, after skimming a bunch of web pages,  
> realized that I didn't really have an answer.  I could send to this  
> page and to this page, but it's not a coherent narrative.  And I  
> don't have a library; most of what I know about this was picked up  
> willy-nilly up in the 1980s and '90s.  For instance knowledge of  
> color separation from writing animations for the Apple II in machine  
>  code, working in a cable TV studio in the 1980s, and from an art  
> history class in college where we discussed 19th C travel photos and  
>  poly/monochromatic film chemistries of the periods.  Later I  
> learned  fluorescence quantification from a cell biologist intent on  
> making  sure we had true linear responses for readouts of f-actin  
> mass and  concentration, I was fortunate to attend both a Woods Hole  
>  microscopy course and at NCSU John Russ's course on image analysis,  
>  and BioRad MRC 600 training course at their offices in Cambridge,  
> MA  covered a lot of material on this too.
>
> So, do people have favorite texts that cover these topics in a  
> manner digestible by new generations of students?  Of biology  
> students.  Keep in mind many don't know what a byte is; when I teach  
>  I ask the class if anyone knows why an image is coded with white as  
>  255 or 4095 or 16383 and typically one or two hands go up but the  
> rest of the class really doesn't know.
> Looking forward to suggestions of easy to read definitive texts!
>
> Thank you.
>
> Cheers-
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use  
> of the intended recipient(s) and may contain information that is  
> proprietary, confidential, and exempt from disclosure under  
> applicable law. Any unauthorized review, use, disclosure, or  
> distribution is prohibited. If you have received this email in error  
>  please notify the sender by return email and delete the original  
> message. Please note, the recipient should check this email and any  
> attachments for the presence of viruses. The organization accepts no  
>  liability for any damage caused by any virus transmitted by this  
> email.
> =================================
>
Nancy Costigliola Nancy Costigliola
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by kspencer007
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike,

I am in industry, and would also really like to have a powerpoint on this for new users. Thanks!
Nancy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kathryn Spencer
Sent: Friday, November 16, 2018 12:32 PM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=DUKV1yCfWn9cWLDKFG46ViE1YatWxZ6e89zok-3__qE&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=BKJbIASgny5fT8vcyLqiawQJQnFihx7oIu9LGbadiMI&e= and include the link in your posting.
*****

Hi Mike;
        Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
        Your PowerPoint should hold more credibility that I obviously have....
        Best;
Kathy



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
Sent: Friday, November 16, 2018 9:00 AM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=DUKV1yCfWn9cWLDKFG46ViE1YatWxZ6e89zok-3__qE&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=BKJbIASgny5fT8vcyLqiawQJQnFihx7oIu9LGbadiMI&e= and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=DUKV1yCfWn9cWLDKFG46ViE1YatWxZ6e89zok-3__qE&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIF-g&c=TzEZu9LIcihmW37vx9Ah6w&r=6yV8Jm8gBKxW3FelQyszarVcemf9nkXOKBO_nLK4fvg&m=Naz6rLjdo8BFyk0-vCP0btGPpFRLROu8ISI-YsK0KyA&s=BKJbIASgny5fT8vcyLqiawQJQnFihx7oIu9LGbadiMI&e= and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================


This email message and any attachments are confidential and intended for use by the addressee(s) only. If you are not the intended recipient, please notify me immediately by replying to this message, and destroy all copies of this message and any attachments. Thank you.
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by kspencer007
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I would also like a copy of your powerpoint slides Michael. They would be useful in our classes.

A couple of articles we've used:
What’s in a picture? The temptation of image manipulation. Mike Rossner and Kenneth M. Yamada
The Journal of Cell Biology, Volume 166, Number 1, July 5, 2004 11–15
and
Avoiding Twisted Pixels: Ethical Guidelines for the Appropriate Use and Manipulation of Scientific Digital Images. Douglas W. Cromey
Science and Engineering Ethics December 2010, Volume 16, Issue 4, pp 639–667

John Russ's Image Processing Handbook is also useful, but probably too expense to have as anything but one copy for a lab reference. There are other imaging books as well.

The Microscopy Society of America's statement on digital imaging:
"The MSA position on digital image processing has been approved as follows:
"Ethical digital imaging requires that the original uncompressed image file be stored on archival media (e.g., CD-R) without any image manipulation or processing operation. All parameters of the production and acquisition of this file, as well as any subsequent processing steps, must be documented and reported to ensure reproducibility.
Generally, acceptable (non-reportable) imaging operations include gamma correction, histogram stretching, and brightness and contrast adjustments. All other operations (such as Unsharp-Masking, Gaussian Blur, etc.) must be directly identified by the author as part of the experimental methodology. However, for diffraction data or any other image data that is used for subsequent quantification, all imaging operations must be reported."
This policy was formulated by the Digital Image Processing & Ethics Group of the MSA Education Committee and was adopted as MSA policy at the Summer Council meeting August 2-3, 2003."
https://www.microscopy.org/resources/digital_imaging.cfm

This hasn't changed since implemented.

As for what file format ... TIFF. Other lossless image format is also OK, and many microscope companys' proprietary formats are really TIFF images with extra stuff in the header. Like Nikon's ND2 format for Elements.
Saving as TIFF (or other lossless format) you just have to pound into people's heads

Phil
-------------
Philip Oshel    
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office
989 774-7567 lab

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Kathryn Spencer <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Friday,  16November, 2018 at 12:32
To: "[hidden email]" <[hidden email]>
Subject: Re: seeking "reliable literature" on correct imaging

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi Mike;
    Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
    Your PowerPoint should hold more credibility that I obviously have....
    Best;
    Kathy
   
   
   
    -----Original Message-----
    From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
    Sent: Friday, November 16, 2018 9:00 AM
    To: [hidden email]
    Subject: Re: seeking "reliable literature" on correct imaging
   
    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi Michael,
   
    We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.
   
    Mike Model
   
    -----Original Message-----
    From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
    Sent: Friday, November 16, 2018 11:50 AM
    To: [hidden email]
    Subject: seeking "reliable literature" on correct imaging
   
    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Increasingly we have found that microscope users do not understand proper imaging.
   
    A few examples:
    Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
    Exposures are changed in fluorescent micrographs to make each one "look good".
    16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
    JPG compression is great because images can be emailed.
    People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
    Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.
   
    A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."
   
    I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.
   
    So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
    Looking forward to suggestions of easy to read definitive texts!
   
    Thank you.
   
    Cheers-
   
   
    ------------------------------------------------------------
    This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
    =================================
   

Adeela Syed Adeela Syed
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by mmodel
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello Mike,

I would like as copy of the PowerPoint as well, this would be very helpful
for our confocal core users!

Kind regards,
Adeela
--
Adeela Syed, Ph.D.
Associate Project Scientist
Optical Biology Facility Manager
Dept. Developmental and Cell Biology
University of California Irvine
Irvine, CA 92697-2300
 
949-824-3856
email: [hidden email]
obc.bio.uci.edu


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf
Of MODEL, MICHAEL
Sent: Friday, November 16, 2018 9:00 AM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of
quantification in microscopy. I can send you a PowerPoint file but don't
have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf
Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper
imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers
intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look
good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot
of saturated and clipped pixels at the extremes and these are what are
considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can
automatically segment and count anything regardless of experimental design,
image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what
magnifications they used.

A student who recognizes these problems and wants to learn the right way
emailed me yesterday asking for "any reliable literature you recommend if
I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized
that I didn't really have an answer.  I could send to this page and to this
page, but it's not a coherent narrative.  And I don't have a library; most
of what I know about this was picked up willy-nilly up in the 1980s and
'90s.  For instance knowledge of color separation from writing animations
for the Apple II in machine code, working in a cable TV studio in the 1980s,
and from an art history class in college where we discussed 19th C travel
photos and poly/monochromatic film chemistries of the periods.  Later I
learned fluorescence quantification from a cell biologist intent on making
sure we had true linear responses for readouts of f-actin mass and
concentration, I was fortunate to attend both a Woods Hole microscopy course
and at NCSU John Russ's course on image analysis, and BioRad MRC 600
training course at their offices in Cambridge, MA covered a lot of material
on this too.

So, do people have favorite texts that cover these topics in a manner
digestible by new generations of students?  Of biology students.  Keep in
mind many don't know what a byte is; when I teach I ask the class if anyone
knows why an image is coded with white as 255 or 4095 or 16383 and typically
one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the
intended recipient(s) and may contain information that is proprietary,
confidential, and exempt from disclosure under applicable law. Any
unauthorized review, use, disclosure, or distribution is prohibited. If you
have received this email in error please notify the sender by return email
and delete the original message. Please note, the recipient should check
this email and any attachments for the presence of viruses. The organization
accepts no liability for any damage caused by any virus transmitted by this
email.
=================================
Carol Heckman Carol Heckman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

It's actually worse than this.  A 16-bit image is stored, high bits and low bits, in different order in different formats.  The best way to be sure you have the same thing when translated into a new software suite is to compare the histograms.  This gets tedious if you have a lot of images to be processed....

Carol Heckman

Center for Microscopy & Microanalysis

Bowling Green State University


________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Cammer, Michael <[hidden email]>
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
LISTSERV 16.0 - CONFOCALMICROSCOPY List at LISTS.UMN.EDU - Listserv Lists at the University of Minnesota<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy


Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Ben Abrams Ben Abrams
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

I routinely cite "Fundamentals of Light Microscopy and Electronic
Imaging"  by Douglas Murphy and Michael Davidson when I lecture.  It
covers a broad range of topics and is accessible.

-Ben

On 11/16/18 8:50 AM, Cammer, Michael wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Increasingly we have found that microscope users do not understand proper imaging.
>
> A few examples:
> Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
> Exposures are changed in fluorescent micrographs to make each one "look good".
> 16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
> JPG compression is great because images can be emailed.
> People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
> Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.
>
> A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."
>
> I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.
>
> So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
> Looking forward to suggestions of easy to read definitive texts!
>
> Thank you.
>
> Cheers-
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
> =================================

--
Benjamin Abrams,  Ph.D.
Facility Manager
UCSC Life Sciences Microscopy Center

University of California, Santa Cruz
1156 High Street
150 Sinsheimer labs
Santa Cruz, CA 95064

Office/voicemail: (831) 459-3999
Mobile: (831) 332-0911

http://stemcell.soe.ucsc.edu/facilities/microscopy

Training Request Form:
http://goo.gl/forms/M9jd1ZHyohTnj0xk1
Glen MacDonald Glen MacDonald
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by John Oreopoulos
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

thanks for posting this excellent list John.  
the ethics articles are helpful, but I think most bad images are made by not understanding how an image is created as well as what constitutes a digital image.  And all the points from sample preparation to quantitation/publication where the informational content may be compromised.  

FWIW, Alan Hibb’s book is dated but the information remains quite useful.    Before his premature death, he had shipped some books to me in preparation of a visit to North America.  I thought all were sold, but I just found one more copy while cleaning my study.  Original price was  US$125, I will sell it for half price plus shipping.  Proceeds are remitted to his family.

regards,
Glen MacDonald

> On Nov 16, 2018, at 9:54 AM, [hidden email] wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Where to begin... Well, here are the most modern books I can think of off the top of my head, but there are other good ones out there. I have an entire library of "basic microscopy" articles collected over the years which I can share with you offline as well. But in terms of a total/coherent narrative...
>
> Start with this paper by Pawley. This will put the fear of changing exposure levels arbitrarily into anyone:
>
> https://www.future-science.com/doi/10.2144/00285bt01
>
>
> Here are some good modern books on microscopy imaging:
>
> https://www.elsevier.com/books/digital-microscopy/sluder/978-0-12-407761-4
>
> https://www.sciencedirect.com/bookseries/methods-in-cell-biology/vol/123
>
> https://onlinelibrary.wiley.com/doi/book/10.1002/9781118382905
>
> https://www.springer.com/gp/book/9780306484681
>
> https://www.springer.com/gp/book/9780387259215
>
>
> And in terms of digital imaging ethics, EVERYONE should read Cromey's article on this topic:
>
> https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4114110/
>
>
> John Oreopoulos
> Staff Scientist
> Andor Technology
>
> PLEASE NOTE OUR NEW ADDRESS:
>
> 250 West Beaver Creek Rd., Unit 1
> Richmond Hill, Ontario Canada L4B 1C7
> Tel: (905) 326-5040 ext. 412
> Email: [hidden email]
> WWW site: www.andor.com
>
> Quoting "Cammer, Michael" <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Increasingly we have found that microscope users do not understand  proper imaging.
>>
>> A few examples:
>> Confocal images are saturated and have the offset set to negative  numbers intentionally "to reduce background".
>> Exposures are changed in fluorescent micrographs to make each one  "look good".
>> 16 bit raw data with no saturation or clipping are set to 8 bits  with a lot of saturated and clipped pixels at the extremes and these  are what are considered raw data.
>> JPG compression is great because images can be emailed.
>> People assume (and get angry when this doesn't work) that computers  can automatically segment and count anything regardless of  experimental design, image quality, etc.
>> Even if they have saved everything as TIF with no metadata and  forgot what magnifications they used.
>>
>> A student who recognizes these problems and wants to learn the right  way emailed me yesterday asking for "any reliable literature you  recommend if I'd like to read more about correct imaging."
>>
>> I did Google searches and, after skimming a bunch of web pages,  realized that I didn't really have an answer.  I could send to this  page and to this page, but it's not a coherent narrative.  And I  don't have a library; most of what I know about this was picked up  willy-nilly up in the 1980s and '90s.  For instance knowledge of  color separation from writing animations for the Apple II in machine  code, working in a cable TV studio in the 1980s, and from an art  history class in college where we discussed 19th C travel photos and  poly/monochromatic film chemistries of the periods.  Later I learned  fluorescence quantification from a cell biologist intent on making  sure we had true linear responses for readouts of f-actin mass and  concentration, I was fortunate to attend both a Woods Hole  microscopy course and at NCSU John Russ's course on image analysis,  and BioRad MRC 600 training course at their offices in Cambridge, MA  covered a lot of material on this too.
>>
>> So, do people have favorite texts that cover these topics in a  manner digestible by new generations of students?  Of biology  students.  Keep in mind many don't know what a byte is; when I teach  I ask the class if anyone knows why an image is coded with white as  255 or 4095 or 16383 and typically one or two hands go up but the  rest of the class really doesn't know.
>> Looking forward to suggestions of easy to read definitive texts!
>>
>> Thank you.
>>
>> Cheers-
>>
>>
>> ------------------------------------------------------------
>> This email message, including any attachments, is for the sole use  of the intended recipient(s) and may contain information that is  proprietary, confidential, and exempt from disclosure under  applicable law. Any unauthorized review, use, disclosure, or  distribution is prohibited. If you have received this email in error  please notify the sender by return email and delete the original  message. Please note, the recipient should check this email and any  attachments for the presence of viruses. The organization accepts no  liability for any damage caused by any virus transmitted by this  email.
>> =================================
>>
Cromey, Douglas W - (dcromey) Cromey, Douglas W - (dcromey)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We see the same problems here. Once a year, right before the fall semester starts, we do a half-day workshop (http://microscopy.arizona.edu/dig_image_workshop/index) on scientific digital images. We start with the basics and work our way up through bit depth, color, file types, ethics and some hands-on demos in Photoshop & ImageJ. We advertise widely, and pull attendees from almost all the science disciplines and engineering (sometimes other colleges as well). Attendees range from undergrads to senior faculty. In the post-workshop evaluations we get a range of responses from complaints about wasting time covering the basics, people wishing there was more in-depth information on a particular topic, and the occasional (and rewarding) "I learned so much! Thank you". The workshop doesn't solve every problem, but I regularly meet people who remember a few key concepts and come to us questions about what they should do.

I have often been inspired by David Piston's commentary in Nature from a few years ago, that I need to help people understand how things work (https://www.nature.com/articles/484440a). Sadly, in these days of "kit biology", the ability to thoughtfully solve problems and learn something new along the way seems to be disappearing.

Doug

FYI: http://microscopy.arizona.edu/learn/digital-image-ethics 

(The workshop wasn't all my idea, a tip of the hat to colleagues Carl Boswell, Chip Hedgecock, David Elliott, Brooke Massani, and Ben Cromey who have all added significantly to the success of the workshop over the 10+ years we have been doing it here at the University of Arizona. It all started when a couple of microscopy core managers got together for lunch...)

------------------------------------------------------------------------------------------
Douglas W. Cromey, M.S. - Associate Scientific Investigator
Dept. of Cellular & Molecular Medicine, University of Arizona
1501 N. Campbell Ave, Tucson, AZ  85724-5044 USA

office:  LSN 463              email: [hidden email]
voice:  520-626-2824       fax:  520-626-2097

http://microscopy.arizona.edu/learn/microscopy-imaging-resources-www
Home of: "Microscopy and Imaging Resources on the WWW"

UA Microscopy Alliance -  http://microscopy.arizona.edu 


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Glen MacDonald
Sent: Friday, November 16, 2018 12:56 PM
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

thanks for posting this excellent list John.  
the ethics articles are helpful, but I think most bad images are made by not understanding how an image is created as well as what constitutes a digital image.  And all the points from sample preparation to quantitation/publication where the informational content may be compromised.  

FWIW, Alan Hibb’s book is dated but the information remains quite useful.    Before his premature death, he had shipped some books to me in preparation of a visit to North America.  I thought all were sold, but I just found one more copy while cleaning my study.  Original price was  US$125, I will sell it for half price plus shipping.  Proceeds are remitted to his family.

regards,
Glen MacDonald

> On Nov 16, 2018, at 9:54 AM, [hidden email] wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Where to begin... Well, here are the most modern books I can think of off the top of my head, but there are other good ones out there. I have an entire library of "basic microscopy" articles collected over the years which I can share with you offline as well. But in terms of a total/coherent narrative...
>
> Start with this paper by Pawley. This will put the fear of changing exposure levels arbitrarily into anyone:
>
> https://www.future-science.com/doi/10.2144/00285bt01
>
>
> Here are some good modern books on microscopy imaging:
>
> https://www.elsevier.com/books/digital-microscopy/sluder/978-0-12-407761-4
>
> https://www.sciencedirect.com/bookseries/methods-in-cell-biology/vol/123
>
> https://onlinelibrary.wiley.com/doi/book/10.1002/9781118382905
>
> https://www.springer.com/gp/book/9780306484681
>
> https://www.springer.com/gp/book/9780387259215
>
>
> And in terms of digital imaging ethics, EVERYONE should read Cromey's article on this topic:
>
> https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4114110/
>
>
> John Oreopoulos
> Staff Scientist
> Andor Technology
>
> PLEASE NOTE OUR NEW ADDRESS:
>
> 250 West Beaver Creek Rd., Unit 1
> Richmond Hill, Ontario Canada L4B 1C7
> Tel: (905) 326-5040 ext. 412
> Email: [hidden email]
> WWW site: www.andor.com
>
> Quoting "Cammer, Michael" <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Increasingly we have found that microscope users do not understand  proper imaging.
>>
>> A few examples:
>> Confocal images are saturated and have the offset set to negative  numbers intentionally "to reduce background".
>> Exposures are changed in fluorescent micrographs to make each one  "look good".
>> 16 bit raw data with no saturation or clipping are set to 8 bits  with a lot of saturated and clipped pixels at the extremes and these  are what are considered raw data.
>> JPG compression is great because images can be emailed.
>> People assume (and get angry when this doesn't work) that computers  can automatically segment and count anything regardless of  experimental design, image quality, etc.
>> Even if they have saved everything as TIF with no metadata and  forgot what magnifications they used.
>>
>> A student who recognizes these problems and wants to learn the right  way emailed me yesterday asking for "any reliable literature you  recommend if I'd like to read more about correct imaging."
>>
>> I did Google searches and, after skimming a bunch of web pages,  realized that I didn't really have an answer.  I could send to this  page and to this page, but it's not a coherent narrative.  And I  don't have a library; most of what I know about this was picked up  willy-nilly up in the 1980s and '90s.  For instance knowledge of  color separation from writing animations for the Apple II in machine  code, working in a cable TV studio in the 1980s, and from an art  history class in college where we discussed 19th C travel photos and  poly/monochromatic film chemistries of the periods.  Later I learned  fluorescence quantification from a cell biologist intent on making  sure we had true linear responses for readouts of f-actin mass and  concentration, I was fortunate to attend both a Woods Hole  microscopy course and at NCSU John Russ's course on image analysis,  and BioRad MRC 600 training course at their offices in Cambridge, MA  covered a lot of material on this too.
>>
>> So, do people have favorite texts that cover these topics in a  manner digestible by new generations of students?  Of biology  students.  Keep in mind many don't know what a byte is; when I teach  I ask the class if anyone knows why an image is coded with white as  255 or 4095 or 16383 and typically one or two hands go up but the  rest of the class really doesn't know.
>> Looking forward to suggestions of easy to read definitive texts!
>>
>> Thank you.
>>
>> Cheers-
>>
>>
>> ------------------------------------------------------------
>> This email message, including any attachments, is for the sole use  of the intended recipient(s) and may contain information that is  proprietary, confidential, and exempt from disclosure under  applicable law. Any unauthorized review, use, disclosure, or  distribution is prohibited. If you have received this email in error  please notify the sender by return email and delete the original  message. Please note, the recipient should check this email and any  attachments for the presence of viruses. The organization accepts no  liability for any damage caused by any virus transmitted by this  email.
>> =================================
>>
Dr. K N Ganesh Dr. K N Ganesh
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Lemasters, John J.
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi
I do visit various institutions to teach basics on microscopy and imaging always struggle convey the right message in the absence of a proper material..
Your presentation may be really a good tool for me to convey correct message and make the next generation scientist fool proof..
Kindly provide me the PPT ..
Reg
Ganesh Kadasoor
Bangalore
India


Sent from my iPhone

> On 16-Nov-2018, at 11:06 PM, Lemasters, John J. <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mike,
>
> I would like a copy of the PowerPoint as well. This is a very important issue that Michael Cammer raises. Trainees (and even more senior folks) continue to make these mistakes even as you emphatically tell them not to.
>
> Thanks again,
>
> John
>
> --
> John J. Lemasters, MD, PhD
> Professor and GlaxoSmithKline Distinguished Endowed Chair
> Director, Center for Cell Death, Injury & Regeneration
> Departments of Drug Discovery & Biomedical Sciences and Biochemistry & Molecular Biology
> Medical University of South Carolina
> DD504 Drug Discovery Building
> 70 President Street, MSC 139
> Charleston, SC 29425
>
> Office: 843-876-2360
> Lab: 843-876-2354
> Fax: 843-876-2353
> Email: [hidden email]
> http://academicdepartments.musc.edu/ccdir
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kathryn Spencer
> Sent: Friday, November 16, 2018 12:32 PM
> To: [hidden email]
> Subject: Re: seeking "reliable literature" on correct imaging
>
> CAUTION: External
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mike;
>        Could I get your PowerPoint as well? I have the same issues with my users. I'm "retraining" all of them with an open-software quiz, to see if they understand basic concepts. Many are complaining and reluctant, because they don't see the issues with their images. This after one 6-year post-doc's data was totally discarded because she saved everything as JPG....I certainly NEVER taught her to do THAT.
>        Your PowerPoint should hold more credibility that I obviously have....
>        Best;
> Kathy
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On Behalf Of MODEL, MICHAEL
> Sent: Friday, November 16, 2018 9:00 AM
> To: [hidden email]
> Subject: Re: seeking "reliable literature" on correct imaging
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Michael,
>
> We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.
>
> Mike Model
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
> Sent: Friday, November 16, 2018 11:50 AM
> To: [hidden email]
> Subject: seeking "reliable literature" on correct imaging
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Increasingly we have found that microscope users do not understand proper imaging.
>
> A few examples:
> Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
> Exposures are changed in fluorescent micrographs to make each one "look good".
> 16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
> JPG compression is great because images can be emailed.
> People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
> Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.
>
> A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."
>
> I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.
>
> So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
> Looking forward to suggestions of easy to read definitive texts!
>
> Thank you.
>
> Cheers-
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
> =================================
>
>
>
>
> -------------------------------------------------------------------------
> This message was secured via TLS by MUSC.
Karissa Tilbury Karissa Tilbury
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Mike,

I too would also appreciate this PowerPoint. I think in the context of the
ethics thread, this is very interesting and useful conversation. My
students are currently in the mandated ethics class and the topics
discussed are crazy extremes that really have their roots in 'smaller' less
obvious mistakes to most people. Focusing the discussion on what they are
doing in the lab to protect the quality of data or understand limitations
seems more important to me, than discussing ethics of CRISPR and human
life.

Karissa

On Fri, Nov 16, 2018, 8:44 PM Ganesh Kadasoor <[hidden email]
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi
> I do visit various institutions to teach basics on microscopy and imaging
> always struggle convey the right message in the absence of a proper
> material..
> Your presentation may be really a good tool for me to convey correct
> message and make the next generation scientist fool proof..
> Kindly provide me the PPT ..
> Reg
> Ganesh Kadasoor
> Bangalore
> India
>
>
> Sent from my iPhone
>
> > On 16-Nov-2018, at 11:06 PM, Lemasters, John J. <[hidden email]>
> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Mike,
> >
> > I would like a copy of the PowerPoint as well. This is a very important
> issue that Michael Cammer raises. Trainees (and even more senior folks)
> continue to make these mistakes even as you emphatically tell them not to.
> >
> > Thanks again,
> >
> > John
> >
> > --
> > John J. Lemasters, MD, PhD
> > Professor and GlaxoSmithKline Distinguished Endowed Chair
> > Director, Center for Cell Death, Injury & Regeneration
> > Departments of Drug Discovery & Biomedical Sciences and Biochemistry &
> Molecular Biology
> > Medical University of South Carolina
> > DD504 Drug Discovery Building
> > 70 President Street, MSC 139
> > Charleston, SC 29425
> >
> > Office: 843-876-2360
> > Lab: 843-876-2354
> > Fax: 843-876-2353
> > Email: [hidden email]
> > http://academicdepartments.musc.edu/ccdir
> >
> > -----Original Message-----
> > From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Kathryn Spencer
> > Sent: Friday, November 16, 2018 12:32 PM
> > To: [hidden email]
> > Subject: Re: seeking "reliable literature" on correct imaging
> >
> > CAUTION: External
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Mike;
> >        Could I get your PowerPoint as well? I have the same issues with
> my users. I'm "retraining" all of them with an open-software quiz, to see
> if they understand basic concepts. Many are complaining and reluctant,
> because they don't see the issues with their images. This after one 6-year
> post-doc's data was totally discarded because she saved everything as
> JPG....I certainly NEVER taught her to do THAT.
> >        Your PowerPoint should hold more credibility that I obviously
> have....
> >        Best;
> > Kathy
> >
> >
> >
> > -----Original Message-----
> > From: Confocal Microscopy List <[hidden email]> On
> Behalf Of MODEL, MICHAEL
> > Sent: Friday, November 16, 2018 9:00 AM
> > To: [hidden email]
> > Subject: Re: seeking "reliable literature" on correct imaging
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Michael,
> >
> > We recently ran a workshop where I gave a 3-hour talk on various kinds
> of quantification in microscopy. I can send you a PowerPoint file but don't
> have a text.
> >
> > Mike Model
> >
> > -----Original Message-----
> > From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Cammer, Michael
> > Sent: Friday, November 16, 2018 11:50 AM
> > To: [hidden email]
> > Subject: seeking "reliable literature" on correct imaging
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Increasingly we have found that microscope users do not understand
> proper imaging.
> >
> > A few examples:
> > Confocal images are saturated and have the offset set to negative
> numbers intentionally "to reduce background".
> > Exposures are changed in fluorescent micrographs to make each one "look
> good".
> > 16 bit raw data with no saturation or clipping are set to 8 bits with a
> lot of saturated and clipped pixels at the extremes and these are what are
> considered raw data.
> > JPG compression is great because images can be emailed.
> > People assume (and get angry when this doesn't work) that computers can
> automatically segment and count anything regardless of experimental design,
> image quality, etc.
> > Even if they have saved everything as TIF with no metadata and forgot
> what magnifications they used.
> >
> > A student who recognizes these problems and wants to learn the right way
> emailed me yesterday asking for "any reliable literature you recommend if
> I'd like to read more about correct imaging."
> >
> > I did Google searches and, after skimming a bunch of web pages, realized
> that I didn't really have an answer.  I could send to this page and to this
> page, but it's not a coherent narrative.  And I don't have a library; most
> of what I know about this was picked up willy-nilly up in the 1980s and
> '90s.  For instance knowledge of color separation from writing animations
> for the Apple II in machine code, working in a cable TV studio in the
> 1980s, and from an art history class in college where we discussed 19th C
> travel photos and poly/monochromatic film chemistries of the periods.
> Later I learned fluorescence quantification from a cell biologist intent on
> making sure we had true linear responses for readouts of f-actin mass and
> concentration, I was fortunate to attend both a Woods Hole microscopy
> course and at NCSU John Russ's course on image analysis, and BioRad MRC 600
> training course at their offices in Cambridge, MA covered a lot of material
> on this too.
> >
> > So, do people have favorite texts that cover these topics in a manner
> digestible by new generations of students?  Of biology students.  Keep in
> mind many don't know what a byte is; when I teach I ask the class if anyone
> knows why an image is coded with white as 255 or 4095 or 16383 and
> typically one or two hands go up but the rest of the class really doesn't
> know.
> > Looking forward to suggestions of easy to read definitive texts!
> >
> > Thank you.
> >
> > Cheers-
> >
> >
> > ------------------------------------------------------------
> > This email message, including any attachments, is for the sole use of
> the intended recipient(s) and may contain information that is proprietary,
> confidential, and exempt from disclosure under applicable law. Any
> unauthorized review, use, disclosure, or distribution is prohibited. If you
> have received this email in error please notify the sender by return email
> and delete the original message. Please note, the recipient should check
> this email and any attachments for the presence of viruses. The
> organization accepts no liability for any damage caused by any virus
> transmitted by this email.
> > =================================
> >
> >
> >
> >
> > -------------------------------------------------------------------------
> > This message was secured via TLS by MUSC.
>
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Here are a few resources I use.
However, I think a good introductory microscopy course is best. Would be great to see these kinds of courses being built into the undergraduate programs for life scientists but it might be a while before this kind of training is broadly offered. However, there are many great courses offered through core facilities worldwide.

Great web resource. They are about to add a module on super resolution imaging. Tests at the end of each module.
https://myscope.training/

Other web resources:
iBiology microscopy courses: instructive link with a lot of information and videos. Excellent for new users looking to get a solid understanding of microscopy basics. http://www.ibiology.org/ibioeducation/taking-courses.html

Good for users with Leica microscopes (somewhat specific to the brand) and some tutorials that are good to understand the basics of fluorescent microscopy. http://www.leica-microsystems.com/science-lab/topics/

Nikon’s Microscopy U.: Lots of articles and tutorials subdivided by application type. A very large amount of information so it’s helpful to know what you are interested in before visiting the site. http://www.microscopyu.com/

Interactive tutorials and educational material on both a basic level (basic primers on microscopy) and an advanced, specialized techniques level.
http://www.olympusmicro.com/

Similar to the above sites, but offered through Zeiss with reviews and interactive tutorials from basic concepts up to specialized techniques.
http://zeiss-campus.magnet.fsu.edu/

Quantitative Microscopy

Swedlow JR. Quantitative fluorescence microscopy and image deconvolution. Methods Cell Biol 2013;114:407–426.

Murray JM. Practical aspects of quantitative confocal microscopy. Methods Cell Biol 2007;81:467–478.

Waters JC. Accuracy and precision in quantitative fluorescence microscopy. J Cell Biol 2009;185:1135–1148.

North, A. J. (2006). Seeing is believing? A beginners' guide to practical pitfalls in image acquisition. J. Cell Biol. 172, 9-18.

Methods Cell Biol. 2014;123:113-34. doi: 10.1016/B978-0-12-420138-5.00007-0. Quantitative confocal microscopy: beyond a pretty picture. Jonkman J, Brown CM, Cole RW.

Fluorescence microscopy - avoiding the pitfalls (Probably needs to be updated) Claire M. Brown Journal of Cell Science 2007 120: 1703-1705; doi: 10.1242/jcs.03433

Good luck!
Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by mmodel
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mike,

I have given talks like this in the rather dim past, but would be very interested in a copy of your Powerpoint file.
Maybe you might have time to make it into a small (review) paper? One of the microscopy journals might publish it - e.g. Journal of Microscopy(?)

thanks,
Rosemary

Dr Rosemary White
CSIRO Black Mountain
GPO Box 1700
Canberra, ACT 2601
Australia

T 61 2 6246 5475
M 61 468 966 713
________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of MODEL, MICHAEL [[hidden email]]
Sent: Saturday, 17 November 2018 4:00 a.m.
To: [hidden email]
Subject: Re: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

We recently ran a workshop where I gave a 3-hour talk on various kinds of quantification in microscopy. I can send you a PowerPoint file but don't have a text.

Mike Model

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Cammer, Michael
Sent: Friday, November 16, 2018 11:50 AM
To: [hidden email]
Subject: seeking "reliable literature" on correct imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Increasingly we have found that microscope users do not understand proper imaging.

A few examples:
Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
Exposures are changed in fluorescent micrographs to make each one "look good".
16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
JPG compression is great because images can be emailed.
People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.

A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."

I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.

So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
Looking forward to suggestions of easy to read definitive texts!

Thank you.

Cheers-


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Steffen Dietzel Steffen Dietzel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: seeking "reliable literature" on correct imaging

In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Michael,

I guess we all face similar problems in that respect with our users. I
put some of my favorite resources on our web site so that I have
something to refer to, when they ask.
(https://www.bioimaging.bmc.med.uni-muenchen.de/learn/index.html).
Paperbook-wise I often recommend Guy Cox's "optical imaging techniques
in cell biology", not because it would be better than the alternatives
(Murphy&Davison, The Handbook) but because it is very concise and to the
point. For those who'd rather listen than read I recommend the iBiology
videos.

Within Germanbioimaging we developed a list of topics that should be
taught to (new) users and provide some teaching materials. The GerBi
website is currently being remodeled but my contribution to the
materials also can be found here:
https://www.bioimaging.bmc.med.uni-muenchen.de/learn/materialdownload/index.html 
These powerpoints do not include my complete lectures, all the
copyrighted materials are omitted. And sometimes slides are redundant.
It is more a collection of slides that may be useful for those who have
to give a lecture on the respective topics.

But I guess much of the important bits we teach during the practical
introduction (always save the original file format, don't overexpose,
etc.) And about once in a year I give a 2x3h lecture on the basics of
microscopy and imaging.

But in some aspects I guess you always will leave room for surprises.
Kathryn's example of the postdoc saving only Jpgs and Ganesh's comment
on making the next generation of scientists fool proof remind me of one
of my favorite Douglas Adam quotes, though: "The problem with making
something completely foolproof is the ingenuity of complete fools."
Meaning: we can hope to diminish the problem, but we will never solve it
completely.

I definitely will have a look into John's list, thank you for that (and
all the other interesting sources mentioned)

Steffen




Am 16.11.2018 um 17:50 schrieb Cammer, Michael:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Increasingly we have found that microscope users do not understand proper imaging.
>
> A few examples:
> Confocal images are saturated and have the offset set to negative numbers intentionally "to reduce background".
> Exposures are changed in fluorescent micrographs to make each one "look good".
> 16 bit raw data with no saturation or clipping are set to 8 bits with a lot of saturated and clipped pixels at the extremes and these are what are considered raw data.
> JPG compression is great because images can be emailed.
> People assume (and get angry when this doesn't work) that computers can automatically segment and count anything regardless of experimental design, image quality, etc.
> Even if they have saved everything as TIF with no metadata and forgot what magnifications they used.
>
> A student who recognizes these problems and wants to learn the right way emailed me yesterday asking for "any reliable literature you recommend if I'd like to read more about correct imaging."
>
> I did Google searches and, after skimming a bunch of web pages, realized that I didn't really have an answer.  I could send to this page and to this page, but it's not a coherent narrative.  And I don't have a library; most of what I know about this was picked up willy-nilly up in the 1980s and '90s.  For instance knowledge of color separation from writing animations for the Apple II in machine code, working in a cable TV studio in the 1980s, and from an art history class in college where we discussed 19th C travel photos and poly/monochromatic film chemistries of the periods.  Later I learned fluorescence quantification from a cell biologist intent on making sure we had true linear responses for readouts of f-actin mass and concentration, I was fortunate to attend both a Woods Hole microscopy course and at NCSU John Russ's course on image analysis, and BioRad MRC 600 training course at their offices in Cambridge, MA covered a lot of material on this too.
>
> So, do people have favorite texts that cover these topics in a manner digestible by new generations of students?  Of biology students.  Keep in mind many don't know what a byte is; when I teach I ask the class if anyone knows why an image is coded with white as 255 or 4095 or 16383 and typically one or two hands go up but the rest of the class really doesn't know.
> Looking forward to suggestions of easy to read definitive texts!
>
> Thank you.
>
> Cheers-
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
> =================================
>
--
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Biomedical Center (BMC)
Head of the Core Facility Bioimaging

Großhaderner Straße 9
D-82152 Planegg-Martinsried
Germany

http://www.bioimaging.bmc.med.uni-muenchen.de
12