Michael Giacomelli |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have been reading up on signal to background ratios in point scanning confocal and spinning disk confocal. There are a number of very detailed papers that characterize the out of focus rejection in terms of the edge response function for a thick fluorescent sample in the one photon case (figure 14, [1] for instance) . What i have not been able to locate is a reasonably thorough treatment of the subject for multiphoton microscopy with non-descanned detection. Instead most of the work I can find only looks at the signal to background ratios with depth into a tissue specimen. This is helpful, but difficult to compare since you can't image deeply with confocal. Does anyone have a good reference they could suggest? [1] Zimmermann, B., Kempe, M., & Wolleschensky, R. (2006). High-speed confocal fluorescence imaging with a novel line scanning microscope. Journal of Biomedical Optics, 11(6), 064011. http://doi.org/10.1117/1.2402110 |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Because the signal is only generated at the focal plane, multiphoton doesn't have the same kind of background issues as confocal. This means there is no out-of-focus signal to reject. The spatial restriction means that your laser's point spread function (in XYZ) is the limit of your detection and your signal generation volume. For instance in the edge response to a thick fluorescent sample, you won't get any signal generated until the PSF of your laser actually starts overlapping with the edge of the slab. Conversely in confocal, you will always have signal being generated and are relying on the pinhole to reject the out-of-plane light. As you take a volume stack into a slab In 2P you will get some entertaining edge effects due to spherical aberration if the index of your fluorescent slab isn't matched to your immersion media, but otherwise 2P just doesn't generate signal unless the sample is within the focal spot. Please keep in mind I am thinking about the Z direction in this, not XY. I hope this helps; I'm really tired and may not be as coherent as usual. :) Craig On Wed, May 11, 2016 at 3:01 PM, Michael Giacomelli <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I have been reading up on signal to background ratios in point > scanning confocal and spinning disk confocal. There are a number of > very detailed papers that characterize the out of focus rejection in > terms of the edge response function for a thick fluorescent sample in > the one photon case (figure 14, [1] for instance) . > > What i have not been able to locate is a reasonably thorough treatment > of the subject for multiphoton microscopy with non-descanned > detection. Instead most of the work I can find only looks at the > signal to background ratios with depth into a tissue specimen. This > is helpful, but difficult to compare since you can't image deeply with > confocal. > > Does anyone have a good reference they could suggest? > > > [1] Zimmermann, B., Kempe, M., & Wolleschensky, R. (2006). High-speed > confocal fluorescence imaging with a novel line scanning microscope. > Journal of Biomedical Optics, 11(6), 064011. > http://doi.org/10.1117/1.2402110 > |
|
Michael Giacomelli |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Craig, I realize that 2 photon excitation is most likely in the focal plane, but the probability of it happening is nonzero everywhere the beam reaches, so you do get some signal from out of focus light as in the confocal case. This is an issue in deep brain imaging for instance. I spent a little while searching on Google scholar but didn't see very much on signal to background ratios derived or measured for the step test though, just deep tissue imaging. For what it's worth, I'm curious because I do see some depth dependent background when Z-stacking the air above a fluorescent slide. It's weak, but I'm curious if what I measure is reasonable, or if I should be double checking my PSF and filter rejection. Mike On Wed, May 11, 2016 at 9:01 PM, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Because the signal is only generated at the focal plane, multiphoton > doesn't have the same kind of background issues as confocal. This means > there is no out-of-focus signal to reject. The spatial restriction means > that your laser's point spread function (in XYZ) is the limit of your > detection and your signal generation volume. For instance in the edge > response to a thick fluorescent sample, you won't get any signal generated > until the PSF of your laser actually starts overlapping with the edge of > the slab. Conversely in confocal, you will always have signal being > generated and are relying on the pinhole to reject the out-of-plane light. > As you take a volume stack into a slab In 2P you will get some entertaining > edge effects due to spherical aberration if the index of your fluorescent > slab isn't matched to your immersion media, but otherwise 2P just doesn't > generate signal unless the sample is within the focal spot. Please keep in > mind I am thinking about the Z direction in this, not XY. > > I hope this helps; I'm really tired and may not be as coherent as usual. :) > Craig > > On Wed, May 11, 2016 at 3:01 PM, Michael Giacomelli <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I have been reading up on signal to background ratios in point >> scanning confocal and spinning disk confocal. There are a number of >> very detailed papers that characterize the out of focus rejection in >> terms of the edge response function for a thick fluorescent sample in >> the one photon case (figure 14, [1] for instance) . >> >> What i have not been able to locate is a reasonably thorough treatment >> of the subject for multiphoton microscopy with non-descanned >> detection. Instead most of the work I can find only looks at the >> signal to background ratios with depth into a tissue specimen. This >> is helpful, but difficult to compare since you can't image deeply with >> confocal. >> >> Does anyone have a good reference they could suggest? >> >> >> [1] Zimmermann, B., Kempe, M., & Wolleschensky, R. (2006). High-speed >> confocal fluorescence imaging with a novel line scanning microscope. >> Journal of Biomedical Optics, 11(6), 064011. >> http://doi.org/10.1117/1.2402110 >> |
|
Michael Giacomelli |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** After some more reading, I found a reference: http://onlinelibrary.wiley.com/doi/10.1046/j.1365-2818.1998.00431.x/abstract My edge response looks reasonable, but the long tail I get is a little too high, so something isn't quite right. Mike On Wed, May 11, 2016 at 9:31 PM, Michael Giacomelli <[hidden email]> wrote: > Hi Craig, > > I realize that 2 photon excitation is most likely in the focal plane, > but the probability of it happening is nonzero everywhere the beam > reaches, so you do get some signal from out of focus light as in the > confocal case. This is an issue in deep brain imaging for instance. > I spent a little while searching on Google scholar but didn't see very > much on signal to background ratios derived or measured for the step > test though, just deep tissue imaging. > > For what it's worth, I'm curious because I do see some depth dependent > background when Z-stacking the air above a fluorescent slide. It's > weak, but I'm curious if what I measure is reasonable, or if I should > be double checking my PSF and filter rejection. > > Mike > > On Wed, May 11, 2016 at 9:01 PM, Craig Brideau <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Because the signal is only generated at the focal plane, multiphoton >> doesn't have the same kind of background issues as confocal. This means >> there is no out-of-focus signal to reject. The spatial restriction means >> that your laser's point spread function (in XYZ) is the limit of your >> detection and your signal generation volume. For instance in the edge >> response to a thick fluorescent sample, you won't get any signal generated >> until the PSF of your laser actually starts overlapping with the edge of >> the slab. Conversely in confocal, you will always have signal being >> generated and are relying on the pinhole to reject the out-of-plane light. >> As you take a volume stack into a slab In 2P you will get some entertaining >> edge effects due to spherical aberration if the index of your fluorescent >> slab isn't matched to your immersion media, but otherwise 2P just doesn't >> generate signal unless the sample is within the focal spot. Please keep in >> mind I am thinking about the Z direction in this, not XY. >> >> I hope this helps; I'm really tired and may not be as coherent as usual. :) >> Craig >> >> On Wed, May 11, 2016 at 3:01 PM, Michael Giacomelli <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> I have been reading up on signal to background ratios in point >>> scanning confocal and spinning disk confocal. There are a number of >>> very detailed papers that characterize the out of focus rejection in >>> terms of the edge response function for a thick fluorescent sample in >>> the one photon case (figure 14, [1] for instance) . >>> >>> What i have not been able to locate is a reasonably thorough treatment >>> of the subject for multiphoton microscopy with non-descanned >>> detection. Instead most of the work I can find only looks at the >>> signal to background ratios with depth into a tissue specimen. This >>> is helpful, but difficult to compare since you can't image deeply with >>> confocal. >>> >>> Does anyone have a good reference they could suggest? >>> >>> >>> [1] Zimmermann, B., Kempe, M., & Wolleschensky, R. (2006). High-speed >>> confocal fluorescence imaging with a novel line scanning microscope. >>> Journal of Biomedical Optics, 11(6), 064011. >>> http://doi.org/10.1117/1.2402110 >>> |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If you are referring to a plot of Z depth to signal intensity, it will be very asymmetric due to spherical aberration caused as you go into the interface. I have also achieved strange results when doing this myself by saturating all the fluorophores within the PSF which will give you very weird Z vs intensity plots. Try keeping your excitation power low to avoid this. Craig On May 11, 2016 8:30 PM, "Michael Giacomelli" <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** After some more reading, I found a reference: http://onlinelibrary.wiley.com/doi/10.1046/j.1365-2818.1998.00431.x/abstract My edge response looks reasonable, but the long tail I get is a little too high, so something isn't quite right. Mike On Wed, May 11, 2016 at 9:31 PM, Michael Giacomelli <[hidden email]> wrote: > Hi Craig, > > I realize that 2 photon excitation is most likely in the focal plane, > but the probability of it happening is nonzero everywhere the beam > reaches, so you do get some signal from out of focus light as in the > confocal case. This is an issue in deep brain imaging for instance. > I spent a little while searching on Google scholar but didn't see very > much on signal to background ratios derived or measured for the step > test though, just deep tissue imaging. > > For what it's worth, I'm curious because I do see some depth dependent > background when Z-stacking the air above a fluorescent slide. It's > weak, but I'm curious if what I measure is reasonable, or if I should > be double checking my PSF and filter rejection. > > Mike > > On Wed, May 11, 2016 at 9:01 PM, Craig Brideau <[hidden email]> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Because the signal is only generated at the focal plane, multiphoton >> doesn't have the same kind of background issues as confocal. This means >> there is no out-of-focus signal to reject. The spatial restriction means >> that your laser's point spread function (in XYZ) is the limit of your >> detection and your signal generation volume. For instance in the edge >> response to a thick fluorescent sample, you won't get any signal >> until the PSF of your laser actually starts overlapping with the edge of >> the slab. Conversely in confocal, you will always have signal being >> generated and are relying on the pinhole to reject the out-of-plane light. >> As you take a volume stack into a slab In 2P you will get some entertaining >> edge effects due to spherical aberration if the index of your fluorescent >> slab isn't matched to your immersion media, but otherwise 2P just doesn't >> generate signal unless the sample is within the focal spot. Please keep in >> mind I am thinking about the Z direction in this, not XY. >> >> I hope this helps; I'm really tired and may not be as coherent as usual. :) >> Craig >> >> On Wed, May 11, 2016 at 3:01 PM, Michael Giacomelli <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> I have been reading up on signal to background ratios in point >>> scanning confocal and spinning disk confocal. There are a number of >>> very detailed papers that characterize the out of focus rejection in >>> terms of the edge response function for a thick fluorescent sample in >>> the one photon case (figure 14, [1] for instance) . >>> >>> What i have not been able to locate is a reasonably thorough treatment >>> of the subject for multiphoton microscopy with non-descanned >>> detection. Instead most of the work I can find only looks at the >>> signal to background ratios with depth into a tissue specimen. This >>> is helpful, but difficult to compare since you can't image deeply with >>> confocal. >>> >>> Does anyone have a good reference they could suggest? >>> >>> >>> [1] Zimmermann, B., Kempe, M., & Wolleschensky, R. (2006). High-speed >>> confocal fluorescence imaging with a novel line scanning microscope. >>> Journal of Biomedical Optics, 11(6), 064011. >>> http://doi.org/10.1117/1.2402110 >>> |