![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. Any suggestions welcome, Thanks alot, -- Philipp Bethge |
![]() ![]() |
Johannes Schindelin |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, On Wed, 21 Mar 2012, Philipp Bethge wrote: > I was wondering if somebody has worked with a 'smart' average > z-projection, that is a average projection of a z-stack that does not > average of the entire stack size but only over the slides where a signal > is actually present. So, for example, one has an oblique structure in a > large stack, lets say the structure has a z-extension of 4 slides but > the whole stack is 30 stacks large. As I understand the regular average > z-projection algorithm in ImageJ, every pixel is summed and devided by > the total number of slides in the stack. Then the signal of only 4 > slides would be averaged over 30 slides, therefore reducing signal > intensity dramatically without further reducing background noise. Of > course, one would have to define 'signal' somehow, but I guess that > could be done manually by binarization of the image slides individually. You could use a weighted sum rather than the mean intensity, for example by multiplying with the squared difference to the mean intensity. Ciao, Johannes |
![]() ![]() |
Glen MacDonald-2 |
![]() |
In reply to this post by Philipp Bethge
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Philipp, Duplicate the stack, threshold for the structure then make the tholded stack binary dividing by 255 converts the binarized stack from 0+255 to 0+1 multiply the 2 stacks and project the result. this function does all this, choosing middle slice for setting the threshold Maybe this is helpful, function thresholdprojection(){ stackid=getImageID; no=nSlices; stackname=getTitle; run("Duplicate...", "title=[dup] duplicate range=1-no"); setSlice((floor(no/2))); run("Threshold..."); waitForUser("Set Threshold then click 'OK'."); setBatchMode(true); getThreshold(lower,upper); run("Make Binary", "black"); run("Divide...", "value=255 stack"); rename("divided"); bindupid=getImageID; imageCalculator("Multiply create stack", stackid ,bindupid); run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); rename(stackname+"sum"); sumid=getImageID; setBatchMode(false); selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode close(); } Regards, Glen Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] On Mar 21, 2012, at 8:20 AM, Philipp Bethge wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. > Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. > > Any suggestions welcome, > > Thanks alot, > > -- > Philipp Bethge |
![]() |
In reply to this post by Philipp Bethge
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, Have you tried the Standard Deviation projection? I really like that it w ill not over saturate, and the projection looks "natural" best James Beals [hidden email] On Mar 21, 2012, at 11:20 AM, Philipp Bethge wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. > Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. > > Any suggestions welcome, > > Thanks alot, > > -- > Philipp Bethge |
![]() ![]() |
Lloyd Donaldson |
![]() |
In reply to this post by Philipp Bethge
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Philipp Calculate the sum as normal but divide only by the number of pixels greater than some threshold - we typically use 5 or 10 which excludes any black/empty pixels. This avoids the projection being dim compared to say a maximum intensity projection, where there are empty regions within the volume. Dr Lloyd Donaldson Senior Scientist, Project Leader - Microscopy/Wood Identification Scion - Forests . Products . Innovation Private Bag 3020, Rotorua 49 Sala Street, Rotorua 3010 New Zealand Ph: 64 7 343 5581 www.scionresearch.com -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philipp Bethge Sent: Thursday, 22 March 2012 4:21 a.m. To: [hidden email] Subject: 'smart' average z-projection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. Any suggestions welcome, Thanks alot, -- Philipp Bethge This e-mail and any attachments may contain information which is confidential or subject to copyright. If you receive this e-mail in error, please delete it. Scion does not accept responsibility for anything in this e-mail which is not provided in the course of Scion's usual business or for any computer virus, data corruption, interference or delay arising from this e-mail. |
![]() ![]() |
Philipp Bethge |
![]() |
In reply to this post by Glen MacDonald-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Users, Wow, thanks alot to everybody for the quick reply and the solutions offered. I tried out the one from Glen and it works very nice! Thanks alot, > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Philipp, > Duplicate the stack, threshold for the structure then make the tholded stack binary > dividing by 255 converts the binarized stack from 0+255 to 0+1 > multiply the 2 stacks and project the result. > this function does all this, choosing middle slice for setting the threshold > > Maybe this is helpful, > > function thresholdprojection(){ > stackid=getImageID; > no=nSlices; > stackname=getTitle; > run("Duplicate...", "title=[dup] duplicate range=1-no"); > setSlice((floor(no/2))); > run("Threshold..."); > waitForUser("Set Threshold then click 'OK'."); > setBatchMode(true); > getThreshold(lower,upper); > run("Make Binary", "black"); > run("Divide...", "value=255 stack"); > rename("divided"); > bindupid=getImageID; > imageCalculator("Multiply create stack", stackid ,bindupid); > run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); > rename(stackname+"sum"); > sumid=getImageID; > setBatchMode(false); > selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode > close(); > } > > Regards, > Glen > Glen MacDonald > Core for Communication Research > Virginia Merrill Bloedel Hearing Research Center > Box 357923 > University of Washington > Seattle, WA 98195-7923 USA > (206) 616-4156 > [hidden email] > > > > > > > > > On Mar 21, 2012, at 8:20 AM, Philipp Bethge wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. >> Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. >> >> Any suggestions welcome, >> >> Thanks alot, >> >> -- >> Philipp Bethge -- Philipp Bethge |
![]() ![]() |
Straatman, Kees (Dr.) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Philipp, Glen and list, I tried the code Glen gave but the macro did not behave as I expected, at least not in my hands. Problem was the line 'run("Make Binary", "black");' Changes in threshold settings were not reflected in the final image. I made a small change what worked for me. Just in case others have the same problem: macro Smart_Average{ stackid=getImageID; no=nSlices; stackname=getTitle; run("Duplicate...", "title=[dup] duplicate range=1-no"); setSlice((floor(no/2))); run("Threshold..."); waitForUser("Set Threshold then click 'OK'."); setBatchMode(true); run("Convert to Mask", " black"); run("Divide...", "value=255 stack"); rename("divided"); bindupid=getImageID; imageCalculator("Multiply create stack", stackid ,bindupid); run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); rename(stackname+"sum"); sumid=getImageID; setBatchMode(false); selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode close(); } Best wishes Kees Dr Ir K.R. Straatman Senior Experimental Officer Centre for Core Biotechnology Services College of Medicine, Biological Sciences and Psychology University of Leicester http://www.le.ac.uk/biochem/microscopy/home.html -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philipp Bethge Sent: 21 March 2012 19:55 To: [hidden email] Subject: Re: 'smart' average z-projection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Users, Wow, thanks alot to everybody for the quick reply and the solutions offered. I tried out the one from Glen and it works very nice! Thanks alot, > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Philipp, > Duplicate the stack, threshold for the structure then make the tholded stack binary > dividing by 255 converts the binarized stack from 0+255 to 0+1 > multiply the 2 stacks and project the result. > this function does all this, choosing middle slice for setting the threshold > > Maybe this is helpful, > > function thresholdprojection(){ > stackid=getImageID; > no=nSlices; > stackname=getTitle; > run("Duplicate...", "title=[dup] duplicate range=1-no"); > setSlice((floor(no/2))); > run("Threshold..."); > waitForUser("Set Threshold then click 'OK'."); > setBatchMode(true); > getThreshold(lower,upper); > run("Make Binary", "black"); > run("Divide...", "value=255 stack"); > rename("divided"); > bindupid=getImageID; > imageCalculator("Multiply create stack", stackid ,bindupid); > run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); > rename(stackname+"sum"); > sumid=getImageID; > setBatchMode(false); > selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode > close(); > } > > Regards, > Glen > Glen MacDonald > Core for Communication Research > Virginia Merrill Bloedel Hearing Research Center > Box 357923 > University of Washington > Seattle, WA 98195-7923 USA > (206) 616-4156 > [hidden email] > > > > > > > > > On Mar 21, 2012, at 8:20 AM, Philipp Bethge wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. >> Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. >> >> Any suggestions welcome, >> >> Thanks alot, >> >> -- >> Philipp Bethge -- Philipp Bethge |
![]() ![]() |
Glen MacDonald-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Kees, did you run this on 8-bit images or 16-bit? 'run("Make Binary", "black");' and 'run("Convert to Mask", " black");' behave similarly on 16-bit images, but run("Make Binary", "black"); on an 8-bit does not set to 0+255. However, dividing the Make Binary 8-bit image by 255 effectively sets the image to 0+1 due to rounding. My macro was written for 16-bit images, but I'll change it to prevent future issues if used with 8-bit. thanks for pointing this out. Regards, Glen On Mar 22, 2012, at 4:28 AM, Straatman, Kees R. (Dr.) wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Philipp, Glen and list, > > I tried the code Glen gave but the macro did not behave as I expected, at least not in my hands. Problem was the line 'run("Make Binary", "black");' Changes in threshold settings were not reflected in the final image. I made a small change what worked for me. Just in case others have the same problem: > > macro Smart_Average{ > stackid=getImageID; > no=nSlices; > stackname=getTitle; > run("Duplicate...", "title=[dup] duplicate range=1-no"); > setSlice((floor(no/2))); > run("Threshold..."); > waitForUser("Set Threshold then click 'OK'."); > setBatchMode(true); > run("Convert to Mask", " black"); > run("Divide...", "value=255 stack"); > rename("divided"); > bindupid=getImageID; > imageCalculator("Multiply create stack", stackid ,bindupid); > run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); > rename(stackname+"sum"); > sumid=getImageID; > setBatchMode(false); > selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode > close(); > } > > Best wishes > > Kees > > > Dr Ir K.R. Straatman > Senior Experimental Officer > Centre for Core Biotechnology Services > College of Medicine, Biological Sciences and Psychology > University of Leicester > > http://www.le.ac.uk/biochem/microscopy/home.html > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philipp Bethge > Sent: 21 March 2012 19:55 > To: [hidden email] > Subject: Re: 'smart' average z-projection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Users, > > Wow, thanks alot to everybody for the quick reply and the solutions offered. > I tried out the one from Glen and it works very nice! > > Thanks alot, > > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear Philipp, >> Duplicate the stack, threshold for the structure then make the tholded stack binary >> dividing by 255 converts the binarized stack from 0+255 to 0+1 >> multiply the 2 stacks and project the result. >> this function does all this, choosing middle slice for setting the threshold >> >> Maybe this is helpful, >> >> function thresholdprojection(){ >> stackid=getImageID; >> no=nSlices; >> stackname=getTitle; >> run("Duplicate...", "title=[dup] duplicate range=1-no"); >> setSlice((floor(no/2))); >> run("Threshold..."); >> waitForUser("Set Threshold then click 'OK'."); >> setBatchMode(true); >> getThreshold(lower,upper); >> run("Make Binary", "black"); >> run("Divide...", "value=255 stack"); >> rename("divided"); >> bindupid=getImageID; >> imageCalculator("Multiply create stack", stackid ,bindupid); >> run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); >> rename(stackname+"sum"); >> sumid=getImageID; >> setBatchMode(false); >> selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode >> close(); >> } >> >> Regards, >> Glen >> Glen MacDonald >> Core for Communication Research >> Virginia Merrill Bloedel Hearing Research Center >> Box 357923 >> University of Washington >> Seattle, WA 98195-7923 USA >> (206) 616-4156 >> [hidden email] >> >> >> >> >> >> >> >> >> On Mar 21, 2012, at 8:20 AM, Philipp Bethge wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear all, >>> >>> I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. >>> Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. >>> >>> Any suggestions welcome, >>> >>> Thanks alot, >>> >>> -- >>> Philipp Bethge > > > -- > Philipp Bethge |
![]() ![]() |
Straatman, Kees (Dr.) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Glen, Indeed, I was using an 8 bit images, the sample Confocal-Series.tif. But I was not aware of this difference in thresholding between 8 bit and 16 bit images... Cheers Kees -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Glen MacDonald Sent: 22 March 2012 18:25 To: [hidden email] Subject: Re: 'smart' average z-projection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Kees, did you run this on 8-bit images or 16-bit? 'run("Make Binary", "black");' and 'run("Convert to Mask", " black");' behave similarly on 16-bit images, but run("Make Binary", "black"); on an 8-bit does not set to 0+255. However, dividing the Make Binary 8-bit image by 255 effectively sets the image to 0+1 due to rounding. My macro was written for 16-bit images, but I'll change it to prevent future issues if used with 8-bit. thanks for pointing this out. Regards, Glen On Mar 22, 2012, at 4:28 AM, Straatman, Kees R. (Dr.) wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Philipp, Glen and list, > > I tried the code Glen gave but the macro did not behave as I expected, at least not in my hands. Problem was the line 'run("Make Binary", "black");' Changes in threshold settings were not reflected in the final image. I made a small change what worked for me. Just in case others have the same problem: > > macro Smart_Average{ > stackid=getImageID; > no=nSlices; > stackname=getTitle; > run("Duplicate...", "title=[dup] duplicate range=1-no"); > setSlice((floor(no/2))); > run("Threshold..."); > waitForUser("Set Threshold then click 'OK'."); > setBatchMode(true); > run("Convert to Mask", " black"); > run("Divide...", "value=255 stack"); > rename("divided"); > bindupid=getImageID; > imageCalculator("Multiply create stack", stackid ,bindupid); > run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); > rename(stackname+"sum"); > sumid=getImageID; > setBatchMode(false); > selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode > close(); > } > > Best wishes > > Kees > > > Dr Ir K.R. Straatman > Senior Experimental Officer > Centre for Core Biotechnology Services > College of Medicine, Biological Sciences and Psychology > University of Leicester > > http://www.le.ac.uk/biochem/microscopy/home.html > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philipp Bethge > Sent: 21 March 2012 19:55 > To: [hidden email] > Subject: Re: 'smart' average z-projection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Users, > > Wow, thanks alot to everybody for the quick reply and the solutions offered. > I tried out the one from Glen and it works very nice! > > Thanks alot, > > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear Philipp, >> Duplicate the stack, threshold for the structure then make the tholded stack binary >> dividing by 255 converts the binarized stack from 0+255 to 0+1 >> multiply the 2 stacks and project the result. >> this function does all this, choosing middle slice for setting the threshold >> >> Maybe this is helpful, >> >> function thresholdprojection(){ >> stackid=getImageID; >> no=nSlices; >> stackname=getTitle; >> run("Duplicate...", "title=[dup] duplicate range=1-no"); >> setSlice((floor(no/2))); >> run("Threshold..."); >> waitForUser("Set Threshold then click 'OK'."); >> setBatchMode(true); >> getThreshold(lower,upper); >> run("Make Binary", "black"); >> run("Divide...", "value=255 stack"); >> rename("divided"); >> bindupid=getImageID; >> imageCalculator("Multiply create stack", stackid ,bindupid); >> run("Z Project...", "start=1 stop=no projection=[Sum Slices]"); >> rename(stackname+"sum"); >> sumid=getImageID; >> setBatchMode(false); >> selectImage(bindupid);//closing and renaming actions must occur outside of the batch mode >> close(); >> } >> >> Regards, >> Glen >> Glen MacDonald >> Core for Communication Research >> Virginia Merrill Bloedel Hearing Research Center >> Box 357923 >> University of Washington >> Seattle, WA 98195-7923 USA >> (206) 616-4156 >> [hidden email] >> >> >> >> >> >> >> >> >> On Mar 21, 2012, at 8:20 AM, Philipp Bethge wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear all, >>> >>> I was wondering if somebody has worked with a 'smart' average z-projection, that is a average projection of a z-stack that does not average of the entire stack size but only over the slides where a signal is actually present. So, for example, one has an oblique structure in a large stack, lets say the structure has a z-extension of 4 slides but the whole stack is 30 stacks large. As I understand the regular average z-projection algorithm in ImageJ, every pixel is summed and devided by the total number of slides in the stack. Then the signal of only 4 slides would be averaged over 30 slides, therefore reducing signal intensity dramatically without further reducing background noise. >>> Of course, one would have to define 'signal' somehow, but I guess that could be done manually by binarization of the image slides individually. >>> >>> Any suggestions welcome, >>> >>> Thanks alot, >>> >>> -- >>> Philipp Bethge > > > -- > Philipp Bethge |
Free forum by Nabble | Edit this page |