Scott, it looks like you have a fully automated system. Personally, I do not rely on autofocus, I would suggest, if you take multi position, multi channels over multiple time point, then instead of one frame position/well also include a small z stack, say 4 images (2 above and 2 below) from the best focus of each position, so latter even if there is a drift, you will have an option to select the best plane out of those four.
Shiv
At 07:56 AM 9/18/2008, you wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
All,
Have new faculty coming aboard and a major application will be 48 hour
timelapse studies in 96 well plates using multichannel collections.
Therefore my question is how critical is it to have a system that has
a perfect focus mechanism to meet this application? Are there decent
work arounds if the perfect focus is not used? Have never actually
worked with this type of system so would like input on this matter
from the list. Thanks.