Dear Matt,

Typically such calculations are done by calculating the maximal optical
resolution in Z-direction for the given objective (NA) and pinhole size
and taking a 2x oversampling. This is how it is done e.g. in the Zeiss
software. The Leica SP2 unfortunately doesn't do this automatically - as
far as I recall it takes the X-Y resolution values also for calculating
the Z-sections. You can kind of cheat the system by setting the
oversampling factor to 2. (Under Settings/Scan options/Sampling factor).

Cheers   Gabor

>