Re: Laser effects during bleaching

Posted by Kevin Braeckmans on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Laser-effects-during-bleaching-tp1515193p1519027.html

Runions signature

Hi John,

 

 

Yes, I appreciate that the maths can be challenging, however they tell us what can and cannot be done. In this particular case you could skip the maths and have a look at the experimental section, this is where it is actually explained without too much maths.

 

Just to make sure, if you want to compare recovery half times between experiments, not only the amount of photobleaching but the entire experimental set-up has to be the same: same size of photobleaching disk, same objective lens, same pinhole setting, same space around the photobleaching area – they all enter the equation. Some argue that this is not so important if you are just interested in a rough measurement. The thing is that with so many unknowns, you don’t have a clue anymore of the accuracy of the values that are obtained.

 

Sorry, don’t want to sound too strict, this is just my take on the subject.

 

Good luck and best regards,

 

Kevin

 

 

 

 

 

Van: Confocal Microscopy List [mailto:[hidden email]] Namens John Runions
Verzonden: woensdag 19 november 2008 11:21
Aan: [hidden email]
Onderwerp: Re: Laser effects during bleaching

 

Whew, I'm glad you jumped in there Kevin.  The Axelrod paper is the paper that explains things but I find the math very dense.  Within a given experiment, we bleach a 2D uniform disk and I do try to control so that the level of photobleaching is approximately the same each time.  But, I don't necessarily  control between experiments and I have been happily letting people compare results from different experiments that used different bleaching conditions and levels.

The type of FRAP we are doing is only semi-quantitative.  In fact, I have been calling the half-time of recovery the 'relative diffusion coefficient'. This is because there are too many unknowns in the cell membranes we study to derive the actual diffusion coefficient.

From now on, I will stress the importance of bleaching to the same approximate level each time.

Thanks for your help.  John.

Kevin Braeckmans wrote:

Hi,

 

In fact, generally speaking, this is not the case. For example, the recovery half-time generally depends on the amount of photobleaching. There is one exception: the photobleaching of a 2-D uniform disk. This is clearly explained in the ‘father of all FRAP articles’: Axelrod et al 1976, Biophys J 16.

 

For the same reason I strongly discourage the normalization to the post-photobleaching level (i.e. putting the post-photobleaching level to 0%)! If you want to compare FRAP curves, they all should have been acquired using exactly the same settings, including the amount of photobleaching. Therefore, there is no need at all to do the post-photobleaching normalization, except to mask the differences between different experiments… I know that in many semi-quantitative FRAP articles this is being done, but I would not recommend this since it is physically incorrect. That being said, for small amounts of photobleaching, small differences do not matter so much (see e.g. fig 7 of the Axelrod paper in case of a 2-D spot photobleached by a stationary Gaussian beam).

 

Hope this helps,

 

Best regards,

 

Kevin

 

 

 

Prof. Dr. Kevin Braeckmans

Lab. General Biochemistry and Physical Pharmacy

Ghent University

Harelbekestraat 72

9000 Ghent

Belgium

Tel: +32 (0)9 264.80.78

Fax: +32 (0)9 264.81.89

 

 

 

 

 

 

 

Van: Confocal Microscopy List [[hidden email]] Namens John Runions
Verzonden: dinsdag 18 november 2008 18:37
Aan: [hidden email]
Onderwerp: Laser effects during bleaching

 

Hi Tom,

When doing FRAP, we generally try to keep the bleaching lasers as low as possible so that a loss of fluorescence results but so that we aren't applying enough laser power to damage other proteins in membranes.  If the lasers were so hot as to denature proteins the whole procedure would be invalidated as we'd be killing the cells (I do often imagine the sound of frying bacon during the bleaching phase and hope that it's not really happening!).

At normal laser power levels for confocal, cells seem to function properly over the relatively long term.  The trick during the bleaching phase is to increase the laser power so that a bit of bleaching results.  If the region to be studied is bleached completely black, then I feel there really is the potential for cellular damage.  I've certainly exploded cells with too much laser power so your specimens should be handled gently.  The small amount of bleaching attained using reasonable laser power is analyzed relativistically, i.e. you normalize post-bleach and pre-bleach intensity between 0-100% intensity so that there is no requirement to bleach the fluorescence completely away.  The idea of fluorescence recovery is that any molecule will move to its equilibrium concentration in time (unless constrained).  The species of molecule that we are analyzing the dynamics of in FRAP are the bleached and unbleached versions of the fluorochome.  FRAP can be done, therefore, with a small population of bleached molecules.

In some recent experiments, we have been FRAPing so that only a very small decrease in fluorescence occurs within a region of interest.  Recovery curves fit from these experiments seem equivalent to those fit to data sets in which much more bleaching of the ROI was visible.  We need to ascertain if they are significantly the same, however.  If any physicists of diffusion are listening it would be nice to have your take on this.

Regards, John.

Donnelly, Tom wrote:

Just to add to the confusion.

 

What happens to the non-fluorescent proteins in the cell when you crank up the power to bleach or image in live cells?

 

Tom

-----Original Message-----
From: Confocal Microscopy List [[hidden email]]On Behalf Of John Runions
Sent: Tuesday, November 18, 2008 9:08 AM
To: [hidden email]
Subject: Re: broadband excitation vs. narrow band



Boy, what a stupid question Carl.  I think we should all fail to dignify it with a response!

Actually, that kind of question is fundamental to us in the FRAP world and it sounds like someone needs to do some good empirical measuring.  When bleaching GFP we will often use all of the 458, 477,488 and 514 lines of the argon laser at the same time.  It works better that just using the 488 and I explain this by saying that it is because we are bleaching with a broader spectrum across the excitation range.  I have never been able (or tried) to confirm if this is the case or if the higher power at the specimen plays a role.

Sorry to be not much help.  John.

Carl Boswell wrote:

Hi all,
I was told once that there are no stupid questions, so let's test that assumption.

The question has to do with photobleaching vs. excitation energy.  To get X photons from a fluor, would there be less photobleaching using a single wavelength excitation source at peak excitation wavelength, or a broadband (20-30nm) light source centered on the peak excitation of the molecule, or would there be no difference?  My assumption is that lower "power" (brightness?) would be needed for the broadband source, but would the overall photon flux be greater to get equivalent output?

To take this one step farther, is there less or more bleaching from "inefficient" excitation, i.e. off-peak excitation, to get the same output? If a fluor is less efficiently excited, is it less efficiently bleached, even though more power may be needed to get equivalent output?

Thanks,
Carl

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709

 

--


(Sent from my cra%#y non-Blackberry electronic device that still has wires)

*********************************
C. John Runions, Ph.D.
School of Life Sciences
Oxford Brookes University
Oxford, UK
OX3 0BP

email:  [hidden email]
phone: +44 (0) 1865 483 964

Runions’ lab web site

 

Visit The Illuminated Plant Cell dot com
Oxford Brookes Master's in Bioimaging with Molecular Technology

 

--


(Sent from my cra%#y non-Blackberry electronic device that still has wires)

 

*********************************
C. John Runions, Ph.D.
School of Life Sciences
Oxford Brookes University
Oxford, UK
OX3 0BP

email:  [hidden email]
phone: +44 (0) 1865 483 964

Runions’ lab web site

 

Visit The Illuminated Plant Cell dot com
Oxford Brookes Master's in Bioimaging with Molecular Technology

 

--

(Sent from my cra%#y non-Blackberry electronic device that still has wires)

 

*********************************
C. John Runions, Ph.D.
School of Life Sciences
Oxford Brookes University
Oxford, UK
OX3 0BP

email:  [hidden email]
phone: +44 (0) 1865 483 964

Runions’ lab web site

 

Visit The Illuminated Plant Cell dot com
Oxford Brookes Master's in Bioimaging with Molecular Technology