Re: Photobleaching of GFP in Spinning Disk Confocal

Posted by Gary Laevsky-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/wasabi-freeware-tp1574097p1576384.html

Hi Shiv,

 

A few options that may or may not have been tried;

 

Turning the EM gain off.  Depending on the intensity of the fluor, you actually may be adding noise.  If you turn the EM gain off, you may get a “better” image, and therefore you may be able to reduce the exposure/laser power.

 

Make sure you have optimized the filters in the path. 

 

Also, I believe you may have additional optics in the emission path that are stealing valuable photons.

 

As stated, “30” (FYI this is in a non-linear scale to 100) is not a very informative number to determine intensity at the sample as it depends size of laser and throughput i.e. optics in the excitation path (this is with a CSU22).  Your measurement of 240um, however, is perfectly objective (excuse the mild pun).  And, again as stated, is very reasonable.

 

Talk soon.

 

 

 

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell         (774) 291 - 9992

Office       (860) 290 - 9211 x219

Fax          (860) 290 - 9566

Web:       www.andor.com


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mayandi Sivaguru
Sent: Monday, November 24, 2008 6:09 PM
To: [hidden email]
Subject: Photobleaching of GFP in Spinning Disk Confocal

 


Hi all, One of our client is using a cell line expressing GFP bleaches very fast (Andor Revolution SDC system), despite using low excitation setting at the 488 line laser (set at 30 at the AOTF and approximately 240 microW on the 100x Olympus UPLANSAPO objective), EM Gain in combination with high amplifier gain. She wants to get two frames per second under two channels she is getting it but the cell bleaches within couple of minutes. Using live cell antioxidants such as Trolox is not an option, I would greatly appreciate any insights in this regard. Will neutral density filters help?

Thanking you in advance
Shiv




Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]