http://confocal-microscopy-list.275.s1.nabble.com/wasabi-freeware-tp1574097p1578586.html
of oxygen (e.g. in the presence of Oxyrase)! And it recovers slowly
once oxygen is present again. I made this observation (and tried to
(I don't know about other variants). If you just put a wet coverslip
>Hi Shiv,
>
>Anything that reduces illumination intensity at the sample will
>help. If you don't need ultimate resolution, you could try binning
>or using a lower power objective. If your samples are in water, you
>may try using a water lens if available to reduce spherical
>aberration, and hence increase detection efficiency. Also, there may
>be other trivial reasons for the bleaching, such as suboptimal pH in
>the sample/medium, and, of course, GFP is very sensitive to solvents
>(nail polish, etc...). You mention Trolox is not an option. How about Oxyrase:
>
><
http://www.oxyrase.com/oxyfluor.html>
http://www.oxyrase.com/oxyfluor.html>
>
>There was a thread on this list a while ago about the merits of
>Catalase/Glucose Oxydase
>
>>10mM -mercaptoethanol, 2.5 mg/ml glucose, 20 g/ml catalase, 0.1
>>mg/ml glucose oxydase
>>
>>The concentrations above are from Chabrillat et al, Molec Biol of the
>>Cell 16, 1640-1650 (2005).
>
>I think there is still some debate as to whether this actually works...
>
>Julio.
>
>
>--
>Julio Vazquez,
>Fred Hutchinson Cancer Research Center
>Seattle, WA 98109-1024
>
>
><
http://www.fhcrc.org>
http://www.fhcrc.org/>
>===
>
>
>On Nov 24, 2008, at 3:08 PM, Mayandi Sivaguru wrote:
>
>>
>>Hi all, One of our client is using a cell line expressing GFP
>>bleaches very fast (Andor Revolution SDC system), despite using low
>>excitation setting at the 488 line laser (set at 30 at the AOTF and
>>approximately 240 microW on the 100x Olympus UPLANSAPO objective),
>>EM Gain in combination with high amplifier gain. She wants to get
>>two frames per second under two channels she is getting it but the
>>cell bleaches within couple of minutes. Using live cell
>>antioxidants such as Trolox is not an option, I would greatly
>>appreciate any insights in this regard. Will neutral density filters help?
>>
>>Thanking you in advance
>>Shiv
>>
>>
>>
>>Mayandi Sivaguru, PhD, PhD
>>Microscopy Facility Manager
>>8, Institute for Genomic Biology
>>University of Illinois at Urbana-Champaign
>>1206 West Gregory Dr.
>>Urbana, IL 61801 USA
>>
>>Office: 217.333.1214
>>Fax: 217.244.2496
>><mailto:
[hidden email]>
[hidden email]
>>
http://core.igb.uiuc.edu