I have been informed I neglected to
mention a very important issue.
Not only should you turn the EM gain off,
but you should also switch to the conventional register.
The EM register, because of it’s amplification
characteristics adds a tremendous amount of noise to the image.
Some EMCCDs allow you to be able to switch
to the conventional register. Your noise will be greatly reduced, and you
can then possibly turn down the laser.
Best,
Gary Laevsky, Ph.D.
Imaging Application Specialist
Andor Technology
discover new ways of seeing
Cell (774)
291 - 9992
Office
(860) 290 - 9211 x219
Fax
(860) 290 - 9566
Web:
www.andor.com
From: Confocal
Microscopy List [mailto:[hidden email]] On Behalf Of Mayandi Sivaguru
Sent: Monday, November 24, 2008
6:09 PM
To:
[hidden email]
Subject: Photobleaching of GFP in
Spinning Disk Confocal
Hi all, One of our client is using a cell line expressing GFP bleaches very
fast (Andor Revolution SDC system), despite using low excitation setting at the
488 line laser (set at 30 at the AOTF and approximately 240 microW on the 100x
Olympus UPLANSAPO objective), EM Gain in combination with high amplifier gain.
She wants to get two frames per second under two channels she is getting it but
the cell bleaches within couple of minutes. Using live cell antioxidants such
as Trolox is not an option, I would greatly appreciate any insights in this
regard. Will neutral density filters help?
Thanking you in advance
Shiv