Re: objective focal length
Posted by
Beat Ludin on
URL: http://confocal-microscopy-list.275.s1.nabble.com/objective-focal-length-tp1588271p1588538.html
Hi Aryeh
Maybe the tube lens is not the original one, so that the "25x"
objective really functions as a 10x in this setup?
Can you give us more info about the setup. A 25x/0.5 air objective
with a 11mm WD must be pretty big...
Beat
At 09:20 28-11-2008, you wrote:
>I am confused with regard to the front focal length of objectives. I
>thought that the magnification of an infinity corrected objective
>will be the distance between the tube lens and the intermediate
>image, divided by the front focal length of the objective (which is
>where I expect the object to be). Since the objective is not really
>a thin lens, I can understand that the actual working distance may
>be less than the focal length, since the focal length may need to be
>measured from inside the objective casing.
>
>However, I have a paper that describes a 25x/NA=0.5 air objective,
>which has an 11mm working distance, as having a 25.1mm focal length,
>while the distance from the tube lens to the intermediate image is
>245mm. This has me confused, and I realize that I do not understand
>something fundamental here.
>
>So, I turn to the optics gurus on the list to clear this up, with
>many thanks in advance.
>
>--aryeh
>--
>Aryeh Weiss
>School of Engineering
>Bar Ilan University
>Ramat Gan 52900 Israel
>
>Ph: 972-3-5317638
>FAX: 972-3-7384050