Re: objective focal length

Posted by Aryeh Weiss on
URL: http://confocal-microscopy-list.275.s1.nabble.com/objective-focal-length-tp1588271p1592872.html

The problem is that an infinity corrected 25x lens with a 25mm focal length
should require a tube lens with a 625mm focal length (25x25). Does such a
microscope exist?

--aryeh


Patrick Van Oostveldt wrote:

> Dear Arey,
>
> Not being an expert I should say: if the tube lens forms an image at
> 245mm (=focal length) and if the lens magnification is 25x the objective
> should have a focal length of 9.8mm. The working distance is smaller
> than this, but is not directly predictable.
>
> However the magnifcation inscribed on the lens (25X) is dependent on a
> specific focal length of the tube lens. For Nikon objectives it is 200mm
> for Zeiss it is 164.5nmm. Probably this can be the point of confusion.
>
> Best regards
>
> Patrick
>
> Quoting Aryeh Weiss <[hidden email]>:
>
>> That is also how I understand it. So I assume that the distance between
>> the tube
>> lens and the intermediate image plane is the focal length of the tube
>> lens.
>>
>> Best regards,
>> --aryeh
>>
>> Guy Cox wrote:
>>> I am absolutely not an optics guru!
>>>
>>> But, as I understand it, an infinity corrected objective will form an
>>> image
>>> at 'infinity', so the distance from the principal plane to the  
>>> object will be
>>> the focal length.  But where the image is formed after the tube lens
>>> will
>>> depend on the focal length of the tube lens and we have to know this
>>> to make
>>> any useful calculation.
>>>
>>> It was all so much simpler in the days of 160mm tube length!
>>>
>>> Guy
>>>
>>> -----Original Message----- From: Confocal Microscopy List
>>> [mailto:[hidden email]] On Behalf Of Aryeh Weiss Sent:
>>> Friday, 28 November 2008 7:20 PM To:  
>>> [hidden email] Subject: objective focal length
>>>
>>> I am confused with regard to the front focal length of objectives. I
>>> thought
>>> that the magnification of an infinity corrected objective will be the
>>> distance between the tube lens and the intermediate image, divided by
>>> the
>>> front focal length of the objective (which is where I expect the
>>> object to
>>> be). Since the objective is not really a thin lens, I can  understand
>>> that the
>>> actual working distance may be less than the focal length, since the
>>> focal
>>> length may need to be measured from inside the objective casing.
>>>
>>> However, I have a paper that describes a 25x/NA=0.5 air objective,
>>> which has
>>> an 11mm working distance, as having a 25.1mm focal length, while  the
>>> distance
>>> from the tube lens to the intermediate image is 245mm. This has me
>>> confused,
>>> and I realize that I do not understand something fundamental here.
>>>
>>> So, I turn to the optics gurus on the list to clear this up, with  
>>> many thanks
>>> in advance.
>>>