Re: Live cell Imaging with fluoview 1000
Posted by
Naomi Book on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Live-cell-Imaging-with-fluoview-1000-tp2242987p2243590.html
Hi,
I have used the FV-1000 many times, for long
(up to 72 hours) live cell imaging, using HeLa cells, cos and other cell lines.
I use 2 excitations 488 and 543. the cells survive with no problem. the LIS
incubator is doing a great job.
Naomi Book
--------------------------------------------
Naomi Melamed-Book, PhD
Bio-Imaging Unit, Life Science
Institute
Hebrew University, Jerusalem, Israel.
Phone: 972-2 6585453
-------------------------------------------------
----- Original Message -----
Sent: Friday, January 30, 2009 5:31
AM
Subject: Live cell Imaging with fluoview
1000
Hi,
I had my first experience today with this system
for live cell imaging and all my cells under the laser didn't make it safely
(the one around were just fine).
So I'm curious to know if some of you used a fluoview 1000 for live
cell imaging of hoechst or Fm4-64? If yes, what kind of range of power and
aquisition repetition (z stack and repetition) did your cells take ? (I did a
stack of 16 images in z every 2 minutes so around 1minute of imaging and 1
minute of resting)
Thanks,
JP
- - - - - - - - - - - - - - - - - - - - - - - - - - - -
- - - - - -