> Dear JP,
> You need to measure thresholded Manders coefficients
(1 for red , 1
> for green!)
> search this list archives for my previous long
posts on this topic.
>
> in short, use ImageJ or easier BioImageXD (both free)
> to calculate threshold using the methods of costes,
> then calculate the Manders coefficents 1 and 2
> then do the costes statistical significance test.
> Publish:
> thresholds, manders coefficents, statistical significance,
> 2D histogram . scatter plots with straight line fit and thresholds.
> You need to read 2 papers at least
> Costes et al
> and Manders et al.
> I will send them to you in a moment
> cheers
> Dan
>
> Begin forwarded message:
> >
> > Date: Wed, 4 Feb 2009 17:05:30 -0800
> > From: Jean-Pierre CLAMME <[hidden email]>
> > Subject: colocalization question
> >
> > This is a multipart message in MIME format.
> > --=_alternative 0005FF9888257554_=
> > Content-Type: text/plain; charset="US-ASCII"
> >
> > Hi,
> >
> > When you do colocalization measurement (in image pro for example)
> > between
> > let say the red and a green channel the co-localization coefficient
> > tells
> > you how much red and green are co-localized. But as much
as I
> > understood,
> > it doesn't give you information on the fact that for example
red is
> > always
> > co-localized with green but green is not always colocalized with
red ?
> > What kind of analysis would I need to compare this between
samples ?
> >
> > Thank you in advance,
> >
> > JP
> >
> >
> > - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
- - -
> > Jean-Pierre CLAMME, PhD
> > Senior Scientist
> > Nitto Denko Technical
> > 501 Via Del Monte
> > Oceanside, CA 92058
> > E-mail: [hidden email]
> > Phone: +760.435.7065
> Dr. Daniel James White BSc. (Hons.) PhD
> Senior Microscopist / Image Processing and Analysis
> Light Microscopy Facility
> Max Planck Institute of Molecular Cell Biology and Genetics
> Pfotenhauerstrasse 108
> 01307 DRESDEN
> Germany
>
> New Mobile Number!!!
> +49 (0)15114966933 (German Mobile)
> +49 (0)351 210 2627 (Work phone at MPI-CBG)
> +49 (0)351 210 1078 (Fax MPI-CBG LMF)
> http://www.bioimagexd.net
> http://www.chalkie.org.uk
> [hidden email]
> ( [hidden email] )