Re: Configuration problem with Zeiss LSM 510
Posted by
G. Esteban Fernandez on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Configuration-problem-with-Zeiss-LSM-510-tp2303218p2303330.html
You need to go into the Maintain > Pinhole menu and optimize the X/Y positions of pinhole 3. The message means that has never been done for that particular configuration of dichroics. The pinhole is probably not centered directly in front of the PMT so you don't get a good image if it is closed down, but you'll probably get an image if you open the pinhole to max.
-Esteban
On Tue, Feb 10, 2009 at 9:54 AM, yee z
<[hidden email]> wrote:
Dear listers,
We have a Zeiss LSM510 confocal microscopy system at our core facility. And I got some problems while writing new configuration.
Here is an example:
I am still be able to take the images regardless the error window.
So does anyone here have the same problem? What does "Pinhole PH3 is not adjusted" mean, any reason for this?
Any suggestion will be appreciated.
Thanks,
Yi
|
--