Re: Configuration problem with Zeiss LSM 510
Posted by
Boswell, Carl A - (cboswell) on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Configuration-problem-with-Zeiss-LSM-510-tp2303218p2303677.html
All I can add to what Estaban said was be sure to
Save the position after you have made the adjustments. Adjust the image
Gain using the Palatte>Range Indicator so there is some saturation in it so
that you have an objective way to see if your changes are making things brighter
or dimmer. Adjusting the columator may also improve your
image.
Carl
Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of
Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
Sent: Tuesday, February 10, 2009 9:12
AM
Subject: Re: Configuration problem with
Zeiss LSM 510
You need to go into the Maintain > Pinhole menu and optimize the X/Y
positions of pinhole 3. The message means that has never been done for
that particular configuration of dichroics. The pinhole is probably not
centered directly in front of the PMT so you don't get a good image if it is
closed down, but you'll probably get an image if you open the
pinhole to max.
-Esteban
On Tue, Feb 10, 2009 at 9:54 AM, yee z
<[hidden email]>
wrote:
|
Dear listers,
We have a Zeiss LSM510 confocal microscopy system at our core
facility. And I got some problems while writing new configuration.
Here is an example:
I am still be able to take the images regardless the error
window.
So does anyone here have the same problem? What does "Pinhole PH3
is not adjusted" mean, any reason for this?
Any suggestion will be appreciated.
Thanks,
Yi
|
--