Re: Configuration problem with Zeiss LSM 510
Posted by
yee z on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Configuration-problem-with-Zeiss-LSM-510-tp2303218p2304260.html
Dear lister,
Thanks a lot for the feedback on my question. From what you guys said, now I know I need to do pinhole adjustment for every new configuration. Learn something new everyday :-)
And I have another question regarding the pinhole adjustment. Jean-Christophe suggested to use a mirror slide while Xuejun suggested to use Zeiss plant cross section slide, which one is better for this purpose? I have a cross cut plant seed slice, can I use this one?
Thanks again!
--- On Tue, 2/10/09, Carl Boswell <[hidden email]> wrote:
From: Carl Boswell <[hidden email]> Subject: Re: Configuration problem with Zeiss LSM 510 To: [hidden email] Date:
Tuesday, February 10, 2009, 12:16 PM
All I can add to what Estaban said was be sure to
Save the position after you have made the adjustments. Adjust the image
Gain using the Palatte>Range Indicator so there is some saturation in it so
that you have an objective way to see if your changes are making things brighter
or dimmer. Adjusting the columator may also improve your
image.
Carl
Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of
Arizona 520-954-7053 FAX 520-621-3709
----- Original Message -----
Sent: Tuesday, February 10, 2009 9:12
AM
Subject: Re: Configuration problem with
Zeiss LSM 510
You need to go into the Maintain > Pinhole menu and optimize the X/Y
positions of pinhole 3. The message means that has never been done for
that particular configuration of dichroics. The pinhole is probably not
centered directly in front of the PMT so you don't get a good image if it is
closed down, but you'll probably get an image if you open the
pinhole to max.
-Esteban
On Tue, Feb 10, 2009 at 9:54 AM, yee z <[hidden email]>
wrote:
|
Dear listers,
We have a Zeiss LSM510 confocal microscopy system at our core
facility. And I got some problems while writing new configuration.
Here is an example:
I am still be able to take the images regardless the error
window.
So does anyone here have the same problem? What does "Pinhole PH3
is not adjusted" mean, any reason for this?
Any suggestion will be appreciated.
Thanks,
Yi
|
-- G. Esteban Fernandez, Ph.D. Associate
Director Molecular Cytology Core Facility University of Missouri 120
Bond Life Sciences Center Columbia, MO 65211
http://www.biotech.missouri.edu/mcc/
573-882-4895 573-884-9395
fax
|