Presumably the problem is that the Yokogawa is designed for x60 NA
1.4 - >
it's the magnification as well as the NA that determines the size
of the Airy >
disk at the pinhole. So a x20 NA 0.5, which is a common lens,
ought to >
work pretty well (one third the mag and ~ one third the
NA). >
Guy >
Optical Imaging Techniques in Cell Biology > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C > Sent: Wednesday, 11 February 2009 7:47 AM > To: [hidden email] > Subject: Re: Zeiss SD >
I tend to agree with Dick here, The pinhole array solutions
while less "mature" than the spinning disk confocals work very well.
We have had a Yokagawa head for ages, and the limitation to a 1.4NA
objective was significant. About a year ago we bought a Prairie
swept field array scanning confocal. It allows multiple pinholes or
slits sizes (7) depending on your specimen and optics used. It
works very well indeed, and in our hands is more flexible than the
Yokogawa head. Its important to note that the yokogawa head I have
is very old and there have been many advances. However without rapid
and simple changing of pinhole size (as is possible with the Prairie or
Visitech solutions) your image will be compromised when NA is
changed. >
S. >
Simon C. Watkins Ph.D, FRCPath >
Professor and Vice Chair, Cell Biology and
Physiology >
Professor, Immunology >
Director, Center for Biologic Imaging >
BSTS 225, University of Pittsburgh >
3500 Terrace St. >
Pittsburgh PA 15261 >
Tel: 412-352-2277 >
Fax:412-648-2797 >
URL: http://www.cbi.pitt.edu >
From: Confocal
Microscopy List [mailto:[hidden email]] On Behalf Of
RICHARD BURRY >
Fred
>
No virus found in this incoming message. > Spam > Not spam > Forget previous vote |
>
No virus found in this outgoing message.
>
Checked by AVG.
>
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date:
10/02/2009 5:44 PM
No virus found in this incoming message.
Checked by
AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date:
10/02/2009 5:44 PM
No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM
| Free forum by Nabble | Edit this page |